From grallion at ncsu.edu Mon Jun 3 08:35:38 2019 From: grallion at ncsu.edu (Greg Allion) Date: Mon, 3 Jun 2019 08:35:38 -0400 Subject: [labnetwork] Gold Deposition over Chrome In-Reply-To: <1a80dabc-cfb9-aa23-6958-7d0e9da2a40b@4dlabs.ca> References: <847898314bf749a3aa68c76b6d1b8db0@draper.com> <1a80dabc-cfb9-aa23-6958-7d0e9da2a40b@4dlabs.ca> Message-ID: Maybe I'm not understanding your process correctly, but couldn't you just write the mask inverted, develop, evaporate your Au, lift-off, then Cr etch? The Au will act as a your mask then for the Ctr etch. -Greg On Fri, May 31, 2019 at 7:37 PM Nathanael Sieb wrote: > A similar interesting idea might be to use a fluoro-silane to make a > fluorinated surface in the glass areas. I've never tried it before for > this application, but it might work. > > Thanks, > > Nathanael > On 2019-05-31 3:40 a.m., Morrison, Richard H., Jr wrote: > > Here?s a crazy idea try vapor HMDS, it will react with the glass but not > the Chrome and it might help. Gold does not really like to stick to glass > how thick is the film? > > > > Rick > > > > > > > > *From:* labnetwork-bounces at mtl.mit.edu [ > mailto:labnetwork-bounces at mtl.mit.edu ] *On > Behalf Of *Jim Carroll > *Sent:* Thursday, May 30, 2019 4:24 PM > *To:* labnetwork at mtl.mit.edu > *Subject:* [labnetwork] Gold Deposition over Chrome > > > > Dear LabNetwork colleagues, > > > > We are trying to add a thin film of gold over patterned chrome on a > standard photomask using e-beam evaporation. We want the gold to > preferentially adhere to the chrome while lifting off the glass. > > > After a few tests, it appears that the gold sticks equally well to both > the glass and chrome. Ultrasonic agitation and swabbing do not selectively > remove the gold from the glass while leaving it on the chrome. Eventually, > after enough pressure is applied with a swab, the gold releases from both > the glass and chrome. > > > > Does anyone have a suggested approach? > > > Thanks, > > Jim Carroll > *PhotomaskPORTAL* > > *We help you make masks* > (415) 448-6275 > > ? [image: Image removed by sender.] > > ------------------------------ > Notice: This email and any attachments may contain proprietary (Draper > non-public) and/or export-controlled information of Draper. If you are not > the intended recipient of this email, please immediately notify the sender > by replying to this email and immediately destroy all copies of this email. > ------------------------------ > > _______________________________________________ > labnetwork mailing listlabnetwork at mtl.mit.eduhttps://mtl.mit.edu/mailman/listinfo.cgi/labnetwork > > > ------------------------------ > *Nathanael Sieb* > Director of Operations and Administration > > 4D LABS, Simon Fraser University 8888 University Drive, Burnaby, B.C. > V5A 1S6 > > E sieb at 4dlabs.ca T 778.782.8084 F 778.782.3765 W www.4dlabs.ca > > > > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork > -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.jpg Type: image/jpeg Size: 463 bytes Desc: not available URL: From rosendo.bindoy at kaust.edu.sa Mon Jun 3 09:58:56 2019 From: rosendo.bindoy at kaust.edu.sa (Rosendo Bindoy) Date: Mon, 3 Jun 2019 13:58:56 +0000 Subject: [labnetwork] Manual Spin/Rinse/Dryer from MicroProcess Technologies Message-ID: <9AAB50C9-F068-4D8B-AC3F-2FF3EE83A739@kaust.edu.sa> Hello Everyone, I?m looking for a service manual of a Spin/Rinse/Dryer (8800SRD) from MicroProcess Technologies. Can somebody with a kind heart share a copy of it. Regards, Rosendo M. Bindoy Equipment Engineer Nanofabrication Core Lab Ibn Sina Bldg. 3 (Sea Side), Level 0 Room 0252 King Abdullah University of Science and Technology Phone: +966 12 808 4413 Mobile: +966 54 251 5257 Email: rosendo.bindoy at kaust.edu.sa -------------- next part -------------- An HTML attachment was scrubbed... URL: From yaofootball at gmail.com Mon Jun 3 10:53:40 2019 From: yaofootball at gmail.com (Football) Date: Mon, 3 Jun 2019 09:53:40 -0500 Subject: [labnetwork] Research Scientist position available at University of Illinois at Urbana-Champaign Message-ID: Dear All, I'm excited to join in the network. I would like to post a Research Scientist position available in Materials Research laboratory at University of Illinois at Urbana-Champaign. Please feel free to share the information to your friends and colleagues. Please find the details about the position below. Thanks. Fubo Rao, Ph.D., Research Scientist, Materials Research Laboratory University of Illinois at Urbana-Champaign ---------------------------------- Research Scientist Materials Research Laboratory University of Illinois at Urbana-Champaign The Materials Research Laboratory (MRL) at the University of Illinois at Urbana-Champaign is seeking a highly motivated Research Scientist to participate and/or actively support research in the areas of x-ray scattering and cryo-magneto analysis of materials. The Research Scientist will provide comprehensive technical support and user training for equipment, procedures, and safety in the broad areas of materials microanalysis and microfabrication. The successful candidate will become a member of the dedicated staff of approximately 20 scientists and engineers, who maintain major research instruments in the MRL?s core facilities for soft materials characterization, scanning probe microscopy, laser and optical spectroscopy, electron microscopy, surface analysis, x-ray diffraction, and nanofabrication facilities including two cleanrooms. Over 1,000 researchers from across our campus as well as other academic institutions, industry, and national laboratories use the facilities, logging more than 100,000 user hours annually. The lab is recognized as one of the premier mid-sized user facilities in the nation. For details, please visit us online at https://mrl.illinois.edu/facilities/. The University of Illinois is an Equal Opportunity, Affirmative Action employer. Minorities, women, veterans and individuals with disabilities are encouraged to apply. For more information, visit http://go.illinois.edu/EEO. Responsibilities include: ? Actively participate in research using x-ray scattering techniques in new materials, such as ceramics, semiconductor multilayers and super lattices, polymer and biological materials, etc. The use of new, high-resolution detector capabilities just added to the x-ray lab will be emphasized. A second set of responsibilities will involve similar work in the MPMS (magnetic property measurement system) and PPMS (physical property measurements systems). ? Prepare and deliver primary and advanced training on the various techniques and instrumentation available in the x-ray analysis laboratory such as XRD, SAXS/WAXS, XRF and Laue camera systems, and in the MPMS and PPMS laboratory in the MRL. ? Formulate, compile and distribute suitable suggestions for documentation improvements for collective staff scientist review. Incorporate approved modifications into training documents and procedures. ? Perform routine preventative maintenance tasks which vary among daily, monthly, and annual basis for assigned laboratory instrumentation. Examples include daily checks to ensure instruments are running correctly; monthly proactive checks to ensure no hazards are present (e.g. x-ray enclosure fail safes are functional); annual maintenance service on vacuum pumps and supporting mechanical equipment. ? Analyze performance and procedures for assigned equipment during regular operation and scheduled maintenance routines to determine if modifications are necessary to improve or augment their reliability, performance, or capabilities. ? Identify hazards and/or potential failure modes by comparing equipment usage and performance to establish safety protocols while conducting user training or performing maintenance. If the equipment is not operating within tolerances or any engineering safety controls are malfunctioning, determine corrective actions to hardware, operating procedures or user training in conjunction with the assigned staff scientist and implement the changes. ? Perform facility-wide safety tasks as assigned. ? Conduct department or campus specified lab inspections for assigned operating areas and participate in reviews of non-assigned areas as requested. ? Assume additional responsibilities to promote the unit?s mission as needed. Minimum Qualifications ? Master?s degree in engineering or physical sciences. ? Three years of hands-on laboratory experience via degree coursework or independent research. ? Three years of instructional/training experience delivering technical information. ? Two years of experience with x-ray analytical tools. ? Excellent oral and written communication skills. Preferred Qualifications ? PhD in engineering or physical sciences. ? Experience working in a multi-user academic research facility. ? Familiarity with a wide variety of analytical tools such as XRD, XRF, SAXS, PPMS and MPMS systems. ? Chemistry background necessary for identifying chemicals for waste processing. This is a full-time, benefits-eligible academic professional position appointed on a 12-month service basis. The expected start date is as soon as possible. Salary is commensurate with experience and qualifications. To apply, please complete a candidate profile at http://jobs.illinois.edu and upload the following as a single file: a cover letter, resume, and the names and contact information for three professional references. To ensure full consideration, all requested information must be submitted by July 5, 2019. Applicants may be interviewed before the closing date; however, no hiring decision will be made until after that date. For further information regarding application procedures, contact Summer Redman at sredman at illinois.edu or 217-300-5400. The University of Illinois conducts criminal background checks on all job candidates upon acceptance of a contingent offer. -------------- next part -------------- An HTML attachment was scrubbed... URL: From jim at photomaskportal.com Mon Jun 3 17:01:53 2019 From: jim at photomaskportal.com (Jim Carroll) Date: Mon, 3 Jun 2019 16:01:53 -0500 Subject: [labnetwork] Gold Deposition over Chrome In-Reply-To: References: <847898314bf749a3aa68c76b6d1b8db0@draper.com> <1a80dabc-cfb9-aa23-6958-7d0e9da2a40b@4dlabs.ca> Message-ID: Hi Greg, Thanks, that sounds like a viable option. In this specific case, since we already have patterned chrome masks, we're going to try the "?Reverse-side exposure Lift-Off? process mentioned by Andrei near the beginning of the thread, plus the addition of an anneal step to punch through the oxide layer and improve adhesion. Thanks, Jim Carroll *PhotomaskPORTAL* *We help you make masks* (415) 448-6275 - On Mon, Jun 3, 2019 at 2:49 PM Greg Allion wrote: > Maybe I'm not understanding your process correctly, but couldn't you just > write the mask inverted, develop, evaporate your Au, lift-off, then Cr > etch? The Au will act as a your mask then for the Ctr etch. > -Greg > > On Fri, May 31, 2019 at 7:37 PM Nathanael Sieb wrote: > >> A similar interesting idea might be to use a fluoro-silane to make a >> fluorinated surface in the glass areas. I've never tried it before for >> this application, but it might work. >> >> Thanks, >> >> Nathanael >> On 2019-05-31 3:40 a.m., Morrison, Richard H., Jr wrote: >> >> Here?s a crazy idea try vapor HMDS, it will react with the glass but not >> the Chrome and it might help. Gold does not really like to stick to glass >> how thick is the film? >> >> >> >> Rick >> >> >> >> >> >> >> >> *From:* labnetwork-bounces at mtl.mit.edu [ >> mailto:labnetwork-bounces at mtl.mit.edu ] *On >> Behalf Of *Jim Carroll >> *Sent:* Thursday, May 30, 2019 4:24 PM >> *To:* labnetwork at mtl.mit.edu >> *Subject:* [labnetwork] Gold Deposition over Chrome >> >> >> >> Dear LabNetwork colleagues, >> >> >> >> We are trying to add a thin film of gold over patterned chrome on a >> standard photomask using e-beam evaporation. We want the gold to >> preferentially adhere to the chrome while lifting off the glass. >> >> >> After a few tests, it appears that the gold sticks equally well to both >> the glass and chrome. Ultrasonic agitation and swabbing do not selectively >> remove the gold from the glass while leaving it on the chrome. Eventually, >> after enough pressure is applied with a swab, the gold releases from both >> the glass and chrome. >> >> >> >> Does anyone have a suggested approach? >> >> >> Thanks, >> >> Jim Carroll >> *PhotomaskPORTAL* >> >> *We help you make masks* >> (415) 448-6275 >> >> ? [image: Image removed by sender.] >> >> ------------------------------ >> Notice: This email and any attachments may contain proprietary (Draper >> non-public) and/or export-controlled information of Draper. If you are not >> the intended recipient of this email, please immediately notify the sender >> by replying to this email and immediately destroy all copies of this email. >> ------------------------------ >> >> _______________________________________________ >> labnetwork mailing listlabnetwork at mtl.mit.eduhttps://mtl.mit.edu/mailman/listinfo.cgi/labnetwork >> >> >> ------------------------------ >> *Nathanael Sieb* >> Director of Operations and Administration >> >> 4D LABS, Simon Fraser University 8888 University Drive, Burnaby, B.C. >> V5A 1S6 >> >> E sieb at 4dlabs.ca T 778.782.8084 F 778.782.3765 W www.4dlabs.ca >> >> >> >> >> _______________________________________________ >> labnetwork mailing list >> labnetwork at mtl.mit.edu >> https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork >> > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork > -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.jpg Type: image/jpeg Size: 463 bytes Desc: not available URL: From aandreib at gmail.com Tue Jun 4 01:48:46 2019 From: aandreib at gmail.com (Andrei Alamariu) Date: Tue, 4 Jun 2019 01:48:46 -0400 Subject: [labnetwork] Gold Deposition over Chrome In-Reply-To: References: <847898314bf749a3aa68c76b6d1b8db0@draper.com> <1a80dabc-cfb9-aa23-6958-7d0e9da2a40b@4dlabs.ca> Message-ID: <23509478-322C-4475-9F27-4FE3BFE3C260@gmail.com> Hello Jim, The Gold - Cr2O3 adherence test you had done applying enough pressure with a swab was very conclusive, and this problem has to be addressed indeed. I think there are some constraints for heating up a mask plate. First Cr itself might get Sinter like alloyed with the Glass/Quartz plate, resulting a rough top surface (pinholes too) for such a thin 1000A film: as the metal is ?suck? like in the substrate. The lines edges might get affected. All these for 15 min at ~ 400-500C : actually it depends on the ?Temperature-Time budget? Au, Cu, Cr, Fe diffuse very fast in Silicon based materials; TiN blocks them/ see the Damascene process. Very possible that Au to significantly diffuse in Cr2O3, or other thermal assisted enhanced adherence reaction. You could try to use RTA. Second is the mask plate flatness issue. The Glass plate is safe limited to ~ 490C to minimize the plastic-gravity deformation. The plate bending due to the metal stress and metal-Glass expansion coefficient mismatch might not be a problem as the metal is very thin and is not continuous being patterned. Also possible you might think of depositing in-situ Cr & Au, or Cr2O3-Au and see how it goes cold or after anneal. Success! Bernard Sent from my iPad > On Jun 3, 2019, at 5:01 PM, Jim Carroll wrote: > > Hi Greg, > > Thanks, that sounds like a viable option. In this specific case, since we already have patterned chrome masks, we're going to try the "?Reverse-side exposure Lift-Off? process mentioned by Andrei near the beginning of the thread, plus the addition of an anneal step to punch through the oxide layer and improve adhesion. > > Thanks, > Jim Carroll > PhotomaskPORTAL > We help you make masks > (415) 448-6275 > > > >> On Mon, Jun 3, 2019 at 2:49 PM Greg Allion wrote: >> Maybe I'm not understanding your process correctly, but couldn't you just write the mask inverted, develop, evaporate your Au, lift-off, then Cr etch? The Au will act as a your mask then for the Ctr etch. >> -Greg >> >>> On Fri, May 31, 2019 at 7:37 PM Nathanael Sieb wrote: >>> A similar interesting idea might be to use a fluoro-silane to make a fluorinated surface in the glass areas. I've never tried it before for this application, but it might work. >>> Thanks, >>> >>> Nathanael >>>> On 2019-05-31 3:40 a.m., Morrison, Richard H., Jr wrote: >>>> Here?s a crazy idea try vapor HMDS, it will react with the glass but not the Chrome and it might help. Gold does not really like to stick to glass how thick is the film? >>>> >>>> >>>> >>>> Rick >>>> >>>> >>>> >>>> >>>> >>>> >>>> >>>> From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Jim Carroll >>>> Sent: Thursday, May 30, 2019 4:24 PM >>>> To: labnetwork at mtl.mit.edu >>>> Subject: [labnetwork] Gold Deposition over Chrome >>>> >>>> >>>> >>>> Dear LabNetwork colleagues, >>>> >>>> >>>> >>>> We are trying to add a thin film of gold over patterned chrome on a standard photomask using e-beam evaporation. We want the gold to preferentially adhere to the chrome while lifting off the glass. >>>> >>>> >>>> After a few tests, it appears that the gold sticks equally well to both the glass and chrome. Ultrasonic agitation and swabbing do not selectively remove the gold from the glass while leaving it on the chrome. Eventually, after enough pressure is applied with a swab, the gold releases from both the glass and chrome. >>>> >>>> >>>> >>>> Does anyone have a suggested approach? >>>> >>>> >>>> Thanks, >>>> >>>> Jim Carroll >>>> PhotomaskPORTAL >>>> >>>> We help you make masks >>>> (415) 448-6275 >>>> >>>> ? >>>> >>>> Notice: This email and any attachments may contain proprietary (Draper non-public) and/or export-controlled information of Draper. If you are not the intended recipient of this email, please immediately notify the sender by replying to this email and immediately destroy all copies of this email. >>>> >>>> >>>> _______________________________________________ >>>> labnetwork mailing list >>>> labnetwork at mtl.mit.edu >>>> https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork >>> >>> Nathanael Sieb >>> Director of Operations and Administration >>> 4D LABS, Simon Fraser University 8888 University Drive, Burnaby, B.C. V5A 1S6 >>> E sieb at 4dlabs.ca T 778.782.8084 F 778.782.3765 W www.4dlabs.ca >>> >>> >>> _______________________________________________ >>> labnetwork mailing list >>> labnetwork at mtl.mit.edu >>> https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork >> _______________________________________________ >> labnetwork mailing list >> labnetwork at mtl.mit.edu >> https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: From skooi at mit.edu Tue Jun 4 09:00:23 2019 From: skooi at mit.edu (Steven Earl Kooi) Date: Tue, 4 Jun 2019 13:00:23 +0000 Subject: [labnetwork] Job posting - Electron and Surface Microscopy Instrumentation Specialist @ MIT Message-ID: Hello, The Institute for Soldier Nanotechnologies (ISN) at MIT has an open position for an "Electron and Surface Microscopy Instrumentation Specialist? The candidate will be responsible for user training, operation, and maintenance of our SEMs and selected other research instruments at the ISN research facility. Responsibilities include providing one-on-one and group training for a variety of instruments, maintaining/troubleshooting research instrumentation including simple repairs, cleanings, routine calibrations, etc., and working with vendors to schedule more complex repairs/maintenance. The candidate should also be interested in learning new techniques and expanding their expertise as additional equipment is added to the facility. Anyone interested can find out more about the position and apply at: https://careers.peopleclick.com/careerscp/client_mit/external/jobDetails/jobDetail.html?jobPostId=14851&localeCode=en-us Best, Steve. Steven Kooi, Ph.D. Principal Research Scientist Institute for Soldier Nanotechnologies Massachusetts Institute of Technology Phone: 617-324-6416 Fax: 617-253-5859 Email: skooi at mit.edu -------------- next part -------------- An HTML attachment was scrubbed... URL: From rmorrison at draper.com Tue Jun 4 10:25:10 2019 From: rmorrison at draper.com (Morrison, Richard H., Jr) Date: Tue, 4 Jun 2019 14:25:10 +0000 Subject: [labnetwork] DI water Question Message-ID: <660fc24b98d04c11a9310ba47810cc25@draper.com> Hi, I have a question about Chloride in DI water, the E1 water spec states 0.1ppm for Chlorides, have any of your operations monitor for this? I had a test done and my6 redcir tank is 10ppm and the RO output is 25 these seem high to me, any data out there Rick Richard H. Morrison Principal Member of the Technical Staff Draper 555 Technology Square Cambridge, MA 02139-3573 Work 617-258-3420 Cell 508-930-3461 www.draper.com ________________________________ Notice: This email and any attachments may contain proprietary (Draper non-public) and/or export-controlled information of Draper. If you are not the intended recipient of this email, please immediately notify the sender by replying to this email and immediately destroy all copies of this email. ________________________________ -------------- next part -------------- An HTML attachment was scrubbed... URL: From lrehn at tamu.edu Tue Jun 4 15:39:02 2019 From: lrehn at tamu.edu (Rehn, Larry A) Date: Tue, 4 Jun 2019 19:39:02 +0000 Subject: [labnetwork] DI water Question In-Reply-To: <660fc24b98d04c11a9310ba47810cc25@draper.com> References: <660fc24b98d04c11a9310ba47810cc25@draper.com> Message-ID: Hello Richard, We recently had our DI system sanitized by our service provider with a chlorine treatment. They used a chlorine color drop test indicator to determine that the chlorine was present at all drops in our system, and then later to determine that it was rinsed to around < 1ppm. After that I was told that the DI resin beds would continue to remove remaining chlorine as indicated by the resistivity getting back into the Mohm range. We experienced a gradual climb in DI resistivity from about 15Mohm up. Our service provider offered the following explanation (of why it did not immediately climb back to 18Mohm), which might be useful to answer your question about chlorine. The numbers he mentions refers to data in the attachment. ....Usually, it means that a longer rinse should have been done on the regeneration, meaning that a very slight amount of regenerant hasn't been fully rinsed out. I've attached one for the purpose of explanation. Six Megohm is equal to 83 ppb and eight megohm is equal to 63 ppb, so we're talking about t 20 ppb difference overnight. The difference between 14 megohm and 16 megohm is only 5ppb. I honestly believe that the quality will continue to climb very slowly and reach it's highest quality in the course of a few days. I've just seen this scenario too many times over the 30 plus years I've been working with high purity water. FYI: When we regenerate the resin in our shop, we recirculate DI water for typically 24-48 hours through the separately regenerated cation and anion beds before we achieve the quality we desire, prior to mixing it. I suspect very strongly that a batch didn't get fully rinsed. Going back to the paragraph above, it comes down to ppb levels of regenerants (HCl and NaOH) that haven't been fully removed through the resin. If you consume water throughout the day (which in essence is the same as rinsing), along with constant recirculation, I truly believe the quality will climb. So, my interpretation is that maintaining high DI resistivity essentially guarantees that you will not have much chlorine in your water. Therefore, there may not be a need to make a separate measurement for Cl-. I agree that the ppm numbers you cite below seem high. Maybe I am missing something. Regards, Larry A Rehn Technical Lab Manager AggieFab Nanofabrication Facility Texas A&M University 979 845-3199 lrehn at tamu.edu [cid:image001.jpg at 01CEC37D.FAF8C9E0] From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Morrison, Richard H., Jr Sent: Tuesday, June 04, 2019 9:25 AM To: labnetwork at mtl.mit.edu Subject: [labnetwork] DI water Question Hi, I have a question about Chloride in DI water, the E1 water spec states 0.1ppm for Chlorides, have any of your operations monitor for this? I had a test done and my6 redcir tank is 10ppm and the RO output is 25 these seem high to me, any data out there Rick Richard H. Morrison Principal Member of the Technical Staff Draper 555 Technology Square Cambridge, MA 02139-3573 Work 617-258-3420 Cell 508-930-3461 www.draper.com ________________________________ Notice: This email and any attachments may contain proprietary (Draper non-public) and/or export-controlled information of Draper. If you are not the intended recipient of this email, please immediately notify the sender by replying to this email and immediately destroy all copies of this email. ________________________________ -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.jpg Type: image/jpeg Size: 13188 bytes Desc: image001.jpg URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: DI Water Conversion Chart with pH correlation.pdf Type: application/pdf Size: 100201 bytes Desc: DI Water Conversion Chart with pH correlation.pdf URL: From lrehn at tamu.edu Tue Jun 4 16:06:27 2019 From: lrehn at tamu.edu (Rehn, Larry A) Date: Tue, 4 Jun 2019 20:06:27 +0000 Subject: [labnetwork] Name Tags on Cleanroom Gowns Message-ID: <20c22f9e607b4948b2390b603cc67779@tamu.edu> All, Do you have a good method to attach name tags for users in your labs? We have standard bunny suit garments with no pockets or loops. We use plastic name tag holders with the metal clips, but these seem to fall off rather easily when stored on hangars of the gowning racks. Best regards, Larry A Rehn Technical Lab Manager AggieFab Nanofabrication Facility Texas A&M University 979 845-3199 lrehn at tamu.edu [cid:image001.jpg at 01CEC37D.FAF8C9E0] -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.jpg Type: image/jpeg Size: 13188 bytes Desc: image001.jpg URL: From rmorrison at draper.com Wed Jun 5 12:59:44 2019 From: rmorrison at draper.com (Morrison, Richard H., Jr) Date: Wed, 5 Jun 2019 16:59:44 +0000 Subject: [labnetwork] DI water Question In-Reply-To: References: <660fc24b98d04c11a9310ba47810cc25@draper.com> Message-ID: <0717924438984df883e708fed53c9b7d@draper.com> I just had the annual PM done I will look at the DI mixed bed rinsing. Rick From: Rehn, Larry A [mailto:lrehn at tamu.edu] Sent: Tuesday, June 04, 2019 3:39 PM To: Morrison, Richard H., Jr ; labnetwork at mtl.mit.edu Subject: RE: DI water Question Hello Richard, We recently had our DI system sanitized by our service provider with a chlorine treatment. They used a chlorine color drop test indicator to determine that the chlorine was present at all drops in our system, and then later to determine that it was rinsed to around < 1ppm. After that I was told that the DI resin beds would continue to remove remaining chlorine as indicated by the resistivity getting back into the Mohm range. We experienced a gradual climb in DI resistivity from about 15Mohm up. Our service provider offered the following explanation (of why it did not immediately climb back to 18Mohm), which might be useful to answer your question about chlorine. The numbers he mentions refers to data in the attachment. ....Usually, it means that a longer rinse should have been done on the regeneration, meaning that a very slight amount of regenerant hasn't been fully rinsed out. I've attached one for the purpose of explanation. Six Megohm is equal to 83 ppb and eight megohm is equal to 63 ppb, so we're talking about t 20 ppb difference overnight. The difference between 14 megohm and 16 megohm is only 5ppb. I honestly believe that the quality will continue to climb very slowly and reach it's highest quality in the course of a few days. I've just seen this scenario too many times over the 30 plus years I've been working with high purity water. FYI: When we regenerate the resin in our shop, we recirculate DI water for typically 24-48 hours through the separately regenerated cation and anion beds before we achieve the quality we desire, prior to mixing it. I suspect very strongly that a batch didn't get fully rinsed. Going back to the paragraph above, it comes down to ppb levels of regenerants (HCl and NaOH) that haven't been fully removed through the resin. If you consume water throughout the day (which in essence is the same as rinsing), along with constant recirculation, I truly believe the quality will climb. So, my interpretation is that maintaining high DI resistivity essentially guarantees that you will not have much chlorine in your water. Therefore, there may not be a need to make a separate measurement for Cl-. I agree that the ppm numbers you cite below seem high. Maybe I am missing something. Regards, Larry A Rehn Technical Lab Manager AggieFab Nanofabrication Facility Texas A&M University 979 845-3199 lrehn at tamu.edu [cid:image001.jpg at 01CEC37D.FAF8C9E0] From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Morrison, Richard H., Jr Sent: Tuesday, June 04, 2019 9:25 AM To: labnetwork at mtl.mit.edu Subject: [labnetwork] DI water Question Hi, I have a question about Chloride in DI water, the E1 water spec states 0.1ppm for Chlorides, have any of your operations monitor for this? I had a test done and my6 redcir tank is 10ppm and the RO output is 25 these seem high to me, any data out there Rick Richard H. Morrison Principal Member of the Technical Staff Draper 555 Technology Square Cambridge, MA 02139-3573 Work 617-258-3420 Cell 508-930-3461 www.draper.com ________________________________ Notice: This email and any attachments may contain proprietary (Draper non-public) and/or export-controlled information of Draper. If you are not the intended recipient of this email, please immediately notify the sender by replying to this email and immediately destroy all copies of this email. ________________________________ ________________________________ Notice: This email and any attachments may contain proprietary (Draper non-public) and/or export-controlled information of Draper. If you are not the intended recipient of this email, please immediately notify the sender by replying to this email and immediately destroy all copies of this email. ________________________________ -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.jpg Type: image/jpeg Size: 13188 bytes Desc: image001.jpg URL: From dlafleur at cns.fas.harvard.edu Wed Jun 5 13:50:18 2019 From: dlafleur at cns.fas.harvard.edu (LaFleur, David W) Date: Wed, 5 Jun 2019 17:50:18 +0000 Subject: [labnetwork] Name Tags on Cleanroom Gowns Message-ID: <8AAA62CE-8360-463D-AEE6-495CBA530BE3@cns.fas.harvard.edu> Hi Larry, Rare earth magnets work well. Metal strip on the tag and magnets inside the garment. Best Regards, Dave LaFleur Harvard CNS [cid:image001.jpg at 01D51BA5.9CB94900] From: "labnetwork-bounces at mtl.mit.edu" on behalf of "Rehn, Larry A" Date: Tuesday, June 4, 2019 at 4:06 PM To: "labnetwork at mtl.mit.edu" Subject: [labnetwork] Name Tags on Cleanroom Gowns All, Do you have a good method to attach name tags for users in your labs? We have standard bunny suit garments with no pockets or loops. We use plastic name tag holders with the metal clips, but these seem to fall off rather easily when stored on hangars of the gowning racks. Best regards, Larry A Rehn Technical Lab Manager AggieFab Nanofabrication Facility Texas A&M University 979 845-3199 lrehn at tamu.edu [id:image001.jpg at 01CEC37D.FAF8C9E0] -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.jpg Type: image/jpeg Size: 80839 bytes Desc: image001.jpg URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image002.jpg Type: image/jpeg Size: 13189 bytes Desc: image002.jpg URL: From lrehn at tamu.edu Wed Jun 5 14:20:00 2019 From: lrehn at tamu.edu (Rehn, Larry A) Date: Wed, 5 Jun 2019 18:20:00 +0000 Subject: [labnetwork] DI water Question In-Reply-To: <0717924438984df883e708fed53c9b7d@draper.com> References: <660fc24b98d04c11a9310ba47810cc25@draper.com> <0717924438984df883e708fed53c9b7d@draper.com> Message-ID: <8a3e355cb844453db7d78ea70a7ed482@tamu.edu> You might be seeing the same thing as we are experiencing. I am still at about 17.5 Mohm and it has been about 2 weeks since they changed out every DI resin bed. But it continues to climb back to the normal 18 Mohm ever so slowly. Larry From: Morrison, Richard H., Jr [mailto:rmorrison at draper.com] Sent: Wednesday, June 05, 2019 12:00 PM To: Rehn, Larry A ; labnetwork at mtl.mit.edu Subject: RE: DI water Question I just had the annual PM done I will look at the DI mixed bed rinsing. Rick From: Rehn, Larry A [mailto:lrehn at tamu.edu] Sent: Tuesday, June 04, 2019 3:39 PM To: Morrison, Richard H., Jr >; labnetwork at mtl.mit.edu Subject: RE: DI water Question Hello Richard, We recently had our DI system sanitized by our service provider with a chlorine treatment. They used a chlorine color drop test indicator to determine that the chlorine was present at all drops in our system, and then later to determine that it was rinsed to around < 1ppm. After that I was told that the DI resin beds would continue to remove remaining chlorine as indicated by the resistivity getting back into the Mohm range. We experienced a gradual climb in DI resistivity from about 15Mohm up. Our service provider offered the following explanation (of why it did not immediately climb back to 18Mohm), which might be useful to answer your question about chlorine. The numbers he mentions refers to data in the attachment. ....Usually, it means that a longer rinse should have been done on the regeneration, meaning that a very slight amount of regenerant hasn't been fully rinsed out. I've attached one for the purpose of explanation. Six Megohm is equal to 83 ppb and eight megohm is equal to 63 ppb, so we're talking about t 20 ppb difference overnight. The difference between 14 megohm and 16 megohm is only 5ppb. I honestly believe that the quality will continue to climb very slowly and reach it's highest quality in the course of a few days. I've just seen this scenario too many times over the 30 plus years I've been working with high purity water. FYI: When we regenerate the resin in our shop, we recirculate DI water for typically 24-48 hours through the separately regenerated cation and anion beds before we achieve the quality we desire, prior to mixing it. I suspect very strongly that a batch didn't get fully rinsed. Going back to the paragraph above, it comes down to ppb levels of regenerants (HCl and NaOH) that haven't been fully removed through the resin. If you consume water throughout the day (which in essence is the same as rinsing), along with constant recirculation, I truly believe the quality will climb. So, my interpretation is that maintaining high DI resistivity essentially guarantees that you will not have much chlorine in your water. Therefore, there may not be a need to make a separate measurement for Cl-. I agree that the ppm numbers you cite below seem high. Maybe I am missing something. Regards, Larry A Rehn Technical Lab Manager AggieFab Nanofabrication Facility Texas A&M University 979 845-3199 lrehn at tamu.edu [cid:image001.jpg at 01CEC37D.FAF8C9E0] From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Morrison, Richard H., Jr Sent: Tuesday, June 04, 2019 9:25 AM To: labnetwork at mtl.mit.edu Subject: [labnetwork] DI water Question Hi, I have a question about Chloride in DI water, the E1 water spec states 0.1ppm for Chlorides, have any of your operations monitor for this? I had a test done and my6 redcir tank is 10ppm and the RO output is 25 these seem high to me, any data out there Rick Richard H. Morrison Principal Member of the Technical Staff Draper 555 Technology Square Cambridge, MA 02139-3573 Work 617-258-3420 Cell 508-930-3461 www.draper.com ________________________________ Notice: This email and any attachments may contain proprietary (Draper non-public) and/or export-controlled information of Draper. If you are not the intended recipient of this email, please immediately notify the sender by replying to this email and immediately destroy all copies of this email. ________________________________ ________________________________ Notice: This email and any attachments may contain proprietary (Draper non-public) and/or export-controlled information of Draper. If you are not the intended recipient of this email, please immediately notify the sender by replying to this email and immediately destroy all copies of this email. ________________________________ -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.jpg Type: image/jpeg Size: 13188 bytes Desc: image001.jpg URL: From derose at caltech.edu Wed Jun 5 16:50:27 2019 From: derose at caltech.edu (DeRose, Guy A.) Date: Wed, 5 Jun 2019 20:50:27 +0000 Subject: [labnetwork] Name Tags on Cleanroom Gowns In-Reply-To: <8AAA62CE-8360-463D-AEE6-495CBA530BE3@cns.fas.harvard.edu> References: <8AAA62CE-8360-463D-AEE6-495CBA530BE3@cns.fas.harvard.edu> Message-ID: <32A1F9F4-32DD-435D-92B3-08F15D67CFC1@caltech.edu> Hello group, We don?t put name tags on our garments, but rather on the hangers. We just use a ?P-Touch? labeler to make a tape with the wearer?s name on it and attach it to the neck of the hanger. It wraps around the hanger wire and sticks to itself, so they know which garments is theirs and where to find it on the garment rack when going in and out of the cleanroom. Maybe this works at Caltech inspired by Sheldon of ?The Big Bang Theory? having his spot. If you are looking for a solution with names on garments to track people on the cameras, I know of at least one lab at another university that has modified their garments such that a fabric strip with the person?s name is attached with snaps to the back like a sports jersey. Guy DeRose, PhD, Member of the Professional Staff Associate Director of Technical Operations, Kavli Nanoscience Institute California Institute of Technology, Pasadena, CA USA (O) 1-626-395-3423 (M) 1-626-676-8529 http://kni.caltech.edu From: on behalf of "LaFleur, David W" Date: Wednesday, June 5, 2019 at 1:08 PM To: "Rehn, Larry A" , "labnetwork at mtl.mit.edu" Subject: Re: [labnetwork] Name Tags on Cleanroom Gowns Hi Larry, Rare earth magnets work well. Metal strip on the tag and magnets inside the garment. Best Regards, Dave LaFleur Harvard CNS [cid:image001.jpg at 01D51BA5.9CB94900] From: "labnetwork-bounces at mtl.mit.edu" on behalf of "Rehn, Larry A" Date: Tuesday, June 4, 2019 at 4:06 PM To: "labnetwork at mtl.mit.edu" Subject: [labnetwork] Name Tags on Cleanroom Gowns All, Do you have a good method to attach name tags for users in your labs? We have standard bunny suit garments with no pockets or loops. We use plastic name tag holders with the metal clips, but these seem to fall off rather easily when stored on hangars of the gowning racks. Best regards, Larry A Rehn Technical Lab Manager AggieFab Nanofabrication Facility Texas A&M University 979 845-3199 lrehn at tamu.edu [id:image001.jpg at 01CEC37D.FAF8C9E0] -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.jpg Type: image/jpeg Size: 80840 bytes Desc: image001.jpg URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image002.jpg Type: image/jpeg Size: 13190 bytes Desc: image002.jpg URL: From sguo18 at yorku.ca Fri Jun 7 12:16:13 2019 From: sguo18 at yorku.ca (Xin (Shane) Guo) Date: Fri, 7 Jun 2019 16:16:13 +0000 Subject: [labnetwork] special canisters of DES and TEOS for PECVD Message-ID: Hi Colleagues, Does anyone know where to buy stainless steel canisters with control valves to supply DES and TEOS to PECVD? Some vendors sell DES in a bottle which cannot be connected to the PECVD. Any feedback is highly appreciated! Best Shane -------------- next part -------------- An HTML attachment was scrubbed... URL: From fnewman at uw.edu Fri Jun 7 16:22:57 2019 From: fnewman at uw.edu (Fred Newman) Date: Fri, 7 Jun 2019 13:22:57 -0700 Subject: [labnetwork] special canisters of DES and TEOS for PECVD In-Reply-To: References: Message-ID: Hi Shane, STREM sells these canisters: https://www.strem.com/catalog/accessories.php Best regards Fred On Fri, Jun 7, 2019 at 1:18 PM Xin (Shane) Guo wrote: > Hi Colleagues, > > Does anyone know where to buy stainless steel canisters with control > valves to supply DES and TEOS to PECVD? Some vendors sell DES in a bottle > which cannot be connected to the PECVD. > > Any feedback is highly appreciated! > > Best > > Shane > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork > -- Fred Newman Research Engineer Washington Nanofabrication Facility (WNF) University of Washington Fluke Hall 115, Box 352143 office 206-616-3534 mobile 505-450-4447 fnewman at uw.edu https://www.wnf.washington.edu/ -------------- next part -------------- An HTML attachment was scrubbed... URL: From tbritton at criticalsystemsinc.com Fri Jun 7 16:58:40 2019 From: tbritton at criticalsystemsinc.com (Tom Britton) Date: Fri, 7 Jun 2019 20:58:40 +0000 Subject: [labnetwork] special canisters of DES and TEOS for PECVD In-Reply-To: References: Message-ID: Hi Shane, We provide these. Can I ask what you are looking for? Tom Tom Britton Director of Sales Critical Systems, Inc. Direct: 208-890-1417 Office: 877-572-5515 Skype: tombrittoncsi www.CriticalSystemsInc.com [logo for email signature png] From: labnetwork-bounces at mtl.mit.edu On Behalf Of Xin (Shane) Guo Sent: Friday, June 07, 2019 10:16 AM To: labnetwork at mtl.mit.edu Subject: [labnetwork] special canisters of DES and TEOS for PECVD Hi Colleagues, Does anyone know where to buy stainless steel canisters with control valves to supply DES and TEOS to PECVD? Some vendors sell DES in a bottle which cannot be connected to the PECVD. Any feedback is highly appreciated! Best Shane This email and any files transmitted with it are confidential and intended solely for the use of the individual or entity to whom they are addressed. If you have received this email in error please notify the system manager. This message contains confidential information and is intended only for the individual named. If you are not the named addressee you should not disseminate, distribute or copy this e-mail. Please notify the sender immediately by e-mail if you have received this e-mail by mistake and delete this e-mail from your system. If you are not the intended recipient you are notified that disclosing, copying, distributing or taking any action in reliance on the contents of this information is strictly prohibited.'. If the disclaimer can't be applied, attach the message to a new disclaimer message. -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.png Type: image/png Size: 9501 bytes Desc: image001.png URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: Critical Systems Liquid Dispense Containers 2016.pdf Type: application/pdf Size: 633569 bytes Desc: Critical Systems Liquid Dispense Containers 2016.pdf URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: CSI Products & Services Overview Q2-19.pdf Type: application/pdf Size: 7504121 bytes Desc: CSI Products & Services Overview Q2-19.pdf URL: From fouad.karouta at anu.edu.au Fri Jun 7 18:51:04 2019 From: fouad.karouta at anu.edu.au (Fouad Karouta) Date: Fri, 7 Jun 2019 22:51:04 +0000 Subject: [labnetwork] special canisters of DES and TEOS for PECVD In-Reply-To: References: Message-ID: Hi Shane, We have an Oxford PECVD with a TEOS module and first canister we purchased from Air Products in USA and the second time we used Pegasus in UK http://pegasuschemicals.com/ Hope this helps, Fouad Karouta On 8 Jun 2019, at 6:20 am, Xin (Shane) Guo > wrote: Hi Colleagues, Does anyone know where to buy stainless steel canisters with control valves to supply DES and TEOS to PECVD? Some vendors sell DES in a bottle which cannot be connected to the PECVD. Any feedback is highly appreciated! Best Shane _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: From ahad.syed at kaust.edu.sa Sat Jun 8 03:49:15 2019 From: ahad.syed at kaust.edu.sa (Ahad Syed) Date: Sat, 8 Jun 2019 07:49:15 +0000 Subject: [labnetwork] special canisters of DES and TEOS for PECVD In-Reply-To: References: Message-ID: <1559980155219.23443@kaust.edu.sa> Hello Shane, We have an Oxford Instruments PlasmaLab100 with the TEOS module. We buy our TEOS canisters from Versum Materials. Our contact is Ramos Alonso, Aida All the best, Ahad -- Nanofabrication Core Lab 4700 King Abdullah University of Science and Technology 0247 Building 3, Thuwal 23955-6900, Kingdom of Saudi Arabia Tel:+966 12 808 2348 Visit The Core Labs ________________________________ From: labnetwork-bounces at mtl.mit.edu on behalf of Xin (Shane) Guo Sent: Friday, June 7, 2019 7:16 PM To: labnetwork at mtl.mit.edu Subject: [labnetwork] special canisters of DES and TEOS for PECVD Hi Colleagues, Does anyone know where to buy stainless steel canisters with control valves to supply DES and TEOS to PECVD? Some vendors sell DES in a bottle which cannot be connected to the PECVD. Any feedback is highly appreciated! Best Shane -------------- next part -------------- An HTML attachment was scrubbed... URL: From sguo18 at yorku.ca Mon Jun 10 10:22:23 2019 From: sguo18 at yorku.ca (Xin (Shane) Guo) Date: Mon, 10 Jun 2019 14:22:23 +0000 Subject: [labnetwork] special canisters of DES and TEOS for PECVD In-Reply-To: References: Message-ID: Hi Colleagues, Thank you all for the kind suggestions. Best Shane On Fri, Jun 7, 2019 at 12:15 PM Xin (Shane) GUO > wrote: Hi Colleagues, Does anyone know where to buy stainless steel canisters with control valves to supply DES and TEOS to PECVD? Some vendors sell DES in a bottle which cannot be connected to the PECVD. Any feedback is highly appreciated! Best Shane -------------- next part -------------- An HTML attachment was scrubbed... URL: From tbritton at criticalsystemsinc.com Mon Jun 10 11:59:38 2019 From: tbritton at criticalsystemsinc.com (Tom Britton) Date: Mon, 10 Jun 2019 15:59:38 +0000 Subject: [labnetwork] Special canisters of DES and TEOS for PECVD Message-ID: Hi Shane, We provide these. Can I ask what you are looking for? Tom Tom Britton Director of Sales Critical Systems, Inc. Direct: 208-890-1417 Office: 877-572-5515 Skype: tombrittoncsi www.CriticalSystemsInc.com [logo for email signature png] From: labnetwork-bounces at mtl.mit.edu > On Behalf Of Xin (Shane) Guo Sent: Friday, June 07, 2019 10:16 AM To: labnetwork at mtl.mit.edu Subject: [labnetwork] special canisters of DES and TEOS for PECVD Hi Colleagues, Does anyone know where to buy stainless steel canisters with control valves to supply DES and TEOS to PECVD? Some vendors sell DES in a bottle which cannot be connected to the PECVD. Any feedback is highly appreciated! Best Shane This email and any files transmitted with it are confidential and intended solely for the use of the individual or entity to whom they are addressed. If you have received this email in error please notify the system manager. This message contains confidential information and is intended only for the individual named. If you are not the named addressee you should not disseminate, distribute or copy this e-mail. Please notify the sender immediately by e-mail if you have received this e-mail by mistake and delete this e-mail from your system. If you are not the intended recipient you are notified that disclosing, copying, distributing or taking any action in reliance on the contents of this information is strictly prohibited.'. If the disclaimer can't be applied, attach the message to a new disclaimer message. -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.png Type: image/png Size: 9501 bytes Desc: image001.png URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: Critical Systems Liquid Dispense Containers 2016.pdf Type: application/pdf Size: 633569 bytes Desc: Critical Systems Liquid Dispense Containers 2016.pdf URL: From noreply at gphjournal18.xyz Tue Jun 11 07:14:53 2019 From: noreply at gphjournal18.xyz (IJO Journal) Date: Tue, 11 Jun 2019 16:44:53 +0530 Subject: [labnetwork] =?utf-8?q?Open_access_Publication_Volume=3A_2=2C_Iss?= =?utf-8?b?dWUgNuKAkyAyMDE5?= Message-ID: CALL FOR PAPER HELLO, It gives an immense pleasure to inform you that IJO JOURNAL, an International Journal in various fields of science and technology, engineering, medical, management and social studies and more invites the researcher to submit their research paper for publication. IJO journal is double blind peer reviewed journal. The review process takes maximum a week to notify the author about his/her article's status. For submission of research paper please mail us at submission at ijojournal.com for more information please visit us at _HTTP://IJOJOURNAL.COM/_ [1] If you are interested in joining editorial board please mail your CV at editor at ijojournal.com CONTACT FOR CONFERENCE AT: - info at ijojournal.com Feel free to contact us for any query. We are happy to help you. Thanks Best Regards Chief Editor You can unsubscribe at any time. Please send a message on unsubscribeijo at gmail.com Links: ------ [1] http://ijojournal.com/%0d -------------- next part -------------- An HTML attachment was scrubbed... URL: From hromans at eng.ucsd.edu Wed Jun 12 14:17:21 2019 From: hromans at eng.ucsd.edu (J Romans) Date: Wed, 12 Jun 2019 11:17:21 -0700 Subject: [labnetwork] Sputtering Mg, MgO, MgF2 Message-ID: Hello Esteemed Colleagues - Does anyone have any experience sputtering Mg, MgO, MgF2? We have a faculty inquiry and need to make a determination on risk/possible complications? Thanks, J. Hal Romans Equipment Engineer Nano3 Cleanroom Facility University of California San Diego ph: (858) 534-6674 -------------- next part -------------- An HTML attachment was scrubbed... URL: From dcchrist at wisc.edu Thu Jun 13 09:00:03 2019 From: dcchrist at wisc.edu (Daniel Christensen) Date: Thu, 13 Jun 2019 13:00:03 +0000 Subject: [labnetwork] UGIM 2020 - Save the date Message-ID: Save the date! UGIM 2020 Symposium Sunday June 14, 2020 thru noon on Wednesday June 17, 2020 University of Wisconsin-Madison Madison, WI 53706 Much more information to come in the near future. Our symposium website has not been generated yet, we will make another announcement when it is up and running. See you there! Dan Christensen and Jerry Hunter Daniel C. Christensen Nanoscale Fabrication Center University of Wisconsin-Madison 1550 Engineering Dr Madison, WI 53706 -------------- next part -------------- An HTML attachment was scrubbed... URL: From rmorrison at draper.com Thu Jun 13 10:13:26 2019 From: rmorrison at draper.com (Morrison, Richard H., Jr) Date: Thu, 13 Jun 2019 14:13:26 +0000 Subject: [labnetwork] temperature measurment Message-ID: <8b8601f09d594b26bb2b6fc2d45dc486@draper.com> Hi Everyone, For years we have been using Sensarray wafers to calibrate ovens and hot plates, the issue is they are silicon and break and at $7.5 k each it is costing me $$ I would rather spend on other things. Do any of you use temperature calibration wafers made of metal, I used to get anodized Aluminum wafers for handler issues from H2Squared are they still around? Thanks Rick Richard H. Morrison Principal Member of the Technical Staff Draper 555 Technology Square Cambridge, MA 02139-3573 Work 617-258-3420 Cell 508-930-3461 www.draper.com ________________________________ Notice: This email and any attachments may contain proprietary (Draper non-public) and/or export-controlled information of Draper. If you are not the intended recipient of this email, please immediately notify the sender by replying to this email and immediately destroy all copies of this email. ________________________________ -------------- next part -------------- An HTML attachment was scrubbed... URL: From james.beall at nist.gov Thu Jun 13 13:09:35 2019 From: james.beall at nist.gov (Beall, James A. (Fed)) Date: Thu, 13 Jun 2019 17:09:35 +0000 Subject: [labnetwork] temperature measurment In-Reply-To: <8b8601f09d594b26bb2b6fc2d45dc486@draper.com> References: <8b8601f09d594b26bb2b6fc2d45dc486@draper.com> Message-ID: <04610CA8-2C47-4E5E-BD26-8713DD492786@nist.gov> My latest contact with H-Squared was: Uina Kubota H-Square Corporation 3100 Patrick Henry Dr. Santa Clara CA 95054 Tel: 1-408-982-9108x104 ukubota at h-square.com www.h-square.com Jim Beall National Institute of Standards and Technology Quantum Sensors Group Mailcode 687.08 325 Broadway 1C-110 Boulder, CO 80305-3328 303-497-5989 303-497-3042 (fax) On Jun 13, 2019, at 8:13 AM, Morrison, Richard H., Jr > wrote: Hi Everyone, For years we have been using Sensarray wafers to calibrate ovens and hot plates, the issue is they are silicon and break and at $7.5 k each it is costing me $$ I would rather spend on other things. Do any of you use temperature calibration wafers made of metal, I used to get anodized Aluminum wafers for handler issues from H2Squared are they still around? Thanks Rick Richard H. Morrison Principal Member of the Technical Staff Draper 555 Technology Square Cambridge, MA 02139-3573 Work 617-258-3420 Cell 508-930-3461 www.draper.com ________________________________ Notice: This email and any attachments may contain proprietary (Draper non-public) and/or export-controlled information of Draper. If you are not the intended recipient of this email, please immediately notify the sender by replying to this email and immediately destroy all copies of this email. ________________________________ _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://gcc01.safelinks.protection.outlook.com/?url=https%3A%2F%2Fmtl.mit.edu%2Fmailman%2Flistinfo.cgi%2Flabnetwork&data=02%7C01%7Cjames.beall%40nist.gov%7Cc80e1a5b2b874d12a4cc08d6f015e72c%7C2ab5d82fd8fa4797a93e054655c61dec%7C1%7C0%7C636960374198045175&sdata=%2FlGB1IiOY8YgxN6qLDROQm89N70h1c1hnLoGsNggyDM%3D&reserved=0 -------------- next part -------------- An HTML attachment was scrubbed... URL: From jim at photomaskportal.com Thu Jun 13 15:20:03 2019 From: jim at photomaskportal.com (Jim Carroll) Date: Thu, 13 Jun 2019 15:20:03 -0400 Subject: [labnetwork] temperature measurment In-Reply-To: <8b8601f09d594b26bb2b6fc2d45dc486@draper.com> References: <8b8601f09d594b26bb2b6fc2d45dc486@draper.com> Message-ID: Hi Rick, We buy mask handling equipment from H-Square. Is that who you are thinking of? http://www.h-square.com/ Jim Carroll PhotomaskPORTAL On Jun 13, 2019, at 9:13 AM, Morrison, Richard H., Jr wrote: Hi Everyone, For years we have been using Sensarray wafers to calibrate ovens and hot plates, the issue is they are silicon and break and at $7.5 k each it is costing me $$ I would rather spend on other things. Do any of you use temperature calibration wafers made of metal, I used to get anodized Aluminum wafers for handler issues from H2Squared are they still around? Thanks Rick Richard H. Morrison Principal Member of the Technical Staff Draper 555 Technology Square Cambridge, MA 02139-3573 Work 617-258-3420 Cell 508-930-3461 www.draper.com ------------------------------ Notice: This email and any attachments may contain proprietary (Draper non-public) and/or export-controlled information of Draper. If you are not the intended recipient of this email, please immediately notify the sender by replying to this email and immediately destroy all copies of this email. ------------------------------ _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: From ahryciw at ualberta.ca Fri Jun 14 12:27:36 2019 From: ahryciw at ualberta.ca (Aaron Hryciw) Date: Fri, 14 Jun 2019 10:27:36 -0600 Subject: [labnetwork] Ag-Bi-Zn films in high vacuum Message-ID: Dear colleagues, I'm hoping you can offer some advice on the wisdom of allowing substrates coated with a Zn-containing thin film into some of our open-access facility's shared high vacuum systems. The material in question is a Ag-Bi-Zn alloy (98.9 at% Ag, 0.1 at% Bi, 1.0 at% Zn), which one of our user groups wants to deposit as a cathode layer for an OLED process. A 50 nm thick film of this material is being deposited elsewhere (not in our shared equipment). However, they would like to continue processing this film in our equipment, which includes two vacuum deposition steps (thermal evaporation for organics, then sputtering for the anode layer, patterned via lift-off). Both of these deposition processes nominally have the substrate at room temperature, although there is no active substrate cooling to offset heating from the process itself. The maximum exposed area of the Ag-Bi-Zn would be ~20 cm? (i.e., equivalent of a quarter of a 4" wafer). I understand that due to its high vapour pressure, having metallic Zn is generally a bad idea in shared systems if you want Zn-free films thereafter, but I wanted to discuss how likely we'd be to see Zn contamination in either the organics evaporator or the sputtering tool, given the relatively low (?) total volume of exposed Zn that would be involved. I am mostly concerned about the organics evaporator, as the Ag-Bi-Zn film would be exposed to high vacuum; in the sputtering tool, the Ag-Bi-Zn would already be coated by several layers of organics. Could contamination risk be mitigated by copious foiling/shielding of the chamber? Does anyone have any recommendations for a (non-destructive) test that could be done to determine the extent of Zn contamination during the organics evaporation step? Any advice would be greatly appreciated. Many thanks. Cheers, ? Aaron Aaron Hryciw, PhD, PEng Fabrication Group Manager University of Alberta - nanoFAB W1-060 ECERF Building 9107 - 116 Street Edmonton, Alberta Canada T6G 2V4 Ph: 780-940-7938 www.nanofab.ualberta.ca -------------- next part -------------- An HTML attachment was scrubbed... URL: From wrightsh at lnf.umich.edu Tue Jun 18 15:41:25 2019 From: wrightsh at lnf.umich.edu (Shawn Wright) Date: Tue, 18 Jun 2019 15:41:25 -0400 Subject: [labnetwork] Cassette Loading Spinner/Developer Tools Message-ID: Hi Everyone, We?re currently looking at replacing our SUSS ACS200 cluster tool, it?s showing it?s age and SUSS no longer supports it for replacement parts. It would be good to hear what people are running for automated spin coating and developing, especially any experience with reliability and vendor support. Specifically we?re looking for tools that can do cassette loading for 4? and 6? wafers that can do vapor prime, photoresist spinning, bake, and aqueous develop. Thanks for your help, Shawn Wright Lead Engineer in Research Lurie Nanofabrication Facility University of Michigan From fouad.karouta at anu.edu.au Tue Jun 18 20:29:44 2019 From: fouad.karouta at anu.edu.au (Fouad Karouta) Date: Wed, 19 Jun 2019 00:29:44 +0000 Subject: [labnetwork] Clean Room Ceiling Message-ID: Dear all, We are in the process of getting a new building that includes a 650 m2 clean room (class 100 and 1000) with a sub-fab to host chiller, pumps, scrubbers, etc. We are looking to possible suppliers of clean room compartmented ceiling to host light, Hepa filters and others like fire sprinklers, etc. Any suggestions for suppliers are welcome. Please also comment on what to include or avoid in ceiling. Thanks, Fouad Karouta ************************************* Manager ANFF ACT Node Australian National Fabrication Facility Research School of Physics and Engineering L. Huxley Building (#56), Mills Road, Room 4.02 Australian National University ACT 2601, Canberra, Australia Tel: + 61 2 6125 7174 Mob: + 61 451 046 412 Email: fouad.karouta at anu.edu.au http://anff-act.anu.edu.au/ -------------- next part -------------- An HTML attachment was scrubbed... URL: From kpayer at mit.edu Tue Jun 18 20:34:31 2019 From: kpayer at mit.edu (Kristofor Robert Payer) Date: Wed, 19 Jun 2019 00:34:31 +0000 Subject: [labnetwork] Cassette Loading Spinner/Developer Tools In-Reply-To: References: Message-ID: Hi Shawn, We (MTL) bought a coater/developer unit from PicoTrack a couple of years ago and have been very happy with the system. Its features were comparable to the Suss cluster tool. They have a highly customizable interface, easy to use software and great support. If I remember correctly PicoTrack was founded by former SVG engineers. I'm happy to answer any specific questions you might have on the system, just send me an email or give me a call offline. Kris Kristofor Payer MIT.nano Massachusetts Institute of Technology 60 Vassar Street 12-4011 Cambridge, MA 02139 617-324-1469 kpayer at mit.edu ?On 6/18/19, 8:18 PM, "labnetwork-bounces at mtl.mit.edu on behalf of Shawn Wright" wrote: Hi Everyone, We?re currently looking at replacing our SUSS ACS200 cluster tool, it?s showing it?s age and SUSS no longer supports it for replacement parts. It would be good to hear what people are running for automated spin coating and developing, especially any experience with reliability and vendor support. Specifically we?re looking for tools that can do cassette loading for 4? and 6? wafers that can do vapor prime, photoresist spinning, bake, and aqueous develop. Thanks for your help, Shawn Wright Lead Engineer in Research Lurie Nanofabrication Facility University of Michigan _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork From mweiler at andrew.cmu.edu Wed Jun 19 10:11:39 2019 From: mweiler at andrew.cmu.edu (Mark Weiler) Date: Wed, 19 Jun 2019 14:11:39 +0000 Subject: [labnetwork] Clean Room Ceiling In-Reply-To: References: Message-ID: Hi Fouad, Take a look at Plascore and their product line? https://www.plascore.com/honeycomb/cleanrooms/semiconductor-system/ceiling-panels/ All the best, Mark ________________________________________________________________ Mark Weiler Equipment & Facilites Manager Clair and John Bertucci Nanotechnology Laboratory Eden Hall Nanofabrication Cleanroom Carnegie Mellon University P: 412-268-2471 http://www.nanofab.ece.cmu.edu [cid:2D2E01E3-CEC1-4F48-A845-224D8D7CED12 at wv.cc.cmu.edu] On Jun 18, 2019, at 8:29 PM, Fouad Karouta > wrote: Dear all, We are in the process of getting a new building that includes a 650 m2 clean room (class 100 and 1000) with a sub-fab to host chiller, pumps, scrubbers, etc. We are looking to possible suppliers of clean room compartmented ceiling to host light, Hepa filters and others like fire sprinklers, etc. Any suggestions for suppliers are welcome. Please also comment on what to include or avoid in ceiling. Thanks, Fouad Karouta ************************************* Manager ANFF ACT Node Australian National Fabrication Facility Research School of Physics and Engineering L. Huxley Building (#56), Mills Road, Room 4.02 Australian National University ACT 2601, Canberra, Australia Tel: + 61 2 6125 7174 Mob: + 61 451 046 412 Email: fouad.karouta at anu.edu.au http://anff-act.anu.edu.au/ _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: images.png Type: image/png Size: 720 bytes Desc: images.png URL: From nnelsonfitzpatrick at uwaterloo.ca Wed Jun 19 12:37:33 2019 From: nnelsonfitzpatrick at uwaterloo.ca (Nathan Nelson - Fitzpatrick) Date: Wed, 19 Jun 2019 16:37:33 +0000 Subject: [labnetwork] Cleaning reactive deposition materials prior to install in UHV tool Message-ID: <4434FE40-9444-437C-B40D-22CF08B53857@uwaterloo.ca> Hi Everyone, I received a question from another lab on campus here and I?m not sure of the correct advice, I was hoping to rely on this experienced community. This lab in question is interested in using reactive metal materials (Ba, Sr, etc..) in one of their UHV deposition tools. Since these materials are reactive in air, they were shipped immersed in an oil. They were asking me about the best way to clean off the oil prior to installing the material in their deposition system. I believe they have already tried IPA and were not too happy with its effectiveness. The vendor recommended ?100% mineral spirits?, but a quick google search indicates that this is a sort of vaguely defined solution of short chain hydrocarbons (alkanes?). I was wondering if anyone has experience they could share regarding this type of problem and how they cleaned their materials? Thank you in advance, -Nathan -- Nathan Nelson-Fitzpatrick PhD Nanofabrication Process & Characterization Engineering Manager Quantum-Nano Fabrication and Characterization Facility (QNFCF) Office of Research University of Waterloo 200 University Avenue West, Waterloo, ON N2L 3G1 P: 519-888-4567 ext. 31796 C: 226-218-3206 https://fab.qnc.uwaterloo.ca [/Users/nnelsonf/Desktop/university-of-waterloo-logo-esig.png] -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.png Type: image/png Size: 11994 bytes Desc: image001.png URL: From sean.rinehart at yale.edu Wed Jun 19 13:15:26 2019 From: sean.rinehart at yale.edu (Rinehart, Sean) Date: Wed, 19 Jun 2019 17:15:26 +0000 Subject: [labnetwork] Argon in a BCl3 cylinder Message-ID: Hello all, Recently our facility's BCl3 cylinder had been charged with argon, increasing the delivery pressure to 25 psi. The cause has been corrected, but now I'm evaluating whether or not the cylinder needs to be replaced, and would appreciate some advice. Since the root issue was fixed, the pressure has slowly dropped from 25 psi to 15 psi with use. Based on this, it seems like with more use the argon should completely deplete from the cylinder, but maybe someone can tell me if that's actually true, and if there's a way to evaluate gas concentrations. Additionally, the plasma glow is whitish instead of pink, so there's definitely BCl3 content in the delivered gas. Thank you, Sean Rinehart Assistant Director, Cleanroom Operations Yale University 203 432 4303 -------------- next part -------------- An HTML attachment was scrubbed... URL: From Dwayne.Chrusch at umanitoba.ca Wed Jun 19 14:28:14 2019 From: Dwayne.Chrusch at umanitoba.ca (Dwayne Chrusch) Date: Wed, 19 Jun 2019 18:28:14 +0000 Subject: [labnetwork] Clean Room Ceiling In-Reply-To: References: Message-ID: <9876E680-AAAF-46EC-95E9-A1419DE41FDF@umanitoba.ca> Hello, A few years ago, we bought a self standing soft wall cleanroom from Terra Universal. Link; https://www.terrauniversal.com/cleanrooms/valuline-softwall-cleanroom.php The ceiling is designed around 2? x 4? (610mm x 1219mm) panels (fans, lights, etc..). The panels are mounted similar to a drop/suspended ceiling that are popular in basements here in Canada/USA. Blank panels could easily be used for sprinkler heads. Terra sells complete (customizable) kits, and/or individual units. Thus, you could simply order the ceiling panels you need, and construct the room in-house. Their distributer in Australia is; Arden Pty. Ltd. Australia Jeff Newing www.arden.com.au Phone: + 61 3 9706 4023 Fax: + 61 3 9706 4034 Email: jeff at arden.com.au Regards, Dwayne D. Chrusch, MSc. Operations Manager, Training & Safety Coordinator Nanosystem Fabrication Lab (NSFL) E3-487 Engineering 75A Chancellor's Circle University of Manitoba Winnipeg, Manitoba Canada, R3T 5V6 Tel: +1 (204) 474-8246 Fax +1 (204) 261-4639 http://home.cc.umanitoba.ca/~shafaic/NSFL_site/NSFL_Home.html > On Jun 18, 2019, at 7:29 PM, Fouad Karouta wrote: > > Dear all, > > We are in the process of getting a new building that includes a 650 m2 clean room (class 100 and 1000) with a sub-fab to host chiller, pumps, scrubbers, etc. > > We are looking to possible suppliers of clean room compartmented ceiling to host light, Hepa filters and others like fire sprinklers, etc. > Any suggestions for suppliers are welcome. > Please also comment on what to include or avoid in ceiling. > > Thanks, > Fouad Karouta > > ************************************* > Manager ANFF ACT Node > Australian National Fabrication Facility > Research School of Physics and Engineering > L. Huxley Building (#56), Mills Road, Room 4.02 > Australian National University > ACT 2601, Canberra, Australia > Tel: + 61 2 6125 7174 > Mob: + 61 451 046 412 > Email: fouad.karouta at anu.edu.au > http://anff-act.anu.edu.au/ > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork From spaolini at cns.fas.harvard.edu Wed Jun 19 16:40:53 2019 From: spaolini at cns.fas.harvard.edu (Paolini, Steven) Date: Wed, 19 Jun 2019 20:40:53 +0000 Subject: [labnetwork] Argon in a BCl3 cylinder In-Reply-To: References: Message-ID: Sean, I am not entirely sure, but it's not likely that the gases will completely disassociate from each other, even if the pressure drops. I'm assuming that because the vapor pressure of BCL3 is a constant because it is a liquid, you are determining that mostly Ar is being consumed because of the pressure drop. I would recommend a complete pump and purge of the entire gas circuit and replacing the cylinder of BCL3. The simplest way to evaluate the gas concentrations is with an RGA but I don't think it's a good idea to expose the filament to a corrosive such as BCL3 Steve "Equipment Dood" Paolini Steve Paolini Principal Equipment Engineer Harvard University Center for Nanoscale Systems 11 Oxford St. Cambridge, MA 02138 617- 496- 9816 spaolini at cns.fas.harvard.edu www.cns.fas.harvard.edu From: labnetwork-bounces at mtl.mit.edu On Behalf Of Rinehart, Sean Sent: Wednesday, June 19, 2019 1:15 PM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Argon in a BCl3 cylinder Hello all, Recently our facility's BCl3 cylinder had been charged with argon, increasing the delivery pressure to 25 psi. The cause has been corrected, but now I'm evaluating whether or not the cylinder needs to be replaced, and would appreciate some advice. Since the root issue was fixed, the pressure has slowly dropped from 25 psi to 15 psi with use. Based on this, it seems like with more use the argon should completely deplete from the cylinder, but maybe someone can tell me if that's actually true, and if there's a way to evaluate gas concentrations. Additionally, the plasma glow is whitish instead of pink, so there's definitely BCl3 content in the delivered gas. Thank you, Sean Rinehart Assistant Director, Cleanroom Operations Yale University 203 432 4303 -------------- next part -------------- An HTML attachment was scrubbed... URL: From wrhess at mit.edu Wed Jun 19 16:47:08 2019 From: wrhess at mit.edu (Whitney Rochelle Hess) Date: Wed, 19 Jun 2019 20:47:08 +0000 Subject: [labnetwork] Cleaning reactive deposition materials prior to install in UHV tool In-Reply-To: <4434FE40-9444-437C-B40D-22CF08B53857@uwaterloo.ca> References: <4434FE40-9444-437C-B40D-22CF08B53857@uwaterloo.ca> Message-ID: <1560977227654.96584@mit.edu> Hi Nathan, Cleaning with anhydrous hexane should work. This should be done in an inert atmosphere glovebox. Best, Whitney _________________________________________ Whitney Hess, PhD EHS Coordinator MIT.nano | Microsystems Technology Laboratories MIT, Room 39-213 Phone: 617-253-8567 Email: wrhess at mit.edu ________________________________ From: labnetwork-bounces at mtl.mit.edu on behalf of Nathan Nelson - Fitzpatrick Sent: Wednesday, June 19, 2019 12:37 PM To: labnetwork at mtl.mit.edu Cc: Saba Sadeghi Subject: [labnetwork] Cleaning reactive deposition materials prior to install in UHV tool Hi Everyone, I received a question from another lab on campus here and I'm not sure of the correct advice, I was hoping to rely on this experienced community. This lab in question is interested in using reactive metal materials (Ba, Sr, etc..) in one of their UHV deposition tools. Since these materials are reactive in air, they were shipped immersed in an oil. They were asking me about the best way to clean off the oil prior to installing the material in their deposition system. I believe they have already tried IPA and were not too happy with its effectiveness. The vendor recommended "100% mineral spirits", but a quick google search indicates that this is a sort of vaguely defined solution of short chain hydrocarbons (alkanes?). I was wondering if anyone has experience they could share regarding this type of problem and how they cleaned their materials? Thank you in advance, -Nathan -- Nathan Nelson-Fitzpatrick PhD Nanofabrication Process & Characterization Engineering Manager Quantum-Nano Fabrication and Characterization Facility (QNFCF) Office of Research University of Waterloo 200 University Avenue West, Waterloo, ON N2L 3G1 P: 519-888-4567 ext. 31796 C: 226-218-3206 https://fab.qnc.uwaterloo.ca [/Users/nnelsonf/Desktop/university-of-waterloo-logo-esig.png] -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.png Type: image/png Size: 11994 bytes Desc: image001.png URL: From Dwayne.Chrusch at umanitoba.ca Wed Jun 19 17:17:02 2019 From: Dwayne.Chrusch at umanitoba.ca (Dwayne Chrusch) Date: Wed, 19 Jun 2019 21:17:02 +0000 Subject: [labnetwork] Argon in a BCl3 cylinder In-Reply-To: References: Message-ID: Hello, If you?re using a 2 stage (or multi-stage) regulator, only the second stage (regulator output) and intermediate stage chambers are contaminated. Here?s a link to how a 2 stage regulator works; https://www.youtube.com/watch?v=UfaucxS5rIc In order to flow gas from the intermediate chamber to the secondary chamber, the secondary pressure must be lower than the intermediate pressure. If the pressure in the secondary is greater than the intermediate, the orifice between the intermediate chamber and the gas cylinder would remain closed. You should be able to purge the contaminated gas out. First, close off the cylinder. Next, use your vacuum system to evacuate the gas from the system. Let the MFC drop to zero flow. Set the gas flow rate to as high as safely possible. This will pull the contaminated gas out of the low side and intermediate chamber of the regulator (assuming it?s dual stage). In theory, there will always be some stubborn Ar gas, but it may be practical to purge ?enough?. You could always recharge the gas line, and then re-purge it. Regards, Dwayne D. Chrusch, MSc. Operations Manager, Training & Safety Coordinator Nanosystem Fabrication Lab (NSFL) E3-487 Engineering 75A Chancellor's Circle University of Manitoba Winnipeg, Manitoba Canada, R3T 5V6 Tel: +1 (204) 474-8246 Fax +1 (204) 261-4639 http://home.cc.umanitoba.ca/~shafaic/NSFL_site/NSFL_Home.html > On Jun 19, 2019, at 12:15 PM, Rinehart, Sean wrote: > > Hello all, > > Recently our facility?s BCl3 cylinder had been charged with argon, increasing the delivery pressure to 25 psi. The cause has been corrected, but now I?m evaluating whether or not the cylinder needs to be replaced, and would appreciate some advice. > > Since the root issue was fixed, the pressure has slowly dropped from 25 psi to 15 psi with use. Based on this, it seems like with more use the argon should completely deplete from the cylinder, but maybe someone can tell me if that?s actually true, and if there?s a way to evaluate gas concentrations. Additionally, the plasma glow is whitish instead of pink, so there?s definitely BCl3 content in the delivered gas. > > Thank you, > > Sean Rinehart > Assistant Director, Cleanroom Operations > Yale University > 203 432 4303 > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork From julia.aebersold at louisville.edu Wed Jun 19 17:54:44 2019 From: julia.aebersold at louisville.edu (Aebersold,Julia W.) Date: Wed, 19 Jun 2019 21:54:44 +0000 Subject: [labnetwork] Argon in a BCl3 cylinder In-Reply-To: References: Message-ID: Hi Sean. I'm not sure if what I have to offer can help, but our cylinder is 100% BCL3 with no mixed argon. We also have heat tape on all the gas delivery lines for it loves to condense at ambient temps. We have noticed that during an oxygen plasma clean in the processing chamber using BCl3 that if the color is not a nice blue then we still have BCl3 trapped in the primary delivery gas line to the chamber. If argon was injected into the cylinder then I would replace it for the concentration may be enough throw off the actual flow rate form your MFC. Another tid bit is that we discovered after processing with BCl3 the photoresist chemistry changes and hardens and is extremely difficult to remove. We are testing with Dyanstrip to see if that will remove the resist. Cheers! Julia Aebersold, Ph.D. MNTC Cleanroom Manager University of Louisville 2210 South Brook Street Shumaker Research Building, Room 233 Louisville, KY 40292 (502) 852-1572 http://louisville.edu/micronano/ ________________________________ From: labnetwork-bounces at mtl.mit.edu on behalf of Rinehart, Sean Sent: Wednesday, June 19, 2019 1:15:26 PM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Argon in a BCl3 cylinder Hello all, Recently our facility?s BCl3 cylinder had been charged with argon, increasing the delivery pressure to 25 psi. The cause has been corrected, but now I?m evaluating whether or not the cylinder needs to be replaced, and would appreciate some advice. Since the root issue was fixed, the pressure has slowly dropped from 25 psi to 15 psi with use. Based on this, it seems like with more use the argon should completely deplete from the cylinder, but maybe someone can tell me if that?s actually true, and if there?s a way to evaluate gas concentrations. Additionally, the plasma glow is whitish instead of pink, so there?s definitely BCl3 content in the delivered gas. Thank you, Sean Rinehart Assistant Director, Cleanroom Operations Yale University 203 432 4303 -------------- next part -------------- An HTML attachment was scrubbed... URL: From fouad.karouta at anu.edu.au Wed Jun 19 19:44:39 2019 From: fouad.karouta at anu.edu.au (Fouad Karouta) Date: Wed, 19 Jun 2019 23:44:39 +0000 Subject: [labnetwork] Argon in a BCl3 cylinder In-Reply-To: References: Message-ID: Hi Sean, I rather agree with Steven Paolini and you should replace the BCl3 cylinder. As the BCl3 is liquid an exposure to another gas like Ar at a pressure > 15 psi would lead to Ar going into the cylinder. We did have a similar issue in the past with a different gas and after solving the leak issue the BCl3 never behaved as before and especially etching results with BCl3 were no longer satisfactory as before the leak. We ended up replacing the cylinder. Kind regards, Fouad ************************************* Manager ANFF ACT Node Australian National Fabrication Facility Research School of Physics and Engineering L. Huxley Building (#56), Mills Road, Room 4.02 Australian National University ACT 2601, Canberra, Australia Tel: + 61 2 6125 7174 Mob: + 61 451 046 412 Email: fouad.karouta at anu.edu.au http://anff-act.anu.edu.au/ -----Original Message----- From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Dwayne Chrusch Sent: Thursday, 20 June 2019 7:17 AM To: Rinehart, Sean Cc: labnetwork at mtl.mit.edu Subject: Re: [labnetwork] Argon in a BCl3 cylinder Hello, If you?re using a 2 stage (or multi-stage) regulator, only the second stage (regulator output) and intermediate stage chambers are contaminated. Here?s a link to how a 2 stage regulator works; https://www.youtube.com/watch?v=UfaucxS5rIc In order to flow gas from the intermediate chamber to the secondary chamber, the secondary pressure must be lower than the intermediate pressure. If the pressure in the secondary is greater than the intermediate, the orifice between the intermediate chamber and the gas cylinder would remain closed. You should be able to purge the contaminated gas out. First, close off the cylinder. Next, use your vacuum system to evacuate the gas from the system. Let the MFC drop to zero flow. Set the gas flow rate to as high as safely possible. This will pull the contaminated gas out of the low side and intermediate chamber of the regulator (assuming it?s dual stage). In theory, there will always be some stubborn Ar gas, but it may be practical to purge ?enough?. You could always recharge the gas line, and then re-purge it. Regards, Dwayne D. Chrusch, MSc. Operations Manager, Training & Safety Coordinator Nanosystem Fabrication Lab (NSFL) E3-487 Engineering 75A Chancellor's Circle University of Manitoba Winnipeg, Manitoba Canada, R3T 5V6 Tel: +1 (204) 474-8246 Fax +1 (204) 261-4639 http://home.cc.umanitoba.ca/~shafaic/NSFL_site/NSFL_Home.html > On Jun 19, 2019, at 12:15 PM, Rinehart, Sean wrote: > > Hello all, > > Recently our facility?s BCl3 cylinder had been charged with argon, increasing the delivery pressure to 25 psi. The cause has been corrected, but now I?m evaluating whether or not the cylinder needs to be replaced, and would appreciate some advice. > > Since the root issue was fixed, the pressure has slowly dropped from 25 psi to 15 psi with use. Based on this, it seems like with more use the argon should completely deplete from the cylinder, but maybe someone can tell me if that?s actually true, and if there?s a way to evaluate gas concentrations. Additionally, the plasma glow is whitish instead of pink, so there?s definitely BCl3 content in the delivered gas. > > Thank you, > > Sean Rinehart > Assistant Director, Cleanroom Operations Yale University > 203 432 4303 > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork From mmoneck at andrew.cmu.edu Wed Jun 19 20:45:33 2019 From: mmoneck at andrew.cmu.edu (Matthew Moneck) Date: Thu, 20 Jun 2019 00:45:33 +0000 Subject: [labnetwork] Clean Room Ceiling In-Reply-To: References: Message-ID: <63c9b617700d4cf1b1c9e968709c4f0d@andrew.cmu.edu> Hi Fouad, We have a Cleanpak ceiling plenum system from Nortek/Hunt Air (http://www.nortekair.com/product/cleanpak-custom-cleanroom-systems/) for our 8,500 sqft Class 10/100 cleanroom. The plenum comes in prefab sections containing the grid for gel mount filter assemblies, integrated piping for fire sprinklers, LED light bars (located in between the filters as opposed to replacing filter locations), and safety shower plumbing/mounts. Above the plenum grid, we have recirculating air handlers mounted to the plenums, also from Hunt Air. Best Regards, Matt -- Matthew T. Moneck, Ph.D. Executive Manager, Claire & John Bertucci Nanotechnology Laboratory Electrical and Computer Engineering | Carnegie Mellon University 5000 Forbes Ave., Pittsburgh, PA 15213-3890 T: 412.268.5430 F: 412.268.3497 www.ece.cmu.edu nanofab.ece.cmu.edu From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Fouad Karouta Sent: Tuesday, June 18, 2019 8:30 PM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Clean Room Ceiling Dear all, We are in the process of getting a new building that includes a 650 m2 clean room (class 100 and 1000) with a sub-fab to host chiller, pumps, scrubbers, etc. We are looking to possible suppliers of clean room compartmented ceiling to host light, Hepa filters and others like fire sprinklers, etc. Any suggestions for suppliers are welcome. Please also comment on what to include or avoid in ceiling. Thanks, Fouad Karouta ************************************* Manager ANFF ACT Node Australian National Fabrication Facility Research School of Physics and Engineering L. Huxley Building (#56), Mills Road, Room 4.02 Australian National University ACT 2601, Canberra, Australia Tel: + 61 2 6125 7174 Mob: + 61 451 046 412 Email: fouad.karouta at anu.edu.au http://anff-act.anu.edu.au/ -------------- next part -------------- An HTML attachment was scrubbed... URL: From shott at stanford.edu Wed Jun 19 20:47:56 2019 From: shott at stanford.edu (John D Shott) Date: Thu, 20 Jun 2019 00:47:56 +0000 Subject: [labnetwork] Argon in a BCl3 cylinder In-Reply-To: References: , Message-ID: Sean: Just to add a bit to the others comments. My guess was that, when full, this was something like a 5 or a 9 lb cylinder ? or maybe 2.5 or 4 kg equivalent. In other words, many moles ? How much Ar did you add? The head volume of the cylinder at a pressure of 25 PSIG minus the vapor pressure of the BCl3. I?m guessing that is a fraction of a mole. Of course, in the vapor it is the partial pressure of the two components that determines the initial mixture you are extracting. However, as you extract gas, argon leaves permanently, whereas the BCl3 is replaced by vaporization. So, in principle, it will get more and more pure ? but never 100%. By estimating the total flow of gas, you should be able to estimate when you get to 99% BCl3 in the vapor, or 99.9%, or whatever you?d like. For example, starting with equal vapor pressures of Ar and BCl3 (that is a 50% mixture), if you extract one ?head volume? of gas, you should be down to 25% Ar. One more head volume of flow, you should be down to 12.5%, etc. How can you make that happen faster? By chilling the BCl3 so that the Ar is initially the more dominant gas in the mixture and thus preferentially removed. At least that is the first order analysis I would consider as an old timer ? Good luck. A word of caution: it might be tempting to put the cylinder in a tub of ice water to reduce the BCl3 vapor pressure and, as a result increase Ar extraction. Remember that BCl3 is water reactive and an external leak of BCl3 would react to form HCl. Why do I think of that? When I was young ? we are talking 1972 or 1973, I think ? we had a ?Coke syrup canister? full of 50 lbs of SiCl4 for an old epitaxial reactor. To keep the deposition rate constant during a long run, the canister lived in a tub of water to minimize vaporization cooling of the SiCl4. One night the canister developed a leak so that the remaining SiCl4 liquid reacted with the water in the tub and released a whole lot of HCl that corroded everything in sight. Fortunately, this all happened in the middle of the night and we did have a crude scrubber ? so nobody was exposed to or injured by HCl. It is, however, not a scenario you want to replicate. Have a good day. John Sent from my iPhone > On Jun 19, 2019, at 4:40 PM, Dwayne Chrusch wrote: > > Hello, > > If you?re using a 2 stage (or multi-stage) regulator, only the second stage (regulator output) and intermediate stage chambers are contaminated. > > Here?s a link to how a 2 stage regulator works; > https://www.youtube.com/watch?v=UfaucxS5rIc > > In order to flow gas from the intermediate chamber to the secondary chamber, the secondary pressure must be lower than the intermediate pressure. If the pressure in the secondary is greater than the intermediate, the orifice between the intermediate chamber and the gas cylinder would remain closed. > > You should be able to purge the contaminated gas out. First, close off the cylinder. Next, use your vacuum system to evacuate the gas from the system. Let the MFC drop to zero flow. Set the gas flow rate to as high as safely possible. This will pull the contaminated gas out of the low side and intermediate chamber of the regulator (assuming it?s dual stage). In theory, there will always be some stubborn Ar gas, but it may be practical to purge ?enough?. You could always recharge the gas line, and then re-purge it. > > > Regards, > > Dwayne D. Chrusch, MSc. > Operations Manager, Training & Safety Coordinator > Nanosystem Fabrication Lab (NSFL) > E3-487 Engineering > 75A Chancellor's Circle > University of Manitoba > Winnipeg, Manitoba > Canada, R3T 5V6 > Tel: +1 (204) 474-8246 > Fax +1 (204) 261-4639 > http://home.cc.umanitoba.ca/~shafaic/NSFL_site/NSFL_Home.html > > >> On Jun 19, 2019, at 12:15 PM, Rinehart, Sean wrote: >> >> Hello all, >> >> Recently our facility?s BCl3 cylinder had been charged with argon, increasing the delivery pressure to 25 psi. The cause has been corrected, but now I?m evaluating whether or not the cylinder needs to be replaced, and would appreciate some advice. >> >> Since the root issue was fixed, the pressure has slowly dropped from 25 psi to 15 psi with use. Based on this, it seems like with more use the argon should completely deplete from the cylinder, but maybe someone can tell me if that?s actually true, and if there?s a way to evaluate gas concentrations. Additionally, the plasma glow is whitish instead of pink, so there?s definitely BCl3 content in the delivered gas. >> >> Thank you, >> >> Sean Rinehart >> Assistant Director, Cleanroom Operations >> Yale University >> 203 432 4303 >> >> _______________________________________________ >> labnetwork mailing list >> labnetwork at mtl.mit.edu >> https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork From quantumrods at gmail.com Thu Jun 20 02:47:28 2019 From: quantumrods at gmail.com (Alexander Fox) Date: Thu, 20 Jun 2019 09:47:28 +0300 Subject: [labnetwork] Clean Room Ceiling In-Reply-To: References: Message-ID: Dear Fouad, Make sure cleanroon ceilings are accessable and walkable from above. If not -- provide catwalks there necessary. Also use FFUs instead of regular HEPA boxes. This way you have better control over the flow and less crowded space above the ceiling. Use LED stripes or 2x4 panels integrated into suspended ceiling (beams), do not use protruding luminairs in clean zones in order to not collect any dust. On Wed, Jun 19, 2019 at 2:50 PM Fouad Karouta wrote: > Dear all, > > > > We are in the process of getting a new building that includes a 650 m2 > clean room (class 100 and 1000) with a sub-fab to host chiller, pumps, > scrubbers, etc. > > > > We are looking to possible suppliers of clean room compartmented ceiling > to host light, Hepa filters and others like fire sprinklers, etc. > > Any suggestions for suppliers are welcome. > > Please also comment on what to include or avoid in ceiling. > > > > Thanks, > > Fouad Karouta > > > > ************************************* > > Manager ANFF ACT Node > > Australian National Fabrication Facility > > Research School of Physics and Engineering > > L. Huxley Building (#56), Mills Road, Room 4.02 > > Australian National University > > ACT 2601, Canberra, Australia > > Tel: + 61 2 6125 7174 > > Mob: + 61 451 046 412 > > Email: fouad.karouta at anu.edu.au > > http://anff-act.anu.edu.au/ > > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork > -- Alexander Fox Clean Room Design Engineer *CSJC ?Elar?* 68 Toreza prospect, Saint-Petersburg, Russia 194223 -------------- next part -------------- An HTML attachment was scrubbed... URL: From feargal.nolan at tyndall.ie Thu Jun 20 08:16:43 2019 From: feargal.nolan at tyndall.ie (Feargal Nolan) Date: Thu, 20 Jun 2019 12:16:43 +0000 Subject: [labnetwork] Argon in a BCl3 cylinder In-Reply-To: References: Message-ID: Hi Sean, We had a similar problem with SiCl4 but in our case it was charged with nitrogen. On our system we were able to use the venturi on the gas panel to evacuate the N2 from the cylinder. We ran several tests afterwards and we did not see any adverse effects in our process. Regards Feargal Nolan Maintenance Manager Tyndall National Institute Lee Maltings, Dyke Parade Cork, T12 R5CP Ireland t +353 21 234 6656 m +353 87 635 7673 From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Rinehart, Sean Sent: 19 June 2019 18:15 To: labnetwork at mtl.mit.edu Subject: [labnetwork] Argon in a BCl3 cylinder Hello all, Recently our facility's BCl3 cylinder had been charged with argon, increasing the delivery pressure to 25 psi. The cause has been corrected, but now I'm evaluating whether or not the cylinder needs to be replaced, and would appreciate some advice. Since the root issue was fixed, the pressure has slowly dropped from 25 psi to 15 psi with use. Based on this, it seems like with more use the argon should completely deplete from the cylinder, but maybe someone can tell me if that's actually true, and if there's a way to evaluate gas concentrations. Additionally, the plasma glow is whitish instead of pink, so there's definitely BCl3 content in the delivered gas. Thank you, Sean Rinehart Assistant Director, Cleanroom Operations Yale University 203 432 4303 -------------- next part -------------- An HTML attachment was scrubbed... URL: From jpalmer at Princeton.EDU Thu Jun 20 09:14:00 2019 From: jpalmer at Princeton.EDU (Joseph E. Palmer) Date: Thu, 20 Jun 2019 09:14:00 -0400 Subject: [labnetwork] Argon in a BCl3 cylinder In-Reply-To: References: Message-ID: Sean, ???? From a research point of view, I would see this as an unacceptable unknown.? I would strongly suggest changing out the cylinder. Regards, Joe Palmer Operations Manager PRISM MNFL Princeton University On 6/19/2019 4:40 PM, Paolini, Steven wrote: > > Sean, > > ? I am not entirely sure, but it?s not likely that the gases will > completely disassociate from each other, even if the pressure drops. > I?m assuming that because the vapor pressure of BCL3 is a constant > because it is a liquid, you are determining that mostly Ar is being > consumed because of the pressure drop. I would recommend a complete > pump and purge of the entire gas circuit and replacing the cylinder of > BCL3. > > The simplest way to evaluate the gas concentrations is with an RGA but > I don?t think it?s a good idea to expose the filament to a corrosive > such as BCL3 > > ??? Steve ?Equipment Dood? Paolini > > Steve Paolini > > Principal Equipment Engineer > > Harvard University Center for Nanoscale Systems > > 11 Oxford St. > > Cambridge, MA 02138 > > 617- 496- 9816 > > spaolini at cns.fas.harvard.edu > > www.cns.fas.harvard.edu > > *From:* labnetwork-bounces at mtl.mit.edu > *On Behalf Of *Rinehart, Sean > *Sent:* Wednesday, June 19, 2019 1:15 PM > *To:* labnetwork at mtl.mit.edu > *Subject:* [labnetwork] Argon in a BCl3 cylinder > > Hello all, > > Recently our facility?s BCl3 cylinder had been charged with argon, > increasing the delivery pressure to 25 psi. The cause has been > corrected, but now I?m evaluating whether or not the cylinder needs to > be replaced, and would appreciate some advice. > > Since the root issue was fixed, the pressure has slowly dropped from > 25 psi to 15 psi with use. Based on this, it seems like with more use > the argon should completely deplete from the cylinder, but maybe > someone can tell me if that?s actually true, and if there?s a way to > evaluate gas concentrations. Additionally, the plasma glow is whitish > instead of pink, so there?s definitely BCl3 content in the delivered gas. > > Thank you, > > Sean Rinehart > > Assistant Director, Cleanroom Operations > > Yale University > > 203 432 4303 > > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: From spb1699 at rit.edu Thu Jun 20 17:17:22 2019 From: spb1699 at rit.edu (Scott Blondell) Date: Thu, 20 Jun 2019 21:17:22 +0000 Subject: [labnetwork] Argon in a BCl3 cylinder In-Reply-To: References: Message-ID: <9987076b5fdb4cfba325938143988ad8@ex04mail01a.ad.rit.edu> Is this a case where a check valve at the BCl3 source failed, or was non-existent? Would a valve cracking pressure of 0.5 psi have prevented this unwanted backflow. Scott P. Blondell Facilities Manager Rochester Institute of Technology Semiconductor & Microsystems Fabrication Laboratory 82 Lomb Memorial Dr. Bldg. 17-2519 Rochester, NY 14623 585 738-4073 c 585 475-2171 o 585 475-5041 f spb1699 at rit.edu www.smfl.rit.edu [http://www.rit.edu/~962www/logos/tiger_walking_rit_color.jpg] From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Feargal Nolan Sent: Thursday, June 20, 2019 8:17 AM To: Rinehart, Sean ; labnetwork at mtl.mit.edu Subject: Re: [labnetwork] Argon in a BCl3 cylinder Hi Sean, We had a similar problem with SiCl4 but in our case it was charged with nitrogen. On our system we were able to use the venturi on the gas panel to evacuate the N2 from the cylinder. We ran several tests afterwards and we did not see any adverse effects in our process. Regards Feargal Nolan Maintenance Manager Tyndall National Institute Lee Maltings, Dyke Parade Cork, T12 R5CP Ireland t +353 21 234 6656 m +353 87 635 7673 From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Rinehart, Sean Sent: 19 June 2019 18:15 To: labnetwork at mtl.mit.edu Subject: [labnetwork] Argon in a BCl3 cylinder Hello all, Recently our facility's BCl3 cylinder had been charged with argon, increasing the delivery pressure to 25 psi. The cause has been corrected, but now I'm evaluating whether or not the cylinder needs to be replaced, and would appreciate some advice. Since the root issue was fixed, the pressure has slowly dropped from 25 psi to 15 psi with use. Based on this, it seems like with more use the argon should completely deplete from the cylinder, but maybe someone can tell me if that's actually true, and if there's a way to evaluate gas concentrations. Additionally, the plasma glow is whitish instead of pink, so there's definitely BCl3 content in the delivered gas. Thank you, Sean Rinehart Assistant Director, Cleanroom Operations Yale University 203 432 4303 -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.jpg Type: image/jpeg Size: 2550 bytes Desc: image001.jpg URL: From fouad.karouta at anu.edu.au Sun Jun 23 19:20:51 2019 From: fouad.karouta at anu.edu.au (Fouad Karouta) Date: Sun, 23 Jun 2019 23:20:51 +0000 Subject: [labnetwork] Clean Room Ceiling In-Reply-To: References: Message-ID: Thank you all for your useful suggestions about my request below. Fouad From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Fouad Karouta Sent: Wednesday, 19 June 2019 10:30 AM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Clean Room Ceiling Dear all, We are in the process of getting a new building that includes a 650 m2 clean room (class 100 and 1000) with a sub-fab to host chiller, pumps, scrubbers, etc. We are looking to possible suppliers of clean room compartmented ceiling to host light, Hepa filters and others like fire sprinklers, etc. Any suggestions for suppliers are welcome. Please also comment on what to include or avoid in ceiling. Thanks, Fouad Karouta ************************************* Manager ANFF ACT Node Australian National Fabrication Facility Research School of Physics and Engineering L. Huxley Building (#56), Mills Road, Room 4.02 Australian National University ACT 2601, Canberra, Australia Tel: + 61 2 6125 7174 Mob: + 61 451 046 412 Email: fouad.karouta at anu.edu.au http://anff-act.anu.edu.au/ -------------- next part -------------- An HTML attachment was scrubbed... URL: From sang-joon.lee at sjsu.edu Mon Jun 24 18:25:24 2019 From: sang-joon.lee at sjsu.edu (John Lee) Date: Mon, 24 Jun 2019 15:25:24 -0700 Subject: [labnetwork] =?utf-8?q?Jay_D=2E_Pinson_Microfabrication_Laborator?= =?utf-8?q?y_Manager_at_San_Jos=C3=A9_State_University_-_position_a?= =?utf-8?q?nnouncement?= Message-ID: All, Please allow me to share this information across this distribution list, and please be welcome to relay to others who might be interested. Details and application instructions are the following link, specifically under Job ID 25024. http://www.sjsu.edu/up/careers/staff_jobs/ Part of the position description is as follows: "The Jay D. Pinson Microfabrication Laboratory Manager will support and enhance capabilities and user experience in the Microscale Process Engineering Laboratory (MPEL) in the Charles W. Davidson College of Engineering at San Jos? State University. The incumbent reports to the Director of General Engineering, and receives work lead direction from a multidisciplinary faculty user group (MPEL Steering Committee) to help ensure that the facility can effectively serve the needs of education and research that depend on hands-on microfabrication capabilities." Thank you, -John ________________________________________ Sang-Joon (John) Lee, Professor Mechanical Engineering, San Jos? State University One Washington Square, San Jos?, CA 95192-0087 408-924-7167 | sang-joon.lee at sjsu.edu http://www.sjsu.edu/people/sang-joon.lee/ From mxm31 at case.edu Mon Jun 24 19:04:05 2019 From: mxm31 at case.edu (Mehran Mehregany) Date: Mon, 24 Jun 2019 16:04:05 -0700 Subject: [labnetwork] SPTS Pro AOE system for sale Message-ID: I have a lightly used, older SPTS Pro AOE in my lab that I would like to sell. It is a complete system and was purchased new and installed in 2006. We used it for a research project early on, but have not turned it on for 9-10 years. Attached please find info on the system. Please e-mail me if interested to know more, Mehran Mehregany, Ph.D. Veale Professor of Wireless Health Innovation Director, Case School of Engineering?San Diego Case Western Reserve University 6540 Lusk Blvd, Suite C272 San Diego, CA 92121 858-869-0100 (direct line) -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: 2018-05-10 STPS Pro AOE for Sale.pdf Type: application/pdf Size: 3271303 bytes Desc: not available URL: -------------- next part -------------- An HTML attachment was scrubbed... URL: From sean.rinehart at yale.edu Tue Jun 25 11:34:18 2019 From: sean.rinehart at yale.edu (Rinehart, Sean) Date: Tue, 25 Jun 2019 15:34:18 +0000 Subject: [labnetwork] Argon in a BCl3 cylinder In-Reply-To: <9987076b5fdb4cfba325938143988ad8@ex04mail01a.ad.rit.edu> References: <9987076b5fdb4cfba325938143988ad8@ex04mail01a.ad.rit.edu> Message-ID: Scott, In this case the check valve at the source was non-existent. The main cylinder valve had become stuck partly open, and it wasn't caught until the damage was done. Everyone, Thank you for your input on this issue. Definitely a lot of food for thought! From: labnetwork-bounces at mtl.mit.edu On Behalf Of Scott Blondell Sent: Thursday, June 20, 2019 5:17 PM To: labnetwork at mtl.mit.edu Subject: Re: [labnetwork] Argon in a BCl3 cylinder Is this a case where a check valve at the BCl3 source failed, or was non-existent? Would a valve cracking pressure of 0.5 psi have prevented this unwanted backflow. Scott P. Blondell Facilities Manager Rochester Institute of Technology Semiconductor & Microsystems Fabrication Laboratory 82 Lomb Memorial Dr. Bldg. 17-2519 Rochester, NY 14623 585 738-4073 c 585 475-2171 o 585 475-5041 f spb1699 at rit.edu www.smfl.rit.edu [http://www.rit.edu/~962www/logos/tiger_walking_rit_color.jpg] From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Feargal Nolan Sent: Thursday, June 20, 2019 8:17 AM To: Rinehart, Sean >; labnetwork at mtl.mit.edu Subject: Re: [labnetwork] Argon in a BCl3 cylinder Hi Sean, We had a similar problem with SiCl4 but in our case it was charged with nitrogen. On our system we were able to use the venturi on the gas panel to evacuate the N2 from the cylinder. We ran several tests afterwards and we did not see any adverse effects in our process. Regards Feargal Nolan Maintenance Manager Tyndall National Institute Lee Maltings, Dyke Parade Cork, T12 R5CP Ireland t +353 21 234 6656 m +353 87 635 7673 From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Rinehart, Sean Sent: 19 June 2019 18:15 To: labnetwork at mtl.mit.edu Subject: [labnetwork] Argon in a BCl3 cylinder Hello all, Recently our facility's BCl3 cylinder had been charged with argon, increasing the delivery pressure to 25 psi. The cause has been corrected, but now I'm evaluating whether or not the cylinder needs to be replaced, and would appreciate some advice. Since the root issue was fixed, the pressure has slowly dropped from 25 psi to 15 psi with use. Based on this, it seems like with more use the argon should completely deplete from the cylinder, but maybe someone can tell me if that's actually true, and if there's a way to evaluate gas concentrations. Additionally, the plasma glow is whitish instead of pink, so there's definitely BCl3 content in the delivered gas. Thank you, Sean Rinehart Assistant Director, Cleanroom Operations Yale University 203 432 4303 -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.jpg Type: image/jpeg Size: 2550 bytes Desc: image001.jpg URL: From jpalmer at Princeton.EDU Wed Jun 26 15:35:59 2019 From: jpalmer at Princeton.EDU (Joseph E. Palmer) Date: Wed, 26 Jun 2019 15:35:59 -0400 Subject: [labnetwork] Fwd: Nanofab staff opening at UIowa In-Reply-To: References: Message-ID: Please excuse me, as I help spread the word on the job opportunity below. - J. Palmer I am reaching out to you today to see if you can help spread the word for a job opening at our nanofabrication facility: https://uiowa.referrals.selectminds.com/jobs/core-facility-research-specialist-3341 which is equipped with these tools: ?Direct-write Electron-Beam Lithography (recently purchased): high speed pattern generatoronthenano-scalewithsub-20nmresolutionpatterningcapabilityandplacement accuracy. ?Electron Beam Lithography Proximity effects control software: GenISys Beamer: control the electron beam dose and electron scattering effects for various pattern geometries and densities. ?Optical lithography: OAI Model 800 mask aligner with back-side alignment capabilities for pattern exposures down to 1?m. ?Thin-film deposition: Electron-beam deposition system (Angstrom Engineering 450) and Thermal evaporator (Edwards Auto 306) for coating conformal thin metal layers down to sub-10 nm thickness with highuniformity. ?Plasma etcher: Reactive ion etchers (RIE) (Oxford PlasmaPro NGP 80) and ICP Plasma Pro S100 for the etching of various semiconductor and insulation materials with features on the micro- and nano-scale using several gas chemistries. These tools offer anisotropic etching on various materials that include glass, silicon, III-V materials, metals with near verticalsidewalls. ?Atomic Layer deposition (ALD): The ALD system (Oxford Opal) provides the ability to precisely coat materials at the atomic level conformally using sequential and self-limiting surface reactions. This tool is used to deposit a wide range of materials, including oxidessuch as TiO_2 , SiO_2 , ZnO, HfO_2 and Nitrides, TiN, Si_3 N_4 and metals such as Ru, Pt, with a typical uniformity of < 2 % and provides excellent step coverage even inside high aspect ratiostructures. ?NanoimprintLithography:Thetool(NanonexNX-1006)isafullwafernanoimprintorwith thermal imprint capability. This tool is capable of imprinting features on the micro- and nano-scaleoverlargeareasinthecentimeterranges?suitableforlow-costandlarge-scale fabrication. ?Thin-film sputtering system: The tool (Intlvac Nanochrome 1) uses any combinations ofRF, AC and DC magnetron sputtering cathodes to produce thin films (including Si, SiO_2 , Si_3 N_4 ) for precision coatings on a wide range ofmaterials. ?Plasma cleaner: The tool (Branson 3100) uses an oxygen plasma to remove thin layers of polymer on semiconductor or insulatingmaterials. ?Wet-processing: Several laminar flow wet-benches and hoods are available for wet- processing using a wide range of solvents and acids for wetetching. The salary and benefits are competitive. For more information, the interested party can contact me directly as well, and I can connect them to the hiring manager. Thanks. Regards, Fatima -------------------------------------------------------------------------------------------------------------------- Fatima Toor, Ph.D. Assistant Professor Electrical & Computer Engineering Physics & Astronomy UI Informatics Initiative (UI3) Center for Computer Aided Design (CCAD) Iowa CREATES and MATFab Facility Affiliate Member, Holden Comprehensive Cancer Center - Experimental Therapeutics Office: 116 Iowa Advanced Technology Laboratories (IATL) 205 N. Madison Street, The University of Iowa, Iowa City, IA 52242-1727, USA Phone: +1 (319) 335-6071 Fax: +1 (319) 335-6028 https://ftoor.lab.uiowa.edu/ Technical Group Development Chair, OSA Board of Meetings Associate Editor, SPIE Journal of Photonics for Energy -------------- next part -------------- An HTML attachment was scrubbed... URL: From sb86922 at usc.edu Thu Jun 27 15:05:59 2019 From: sb86922 at usc.edu (Shivakumar Bhaskaran) Date: Thu, 27 Jun 2019 19:05:59 +0000 Subject: [labnetwork] Dry pump in Savannah ALD & Pd deposition Message-ID: Hello, We are looking into the options of buying ALD; one of the tools that we are considering is Savannah ALD. We are planning to use the tool with a dry pump; the vendor mentioned that most of their recipes work well with the rotary vane pump. And we might have to tweak the recipe. Has anyone tried using the dry pump, if so what was your experience and recommendation on what dry pump we should buy. And has anyone tried depositing Pd using ALD, based on literature survey I see Pd(hfac)2 precursor? Anyone has experience in using this precursor and if so would you mind sharing your recipes. Thanks, Shiva Shivakumar Bhaskaran, Ph.D., John D. O'Brien Nanofabrication Laboratory Michelson Hall, 1002 Childs Way, MCB LL121, Los Angeles, California 90089, 213 821 2374 [Sign] -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.jpg Type: image/jpeg Size: 2080 bytes Desc: image001.jpg URL: