[labnetwork] LN2 vs N2 generation on site
John Shott
shott at stanford.edu
Tue Mar 31 09:26:04 EDT 2015
Rick:
In an earlier message, you had commented about the large volume
difference between what you use and what gets delivered. To me,
something seems strange if you are getting 2.6M cuft per month but are
using only a bit more than half of that.
I believe that large LN2 tanks are noticeably more efficient than that
for gaseous usage and would expect that the amount that you use to be at
least 90% of what is delivered.
Here is my understanding ... and I've tried to check these numbers out
with our LN2 supplier:
A big tank that is just sitting there (that is with little or no usage)
would lose something like 0.5-1.0% per day due to heat transfer and gas
that ends up getting vented but not used. However, for any system that
is using a significant volume of product per day ... and yours clearly
is ... I would expect that venting losses would be very close to zero.
When you are using a significant volume of product, that would tend to
reduce the pressure in the tank, the pressure building circuit kicks in
... but that doesn't result in any loss because the liquid that is
vaporized to build pressure goes into your tank.
There is certainly a significant loss anytime that you have a delivery
... and, if your vendor is like ours, you pay for all of that nitrogen.
I believe that the largest fraction of that consumption is the nitrogen
that is "depressurized" from the tanker following your fill to reduce
the tanker pressure from above your tank pressure down to the legal
limit for them to drive again on the highways which is, I think, 10
PSIG. The nitrogen consumed during your fill that never goes into your
tank including pre- and post-fill purges and this tanker "blow down" is
significant. Others may have a better number than I, but I count on
losing at least 20,000 cuft of gas per delivery. On average, we get 8
deliveries per month of 600,000 cuft per delivery for a total delivered
volume of about 4.8M cuft. So, losing 20,000 cuft per delivery is only
a 3% loss. Of course, with a smaller tank, the per-fill loss percentage
goes up assuming that they are still filling with a 53' trailer.
Note: you will have higher losses if you are using liquid withdrawal to
any degree because there is a fair volume of nitrogen that is "lost" as
gaseous nitrogen for every liquid gallon that you extract. If your tank
pressure is set up for gaseous usage with a tank pressure of 120-150
PSIG, then any liquid usage is probably not terribly efficient because
of flash losses, etc. ... basically the gas that is vaporized in the
process of getting things cooled down so that you get good liquid
extraction. I don't have much experience in that area so can't really
offer a precise number for loss during liquid withdrawal but that is
likely to be significantly lower than the 90% number that I think you
should expect for gaseous withdrawal. However, if your system is
predominantly used as a source of gaseous nitrogen, I would expect that
your overall efficiency should be closer to 90%.
How well are your flow meters calibrated? Do you have flow meters on
all possible usage? Are your flow meters fully temperature compensated
or are you measuring gas that is noticeably colder than room
temperature? Anecdotally, we just sent our two main flow meters out for
calibration: when they came back, they are now giving readings that
indicate that we are consuming more nitrogen than our vendor is
supplying. In short, I'm not sure that I trust the absolute readings
that I get off these big flow meters for an absolute reading. In short,
if your tank is used predominantly for gaseous nitrogen, I would expect
that your overall tank efficiency should be closer to 90% and would look
carefully at whether I was monitoring all usage and whether my usage
flow meters were giving me believable results. If your liquid
withdrawal and usage are significant everything that I have said may be
off base because I expect that liquid withdrawal is an inherently lower
efficiency process.
I trust that my colleagues will correct me if they believe that I've
badly misstated things here.
Good luck,
John
On 3/27/2015 11:08 AM, Morrison, Richard H., Jr. wrote:
> Thanks for your comments so far and I looked forward to more
> discussion. BTW after fees and surcharges I pay $1.23 per 100cuft and
> I use 1.45million cuft per month (data from flow meters). The company
> delivers the equivalent of 2.6million cuft so I lose to evaporation
> and cooling 1million cuft per month, does this sound right or do I
> have a big leak some place?