Jerry et al:
As we have a newly commissioned toxic gas monitoring system here at the
Stanford Nanofabrication Facility, let me tell you what we ended up with
to complement all of the other good input you are receiving.
In February of this year, we replaced an aging system that had about 80
monitored points. Our new system monitors nearly 150 points with gas
detectors. Ny guess is that about 125 of those points are monitoring
our main shared facility and the remaining 25 are monitoring points in
private labs that share our building. In our case, we went with "draw
type" sensors. We have 16 channels of paper tape monitoring of hydrides
using a DOD Technologies CL-96 unit. The remainder of our sensors are
DOD Technologies PS-7 electrochemical sensors that are also draw type
sensors. Note: we would have likely only used the PS-7 electrochemical
sensors were it not for the fact that we have an epitaxial reactor that
uses arsine. To my knowledge, none of the electochemical sensors can
reliably detect the 50 ppb PEL level required for arsine gas.
We have also added digital input fire eyes in all of our silane
cylinders and in our silane VMBs. Note: all of our sensors put out
old-fashioned 4-20 mA signals that are monitored by analog input modules
in the PLC. I know that there are a number of detectors that have web
interfaces, etc., but we chose not to go that route.
We do like draw-type sensors:
They don't tend to dry out if they are sensing a point in either a
high-velocity or high-temperature area.
You can look at a non-zero reading without being nearly so close to an
actual leak.
Being able to cluster a bunch of sensor bodies in one area makes annual
live-gas testing easier and allows you to get a better idea if more than
one sensor is "smelling" something.
Our system is PLC based with inputs going to a GE 3Xi PLC. HMI in our
case is provided through a WonderWare interface. Our TGO computer is in
the FACP control room which is, we believe, out of harms way in the
event of a gas leak. Our fire department always goes there first and is
sufficiently well trained that they can get a pretty good idea of what
is going on even if none of us is there yet. We also have VPN remote
access to the HMI as well as synthesized voice phone dialer and email
messaging. The voice and email options provide a listing of the sensor
involved and whether it is reporting a fault, warning level, or alarm.
All of our sensors are set up to provide a warning signal at about 50%
of the alarm level. The warning level does not evacuate the building or
call the fire department.
In general, even though legal requirements only require alarms in
exhausted spaces at 1/2 IDLH, we have all of our alarms set at PEL for
several reasons:
1. Dilution in most gas cabinets and exhausted enclosures is high. In
many cases it would be quite a large leak to reach 1/2 IDLH in those spaces.
2. If our detectors are set at PEL in exhausted enclosures, we can also
use that sensor to monitor the nearby breathing air as any gas outside
the enclosure will soon get drawn in to the enclosed space.
3. We'd rather find and deal with a problem as soon as possible. Little
leaks seem to have an annoying habit of becoming larger leaks if left
unattended.
Note: in the same vein, we have also installed 0-1000 ppm hydrogen
sensors rather than 0-100% LEL sensors.
I recommend early discussions with your Authority Having Jurisdiction.
Their interpretation and preferences can have a large bearing on what
you may end up with. For example, in our case, Santa Clara county would
categorically be opposed to any approach that was not fully designed and
stamped by a Registered PE including comparatively minor
changes/upgrades to an existing system. While I know that there are a
number of examples of very nice in-house designed and built systems ...
and, at some level, I envy those that have full control ... that is
something that simply wouldn't fly with our AHJ.
Good luck, there are a lot of choices and options. You are welcome to
contact me should you care for any additional detail.
Thanks,
John