[labnetwork] Pegasus - Rapier Esc question
Beall, James A. (Fed)
james.beall at nist.gov
Tue Jun 14 14:15:42 EDT 2016
David,
Our 2009 Pegasus with 150 mm TDESC typically runs 40-100 mT/min leak up at 10 Torr, 10.5-12 sccm He.
Our warning is at 125 and alarm is at 150.
- Jim
NIST Boulder Microfabrication Facility
On Jun 13, 2016, at 12:03 PM, LaFleur, David W <dlafleur at cns.fas.harvard.edu<mailto:dlafleur at cns.fas.harvard.edu>> wrote:
To Engineers who maintain Pegasus – Rapier Systems with Esc
Where do you set your 15 Torr leak up rate pass/fail limit for the Esc. Does anyone have success in achieving the factory spec of 50mt/min on a 150mm chuck for 15 torr @15degC ?
What is your max flow you like to see for 15 Torr?
Thanks in advance.
David LaFleur
Equipment Engineer
Harvard University