[labnetwork] Diced Samples Glued to UV Release Tape

Paolini, Steven spaolini at cns.fas.harvard.edu
Thu May 16 16:59:23 EDT 2019


Sa'ad,
  Are you applying your tape manually on the hotplate at temperature? If so, UV tape adhesive will cross link at higher temperatures than your UV source delivers. Another possibility is overexposure. Best way to handle it is with a D.O.E using different exposure times. You will find that the window between underexposing and overexposing is fairly generous. I do suspect the elevated temperature on the hotplate being the root cause if that's what you're doing.
  I hope this helps,
    Steve (Equipment Dood)

Steve Paolini
Principal Equipment Engineer
Harvard University Center for Nanoscale Systems
11 Oxford St.
Cambridge, MA 02138
617- 496- 9816
spaolini at cns.fas.harvard.edu
www.cns.fas.harvard.edu

From: labnetwork-bounces at mtl.mit.edu <labnetwork-bounces at mtl.mit.edu> On Behalf Of Hassan, Sa'ad
Sent: Thursday, May 16, 2019 1:40 PM
To: labnetwork at mtl.mit.edu
Subject: [labnetwork] Diced Samples Glued to UV Release Tape

Hello All,

As part of the last step of processing, I take wafers (Si and SiC), with photoresist on top,  dice them, die expand them on a hotplate at 55 degrees Celsius. And finally UV release the tape. I've noticed that sporadically, rather than the tape losing  adhesion, the opposite happens: the die end up cemented to the tape. Has anyone experienced anything like this before?

Cheers,
Sa'ad