Hi Mike,
I used to work with RZ Enterprises, and we did service all GCA tools,
including pattern generators…In fact I still have one that I take care of
here at Berkeley. Let me put together some information later today to see
if I can help you through this, and also I will put out feelers on who is
still servicing them.
Best,
G~
On Fri, Dec 4, 2020 at 8:54 AM Michael Call <michael.call at maine.edu> wrote:
> We have an issue with "Focus Signal lost" and "time out" errors on our
> pattern generator. Sometimes the column will attempt to focus off of the
> glass plate(laser system losing counts?). I've cleaned and oiled the
> stages, changed plates and all of the usual, but with no luck. I know that
> Bob Scruton passed away a few years ago. Has anyone taken over the service
> of these machines?
>
> Thanks,
>
> --
> Mike Call
> Research Engineer
> University of Maine
> Frontier Institute for Research in Sensor Technologies (FIRST)
> Coordinated Research entities Cleanroom
> ESRB Barrows Hall room 294
> 207-581-3382
> [image: image.png]
> _______________________________________________
> labnetwork mailing list
> labnetwork at mtl.mit.edu
> https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
>
--
Greg Mullins
R&D Engineer 4
Marvell Nanofabrication Laboratory
University of California at Berkeley
520 Sutardja Dai Hall, MS 1754
Berkeley, CA 94720-1754
Office: (510) 809-8625
Email: gregmullins at berkeley.edu
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20201204/8d4b037a/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image.png
Type: image/png
Size: 53624 bytes
Desc: not available
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20201204/8d4b037a/attachment.png>
More information about the labnetwork
mailing list