[labnetwork] Stationary Particle Counters
Paolini, Steven
spaolini at cns.fas.harvard.edu
Thu Sep 26 12:26:22 EDT 2019
Devin,
I would suggest getting a baseline of your entire fab by a company that specializes in this. Once a year, we have our fab re-certified and as they perform this, they examine the entire HEPA coverage with a sonic detector that will show leaks, and problem areas. This company hires a HEPA installation/service person to follow them around to fix problems as they are identified.
After you have this baseline map, I would also recommend a permanent particle counter in the middle of all the chase areas if your clean room is a sidewinder, or in the floor if it’s a ballroom style raised floor. This will provide real time averaging and will provide feedback if you wish to have a closed loop system that communicate to the HEPA fan VFD, This type of arrangement will be capable of slowing down your air changes/Hr. during periods of inactivity. This is a tremendous energy conservation measure as well as having the ability to speed up the HEPA air during higher particulate generation times.
I realize the large expense of doing this but if you want to get serious about maintaining a truly clean room without just dumping money on other measures, it’s a good way to go. I am curious to hear what other fabbers use as a method.
Best of luck,
Equipment Dood
Steve Paolini
Principal Equipment Engineer
Harvard University Center for Nanoscale Systems
11 Oxford St.
Cambridge, MA 02138
617- 496- 9816
spaolini at cns.fas.harvard.edu
www.cns.fas.harvard.edu
From: labnetwork-bounces at mtl.mit.edu <labnetwork-bounces at mtl.mit.edu> On Behalf Of Devin Fortier
Sent: Wednesday, September 25, 2019 11:49 AM
To: labnetwork at mtl.mit.edu
Subject: [labnetwork] Stationary Particle Counters
Dear Colleagues,
We are looking to improve the way we do particle counts in our cleanroom and are looking at installing multiple stationary particle counters as a solution. We would also want these devices to tie in with our building monitoring system that uses BACnet.
We currently have a portable particle counter that is moved to various set locations and left to take its measurements. This is being done by a staff member and the whole procedure takes up almost half a day of one person's time per week. Aside from taking so much time, the counter is cumbersome and gets in the way of our users in the lab.
My question to the group is what are you currently using for your particle counts? Does anyone use stationary particle counters and process the data to get an average over the whole cleanroom area?
Thank you,
--
Devin Fortier
Systems/Infrastructure Specialist
University of Alberta - nanoFAB
W1-028A ECERF Building
9107 - 116 Street
Edmonton, Alberta
Canada T6G 2V4
www.nanofab.ualberta.ca<https://urldefense.proofpoint.com/v2/url?u=http-3A__www.nanofab.ualberta.ca_&d=DwMFaQ&c=WO-RGvefibhHBZq3fL85hQ&r=-H4Z_jeDfTYjnKPXor15vLwzBllmg8gFrb9m_k9OGks&m=mc7Q3G2DP5uG_q1s2-5QJgVjUst8hvgaQ00NEZhrcbQ&s=iksD82WAh8TdNhNP_B3a0oqer7moRv2_PtOUOF-8msQ&e=> Ph: 780-868-6480
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20190926/1a0c3ca7/attachment.html>
More information about the labnetwork
mailing list