[labnetwork] Metallization on rough ceramic substrates
Fouad Karouta
fouad.karouta at anu.edu.au
Sun Jan 6 17:54:31 EST 2019
Hi Adithi,
Adding to all other comments and in particular you need to pay attention to the profile of the photoresist: for a good lift-off you need some undercut. You didn’t mention the feature size of the metallisation pads but assuming it is more than 3 µm in any dimension you may need to consider using a negative resist as the undercut can be tailored to your need using longer development time. We use maN 1420 (about 2 ↨0m thick) for lift-off processes and that works fine.
Regards,
Fouad Karouta
*************************************
Manager ANFF ACT Node
Australian National Fabrication Facility
Research School of Physics and Engineering
L. Huxley Building (#56), Mills Road, Room 4.02
Australian National University
ACT 2601, Canberra, Australia
Tel: + 61 2 6125 7174
Mob: + 61 451 046 412
Email: fouad.karouta at anu.edu.au<mailto:fouad.karouta at anu.edu.au>
http://anff-act.anu.edu.au/
From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of mdwyer87 at gmail.com
Sent: Monday, 7 January 2019 1:48 AM
To: 'ADITHI U' <uadithi at iisc.ac.in>
Cc: labnetwork at mtl.mit.edu
Subject: Re: [labnetwork] Metallization on rough ceramic substrates
Hi Adithi,
In addition to the good advice already given, you might also consider thermal issues if your problem is the metal not lifting off. Evaporating Pt (especially up to 250 nm) can result in significant substrate heating, particularly if your samples are not heatsinked. I clamp my samples to a large block of aluminum flattened on a surface plate and that solved my thermal issues (along with sensible deposition rates).
Matt
Postdoc
van der Weide Group
UW-Madison
From: labnetwork-bounces at mtl.mit.edu<mailto:labnetwork-bounces at mtl.mit.edu> <labnetwork-bounces at mtl.mit.edu<mailto:labnetwork-bounces at mtl.mit.edu>> On Behalf Of Noah Clay
Sent: Friday, January 4, 2019 1:17 PM
To: ADITHI U <uadithi at iisc.ac.in<mailto:uadithi at iisc.ac.in>>
Cc: labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>
Subject: Re: [labnetwork] Metallization on rough ceramic substrates
Hello Adithi,
As you know, ceramics can be quite porous and metal adhesion can be diminished by surface/internally adsorbed water, organics, etc. Therefore, you may consider deep solvent cleaning, other surface treatments as appropriate (e.g., plasma) for your material and a lengthy dehydration bake prior to resist application. If pre-resist cleaning proves ineffective, then after resist development, you may consider O2 plasma or, if available, in-situ ion bombardment. Lastly, given the rough topography of your substrate, there may be undeveloped or underexposed pockets of resist on the substrate, which necessitate optimizing bake, exposure and develop conditions.
As for HMDS (mainly used for Si), without knowing the type of ceramic you’re using, I can’t say that it will be effective as an adhesion promoter. I very humbly/respectfully disagree with previous comments that HMDS is used mostly for etching, but agree that to facilitate liftoff, a relatively thick LOR layer should be spun and baked before application of imaging resist.
Best,
Noah Clay
Singh Center for Nanotechnology
University of Pennsylvania
Philadelphia, PA
Sent from my iPhone
On Jan 4, 2019, at 08:51, Hitesh Kamble <hit.kamble at gmail.com<mailto:hit.kamble at gmail.com>> wrote:
Hello Adithi,
In case of HMDS ,it will improve adhesion but resists will not remove easily from surafce.HMDS is mostly use in etching. Since you're having rough surface, Instead of HMDS use LOR 3B or say liftOFF resists which helps to improve liftoff process.
Also for 250nm metallization you need more than 2um thick resists.
We did not do this on ceramic samples exactly but successfully did liftoff on etched/textured glass and etched samples(roughness more than~400nm).
Following are my suggestions:
-Time Interval between litho and deposition should be lowest as possible
- Dehydration at higher temp and for longer time.
-LOR 3B coating at lower rpm and gradually increase rpm.
-Keep sample in Remover PG and heat at 80C for 5min or more if metal not lifted,(Sonication not recquired in most cases)
Thanks,
Hitesh Kamble
IITBNF-IIT Bombay
On Fri 4 Jan, 2019, 6:08 PM ADITHI U <uadithi at iisc.ac.in<mailto:uadithi at iisc.ac.in> wrote:
Dear All,
We are having difficulties in doing metal(Ti/Pt) lift off on ceramic substrates. Does any of you have experience on fabricating metal patterns on the rough Ceramic Substrates. Below are the details of the process done.
The roughness of the ceramic substrates were approximately 400nm. We had used HMDS as adhesion promoter for lithography, the thickness of the Photoresist was ~ 1.5um.The thickness of the Ti/Pt layer was ~ 250nm. Immediately after lithography the deposition was done followed by 48 hours of lift off(Acetone) process. Ultrasonication(5-10mins) in Acetone was done after 48hours.
Best Regards
Adithi.U
Senior Facility Technologist,
National Nano Fabrication Centre,
CeNSE, Indian Institute of Science
Bangalore
_______________________________________________
labnetwork mailing list
labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>
https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
_______________________________________________
labnetwork mailing list
labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>
https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20190106/03c72d36/attachment.html>
More information about the labnetwork
mailing list