[labnetwork] STS DRIE Multiplex Lip Seal failures
Mark Weiler
mweiler at andrew.cmu.edu
Tue Mar 5 15:44:32 EST 2019
Hello Everyone,
We have gone through seven new lip seals purchased form Orbotech/SPTS. They are often failing before we even finish qualifying the system, or within a month thereafter.
Our our process is stable with power and parameters not deviating over the past decade. However, I ordered the most recent batch of lip seals because the wafer seals we had been using were coming out with black residue after only a few runs… however, the current ones do the same. Not only have the lips seals disintegrated, but the chamber lid o-ring and bottom ceramic spool o-ring have also failed with black rubber material shedding off. It’s as if they are made of Buna and not meant for this application.
Our process is typical Bosch with 100 sccm C4F8 and 20 sccm O2 for etching switching off with ~50 Sccm SF6 for passivation at a processing pressure of 15 mTorr, Coil power 600-800W, Platen power 100-150W, Bias voltage 50-100v (up to 200 peak-to-peak), Platen temp at 19 degrees, Lid temp 41 degrees. Qual wafers are new bare Silicon. Etch rates are normal and stable with 10+ years of data…. we just can’t complete the work due to failing seals.
Our chamber base pressure is between 1E-8 and 5E-8 Torr each morning. When we put in a brand new seal and run only the LUR, it passes with 0.00 mTorr/minute leak rate. That jumps to 3.00 mTorr/m after only 30 minutes of the SPTS recommended O2 Clean… then rises as time and more runs progress. Wafers are coming out with black rings on their backsides.
This should not be happening, and I believe it is due to incorrect material of the lip seals and o-rings. I am pinging the network, though, to see if there might be a parameter change we need to effect that may assist us.
Have any of you seen this before?
Best regards,
Mark
________________________________________________________________
Mark Weiler
Equipment & Facilites Manager
Clair and John Bertucci Nanotechnology Laboratory
Eden Hall Nanofabrication Cleanroom
Carnegie Mellon University
P: 412-268-2471
http://www.nanofab.ece.cmu.edu
[cid:2D2E01E3-CEC1-4F48-A845-224D8D7CED12 at wv.cc.cmu.edu] <https://www.linkedin.com/in/marksbasics>
<https://www.linkedin.com/in/marksbasics>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20190305/76fae632/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: images.png
Type: image/png
Size: 720 bytes
Desc: images.png
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20190305/76fae632/attachment.png>
More information about the labnetwork
mailing list