From tbritton at criticalsystemsinc.com Sun Apr 2 11:06:27 2017 From: tbritton at criticalsystemsinc.com (Tom Britton) Date: Sun, 2 Apr 2017 15:06:27 +0000 Subject: [labnetwork] Handheld Hydrogen and Hydride (Phosphine, Germane, diborane, silane) detectors In-Reply-To: References: <9C4552F2D4B19C44970F1CB90360C99194260EF9@CSGMBX211W.pu.win.princeton.edu>, Message-ID: <9A7223F1-6C7C-4164-A9B5-7B9CB38A65F5@criticalsystemsinc.com> Hi Jim, I'm with Noah on this. Our guys have used the DOD instruments in the past with good results. Good luck, Tom Britton Critical Systems, Inc. Sent from my iPhone On Apr 1, 2017, at 6:41 AM, Noah Clay > wrote: Jim, This may be a good fit: http://www.dodtec.com/clpx/ Or http://www.dodtec.com/xp-703dhydrides/ Best, Noah Noah Clay Director, Quattrone Nanofabrication Facility University of Pennsylvania On Mar 31, 2017, at 15:59, James C. Sturm > wrote: I am looking for handheld detectors to ?sniff? at VCR or swagelock joints for leaks of hydrogen and also for common hydrides (silane, germane, diborane, phosphine?). The lab has a central system already, but I am looking for something to sniff at joints with after making connections and repressurizing. (Yes we will Helium leak-check too, but this is as a followup with less likelihood of someone doing it wrong). Many years (25?) we had a handheld unit for hydrides (about 7? x 3 ? x 1?, with a 4? long rubber sniffer port). Probably made by Matheson but can?t seem to find anything similar from them now (although I didn?t look very hard). Suggestions? Thanks, Jim Sturm _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: From schweig at umich.edu Mon Apr 3 10:44:19 2017 From: schweig at umich.edu (Dennis Schweiger) Date: Mon, 3 Apr 2017 10:44:19 -0400 Subject: [labnetwork] Handheld Hydrogen and Hydride (Phosphine, Germane, diborane, silane) detectors Message-ID: James, as everyone has previously mentioned, all of the DOD stuff works really well, as do the Matheson single point hand-held units. You don't say what detection system you're using for the rest of your fab, but I'd check with that vendor to see what they offer. I'd also suggest that you look at a MultiRAE unit (good for monitoring up to six different gas chemistries), and the Honeywell SPMFlex (extractive tapes required). Here at UofMichigan we have a variety of handheld detection systems. What you'll find is that there isn't a "silver bullet" to cover every single thing when you look at multi-use sensors. There will have to be some trade offs. Dennis Schweiger University of Michigan/LNF Facilities 734.647.2055 Ofc On Fri, Mar 31, 2017 at 3:59 PM, James C. Sturm wrote: > I am looking for handheld detectors to ?sniff? at VCR or swagelock joints > for leaks of hydrogen and also for common hydrides (silane, germane, > diborane, phosphine?). The lab has a central system already, but I am > looking for something to sniff at joints with after making connections and > repressurizing. (Yes we will Helium leak-check too, but this is as a > followup with less likelihood of someone doing it wrong). > > > > Many years (25?) we had a handheld unit for hydrides (about 7? x 3 ? x > 1?, with a 4? long rubber sniffer port). Probably made by Matheson but > can?t seem to find anything similar from them now (although I didn?t look > very hard). > > > > Suggestions? > > > > Thanks, Jim Sturm > > > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork > > -------------- next part -------------- An HTML attachment was scrubbed... URL: From beaudoin at physics.ubc.ca Mon Apr 3 12:52:55 2017 From: beaudoin at physics.ubc.ca (Mario Beaudoin) Date: Mon, 3 Apr 2017 09:52:55 -0700 Subject: [labnetwork] Handheld Hydrogen and Hydride (Phosphine, Germane, diborane, silane) detectors In-Reply-To: <9C4552F2D4B19C44970F1CB90360C99194260EF9@CSGMBX211W.pu.win.princeton.edu> References: <9C4552F2D4B19C44970F1CB90360C99194260EF9@CSGMBX211W.pu.win.princeton.edu> Message-ID: I used to have something like this made by Draeger. M. Le 31/03/2017 ? 12:59 PM, James C. Sturm a ?crit : > > I am looking for handheld detectors to ?sniff? at VCR or swagelock > joints for leaks of hydrogen and also for common hydrides (silane, > germane, diborane, phosphine?). The lab has a central system already, > but I am looking for something to sniff at joints with after making > connections and repressurizing. (Yes we will Helium leak-check > too, but this is as a followup with less likelihood of someone doing > it wrong). > > Many years (25?) we had a handheld unit for hydrides (about 7? x 3 ? > x 1?, with a 4? long rubber sniffer port). Probably made by Matheson > but can?t seem to find anything similar from them now (although I > didn?t look very hard). > > Suggestions? > > Thanks, Jim Sturm > > > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork -- *Mario Beaudoin, Ph.D.* Research Associate, Manager Advanced Nanofabrication Facility (ANF) Advanced Materials and Process Engineering Lab (AMPEL) University of British Columbia 2355 East Mall, Vancouver, BC V6T 1Z4 Tel: 604-822-1853 web: www.nanofab.ubc.ca Reply To: _beaudoin at physics.ubc.ca _ -------------- next part -------------- An HTML attachment was scrubbed... URL: From frank.gambino at dodtec.com Mon Apr 3 13:24:49 2017 From: frank.gambino at dodtec.com (Frank Gambino) Date: Mon, 3 Apr 2017 17:24:49 +0000 Subject: [labnetwork] Handheld Hydrogen and Hydride (Phosphine, Germane, diborane, silane) detectors In-Reply-To: <9C4552F2D4B19C44970F1CB90360C99194260EF9@CSGMBX211W.pu.win.princeton.edu> References: <9C4552F2D4B19C44970F1CB90360C99194260EF9@CSGMBX211W.pu.win.princeton.edu> Message-ID: <2D3ADBBA44F2E44889FEF73BBBF7AC140127687E79@SBS2011.DOD.local> Hi Jim, The unit you were using was the Matheson 5057. This was actually made by New Cosmos for Matheson. The New Cosmos model was XP-703D. DOD Technologies is partnered with New Cosmos. [cid:image003.png at 01D2AC75.23990E30] The XP-703D was replaced by a newer version called the XP-703D-III. The XP-703D-III like the XP-703D is a general leak detector and basically detects the flammable part for these gases. It is strictly a qualitative and not quantitative unit. It can read arsine down to 50ppb and H2 as low as 1ppm. http://www.dodtec.com/xp-703dhydrides/ Another portable option is the XPS-7 which uses interchangeable EC sensors can read arsine down to 12ppb. The XPS-7 provides quantitative data where the XP-703D-III does not. http://www.dodtec.com/xps7/ A 3rd portable option is the CLPX which uses colorimetric technology (tape) and can detect arsine down to 1.6ppb. http://www.dodtec.com/clpx/ Regards, Frank Gambino DOD Technologies, Inc. 815-788-5200 office 815-788-5300 fax 847-722-8274 mobile frank.gambino at dodtec.com www.dodtec.com From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of James C. Sturm Sent: Friday, March 31, 2017 3:00 PM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Handheld Hydrogen and Hydride (Phosphine, Germane, diborane, silane) detectors I am looking for handheld detectors to "sniff" at VCR or swagelock joints for leaks of hydrogen and also for common hydrides (silane, germane, diborane, phosphine...). The lab has a central system already, but I am looking for something to sniff at joints with after making connections and repressurizing. (Yes we will Helium leak-check too, but this is as a followup with less likelihood of someone doing it wrong). Many years (25?) we had a handheld unit for hydrides (about 7" x 3 " x 1", with a 4" long rubber sniffer port). Probably made by Matheson but can't seem to find anything similar from them now (although I didn't look very hard). Suggestions? Thanks, Jim Sturm -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image003.png Type: image/png Size: 143254 bytes Desc: image003.png URL: From nannini.matthieu at gmail.com Wed Apr 5 14:11:18 2017 From: nannini.matthieu at gmail.com (Matthieu Nannini) Date: Wed, 5 Apr 2017 14:11:18 -0400 Subject: [labnetwork] XRR mapping services Message-ID: Dear all, I'm looking for a XRR service lab with mapping capabilities. Any names ? Thanks, Matthieu -------------- next part -------------- An HTML attachment was scrubbed... URL: From Vincent.Luciani at nist.gov Wed Apr 5 21:03:24 2017 From: Vincent.Luciani at nist.gov (Luciani, Vincent (Fed)) Date: Thu, 6 Apr 2017 01:03:24 +0000 Subject: [labnetwork] XRR mapping services In-Reply-To: References: Message-ID: Hello Mathieu, We have the Rigaku SmartLab system. You can find details here: https://www.nist.gov/laboratories/tools-instruments/nanofab-tool-rigaku-smartlab-x-ray-diffraction [https://www.nist.gov/sites/default/files/styles/480_x_480_limit/public/images/cnst/nanofab/rigaku_smartlab_1.jpg?itok=MYGx7Ls0] NanoFab Tool: Rigaku SmartLab X-Ray Diffraction | NIST www.nist.gov The Rigaku SmartLab X-ray diffraction system (XRD) is a fully automated, modular system for advanced x-ray diffraction measurements on a wide range of materials ... I am also copying Dr. Kerry Siebein who is our XRD/XRR expert. Feel free to contact her directly with any questions. Best, Vince ________________________________ From: labnetwork-bounces at mtl.mit.edu on behalf of Matthieu Nannini Sent: Wednesday, April 5, 2017 2:11:18 PM To: labnetwork at mtl.mit.edu Subject: [labnetwork] XRR mapping services Dear all, I'm looking for a XRR service lab with mapping capabilities. Any names ? Thanks, Matthieu -------------- next part -------------- An HTML attachment was scrubbed... URL: From bshiva at stanford.edu Wed Apr 5 22:02:59 2017 From: bshiva at stanford.edu (Shivakumar Bhaskaran) Date: Thu, 6 Apr 2017 02:02:59 +0000 Subject: [labnetwork] Blisters on PMMA resists Message-ID: Hi All, We are getting blisters or metal film peeling off when depositing metals (Ti/Au/Al) on PMMA resists using EBeam Evaporation. Did anyone had this issue in your EBeam Evaporation tool. If so how did you fix this? -thanks -Shiva SNSF http://snsf.stanford.edu/ -------------- next part -------------- An HTML attachment was scrubbed... URL: From jsmith at tystar.com Wed Apr 5 17:43:19 2017 From: jsmith at tystar.com (Jim Smith) Date: Wed, 5 Apr 2017 14:43:19 -0700 Subject: [labnetwork] Field Service Opening Message-ID: <023701d2ae55$a44404e0$eccc0ea0$@tystar.com> Dear All, Tystar Corporation announces a job opening for a Field Service Engineer in our Diffusion Furnace Division. Please contact careers at tystar.com for further details on the position. Best Regards, Jim Smith Business Development Mgr. Tystar Corporation 7050 Lampson Ave. Garden Grove, Calif. 92841 * Office: 310-781-9219 ext 216 * Cell: 310-618-4018 * Fax: 310-781-9438 * jsmith at tystar.com http://www.tystar.com This Document is confidential and it is intended only for the addressee(s). Any review, dissemination, distribution, or copying of this message by persons or entities other than the intended recipient is strictly prohibited. -------------- next part -------------- An HTML attachment was scrubbed... URL: From khbeis at uw.edu Thu Apr 6 08:16:57 2017 From: khbeis at uw.edu (Michael Khbeis) Date: Thu, 6 Apr 2017 07:16:57 -0500 Subject: [labnetwork] Plea for Peer Review - Point of Use Abatement (pyrophoric) systems Message-ID: <15C3AC99-A75D-4E3F-85FC-8EDD5873CE4E@uw.edu> Colleagues, I continue to rely on and value your assistance in making the case or finding alternative resolution to issues imposed by unilateral decisions my campus construction folks are making and forcing upon us. Most recently, an 11th hour submittal of this EDOC local abatement unit (3-4 in total - see operating manual attachment) in lieu of running 5 - 1.5" forelines from the pumps to the existing CDO burn box. These units mix air into the effluent stream to facilitate pyrolysis and collect the dust until they clog at which time, my personnel are supposed to wear full PPE and respiratory protection and open up and scrape, vacuum, and even chisel the inside of the boxes clean. As per the attached manual, all the constituents would be considered hazardous. My issues are that the construction folks and this vendor haven't done a thorough hazard review. We are running diborane, phosphine, and MOCVD precursors in addition to silane. I am concerned about scraping phosphine residue as when I was a HAZMAT tech, I personally experienced spontaneous combustion when scraping phosphorus residue from an implanter (and simultaneously experienced the ill effects of when people don't review potential hazards before sending people into harms way). Furthermore, the MOCVD precursors are particularly nasty - for example I watched a used VCR gasket that was exposed to CUPRASELECT spontaneously start to fizz, corrode, and jump around on the base of the tool as soon as it was exposed to air. Now this proposal suggests that we purposely expose this material to air. The vendor doesn't even have a list of the liquid precursors we are using - because they didn't even request it to do a hazard review. Finally, even if I get my folks HAZWOPR certified to do abatement, have them fully geared, the units are open to my service chase and would potentially have these hazardous dusts in the recirculating air flow path. When I agreed to localized abatement - I meant something like an Edwards GRC that has a cartridge that captures the byproducts not something that hangs onto it and dumps it out on the floor once it clogs and we open it. Finally, the vendor claims it would require service about once every 6 months but that of course we would need to establish a statistical idea of how often they need service based on our use. Given that in the recent past, we clogged a foreline in <2 years by sharing with multiple tools (so silane and oxygen were doing this in the unions), I'm assuming that we will be generating a decent amount of sand and likely need to clean out more often. Even then, if I use the 6 month PM figure - that's 8 HAZMAT remediations a year. If we hire out a company like Clean Harbors in lieu of putting my staff in harms way - that's a serious additional expense we would have to adsorb. I certainly don't want to be close minded so if you have experience with these or similar units and can enlighten me on the realistic operating and mitigation conditions, I'd be happy to hear the feedback. Most gratefully, Dr. Michael Khbeis Washington Nanofab Facility University of Washington Fluke Hall, Box 352143 (O) 206.543.5101 (C) 443.254.5192 khbeis at uw.edu -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: EDOC-Altair Manual V2-SERVICE REV 7.doc Type: application/msword Size: 5472768 bytes Desc: not available URL: -------------- next part -------------- An HTML attachment was scrubbed... URL: From myoung6 at nd.edu Thu Apr 6 09:03:55 2017 From: myoung6 at nd.edu (Mike Young) Date: Thu, 6 Apr 2017 09:03:55 -0400 Subject: [labnetwork] Blisters on PMMA resists In-Reply-To: References: Message-ID: <82558C06-DE83-4FFB-86AD-B572CDA6F3C8@nd.edu> Hi Shiva. We had this problem a while back. Turned out to be poor thermal contact between the substrate and its mount in the evaporator - we had to tighten the mount springs. The PMMA is getting too hot! Good luck! --Mike > On Apr 5, 2017, at 10:02 PM, Shivakumar Bhaskaran wrote: > > Hi All, > > We are getting blisters or metal film peeling off when depositing metals (Ti/Au/Al) on PMMA resists using EBeam Evaporation. > > Did anyone had this issue in your EBeam Evaporation tool. If so how did you fix this? > > -thanks > -Shiva > > SNSF > http://snsf.stanford.edu/ > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork -- Michael P. Young (574) 631-3268 (office) Nanofabrication Specialist (574) 631-4393 (fax) Department of Electrical Engineering (765) 637-3784 (cell) University of Notre Dame mike.young at nd.edu B-38 Stinson-Remick Hall Notre Dame, IN 46556-5637 -------------- next part -------------- An HTML attachment was scrubbed... URL: From jguo5 at Central.UH.EDU Thu Apr 6 10:33:07 2017 From: jguo5 at Central.UH.EDU (Guo, Jing) Date: Thu, 6 Apr 2017 09:33:07 -0500 Subject: [labnetwork] CoFeO and BaTiO sputtering Message-ID: Good morning, Does anyone have experience with sputtering CoFeO and BaTiO? Is it possible to reactive sputter those materials? And any issues to worry about? Thanks. Best Regards, Jing Research Lab Supervisor Nanofabrication Facility University of Houston NANOFABRICATION FACILITY 3517 CULLEN BLVD, RM 1011 HOUSTON TX 77204-5062 Phone: 832-842-8822 Email: jguo5 at central.uh.edu Web: www.uh.edu/nanofab -------------- next part -------------- An HTML attachment was scrubbed... URL: From price.798 at osu.edu Thu Apr 6 09:31:44 2017 From: price.798 at osu.edu (Price, Aimee) Date: Thu, 6 Apr 2017 13:31:44 +0000 Subject: [labnetwork] Blisters on PMMA resists In-Reply-To: References: Message-ID: <83583687862B8444A85296A4944147DF95553943@CIO-TNC-D2MBX03.osuad.osu.edu> Hi Shiva, I believe we've had a thread on this subject in the last year or so. Can someone remind me how to find the old/archived threads (or do we just look in our email history)? Past that, we have had this several different times on a couple different evaporators, almost always with Ti/Au or Ni/Au ... Assuming we've baked the PMMA sufficiently it appears to be heating, though we've tried thermal dots alongside and not found very high temps. Our solution has been to slow the deposition rate and put a "wait" step in during thick Au, Ni, sometimes Pt steps. The rate we use is 1-2A/s in most cases and the wait times vary but are usually halfway through the Au step. Lastly, we have a lot of people working with piezoelectric materials and this happens sometimes with those. That's a separate topic. I'm assuming you are working with more "standard" semiconductor substrates and the issue is caused by the metal/PMMA not the substrate itself? Feel free to reach out to me off-line at price.798 at osu.edu if you want more details. We did a lot of work with this at two different times, about 3 years ago and 10 years ago. We even had students travel back to UlUC to use their evaporator when we were diagnosing this issue the first time. Best of luck. Aimee From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Shivakumar Bhaskaran Sent: Wednesday, April 05, 2017 10:03 PM To: LabNet (labnetwork at mtl.mit.edu) Subject: [labnetwork] Blisters on PMMA resists Hi All, We are getting blisters or metal film peeling off when depositing metals (Ti/Au/Al) on PMMA resists using EBeam Evaporation. Did anyone had this issue in your EBeam Evaporation tool. If so how did you fix this? -thanks -Shiva SNSF http://snsf.stanford.edu/ -------------- next part -------------- An HTML attachment was scrubbed... URL: From tbritton at criticalsystemsinc.com Thu Apr 6 12:11:51 2017 From: tbritton at criticalsystemsinc.com (Tom Britton) Date: Thu, 6 Apr 2017 16:11:51 +0000 Subject: [labnetwork] Plea for Peer Review - Point of Use Abatement (pyrophoric) systems In-Reply-To: <15C3AC99-A75D-4E3F-85FC-8EDD5873CE4E@uw.edu> References: <15C3AC99-A75D-4E3F-85FC-8EDD5873CE4E@uw.edu> Message-ID: Hi Michael, You should really look at a dry bed system for this. The Jupiter Callisto is used by many of your fellow Universities and doesn?t need to be cleaned out. You run to it until the canister tells you it is time for a change, and then you wheel it out and wheel in a new one. Veolia or Clean Harbors can take pick these up and incinerate them for you. No operator exposure, minimal utilities, no water or fuel. The canisters now have a DOT rating on them so you can ship them on a pallet without putting them in a hazardous waste container first. I can call you to discuss if you would like. Thanks sir, Tom Tom Britton Director of Sales Critical Systems, Inc. Direct: 208-890-1417 Office: 877-572-5515 www.CriticalSystemsInc.com [logo for email signature png] From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Michael Khbeis Sent: Thursday, April 06, 2017 6:17 AM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Plea for Peer Review - Point of Use Abatement (pyrophoric) systems Colleagues, I continue to rely on and value your assistance in making the case or finding alternative resolution to issues imposed by unilateral decisions my campus construction folks are making and forcing upon us. Most recently, an 11th hour submittal of this EDOC local abatement unit (3-4 in total - see operating manual attachment) in lieu of running 5 - 1.5" forelines from the pumps to the existing CDO burn box. These units mix air into the effluent stream to facilitate pyrolysis and collect the dust until they clog at which time, my personnel are supposed to wear full PPE and respiratory protection and open up and scrape, vacuum, and even chisel the inside of the boxes clean. As per the attached manual, all the constituents would be considered hazardous. My issues are that the construction folks and this vendor haven't done a thorough hazard review. We are running diborane, phosphine, and MOCVD precursors in addition to silane. I am concerned about scraping phosphine residue as when I was a HAZMAT tech, I personally experienced spontaneous combustion when scraping phosphorus residue from an implanter (and simultaneously experienced the ill effects of when people don't review potential hazards before sending people into harms way). Furthermore, the MOCVD precursors are particularly nasty - for example I watched a used VCR gasket that was exposed to CUPRASELECT spontaneously start to fizz, corrode, and jump around on the base of the tool as soon as it was exposed to air. Now this proposal suggests that we purposely expose this material to air. The vendor doesn't even have a list of the liquid precursors we are using - because they didn't even request it to do a hazard review. Finally, even if I get my folks HAZWOPR certified to do abatement, have them fully geared, the units are open to my service chase and would potentially have these hazardous dusts in the recirculating air flow path. When I agreed to localized abatement - I meant something like an Edwards GRC that has a cartridge that captures the byproducts not something that hangs onto it and dumps it out on the floor once it clogs and we open it. Finally, the vendor claims it would require service about once every 6 months but that of course we would need to establish a statistical idea of how often they need service based on our use. Given that in the recent past, we clogged a foreline in <2 years by sharing with multiple tools (so silane and oxygen were doing this in the unions), I'm assuming that we will be generating a decent amount of sand and likely need to clean out more often. Even then, if I use the 6 month PM figure - that's 8 HAZMAT remediations a year. If we hire out a company like Clean Harbors in lieu of putting my staff in harms way - that's a serious additional expense we would have to adsorb. I certainly don't want to be close minded so if you have experience with these or similar units and can enlighten me on the realistic operating and mitigation conditions, I'd be happy to hear the feedback. Most gratefully, Dr. Michael Khbeis Washington Nanofab Facility University of Washington Fluke Hall, Box 352143 (O) 206.543.5101 (C) 443.254.5192 khbeis at uw.edu -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.png Type: image/png Size: 9501 bytes Desc: image001.png URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: Callisto Literature Q316.pdf Type: application/pdf Size: 728177 bytes Desc: Callisto Literature Q316.pdf URL: From boning at mtl.mit.edu Thu Apr 6 13:29:18 2017 From: boning at mtl.mit.edu (Duane Boning) Date: Thu, 6 Apr 2017 13:29:18 -0400 Subject: [labnetwork] Archive and search on old labnetwork posts In-Reply-To: <83583687862B8444A85296A4944147DF95553943@CIO-TNC-D2MBX03.osuad.osu.edu> References: <83583687862B8444A85296A4944147DF95553943@CIO-TNC-D2MBX03.osuad.osu.edu> Message-ID: <0ae3b3a3-8573-a467-0b2a-3021075fa0ba@mtl.mit.edu> Hi All, In response to Aimee's question below, the labnetwork archives can be found at: https://www-mtl.mit.edu/pipermail/labnetwork/ They are not easily searchable in the standard Mailman format at that link, but we've also enabled them to be indexed by existing search engines. I've found Google searches of the form "[labnetwork] whatever my topic is" usually works to find previous labnetwork posts. Your moderator, /Duane ------- Duane S. Boning Clarence J. LeBel Professor, Electrical Engineering and Computer Science Associate Director, MTL Microsystems Technology Laboratories MIT, Room 39-415, Cambridge, MA 02139 Phone: 617-253-0931, boning at mtl.mit.edu On 4/6/2017 9:31 AM, Price, Aimee wrote: > > Hi Shiva, > > I believe we?ve had a thread on this subject in the last year or so. > Can someone remind me how to find the old/archived threads (or do we > just look in our email history)? > [...] -------------- next part -------------- An HTML attachment was scrubbed... URL: From mark.chiappa at ntnu.no Thu Apr 6 13:37:19 2017 From: mark.chiappa at ntnu.no (Mark Giulio Chiappa) Date: Thu, 6 Apr 2017 17:37:19 +0000 Subject: [labnetwork] Blisters on PMMA resists In-Reply-To: <83583687862B8444A85296A4944147DF95553943@CIO-TNC-D2MBX03.osuad.osu.edu> References: , <83583687862B8444A85296A4944147DF95553943@CIO-TNC-D2MBX03.osuad.osu.edu> Message-ID: Hi Shiva, I agree it's probably a thermal issue. When we had the same issue we experimented with PMMA from different batches, baking times and temps as well as thermally conductive mounts (mung etc) to no avail. Interestingly we solved the issue by increasing the deposition rate. So it seems there is no definite answer I suppose it is dependant in the distance between source an substrate. I don't remember how much we experimented with wait times but that sounds like a good idea. Kind regards Mark. Sent from my iPhone On 6 Apr 2017, at 18:23, Price, Aimee > wrote: Hi Shiva, I believe we've had a thread on this subject in the last year or so. Can someone remind me how to find the old/archived threads (or do we just look in our email history)? Past that, we have had this several different times on a couple different evaporators, almost always with Ti/Au or Ni/Au ... Assuming we've baked the PMMA sufficiently it appears to be heating, though we've tried thermal dots alongside and not found very high temps. Our solution has been to slow the deposition rate and put a "wait" step in during thick Au, Ni, sometimes Pt steps. The rate we use is 1-2A/s in most cases and the wait times vary but are usually halfway through the Au step. Lastly, we have a lot of people working with piezoelectric materials and this happens sometimes with those. That's a separate topic. I'm assuming you are working with more "standard" semiconductor substrates and the issue is caused by the metal/PMMA not the substrate itself? Feel free to reach out to me off-line at price.798 at osu.edu if you want more details. We did a lot of work with this at two different times, about 3 years ago and 10 years ago. We even had students travel back to UlUC to use their evaporator when we were diagnosing this issue the first time. Best of luck. Aimee From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Shivakumar Bhaskaran Sent: Wednesday, April 05, 2017 10:03 PM To: LabNet (labnetwork at mtl.mit.edu) Subject: [labnetwork] Blisters on PMMA resists Hi All, We are getting blisters or metal film peeling off when depositing metals (Ti/Au/Al) on PMMA resists using EBeam Evaporation. Did anyone had this issue in your EBeam Evaporation tool. If so how did you fix this? -thanks -Shiva SNSF http://snsf.stanford.edu/ _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: From bshiva at stanford.edu Thu Apr 6 13:51:48 2017 From: bshiva at stanford.edu (Shivakumar Bhaskaran) Date: Thu, 6 Apr 2017 17:51:48 +0000 Subject: [labnetwork] Blisters on PMMA resists In-Reply-To: References: , <83583687862B8444A85296A4944147DF95553943@CIO-TNC-D2MBX03.osuad.osu.edu> Message-ID: Yes, higher deposition rate helped in reducing the blisters or completely gone. However, some users want to do at a low deposition rate. Sometimes we start at low rate and change to a higher rate. -thanks -Shiva From: Mark Giulio Chiappa [mailto:mark.chiappa at ntnu.no] Sent: Thursday, April 6, 2017 10:37 AM To: Price, Aimee Cc: Shivakumar Bhaskaran ; LabNet (labnetwork at mtl.mit.edu) Subject: Re: [labnetwork] Blisters on PMMA resists Hi Shiva, I agree it's probably a thermal issue. When we had the same issue we experimented with PMMA from different batches, baking times and temps as well as thermally conductive mounts (mung etc) to no avail. Interestingly we solved the issue by increasing the deposition rate. So it seems there is no definite answer I suppose it is dependant in the distance between source an substrate. I don't remember how much we experimented with wait times but that sounds like a good idea. Kind regards Mark. Sent from my iPhone On 6 Apr 2017, at 18:23, Price, Aimee > wrote: Hi Shiva, I believe we've had a thread on this subject in the last year or so. Can someone remind me how to find the old/archived threads (or do we just look in our email history)? Past that, we have had this several different times on a couple different evaporators, almost always with Ti/Au or Ni/Au ... Assuming we've baked the PMMA sufficiently it appears to be heating, though we've tried thermal dots alongside and not found very high temps. Our solution has been to slow the deposition rate and put a "wait" step in during thick Au, Ni, sometimes Pt steps. The rate we use is 1-2A/s in most cases and the wait times vary but are usually halfway through the Au step. Lastly, we have a lot of people working with piezoelectric materials and this happens sometimes with those. That's a separate topic. I'm assuming you are working with more "standard" semiconductor substrates and the issue is caused by the metal/PMMA not the substrate itself? Feel free to reach out to me off-line at price.798 at osu.edu if you want more details. We did a lot of work with this at two different times, about 3 years ago and 10 years ago. We even had students travel back to UlUC to use their evaporator when we were diagnosing this issue the first time. Best of luck. Aimee From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Shivakumar Bhaskaran Sent: Wednesday, April 05, 2017 10:03 PM To: LabNet (labnetwork at mtl.mit.edu) Subject: [labnetwork] Blisters on PMMA resists Hi All, We are getting blisters or metal film peeling off when depositing metals (Ti/Au/Al) on PMMA resists using EBeam Evaporation. Did anyone had this issue in your EBeam Evaporation tool. If so how did you fix this? -thanks -Shiva SNSF http://snsf.stanford.edu/ _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: From eabelev at pitt.edu Thu Apr 6 21:28:53 2017 From: eabelev at pitt.edu (Abelev, Esta) Date: Fri, 7 Apr 2017 01:28:53 +0000 Subject: [labnetwork] sound absorbing wall panels Message-ID: Dear All, We are looking for sound absorbing open cell foam wall panels for our JEOL 2100F STEM room. Any recommendations for vendor and type. Thanks, Esta ----------------------- Esta Abelev, PhD Technical Director, Petersen Institute of NanoScience and Engineering University of Pittsburgh | 3700 O'Hara Street | 636| Pittsburgh, PA 15261 412-383-4096 | eabelev at pitt.edu | nano.pitt.edu -------------- next part -------------- An HTML attachment was scrubbed... URL: From Vincent.Luciani at nist.gov Fri Apr 7 09:33:26 2017 From: Vincent.Luciani at nist.gov (Luciani, Vincent (Fed)) Date: Fri, 7 Apr 2017 13:33:26 +0000 Subject: [labnetwork] sound absorbing wall panels In-Reply-To: References: Message-ID: Hello Esta, I am by no means an expert on this topic but I have one lesson learned I will share with you. We have used a type of open cell foam on our walls in our imaging labs and, by in large, it does what it is supposed to do with regard to sound absorption. My only complaint is that it dries out and sheds particles. Places like around doors, door handles, light switches or anywhere there may be friction or any place that one may come in contact with it will crack and shed. The picture attached is an example of the foam where it is constantly touched. If I were to do it again, I am sure this problem can be engineered away somehow. Best, Vince Vincent K. Luciani NanoFab Manager Center for Nanoscale Science and Technology National Institute of Standards and Technology 100 Bureau Drive, MS 6201 Gaithersburg, MD 20899-6200 USA +1-301-975-2886 From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Abelev, Esta Sent: Thursday, April 06, 2017 9:29 PM To: labnetwork at mtl.mit.edu Subject: [labnetwork] sound absorbing wall panels Dear All, We are looking for sound absorbing open cell foam wall panels for our JEOL 2100F STEM room. Any recommendations for vendor and type. Thanks, Esta ----------------------- Esta Abelev, PhD Technical Director, Petersen Institute of NanoScience and Engineering University of Pittsburgh | 3700 O'Hara Street | 636| Pittsburgh, PA 15261 412-383-4096 | eabelev at pitt.edu | nano.pitt.edu -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: 20170407_090458.jpg Type: image/jpeg Size: 3478261 bytes Desc: 20170407_090458.jpg URL: From spaolini at cns.fas.harvard.edu Fri Apr 7 17:15:17 2017 From: spaolini at cns.fas.harvard.edu (Paolini, Steven) Date: Fri, 7 Apr 2017 21:15:17 +0000 Subject: [labnetwork] sound absorbing wall panels In-Reply-To: References: Message-ID: I'm sorry to not be able to offer any particular product but I would be very concerned with the product's ability to resist flame spread as defined by the NFPA. At the risk of bringing up a gruesome event, The Station night club tragedy in Connecticut some years ago was exacerbated by the fact that the acoustical foam attached to the walls in the stage area was quite flammable. I would venture to say that Federal, State, and local ordinances have more than likely been modified to include a restriction on any acoustic foam that is flammable or does not resist flame spread. Thanks for listening. Steve Paolini Principal Equipment Engineer Harvard University Center for Nanoscale Systems 11 Oxford St. Cambridge, MA 02138 617- 496- 9816 spaolini at cns.fas.harvard.edu www.cns.fas.harvard.edu From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Abelev, Esta Sent: Thursday, April 06, 2017 9:29 PM To: labnetwork at mtl.mit.edu Subject: [labnetwork] sound absorbing wall panels Dear All, We are looking for sound absorbing open cell foam wall panels for our JEOL 2100F STEM room. Any recommendations for vendor and type. Thanks, Esta ----------------------- Esta Abelev, PhD Technical Director, Petersen Institute of NanoScience and Engineering University of Pittsburgh | 3700 O'Hara Street | 636| Pittsburgh, PA 15261 412-383-4096 | eabelev at pitt.edu | nano.pitt.edu -------------- next part -------------- An HTML attachment was scrubbed... URL: From steffen.8 at osu.edu Mon Apr 10 10:53:58 2017 From: steffen.8 at osu.edu (Steffen, Paul) Date: Mon, 10 Apr 2017 14:53:58 +0000 Subject: [labnetwork] sound absorbing wall panels In-Reply-To: References: Message-ID: Esta, When I was working for a company that made automated systems to load and unload packaged chips from burn-in boards, we had a noise issue from vibration resonances in the burn-in board elevator and used sound-absorbing foam from a company called EAR. It looks like they are a part of 3M now, https://earglobal.com/en/. I recall that they offered pretty decent technical support. -Paul [The Ohio State University] Paul Steffen Lab Manager Institute for Materials Research Nanotech West Lab 100 Science Village, 1381 Kinnear Road, Columbus, OH 43212 614-688-3546 Office / 614-688-3379 Fax steffen.8 at osu.edu nanotech.osu.edu From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Paolini, Steven Sent: Friday, April 07, 2017 5:15 PM To: Abelev, Esta ; labnetwork at mtl.mit.edu Subject: Re: [labnetwork] sound absorbing wall panels I'm sorry to not be able to offer any particular product but I would be very concerned with the product's ability to resist flame spread as defined by the NFPA. At the risk of bringing up a gruesome event, The Station night club tragedy in Connecticut some years ago was exacerbated by the fact that the acoustical foam attached to the walls in the stage area was quite flammable. I would venture to say that Federal, State, and local ordinances have more than likely been modified to include a restriction on any acoustic foam that is flammable or does not resist flame spread. Thanks for listening. Steve Paolini Principal Equipment Engineer Harvard University Center for Nanoscale Systems 11 Oxford St. Cambridge, MA 02138 617- 496- 9816 spaolini at cns.fas.harvard.edu www.cns.fas.harvard.edu From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Abelev, Esta Sent: Thursday, April 06, 2017 9:29 PM To: labnetwork at mtl.mit.edu Subject: [labnetwork] sound absorbing wall panels Dear All, We are looking for sound absorbing open cell foam wall panels for our JEOL 2100F STEM room. Any recommendations for vendor and type. Thanks, Esta ----------------------- Esta Abelev, PhD Technical Director, Petersen Institute of NanoScience and Engineering University of Pittsburgh | 3700 O'Hara Street | 636| Pittsburgh, PA 15261 412-383-4096 | eabelev at pitt.edu | nano.pitt.edu -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.png Type: image/png Size: 3605 bytes Desc: image001.png URL: From ACarlin at sheaumann.com Tue Apr 11 09:18:45 2017 From: ACarlin at sheaumann.com (Andrew Carlin) Date: Tue, 11 Apr 2017 09:18:45 -0400 Subject: [labnetwork] Benchmark SM8000 lid sealer Message-ID: Hello, We have an old benchmark SM8000 lid sealer, which was still in use until the computer hard drive failed a few weeks ago. The hard drive has been replaced, however we do NOT have a backup of the software. We have reached out to the company, but they don't have anyone with knowledge of these units. Does anyone have access to one of these units and would be willing to make a copy of the software? Regards, Andrew Andrew Carlin Senior Semiconductor Laser Designer Sheaumann Laser, Inc. 45 Bartlett Street Marlborough, MA 01752 USA Tel: 508-970-0600 ext. 267 Web: www.sheaumann.com e-mail: acarlin at sheaumann.com -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.jpg Type: image/jpeg Size: 2774 bytes Desc: image001.jpg URL: From stieg at cnsi.ucla.edu Tue Apr 11 11:48:04 2017 From: stieg at cnsi.ucla.edu (Stieg, Adam) Date: Tue, 11 Apr 2017 15:48:04 +0000 Subject: [labnetwork] Perspectives on equipment management software Message-ID: <86E466A1-90C6-408A-92EE-B84E097C432F@cnsi.ucla.edu> Hi All, We are in the process of reviewing our approach to managing reservations, access, billing, etc. for equipment throughout the CNSI at UCLA. To date, we have relied on a system developed in-house that is no longer viable. There is a significant push amongst the shared resources on campus toward iLabs. We have also begun looking into FOM (fomnetworks.com) as an alternative option. I know this may not be the best venue for a dialogue on the strengths/weaknesses of these specific solutions (or any other for that matter), but if any of you are willing to share your perspectives or experiences working with either of these vendors it would be greatly appreciated. All the best, Adam Adam Z. Stieg Ph.D. Associate Director, California NanoSystems Institute at UCLA Director, Nano and Pico Characterization Laboratory Technical Director, Integrated Systems Nanofabrication Cleanroom (p) 310.206.2902 (e) stieg at cnsi.ucla.edu -------------- next part -------------- An HTML attachment was scrubbed... URL: From Vincent.Luciani at nist.gov Tue Apr 11 13:26:38 2017 From: Vincent.Luciani at nist.gov (Luciani, Vincent (Fed)) Date: Tue, 11 Apr 2017 17:26:38 +0000 Subject: [labnetwork] Cost of cleanroom air balancing Message-ID: Hello All, It is 85?F here in Gaithersburg, MD today and it has thinking about all the spring-time chores. One of them I hope to get done is a complete air-rebalancing of our cleanroom. It is about 750 M2 (8000 Ft.2) of class 100 plus all chase space. I know this comes in many flavors such as with or without 100% individual HEPA filter checks, certifications( if required) etc. I am performing some market research in preparation for a solicitation. Has anyone done this recently or have a relevant data point to add to my search? Thanks, Vince Vincent K. Luciani NanoFab Manager Center for Nanoscale Science and Technology National Institute of Standards and Technology 100 Bureau Drive, MS 6201 Gaithersburg, MD 20899-6200 USA +1-301-975-2886 -------------- next part -------------- An HTML attachment was scrubbed... URL: From michael.rooks at yale.edu Tue Apr 11 16:54:58 2017 From: michael.rooks at yale.edu (Michael Rooks) Date: Tue, 11 Apr 2017 16:54:58 -0400 Subject: [labnetwork] Perspectives on equipment management software In-Reply-To: <86E466A1-90C6-408A-92EE-B84E097C432F@cnsi.ucla.edu> References: <86E466A1-90C6-408A-92EE-B84E097C432F@cnsi.ucla.edu> Message-ID: <58ED42A2.8080209@yale.edu> You could also consider Badger, which is the commercial version of Coral. See badgerlms.com -------------------------------- Michael Rooks Yale Institute of Nanoscience and Quantum Engineering nano.yale.edu On 04/11/2017 11:48 AM, Stieg, Adam wrote: > > Hi All, > > We are in the process of reviewing our approach to managing > reservations, access, billing, etc. for equipment throughout the CNSI > at UCLA. To date, we have relied on a system developed in-house that > is no longer viable. There is a significant push amongst the shared > resources on campus toward iLabs. We have also begun looking into FOM > (fomnetworks.com) as an alternative option. I know this may not be the > best venue for a dialogue on the strengths/weaknesses of these > specific solutions (or any other for that matter), but if any of you > are willing to share your perspectives or experiences working with > either of these vendors it would be greatly appreciated. > > All the best, > > Adam > > __ > > Adam Z. Stieg Ph.D. > > Associate Director, California NanoSystems Institute at UCLA > > Director, Nano and Pico Characterization Laboratory > > Technical Director, Integrated Systems Nanofabrication Cleanroom > > (p) 310.206.2902 > > (e) stieg at cnsi.ucla.edu > > > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: From julia.aebersold at louisville.edu Tue Apr 11 17:15:34 2017 From: julia.aebersold at louisville.edu (julia.aebersold at louisville.edu) Date: Tue, 11 Apr 2017 21:15:34 +0000 Subject: [labnetwork] Perspectives on equipment management software In-Reply-To: <86E466A1-90C6-408A-92EE-B84E097C432F@cnsi.ucla.edu> References: <86E466A1-90C6-408A-92EE-B84E097C432F@cnsi.ucla.edu> Message-ID: We use FOM at our facility and it is an economical package that serves our cleanroom well. Cheers! Julia Aebersold Manager, Micro/Nano Technology Center University of Louisville 2210 South Brook Street Louisville, KY 40292 (502)852-1572 http://louisville.edu/micronano/ From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Stieg, Adam Sent: Tuesday, April 11, 2017 11:48 AM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Perspectives on equipment management software Hi All, We are in the process of reviewing our approach to managing reservations, access, billing, etc. for equipment throughout the CNSI at UCLA. To date, we have relied on a system developed in-house that is no longer viable. There is a significant push amongst the shared resources on campus toward iLabs. We have also begun looking into FOM (fomnetworks.com) as an alternative option. I know this may not be the best venue for a dialogue on the strengths/weaknesses of these specific solutions (or any other for that matter), but if any of you are willing to share your perspectives or experiences working with either of these vendors it would be greatly appreciated. All the best, Adam Adam Z. Stieg Ph.D. Associate Director, California NanoSystems Institute at UCLA Director, Nano and Pico Characterization Laboratory Technical Director, Integrated Systems Nanofabrication Cleanroom (p) 310.206.2902 (e) stieg at cnsi.ucla.edu -------------- next part -------------- An HTML attachment was scrubbed... URL: From ahad.syed at kaust.edu.sa Wed Apr 12 03:02:24 2017 From: ahad.syed at kaust.edu.sa (Ahad Syed) Date: Wed, 12 Apr 2017 07:02:24 +0000 Subject: [labnetwork] Perspectives on equipment management software In-Reply-To: <58ED42A2.8080209@yale.edu> References: <86E466A1-90C6-408A-92EE-B84E097C432F@cnsi.ucla.edu>, <58ED42A2.8080209@yale.edu> Message-ID: <1491980543604.19601@kaust.edu.sa> We use Badger at KAUST for the past 4 years and I must say it was one of our better investments. Brought down abuse considerably and enables us to pull reports almost instantly. We started using it at the Core Labs and the success has been such that the University is planning to implement it to all the labs. All the best. Ahad -- Nanofabrication Core Lab 4700 King Abdullah University of Science and Technology 0250-WS02 Building 3, Thuwal 23955-6900, Kingdom of Saudi Arabia Tel:+966 12 808 2348 ________________________________ From: labnetwork-bounces at mtl.mit.edu on behalf of Michael Rooks Sent: Tuesday, April 11, 2017 11:54 PM To: labnetwork at mtl.mit.edu Subject: Re: [labnetwork] Perspectives on equipment management software You could also consider Badger, which is the commercial version of Coral. See badgerlms.com -------------------------------- Michael Rooks Yale Institute of Nanoscience and Quantum Engineering nano.yale.edu On 04/11/2017 11:48 AM, Stieg, Adam wrote: Hi All, We are in the process of reviewing our approach to managing reservations, access, billing, etc. for equipment throughout the CNSI at UCLA. To date, we have relied on a system developed in-house that is no longer viable. There is a significant push amongst the shared resources on campus toward iLabs. We have also begun looking into FOM (fomnetworks.com) as an alternative option. I know this may not be the best venue for a dialogue on the strengths/weaknesses of these specific solutions (or any other for that matter), but if any of you are willing to share your perspectives or experiences working with either of these vendors it would be greatly appreciated. All the best, Adam Adam Z. Stieg Ph.D. Associate Director, California NanoSystems Institute at UCLA Director, Nano and Pico Characterization Laboratory Technical Director, Integrated Systems Nanofabrication Cleanroom (p) 310.206.2902 (e) stieg at cnsi.ucla.edu _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork ________________________________ This message and its contents including attachments are intended solely for the original recipient. If you are not the intended recipient or have received this message in error, please notify me immediately and delete this message from your computer system. Any unauthorized use or distribution is prohibited. Please consider the environment before printing this email. -------------- next part -------------- An HTML attachment was scrubbed... URL: From aju-jugessur at uiowa.edu Wed Apr 12 07:36:24 2017 From: aju-jugessur at uiowa.edu (Jugessur, Aju S) Date: Wed, 12 Apr 2017 11:36:24 +0000 Subject: [labnetwork] Perspectives on equipment management software In-Reply-To: <86E466A1-90C6-408A-92EE-B84E097C432F@cnsi.ucla.edu> References: <86E466A1-90C6-408A-92EE-B84E097C432F@cnsi.ucla.edu> Message-ID: Adam We have just started using Bookitlab. We have it customized for our needs. There are some glitches but I do think it will be resolved. I went with bookitlab mainly because they are relatively inexpensive. Let me know if you need more detailed feedback. Thanks Aju Jugessur Ph.D. Director, UI Microfabrication Facility Sent from my iPhone On Apr 11, 2017, at 3:55 PM, Stieg, Adam > wrote: Hi All, We are in the process of reviewing our approach to managing reservations, access, billing, etc. for equipment throughout the CNSI at UCLA. To date, we have relied on a system developed in-house that is no longer viable. There is a significant push amongst the shared resources on campus toward iLabs. We have also begun looking into FOM (fomnetworks.com) as an alternative option. I know this may not be the best venue for a dialogue on the strengths/weaknesses of these specific solutions (or any other for that matter), but if any of you are willing to share your perspectives or experiences working with either of these vendors it would be greatly appreciated. All the best, Adam Adam Z. Stieg Ph.D. Associate Director, California NanoSystems Institute at UCLA Director, Nano and Pico Characterization Laboratory Technical Director, Integrated Systems Nanofabrication Cleanroom (p) 310.206.2902 (e) stieg at cnsi.ucla.edu _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: From sandrine at umich.edu Wed Apr 12 08:04:03 2017 From: sandrine at umich.edu (Sandrine Martin) Date: Wed, 12 Apr 2017 12:04:03 +0000 Subject: [labnetwork] Perspectives on equipment management software In-Reply-To: <86E466A1-90C6-408A-92EE-B84E097C432F@cnsi.ucla.edu> References: <86E466A1-90C6-408A-92EE-B84E097C432F@cnsi.ucla.edu> Message-ID: Hello Adam, We are in a very similar situation at the University of Michigan. Following some discussions at last year's UGIM, we have also looked at some of these (iLab is already used here at the medical school, FOM is implemented in the electron microscopy/characterization facility). We have also looked at Stratocore PPMS. Several universities are also pushing for compliance issues to be considered. We have not yet made our decision but here are some of the key points that we are considering: 1- Scheduler and equipment interlocks 2- Administration/configuration - role management (including tool-specific) - web/mobile compatibility - cost model configuration (room + tool charges, consumables charged with certain tools) - support of processing services 3- Documentation (tool manuals, safety/lab training) 4- Communication - to users/staff/PIs/admins (opt in, opt out, preferences) - tool status and ticket system for request tracking - direct communication between users (notification of tool down after hours for instance) 5- Budget and financial tracking - placing external orders by staff (with electronic management approval) - expenses assigned to specific tools / lab budget categories - labor hours assigned to specific tools / lab budget categories - revenue tracking per tool 6- Reporting - tool utilization - revenue, expenses by category/user groups/tools - for staff, admin, users, PIs 7- Support - ongoing operation support - long term reliability (code development/ database structure for instance) 9- External add-ons / interfaces with facility-specific applications Good luck!! Sandrine On Tue, Apr 11, 2017 at 16:48 Stieg, Adam wrote: Hi All, We are in the process of reviewing our approach to managing reservations, access, billing, etc. for equipment throughout the CNSI at UCLA. To date, we have relied on a system developed in-house that is no longer viable. There is a significant push amongst the shared resources on campus toward iLabs. We have also begun looking into FOM (fomnetworks.com) as an alternative option. I know this may not be the best venue for a dialogue on the strengths/weaknesses of these specific solutions (or any other for that matter), but if any of you are willing to share your perspectives or experiences working with either of these vendors it would be greatly appreciated. All the best, Adam Adam Z. Stieg Ph.D. Associate Director, California NanoSystems Institute at UCLA Director, Nano and Pico Characterization Laboratory Technical Director, Integrated Systems Nanofabrication Cleanroom (p) 310.206.2902 (e) stieg at cnsi.ucla.edu _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork -- Sandrine Martin, Ph.D. University of Michigan LNF Managing Director 1246D EECS, 1301 Beal Ave Ann Arbor, MI 48109 Cell 734-277-2365 Fax 734-647-1781 www.LNF.umich.edu @LurieNanofab -------------- next part -------------- An HTML attachment was scrubbed... URL: From na2661 at columbia.edu Wed Apr 12 09:07:06 2017 From: na2661 at columbia.edu (Nava Ariel-Sternberg) Date: Wed, 12 Apr 2017 09:07:06 -0400 Subject: [labnetwork] Perspectives on equipment management software In-Reply-To: <58ED42A2.8080209@yale.edu> References: <86E466A1-90C6-408A-92EE-B84E097C432F@cnsi.ucla.edu> <58ED42A2.8080209@yale.edu> Message-ID: <028a01d2b38d$b01a9b20$104fd160$@columbia.edu> We also use Badger and are happy with it. -Nava Nava Ariel-Sternberg, Ph.D. Director of CNI Shared Facilities Columbia University 530 w120th st., NY 10027 Room 1015/MC 8903 Office: 212-854-9927 Cell: 201-562-7600 ? From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Michael Rooks Sent: Tuesday, April 11, 2017 4:55 PM To: labnetwork at mtl.mit.edu Subject: Re: [labnetwork] Perspectives on equipment management software You could also consider Badger, which is the commercial version of Coral. See badgerlms.com -------------------------------- Michael Rooks Yale Institute of Nanoscience and Quantum Engineering nano.yale.edu On 04/11/2017 11:48 AM, Stieg, Adam wrote: Hi All, We are in the process of reviewing our approach to managing reservations, access, billing, etc. for equipment throughout the CNSI at UCLA. To date, we have relied on a system developed in-house that is no longer viable. There is a significant push amongst the shared resources on campus toward iLabs. We have also begun looking into FOM (fomnetworks.com) as an alternative option. I know this may not be the best venue for a dialogue on the strengths/weaknesses of these specific solutions (or any other for that matter), but if any of you are willing to share your perspectives or experiences working with either of these vendors it would be greatly appreciated. All the best, Adam Adam Z. Stieg Ph.D. Associate Director, California NanoSystems Institute at UCLA Director, Nano and Pico Characterization Laboratory Technical Director, Integrated Systems Nanofabrication Cleanroom (p) 310.206.2902 (e) stieg at cnsi.ucla.edu _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: From stieg at cnsi.ucla.edu Wed Apr 12 13:07:26 2017 From: stieg at cnsi.ucla.edu (Stieg, Adam) Date: Wed, 12 Apr 2017 17:07:26 +0000 Subject: [labnetwork] Perspectives on equipment management software In-Reply-To: <028a01d2b38d$b01a9b20$104fd160$@columbia.edu> References: <86E466A1-90C6-408A-92EE-B84E097C432F@cnsi.ucla.edu> <58ED42A2.8080209@yale.edu> <028a01d2b38d$b01a9b20$104fd160$@columbia.edu> Message-ID: Thanks to everyone for the robust response to my inquiry. As expected, the scope of variables that need be addressed is only matched by the breadth of pespective on the best approach to do so. Part of the challenge we face, which I failed to mention in my initial message, stems from the fact that we house 6 different shared faciltiis at CNSI that range from optical/electron/scanning probe microscopy to cleanroom operations and high-throughput screening. We currently use a single LMS solution for all these labs as there is extensive overlap in the user base of each, with a total of over 5000 active accounts and 1000 unique users per year. Nevertheless, I greatly aprpeciate all the information provided by this channel and will be reaching out to those of you who have provided specific feedback of interest. Kind regards, Adam On 04/11/2017 11:48 AM, Stieg, Adam wrote: Hi All, We are in the process of reviewing our approach to managing reservations, access, billing, etc. for equipment throughout the CNSI at UCLA. To date, we have relied on a system developed in-house that is no longer viable. There is a significant push amongst the shared resources on campus toward iLabs. We have also begun looking into FOM (fomnetworks.com) as an alternative option. I know this may not be the best venue for a dialogue on the strengths/weaknesses of these specific solutions (or any other for that matter), but if any of you are willing to share your perspectives or experiences working with either of these vendors it would be greatly appreciated. All the best, Adam Adam Z. Stieg Ph.D. Associate Director, California NanoSystems Institute at UCLA Director, Nano and Pico Characterization Laboratory Technical Director, Integrated Systems Nanofabrication Cleanroom (p) 310.206.2902 (e) stieg at cnsi.ucla.edu _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: From rmorrison at draper.com Thu Apr 13 10:44:52 2017 From: rmorrison at draper.com (Morrison, Richard H., Jr.) Date: Thu, 13 Apr 2017 14:44:52 +0000 Subject: [labnetwork] Cost of cleanroom air balancing In-Reply-To: References: Message-ID: <321224d4cf7d4b2f92c7d1912f6c6f80@draper.com> HI, I had it done last year a complete rebalance of the Draper Cleanroom complex, about 15000 sqft spread over 3 floors involing fans on 2 roof tops it was around $29K. It took an entire week to get done. Balance AHU, 8 exhaust fans, makeup dapers and equipment dampers. Equipment was about 75 tools. Rick ________________________________ From: labnetwork-bounces at mtl.mit.edu on behalf of Luciani, Vincent (Fed) Sent: Tuesday, April 11, 2017 1:26:38 PM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Cost of cleanroom air balancing Hello All, It is 85?F here in Gaithersburg, MD today and it has thinking about all the spring-time chores. One of them I hope to get done is a complete air-rebalancing of our cleanroom. It is about 750 M2 (8000 Ft.2) of class 100 plus all chase space. I know this comes in many flavors such as with or without 100% individual HEPA filter checks, certifications( if required) etc. I am performing some market research in preparation for a solicitation. Has anyone done this recently or have a relevant data point to add to my search? Thanks, Vince Vincent K. Luciani NanoFab Manager Center for Nanoscale Science and Technology National Institute of Standards and Technology 100 Bureau Drive, MS 6201 Gaithersburg, MD 20899-6200 USA +1-301-975-2886 ________________________________ Notice: This email and any attachments may contain proprietary (Draper non-public) and/or export-controlled information of Draper. If you are not the intended recipient of this email, please immediately notify the sender by replying to this email and immediately destroy all copies of this email. ________________________________ -------------- next part -------------- An HTML attachment was scrubbed... URL: From Thomas_Ferraguto at uml.edu Thu Apr 13 13:47:54 2017 From: Thomas_Ferraguto at uml.edu (Ferraguto, Thomas) Date: Thu, 13 Apr 2017 17:47:54 +0000 Subject: [labnetwork] Card Readers for equipment interlocks Message-ID: <9636e814204a4da4be4e627551a91881@uml.edu> Colleagues, Has anyone integrated RFID card readers to their equipment interlocks. I'd like to set up a system where the "Tap in" & "Tap Out" ID the user on opens the interlock on the tool. Thanks in advance.... Best Regards... Tom Ferraguto Thomas S. Ferraguto Saab ETIC Nanofabrication Laboratory Director Saab ETIC Building Director 1 University Avenue Lowell MA 01854 Mobile 617-755-0910 Land 978-934-1809 Fax 978-934-1014 -------------- next part -------------- An HTML attachment was scrubbed... URL: From mike.young at nd.edu Fri Apr 14 10:10:26 2017 From: mike.young at nd.edu (Mike Young) Date: Fri, 14 Apr 2017 10:10:26 -0400 Subject: [labnetwork] Card Readers for equipment interlocks In-Reply-To: <9636e814204a4da4be4e627551a91881@uml.edu> References: <9636e814204a4da4be4e627551a91881@uml.edu> Message-ID: <953e1093-b5a0-2e9e-e1c8-8be6d56042f9@nd.edu> Hi Tom. We use Coral, along with a mix of internet-based interlocks. We also use RFID cleanroom badges. I have made a first-cut implementation of this for one of our resist spinners, using a Raspberry Pi and an RFID reader. It's a bit of a kludge, I think I can do better, but it is functional. I'd extend the implementation to other tools also, if the darn readers weren't so expensive (~$100 each)! What I have not addressed, is how to do this for equipment which collects run data at disable. I suppose if coral could download the run data from the tool at disable, via SECS or some other interface... That's how the Big Boys do it... Good luck, and contact me offline if you'd like more of the gory details. --Mike On 4/13/2017 1:47 PM, Ferraguto, Thomas wrote: > > Colleagues, > > Has anyone integrated RFID card readers to their equipment interlocks. > I?d like to set up a system where the ?Tap in? & ?Tap Out? ID the user > on opens the interlock on the tool. > > Thanks in advance?. > > Best Regards? > > Tom Ferraguto > > Thomas S. Ferraguto > > Saab ETIC Nanofabrication Laboratory Director > > Saab ETIC Building Director > > 1 University Avenue > > Lowell MA 01854 > > Mobile 617-755-0910 > > Land 978-934-1809 > > Fax 978-934-1014 > > > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork -- Michael P. Young (574) 631-3268 (office) Nanofabrication Specialist (574) 631-4393 (fax) Department of Electrical Engineering (765) 637-3784 (cell) University of Notre Dame mike.young at nd.edu B-38 Stinson-Remick Hall Notre Dame, IN 46556-5637 -------------- next part -------------- An HTML attachment was scrubbed... URL: From Vincent.Luciani at nist.gov Fri Apr 14 13:32:46 2017 From: Vincent.Luciani at nist.gov (Luciani, Vincent (Fed)) Date: Fri, 14 Apr 2017 17:32:46 +0000 Subject: [labnetwork] Perspectives on equipment management software In-Reply-To: <86E466A1-90C6-408A-92EE-B84E097C432F@cnsi.ucla.edu> References: <86E466A1-90C6-408A-92EE-B84E097C432F@cnsi.ucla.edu> Message-ID: Hello Adam and LabNetwork, Here at CNST we have written our own program called NEMO (NanoFab Equipment Management Operation). Staff and users are very pleased with it. It was written with Python and uses Django (a high-level Python Web framework). We are currently working on packaging it to make it available to others but release is still months away. I?ll post updates on this network as we get closer to release. If you would like more details let me know and I will add you to our mailing list. Thanks, Vince Vincent K. Luciani NanoFab Manager Center for Nanoscale Science and Technology National Institute of Standards and Technology 100 Bureau Drive, MS 6201 Gaithersburg, MD 20899-6200 USA +1-301-975-2886 From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Stieg, Adam Sent: Tuesday, April 11, 2017 11:48 AM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Perspectives on equipment management software Hi All, We are in the process of reviewing our approach to managing reservations, access, billing, etc. for equipment throughout the CNSI at UCLA. To date, we have relied on a system developed in-house that is no longer viable. There is a significant push amongst the shared resources on campus toward iLabs. We have also begun looking into FOM (fomnetworks.com) as an alternative option. I know this may not be the best venue for a dialogue on the strengths/weaknesses of these specific solutions (or any other for that matter), but if any of you are willing to share your perspectives or experiences working with either of these vendors it would be greatly appreciated. All the best, Adam Adam Z. Stieg Ph.D. Associate Director, California NanoSystems Institute at UCLA Director, Nano and Pico Characterization Laboratory Technical Director, Integrated Systems Nanofabrication Cleanroom (p) 310.206.2902 (e) stieg at cnsi.ucla.edu -------------- next part -------------- An HTML attachment was scrubbed... URL: From info at fabsurplus.com Sat Apr 15 02:47:27 2017 From: info at fabsurplus.com (Stephen Howe) Date: Sat, 15 Apr 2017 08:47:27 +0200 Subject: [labnetwork] Perspectives on equipment management software In-Reply-To: References: <86E466A1-90C6-408A-92EE-B84E097C432F@cnsi.ucla.edu> Message-ID: <1492238847.2354.11.camel@fabsurplus.com> Dear Vincent, My company is engaged in developing software which will allow us to obtain and store technical data about equipment ?in an efficient way . Therefore, I'd like to take this opportunity to reach out to make contact with persons who are working in this area as we'd be interested to explore ways of working together to achieve our goals. ?Yours sincerely, Stephen Howe Company Owner SDI Fabsurplus Group +1 830 388 1071 (Mobile) +39 335 710 7756 (Mobile, Italy) Skype: Stephencshowe email: info at fabsurplus.com WWW.FABSURPLUS.COM Facebook: www.facebook.com/fabsurplus Twitter: http://twitter.com/fabsurplus Google+: https://plus.google.com/+Fabsurplus On Fri, 2017-04-14 at 17:32 +0000, Luciani, Vincent (Fed) wrote: > Hello Adam and LabNetwork, > ? > Here at CNST we have written our own program called NEMO (NanoFab > Equipment Management Operation).? ?Staff and users are very pleased > with it.? It was written with Python and uses Django (a high-level > Python Web framework). ?We are currently working on packaging it to > make it available to others but release is still months away.? ?I?ll > post updates on this network as we get closer to release.? If you > would like more details let me know and I will add you to our mailing > list. > ? > Thanks, > Vince > ? > ? > Vincent K. Luciani > NanoFab Manager > Center for Nanoscale Science and Technology > National Institute of Standards and Technology > 100 Bureau Drive, MS 6201 > Gaithersburg, MD 20899-6200 USA > +1-301-975-2886 > ? > ? > ? > ? > ? > ? > ? > ? > ? > ? > From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.m > it.edu] On Behalf Of Stieg, Adam > Sent: Tuesday, April 11, 2017 11:48 AM > To: labnetwork at mtl.mit.edu > Subject: [labnetwork] Perspectives on equipment management software > ? > Hi All, > ? > We are in the process of reviewing our approach to managing > reservations, access, billing, etc. for equipment throughout the CNSI > at UCLA. To date, we have relied on a system developed in-house that > is no longer viable. There is a significant push amongst the shared > resources on campus toward iLabs. We have also begun looking into FOM > (fomnetworks.com) as an alternative option. I know this may not be > the best venue for a dialogue on the strengths/weaknesses of these > specific solutions (or any other for that matter), but if any of you > are willing to share your perspectives or experiences working with > either of these vendors it would be greatly appreciated. > ? > All the best, > Adam > ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ?????????? ? ? ? ? ? ? > ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ?? > ? > Adam Z. Stieg Ph.D. ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? > Associate Director,?California NanoSystems Institute at UCLA > Director, Nano and Pico Characterization Laboratory > Technical Director, Integrated Systems Nanofabrication Cleanroom > (p) 310.206.2902 > (e) stieg at cnsi.ucla.edu > ? > ? > ? > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: From vamsinittala at gmail.com Mon Apr 17 05:17:14 2017 From: vamsinittala at gmail.com (N P VAMSI KRISHNA) Date: Mon, 17 Apr 2017 14:47:14 +0530 Subject: [labnetwork] Diamatix DMP-2831 timing belt Message-ID: Dear All, We have a Diamatix Ink jet printer DMP-2831. The timing belt of it gone bad and the vendors are demanding too high for the belt replacement. Does any one have the manual to replace the timing belt? Also kindly let me know if any one had replaced it. -- -- Thanks & Best Regards, ----------------- *N.P.Vamsi Krishna* *?Micro Nano Fluidics lab?*Center for Nano Science and Engineering (CeNSE), Indian Institute of Science(IISc), Bangalore. INDIA-560012 ?Educating the mind without educating the heart is no education at all.? -Aristotle -------------- next part -------------- An HTML attachment was scrubbed... URL: From dylan.klomparens at nist.gov Tue Apr 18 10:33:28 2017 From: dylan.klomparens at nist.gov (Klomparens, Dylan L. (Fed)) Date: Tue, 18 Apr 2017 14:33:28 +0000 Subject: [labnetwork] Card Readers for equipment interlocks In-Reply-To: <953e1093-b5a0-2e9e-e1c8-8be6d56042f9@nd.edu> References: <9636e814204a4da4be4e627551a91881@uml.edu> <953e1093-b5a0-2e9e-e1c8-8be6d56042f9@nd.edu> Message-ID: Hi Tom, CNST has "tap in/out" system to control physical access to the cleanroom and tool interlocks. We use wall mounted Windows tablets that communicate with our lab management software, NEMO. I've attached a picture and parts list of one of the entryway tablets. I have received many positive comments about this system, the users really enjoy the convenience and speed of logging in. It's been working well for over a year now. I enjoyed reading Mike Young's response, since we're also working on improved interlocks using the Raspberry Pi. We have a working prototype (read: kludge) for a single tool. The eventual goal will be to polish it to product level quality. What's nice about the Raspberry Pi is that it has the potential to enable us to collect data from sensors, or create a small embedded systems controller for tools. While prototyping, I've successfully programmed a Raspberry Pi to read an ID number from my government badge. I'd be interested in continuing a conversation about this off-list if anyone is interested in collaborating on Raspberry Pi interlocks. -- Dylan Klomparens From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Mike Young Sent: Friday, April 14, 2017 10:10 AM To: Ferraguto, Thomas ; labnetwork at mtl.mit.edu Subject: Re: [labnetwork] Card Readers for equipment interlocks Hi Tom. We use Coral, along with a mix of internet-based interlocks. We also use RFID cleanroom badges. I have made a first-cut implementation of this for one of our resist spinners, using a Raspberry Pi and an RFID reader. It's a bit of a kludge, I think I can do better, but it is functional. I'd extend the implementation to other tools also, if the darn readers weren't so expensive (~$100 each)! What I have not addressed, is how to do this for equipment which collects run data at disable. I suppose if coral could download the run data from the tool at disable, via SECS or some other interface... That's how the Big Boys do it... Good luck, and contact me offline if you'd like more of the gory details. --Mike On 4/13/2017 1:47 PM, Ferraguto, Thomas wrote: Colleagues, Has anyone integrated RFID card readers to their equipment interlocks. I'd like to set up a system where the "Tap in" & "Tap Out" ID the user on opens the interlock on the tool. Thanks in advance.... Best Regards... Tom Ferraguto -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: Idle_tablet.jpg Type: image/jpeg Size: 359800 bytes Desc: Idle_tablet.jpg URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: Card_swipe.jpg Type: image/jpeg Size: 318967 bytes Desc: Card_swipe.jpg URL: -------------- next part -------------- An embedded and charset-unspecified text was scrubbed... Name: Parts_list.txt URL: From aju-jugessur at uiowa.edu Tue Apr 18 10:54:34 2017 From: aju-jugessur at uiowa.edu (Jugessur, Aju S) Date: Tue, 18 Apr 2017 14:54:34 +0000 Subject: [labnetwork] Etching Alumina (ALD) using BCl3 Message-ID: Hi, Can anyone offer some advice on Alumina etch? We do have BCl3/Cl2/O2 etch on our Oxford PlasmaPro. We plan to use S1818 as the mask as we intend etching less than 30 nm of the Alumina: will this work? I am not sure what is the etch selectivity. We may also use an intermediate metal mask. I am aware that ChF3 and C4F8 also etches Alumina. Can anyone advise us on this process and a benchmark recipe will help? Also, the tool is currently used for III-V etching. Can we safely use the tool to etch Alumina without the risk of process contamination, assuming we do a clean up step each time? Thanks Regards Aju Aju Jugessur Ph.D. Director, University of Iowa Microfabrication Facility Professor (Adj.), Physics and Astronomy OSTC, Iowa Advanced Technology Labs University of Iowa 205 N. Madison St Iowa City, IA 52242 319 -353-2342 aju-jugessur at uiowa.edu http://ostc.uiowa.edu/uimf -------------- next part -------------- An HTML attachment was scrubbed... URL: From whipp003 at umn.edu Tue Apr 18 12:48:29 2017 From: whipp003 at umn.edu (Tony Whipple) Date: Tue, 18 Apr 2017 11:48:29 -0500 Subject: [labnetwork] Etching Alumina (ALD) using BCl3 In-Reply-To: References: Message-ID: Hello Aji; We etch ALD Alumina film with our Oxford system which is patterned by resist. Since most our films are thin we do not have much problem with selectivity to resist which is about 1:1 for those films. The recipe we use is: BCl3 30 sccm ICP 625 W RF 50 W 2 mtorr We get ~ 650 Ang/ min rate for both the Alumina film and resist. The recipe has not been optimized but should be. One of the items that we should work on is to have the recipe etch slower. Our users have said it etches too fast. We have since looked at reducing the RF platten /chuck power while keeping the ICP power the same and this helps in lowering the etch rate. Regards, Tony W. On 4/18/2017 9:54 AM, Jugessur, Aju S wrote: > > Hi, > > Can anyone offer some advice on Alumina etch? > > We do have BCl3/Cl2/O2 etch on our Oxford PlasmaPro. > We plan to use S1818 as the mask as we intend etching less than 30 nm > of the Alumina: will this work? I am not sure what is the etch > selectivity. > We may also use an intermediate metal mask. I am aware that ChF3 and > C4F8 also etches Alumina. > > Can anyone advise us on this process and a benchmark recipe will help? > > Also, the tool is currently used for III-V etching. Can we safely use > the tool to etch Alumina without the risk of process contamination, > assuming we do a clean up step each time? > > Thanks > Regards > Aju > > > > Aju Jugessur /Ph.D./ > Director, University of Iowa Microfabrication Facility > Professor (Adj.), Physics and Astronomy > OSTC, Iowa Advanced Technology Labs > University of Iowa > > 205 N. Madison St > Iowa City, IA 52242 > 319 -353-2342 > aju-jugessur at uiowa.edu > http://ostc.uiowa.edu/uimf > > > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: From vamsinittala at gmail.com Tue Apr 18 13:25:33 2017 From: vamsinittala at gmail.com (N P VAMSI KRISHNA) Date: Tue, 18 Apr 2017 22:55:33 +0530 Subject: [labnetwork] Etching Alumina (ALD) using BCl3 In-Reply-To: References: Message-ID: Dear Aju, I had used the ALD Al2O3 long back. Here is the recipe I used in Plasmalab 100 OPT tool: BCl3 50 sccm Pressure 2mT ICP Power 1250 W Electrode 300 W Temperature 10?C Backside He 15 Torr Etch Rate: 36-38 nm/min Selectivity to PR: ~ 0.4:1 PR I used was AZ5214E for which selectivity was average, so I had used the thick AZ4562 which has very good selectivity. In your case as the Alumina is only 30 nm, S1813 may survive. You may want try spinning at slightly lower rpm (~3K). Hope it helps. Thanks & best regards, vamsi On Tue, Apr 18, 2017 at 8:24 PM, Jugessur, Aju S wrote: > > Hi, > > Can anyone offer some advice on Alumina etch? > > We do have BCl3/Cl2/O2 etch on our Oxford PlasmaPro. > We plan to use S1818 as the mask as we intend etching less than 30 nm of > the Alumina: will this work? I am not sure what is the etch selectivity. > We may also use an intermediate metal mask. I am aware that ChF3 and C4F8 > also etches Alumina. > > Can anyone advise us on this process and a benchmark recipe will help? > > Also, the tool is currently used for III-V etching. Can we safely use the > tool to etch Alumina without the risk of process contamination, assuming we > do a clean up step each time? > > Thanks > Regards > Aju > > > > Aju Jugessur *Ph.D.* > Director, University of Iowa Microfabrication Facility > Professor (Adj.), Physics and Astronomy > OSTC, Iowa Advanced Technology Labs > University of Iowa > > 205 N. Madison St > Iowa City, IA 52242 > 319 -353-2342 <03193%20532%20342> > aju-jugessur at uiowa.edu > http://ostc.uiowa.edu/uimf > > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork > > -- -- Thanks & Best Regards, ----------------- *N.P.Vamsi Krishna* Center for Nano Science and Engineering (CeNSE), Indian Institute of Science(IISc), Bangalore. INDIA-560012 ?Educating the mind without educating the heart is no education at all.? -Aristotle -------------- next part -------------- An HTML attachment was scrubbed... URL: From price.798 at osu.edu Tue Apr 18 13:30:14 2017 From: price.798 at osu.edu (Price, Aimee) Date: Tue, 18 Apr 2017 17:30:14 +0000 Subject: [labnetwork] Card Readers for equipment interlocks In-Reply-To: References: <9636e814204a4da4be4e627551a91881@uml.edu> <953e1093-b5a0-2e9e-e1c8-8be6d56042f9@nd.edu> Message-ID: <83583687862B8444A85296A4944147DF95558152@CIO-TNC-D2MBX03.osuad.osu.edu> We have a home built system that utilizes Raspberry Pi. Users fob in and out of the equipment and it talks to our database to ensure they are certified to use that equipment. We use it for billing as well. Our engineer who designed it is out this week and much of next, but if you contact me I can put you in touch with him. I'm not sure if he's on this list or not. He and our IT engineer worked through a lot of the issues and bugs. We are happy with it for the most part. Aimee From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Klomparens, Dylan L. (Fed) Sent: Tuesday, April 18, 2017 10:33 AM To: Mike Young; Ferraguto, Thomas; labnetwork at mtl.mit.edu Subject: Re: [labnetwork] Card Readers for equipment interlocks Hi Tom, CNST has "tap in/out" system to control physical access to the cleanroom and tool interlocks. We use wall mounted Windows tablets that communicate with our lab management software, NEMO. I've attached a picture and parts list of one of the entryway tablets. I have received many positive comments about this system, the users really enjoy the convenience and speed of logging in. It's been working well for over a year now. I enjoyed reading Mike Young's response, since we're also working on improved interlocks using the Raspberry Pi. We have a working prototype (read: kludge) for a single tool. The eventual goal will be to polish it to product level quality. What's nice about the Raspberry Pi is that it has the potential to enable us to collect data from sensors, or create a small embedded systems controller for tools. While prototyping, I've successfully programmed a Raspberry Pi to read an ID number from my government badge. I'd be interested in continuing a conversation about this off-list if anyone is interested in collaborating on Raspberry Pi interlocks. -- Dylan Klomparens From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Mike Young Sent: Friday, April 14, 2017 10:10 AM To: Ferraguto, Thomas >; labnetwork at mtl.mit.edu Subject: Re: [labnetwork] Card Readers for equipment interlocks Hi Tom. We use Coral, along with a mix of internet-based interlocks. We also use RFID cleanroom badges. I have made a first-cut implementation of this for one of our resist spinners, using a Raspberry Pi and an RFID reader. It's a bit of a kludge, I think I can do better, but it is functional. I'd extend the implementation to other tools also, if the darn readers weren't so expensive (~$100 each)! What I have not addressed, is how to do this for equipment which collects run data at disable. I suppose if coral could download the run data from the tool at disable, via SECS or some other interface... That's how the Big Boys do it... Good luck, and contact me offline if you'd like more of the gory details. --Mike On 4/13/2017 1:47 PM, Ferraguto, Thomas wrote: Colleagues, Has anyone integrated RFID card readers to their equipment interlocks. I'd like to set up a system where the "Tap in" & "Tap Out" ID the user on opens the interlock on the tool. Thanks in advance.... Best Regards... Tom Ferraguto -------------- next part -------------- An HTML attachment was scrubbed... URL: From daniel.woodie at cornell.edu Tue Apr 18 14:14:11 2017 From: daniel.woodie at cornell.edu (Dan P. Woodie) Date: Tue, 18 Apr 2017 18:14:11 +0000 Subject: [labnetwork] How do you clean your cleanroom? Message-ID: Lab Network, We are in the middle of discussions with our janitorial staff on the proper procedures for cleaning our Class 1,000 cleanroom and are looking to see what others are doing for their academic cleanrooms. If any of you have written cleaning steps and frequencies they are willing to share we would appreciate it. Thank you. Dan Daniel Woodie Safety Manager, College of Engineering Cornell University 344 Duffield Hall Ithaca, NY 14853-2700 (607)254-4891 Off Hour Emergency # - (607)227-2993 -------------- next part -------------- An HTML attachment was scrubbed... URL: From barbjn at sfsu.edu Tue Apr 18 18:35:54 2017 From: barbjn at sfsu.edu (Barbara J Neuhauser) Date: Tue, 18 Apr 2017 22:35:54 +0000 Subject: [labnetwork] Diamatix DMP-2831 timing belt In-Reply-To: References: Message-ID: <1492554954774.71358@sfsu.edu> Dr. Krishna --- We are considering purchasing a DMP-2831 for the purpose of printing solar cell front contacts. How long did your printer function before the timing belt failed? Barbara Neuhauser The Phoenix Project San Francisco State University barbjn at sfsu.edu ________________________________ From: labnetwork-bounces at mtl.mit.edu on behalf of N P VAMSI KRISHNA Sent: Monday, April 17, 2017 2:17 AM To: Labnetwork Cc: Prosenjit Sen Subject: [labnetwork] Diamatix DMP-2831 timing belt Dear All, We have a Diamatix Ink jet printer DMP-2831. The timing belt of it gone bad and the vendors are demanding too high for the belt replacement. Does any one have the manual to replace the timing belt? Also kindly let me know if any one had replaced it. -- -- Thanks & Best Regards, ----------------- N.P.Vamsi Krishna ?Micro Nano Fluidics lab? Center for Nano Science and Engineering (CeNSE), Indian Institute of Science(IISc), Bangalore. INDIA-560012 "Educating the mind without educating the heart is no education at all." -Aristotle -------------- next part -------------- An HTML attachment was scrubbed... URL: From vamsinittala at gmail.com Wed Apr 19 00:47:43 2017 From: vamsinittala at gmail.com (N P VAMSI KRISHNA) Date: Wed, 19 Apr 2017 10:17:43 +0530 Subject: [labnetwork] Diamatix DMP-2831 timing belt In-Reply-To: <1492554954774.71358@sfsu.edu> References: <1492554954774.71358@sfsu.edu> Message-ID: Dear Prof. Neuhauser, Thanks for the mail. Its ~4 years old tool. The tool was idle for quite long time, recently (6 months) we started using it extensively for printing Silver ink interconnects in our 3D devices. Thanks & best regards, vamsi krishna On Wed, Apr 19, 2017 at 4:05 AM, Barbara J Neuhauser wrote: > Dr. Krishna --- > > > We are considering purchasing a DMP-2831 for the purpose of printing solar > cell front contacts. How long did your printer function before the timing > belt failed? > > > Barbara Neuhauser > > The Phoenix Project > > San Francisco State University > > barbjn at sfsu.edu > > > > ------------------------------ > *From:* labnetwork-bounces at mtl.mit.edu > on behalf of N P VAMSI KRISHNA > *Sent:* Monday, April 17, 2017 2:17 AM > *To:* Labnetwork > *Cc:* Prosenjit Sen > *Subject:* [labnetwork] Diamatix DMP-2831 timing belt > > Dear All, > > We have a Diamatix Ink jet printer DMP-2831. The timing belt of it gone > bad and the vendors are demanding too high for the belt replacement. > Does any one have the manual to replace the timing belt? Also kindly let > me know if any one had replaced it. > > > -- > -- > Thanks & Best Regards, > ----------------- > *N.P.Vamsi Krishna* > > * ?Micro Nano Fluidics lab? *Center for Nano Science and Engineering > (CeNSE), > Indian Institute of Science(IISc), Bangalore. > INDIA-560012 > > ?Educating the mind without educating the heart is no education at all.? > -Aristotle > -- -- Thanks & Best Regards, ----------------- *N.P.Vamsi Krishna* Center for Nano Science and Engineering (CeNSE), Indian Institute of Science(IISc), Bangalore. INDIA-560012 ?Educating the mind without educating the heart is no education at all.? -Aristotle -------------- next part -------------- An HTML attachment was scrubbed... URL: From rmorrison at draper.com Wed Apr 19 14:28:47 2017 From: rmorrison at draper.com (Morrison, Richard H., Jr.) Date: Wed, 19 Apr 2017 18:28:47 +0000 Subject: [labnetwork] How do you clean your cleanroom? In-Reply-To: References: Message-ID: HI, At Draper we do the following: Empty the trash every day, Using 70% DI AND 30% IPA we wipe the following items daily Entrance door Trash can Chair legs Weekly we wipe the table tops, equipment, anything on the walls that stick out Qtr walls and ceiling Rick ________________________________ From: labnetwork-bounces at mtl.mit.edu on behalf of Dan P. Woodie Sent: Tuesday, April 18, 2017 2:14 PM To: labnetwork at mtl.mit.edu Subject: [labnetwork] How do you clean your cleanroom? Lab Network, We are in the middle of discussions with our janitorial staff on the proper procedures for cleaning our Class 1,000 cleanroom and are looking to see what others are doing for their academic cleanrooms. If any of you have written cleaning steps and frequencies they are willing to share we would appreciate it. Thank you. Dan Daniel Woodie Safety Manager, College of Engineering Cornell University 344 Duffield Hall Ithaca, NY 14853-2700 (607)254-4891 Off Hour Emergency # - (607)227-2993 ________________________________ Notice: This email and any attachments may contain proprietary (Draper non-public) and/or export-controlled information of Draper. If you are not the intended recipient of this email, please immediately notify the sender by replying to this email and immediately destroy all copies of this email. ________________________________ -------------- next part -------------- An HTML attachment was scrubbed... URL: From r.khanna at ucl.ac.uk Mon Apr 24 06:33:00 2017 From: r.khanna at ucl.ac.uk (Khanna, Rohit) Date: Mon, 24 Apr 2017 10:33:00 +0000 Subject: [labnetwork] Card Readers for equipment interlocks In-Reply-To: References: <9636e814204a4da4be4e627551a91881@uml.edu> <953e1093-b5a0-2e9e-e1c8-8be6d56042f9@nd.edu> <83583687862B8444A85296A4944147DF95558152@CIO-TNC-D2MBX03.osuad.osu.edu>, Message-ID: Greeting Mike, Hope you are doing good. A brief introduction, I am an Electronics engineer working with UCL (University college London). We have a Clean room facility at London Center for NANO Tech. Incidentally I was directed to this mailing list by my manager Steve Etienne as we were also exploring option for card access for clean room equipment. We had also developed a custom solution for cleanroom equipment control. The hardware prototypes have been developed by me and has been under trial for a few months with few simple equipment we have. So far the hardware seems to be working reasonably well. The system is based on wired ethernet or WiFi (there have been two separate version of the hardware). The current firmware supports usage logging only as I didn't have the requisite support/expertise for developing the PC side application/middle ware which would accept UDP/TCP requests from multiple control units and run verification requests with the data base and send back confirmation UDP/TCP packets to the hardware units to unlock equipment. Here is a youtube link for the system demo: https://www.youtube.com/watch?v=S9hz3KTrfno&t=30s [https://i.ytimg.com/vi/S9hz3KTrfno/hqdefault.jpg] WiFi Access Control System www.youtube.com This is a demonstration of Access control system developed for equipment control. The system connects to PC/Database using WiFi network and validates/logs th... If this seems interesting we could discuss possibility of developing the system further and share designs for mutual benefit. Also, I am attaching the system overview block diagram for the same. Warm Regards Rohit Khanna Electronic Test & Measurement Engineer London Center for NANO Technology, UCL Ph:+44-020-7679984 Int Ext: 39984 Mob: +44 7456265557 ________________________________ From: Etienne, Steve Sent: 24 April 2017 10:55 To: Khanna, Rohit Subject: FW: [labnetwork] Card Readers for equipment interlocks Rohit, Are you subscribed to this mailing list? It has some information about access systems from time to time. You could also contribute if you wish. Steve From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Price, Aimee Sent: 18 April 2017 18:30 To: Klomparens, Dylan L. (Fed) ; Mike Young ; Ferraguto, Thomas ; labnetwork at mtl.mit.edu Subject: Re: [labnetwork] Card Readers for equipment interlocks We have a home built system that utilizes Raspberry Pi. Users fob in and out of the equipment and it talks to our database to ensure they are certified to use that equipment. We use it for billing as well. Our engineer who designed it is out this week and much of next, but if you contact me I can put you in touch with him. I?m not sure if he?s on this list or not. He and our IT engineer worked through a lot of the issues and bugs. We are happy with it for the most part. Aimee From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Klomparens, Dylan L. (Fed) Sent: Tuesday, April 18, 2017 10:33 AM To: Mike Young; Ferraguto, Thomas; labnetwork at mtl.mit.edu Subject: Re: [labnetwork] Card Readers for equipment interlocks Hi Tom, CNST has ?tap in/out? system to control physical access to the cleanroom and tool interlocks. We use wall mounted Windows tablets that communicate with our lab management software, NEMO. I?ve attached a picture and parts list of one of the entryway tablets. I have received many positive comments about this system, the users really enjoy the convenience and speed of logging in. It?s been working well for over a year now. I enjoyed reading Mike Young?s response, since we?re also working on improved interlocks using the Raspberry Pi. We have a working prototype (read: kludge) for a single tool. The eventual goal will be to polish it to product level quality. What?s nice about the Raspberry Pi is that it has the potential to enable us to collect data from sensors, or create a small embedded systems controller for tools. While prototyping, I?ve successfully programmed a Raspberry Pi to read an ID number from my government badge. I?d be interested in continuing a conversation about this off-list if anyone is interested in collaborating on Raspberry Pi interlocks. -- Dylan Klomparens From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Mike Young Sent: Friday, April 14, 2017 10:10 AM To: Ferraguto, Thomas >; labnetwork at mtl.mit.edu Subject: Re: [labnetwork] Card Readers for equipment interlocks Hi Tom. We use Coral, along with a mix of internet-based interlocks. We also use RFID cleanroom badges. I have made a first-cut implementation of this for one of our resist spinners, using a Raspberry Pi and an RFID reader. It's a bit of a kludge, I think I can do better, but it is functional. I'd extend the implementation to other tools also, if the darn readers weren't so expensive (~$100 each)! What I have not addressed, is how to do this for equipment which collects run data at disable. I suppose if coral could download the run data from the tool at disable, via SECS or some other interface... That's how the Big Boys do it... Good luck, and contact me offline if you'd like more of the gory details. --Mike On 4/13/2017 1:47 PM, Ferraguto, Thomas wrote: Colleagues, Has anyone integrated RFID card readers to their equipment interlocks. I?d like to set up a system where the ?Tap in? & ?Tap Out? ID the user on opens the interlock on the tool. Thanks in advance?. Best Regards? Tom Ferraguto -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: Smart access system.pdf Type: application/pdf Size: 645169 bytes Desc: Smart access system.pdf URL: From cekendri at mtu.edu Thu Apr 27 10:34:58 2017 From: cekendri at mtu.edu (Chito Kendrick) Date: Thu, 27 Apr 2017 10:34:58 -0400 Subject: [labnetwork] Hot embossing of PMMA Message-ID: <09e573a0-f87c-aedd-9087-fea5e871054e@mtu.edu> I have a user that wants to replace there PDMS chamber with a PMMA chamber for a microfluidic cell. They went and brought a low cost resin press to try and hot emboss it with using either a 50-70 um SU-8/Si master or a FR4 master (pattern etched into the Copper). They tried a DRIE silicon master and broke it first press. So far their PMMA print is distorted across the either 4" wafer due to shrinkage. They are looking into alternative polymers, but yet to find one. Does anyone have experience with this type to work? I am considering trying to etch into the PMMA using an RIE, but 50 um would be a long etch for our systems. Regards, Chito Kendrick -- Chito Kendrick Ph.D. Managing Director of the Microfabrication Facility Research Assistant Professor Electrical and Computer Engineering Michigan Technological University Room 436 M&M Building 1400 Townsend Dr. Houghton, Michigan 49931-1295 814-308-4255 -------------- next part -------------- An HTML attachment was scrubbed... URL: