[labnetwork] Toxic gases
Nava Ariel-Sternberg
na2661 at columbia.edu
Mon Dec 7 12:17:40 EST 2020
Same for Columbia.
-Nava
Nava Ariel-Sternberg, Ph.D.
Director of CNI Shared Labs
Columbia University
CEPSR/MC 8903
530 west 120th st. NY
NY 10027
Office: 212-8549927
Cell: 201-5627600
From: labnetwork-bounces at mtl.mit.edu <labnetwork-bounces at mtl.mit.edu> On
Behalf Of Aebersold,Julia W.
Sent: Monday, December 7, 2020 11:01 AM
To: Manish Keswani <manish.keswani01 at gmail.com>; labnetwork at mtl.mit.edu
Subject: Re: [labnetwork] Toxic gases
The Micro/Nano Technology Center at the University of Louisville has the
same protocol. We do not turn off our cylinders, but they controlled by our
gas cabinets and TGMS system.
Cheers!
Julia Aebersold, Ph.D.
Manager, Micro/Nano Technology Center
University of Louisville
2210 South Brook Street
Shumaker Research Building, Room 233
Louisville, KY 40292
(502) 852-1572
http://louisville.edu/micronano/ <http://louisville.edu/micronano>
_____
From: labnetwork-bounces at mtl.mit.edu <mailto:labnetwork-bounces at mtl.mit.edu>
<labnetwork-bounces at mtl.mit.edu <mailto:labnetwork-bounces at mtl.mit.edu> > on
behalf of Manish Keswani <manish.keswani01 at gmail.com
<mailto:manish.keswani01 at gmail.com> >
Sent: Sunday, December 6, 2020 3:26 PM
To: labnetwork at mtl.mit.edu <mailto:labnetwork at mtl.mit.edu>
<labnetwork at mtl.mit.edu <mailto:labnetwork at mtl.mit.edu> >
Subject: Re: [labnetwork] Toxic gases
CAUTION: This email originated from outside of our organization. Do not
click links, open attachments, or respond unless you recognize the sender's
email address and know the contents are safe.
Thank you so much everyone for such an enthusiastic response. It seems
clear that at most facilities (if not all), the gas cylinders are always
kept open. We also have TGMS in our cleanroom which allows monitoring of the
toxic gases at multiple locations (source and delivery point) and will shut
off the ESO valve in the event of a leak.
We will work with our safety and ES&H teams to change our current practice
of turning off the gases when not in use.
Regards,
Manish
On Fri, Dec 4, 2020 at 2:11 PM Manish Keswani <manish.keswani01 at gmail.com
<mailto:manish.keswani01 at gmail.com> > wrote:
Happy Friday everyone,
I have a question related to toxic gases in our nano fabrication center.
We typically shut off our toxic gases at the source when not in use, using
the in line valves on the Safety Manifolds, the ESO valves, and the cylinder
valve. This is to comply with the administrative control stated in our gas
safety notes. The following question came up for toxic gases. I would like
to understand how this is being done at other facilities and the reasoning
behind it.
"Is there a good reason we turn off the toxic gases at the end of the day
instead of just leaving them on? It seems like if there is a risk, it would
be wear and tear on the valves etc, plus someone going in vault often to do
it."
Thanks in advance,
Manish Keswani
Lawrence Livermore National Laboratory
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20201207/860a37fa/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image001.png
Type: image/png
Size: 105149 bytes
Desc: not available
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20201207/860a37fa/attachment.png>
More information about the labnetwork
mailing list