[labnetwork] Steam Humidifier system question
Tribble, Thomas
tribble at fas.harvard.edu
Fri Sep 14 15:56:37 EDT 2012
Harvard FAS does both. We feed our Steam-to-Steam humidification boilers with 1 to 1.5 megohm RO water. These are AHU sized devices (up to 40,000 CFM) used for general ventilation. We feed a critical lab-specific ultrasonic (piezoelectric) duct-mounted humidifier (Stultz) with 18 megohm RO/DI. The RO/DI is piped in stainless by virtue of its corrosive properties as alluded to.
Tom Tribble
From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Morrison, Richard H., Jr.
Sent: Friday, September 14, 2012 3:20 PM
To: labnetwork at mtl.mit.edu
Subject: [labnetwork] Steam Humidifier system question
Hi All,
Facilities at my labs want to run 18megohm to a boiler for humidity control in another lab. I think it should be RO water as 18Megohm water is very corrosive. Any thoughts on this?
Their suggested design would draw 0.5gmp from my supply loop and return it to the DI tank? My belief is that this return water would be contaminated by the boiler system. What do you guys think?
Rick
Draper Laboratory
Group Leader Microfabrication Operations
555 Technology Square
Cambridge Ma, 02139-3563
www.draper.com<http://www.draper.com>