[labnetwork] joining long PVDF pipe runs
Abbie Gregg
agregg at abbiegregg.com
Fri Jul 6 15:39:59 EDT 2012
Hi Richard,
Has the piping been certified to deliver E-1 grade water yet? Or some other specific criteria? Everyone is interested to know if your material and method are adequate....
Best Regards,
Abbie Gregg
President
Abbie Gregg, Inc.
1130 East University Drive, Suite 105
Tempe, Arizona 85281
Phone 480 446-8000 x 107
Cell 480-577-5083
FAX 480-446-8001
email agregg at abbiegregg.com
website www.abbiegregg.com
CONFIDENTIALITY NOTICE: All information contained in or attached to this email constitutes confidential information belonging to Abbie Gregg, Inc., its affiliates and subsidiaries and/or its clients. This email and any attachments are proprietary and/or confidential and are intended for business use of the addressee(s) only. All other uses or disclosures are strictly prohibited. If the reader is not the intended recipient you are hereby notified that the perusal, copying or dissemination of this email is strictly prohibited. If you have received this communication in error, please notify the sender, and delete all copies of this message and its attachments immediately.
-----Original Message-----
From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Morrison, Richard H., Jr.
Sent: Friday, July 06, 2012 4:12 AM
To: Iulian Codreanu; Fab Network
Subject: Re: [labnetwork] joining long PVDF pipe runs
Hi,
We are just finishing construction on a new cleanroom here at Draper. Our DI water piping is natural polypropylene, IR fusion welded with butt joints. We felt that this could provide us a DI system that meets the semiconductor standard. The pipe supplier is Asahi/America. The pipe is weld in an on-site cleanroom then the pieces are assembled in the air on overhead racks resting in a pipe trough. To make assembly a little easier the pipe sections are joined by clamps.
This was done to keep assembly cost reasonable and purity sufficient for our operation.
Hopes this helps.
Rick
Rick Morrison
Senior Member Technical Staff
Group Leader Microfabrication Operations
Draper Laboratory
555 Technology Square
Cambridge, MA 02139
W 617-258-3420
C 508-930-3461
-----Original Message-----
From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Iulian Codreanu
Sent: Thursday, July 05, 2012 8:34 AM
To: Fab Network
Subject: [labnetwork] joining long PVDF pipe runs
Good Morning.
We are trying to determine how to best join long runs of PVDF pipe on
our ultrapure water system (we are trying to achieve Type E-1.1 per ASTM
D5127). We are aware of the following methods:
- fully welded system
- systems using flanges
- systems using unions
- systems using GF sanitary couplings with clamps
Knowing how other University Fabs do it would be very helpful in making
a decision (affecting water quality, cost, and schedule) and thus your
input will be much appreciated.
Iulian
--
iulian Codreanu, Ph.D.
Director, Penn NanoFab
200 South 33rd Street
Room 376 GRW Bldg
Philadelphia, PA 19104-6314
P: 215-898-9308
F: 215-573-2068
www.seas.upenn.edu/~nanofab
_______________________________________________
labnetwork mailing list
labnetwork at mtl.mit.edu
https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork
_______________________________________________
labnetwork mailing list
labnetwork at mtl.mit.edu
https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork
More information about the labnetwork
mailing list