> Valli,
>
> Depending on the temperature and pressure in your gas cabinet it is
> possible to have liquid in the pig tail, regulator, or delivery line. This
> could cause unpredictable pump down times. Since I do not know your
> specific cabinet design and configuration this is only a suggestion on my
> part.
>
> Jay.
>
> James Goodman
> Equipment Manager, Saab/ETIC Nanofabrication Laboratory
> University of Massachusetts, Lowell
> 40 University Ave.
> Room 121
> Lowell, MA -01854
> Office (978) 934-3469
> Cell (603) 235-1496
>
> -----Original Message-----
> From: labnetwork-bounces at mtl.mit.edu
> [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Satyavalli Paluri
> Sent: Monday, March 25, 2013 8:49 AM
> To: labnetwork at mtl.mit.edu
> Cc: xandeeps at gmail.com; vanaparthy at ee.iitb.ac.in; Hemant Ghadi
> Subject: [labnetwork] Query regarding ICPRIE system-IIT Bombay_India
>
>
> Hello ,
>
> We have a Sentech ICPRIE system. BCl3 gas line is connected to it for
> etching purpose. We are facing a problem while evacuating the line
> sometimes. It takes longer time than normal for evacuation.
> Can someone please advise.
>
>
> Thanks & regards
> Valli
>
> --
> Assistant Lab Manager,
> CEN, Electrical Engg Dept, Annexe,
> IIT Bombay, Mumbai, Powai- 400076
> Tel No. 02225764435/09820856337(M)
>
>
>
>
>
>
>>
>>
>> ---------------------------- Original Message
>> ----------------------------
>> Subject: Re: [labnetwork] Toxic Gas Monitoring Systems
>> From:    "Fouad Karouta" <fouad.karouta at anu.edu.au>
>> Date:    Thu, August 9, 2012 5:15 am
>> To:      "'Bowser, Jerry'" <jerry.bowser at nist.gov>
>>          labnetwork at mtl.mit.edu
>> --------------------------------------------------------------------------
>>
>> Hi Jerry,
>>
>>
>>
>> At our facility we have gas monitoring using sensors from Bionics (Cl2,
>> BCl3, SiH4, NH3, flammables). All sensors are monitored by a home built
>> monitoring cabinet based on a quite flexible software that we kept
>> adapting
>> to meet our needs. All sensors have two levels of alarm:
>>
>> 1-      Alarm 1: just give a warning at the local level (monitoring
>> cabinet)
>>
>> 2-      Alarm 2: here we differentiate between sensors in open area
>> towards
>> sensors in exhaust ducting. The open lab sensors will trigger a general
>> alarm: sirens, shutting down all gases, evacuation and alert fire Dept;
>> while the exhaust detectors will initiate shutting down the
>> corresponding
>> gas and the system using it, shutting the exhaust fan plus a local alarm
>> at
>> the monitoring cabinet level/lab.
>>
>>
>>
>> Before taking my actual position in Australia I worked at Technical
>> University Eindhoven where an 800m2 clean room was built in 2000-2002
>> with
>> also Bionics sensors used to monitor gases like AsH3, PH3, SiH4, NH3,
>> Cl2,
>> flammables. Bionics did deliver a complete system with the interfacing
>> computer that controls all sensors.
>>
>>
>>
>> We get the sensors calibrated twice by representatives of Bionics in
>> Australia.
>>
>>
>>
>> Kind regards,
>>
>> Fouad Karouta
>>
>>
>>
>> *********************************
>>
>> Facility Manager ANFF ACT Node
>>
>> Research School of Physics and Engineering
>>
>> Australian National University
>>
>> ACT 0200, Canberra, Australia
>>
>> Tel: + 61 2 6125 7174
>>
>> Mob: + 61 451 046 412
>>
>> Email: fouad.karouta at anu.edu.au
>>
>>
>>
>> From: labnetwork-bounces at mtl.mit.edu
>> [mailto:labnetwork-bounces at mtl.mit.edu]
>> On Behalf Of Bowser, Jerry
>> Sent: Thursday, 9 August 2012 3:14 AM
>> To: 'labnetwork at mtl.mit.edu'
>> Subject: [labnetwork] Toxic Gas Monitoring Systems
>>
>>
>>
>> Hello All,
>>
>>
>>
>> I wanted to survey the group for information regarding the type and
>> brands
>> of toxic gas monitoring control systems in use.
>>
>>
>>
>> Here at the CNST, we use various gas sensors (about 55) connected to a
>> custom PLC based system that is just about at capacity.  If a gas is
>> detected, the system shuts down gas cabinets, sounds alarms, and alerts
>> our
>> fire department.  The major shortfall of the system is that the
>> detection
>> level can be read at the sensor in the detection zone but is not
>> displayed
>> at the control interface.
>>
>>
>>
>> Before we decide on upgrading or replacing our control system, I thought
>> I
>> would see what others are currently using.  Thanks for your input.
>>
>>
>>
>> Jerry
>>
>>
>>
>>
>>
>>
>>
>> *******************************************
>>
>> Jerry Bowser
>>
>> NanoFab Operations Group
>>
>> Center for Nanoscale Science and Technology
>>
>> National Institute of Standards and Technology
>>
>> Phone: (301) 975-8187
>>
>>
>>
>> _______________________________________________
>> labnetwork mailing list
>> labnetwork at mtl.mit.edu
>> https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork
>>
>>
>> --
>>
>
>
>
>
> --
>
>
>
>
>
>
>
> _______________________________________________
> labnetwork mailing list
> labnetwork at mtl.mit.edu
> https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork
>








More information about the labnetwork mailing list