[labnetwork] EV620 not initializing

Hitesh Kamble hit.kamble at gmail.com
Thu Jan 30 14:41:41 EST 2020


Hello Chito,

What i understood from your description is optics initialise, cover down
but it doesn't reach to upper position? And gives initialisation error at
"UC04:*" ?

This is related to Way measurements system which is combination of Z-motor
control and hardware settings.

To identify issue>> access software in service mode: At login window, enter
id password and click 3 times on EVG logo. You'll bypass initialisation and
can access tool in High-level mode. There try to operate cover up/down. If
this function does not work then try below troubleshooting.


Checks1 : Open electronic box where you will see Z-motor controller. It
should not be "F" on 7-segment display. If yes battery could have drained
completely or firmware lost.
Solution: Replace battery and install firmware again.

Checks 2: There is a limiting screw in the Top side of the cover.It´s a M10
Nut screw near the center of the cover when you look onto the cover.

When this screw is loosen, the cover can´t reach it´s upper position for
initialization.





Thanks,
Hitesh Kamble,
CEN, IIT Bombay, India.



On Thu, Jan 30, 2020, 11:17 PM Chito Kendrick <cekendri at mtu.edu> wrote:

> We have a EV620 Aligner at MTU and this tuesday it did not complete the
> initialization process after logging in, all reboots and attempts have
> failed. We have software version 2.64 and we are still running a Win98
> system. I am talking to EVG about this problem to get a solution, but being
> an older system and software version they are still working on it.
> Therefore I was hoping this community might have faced this issue and have
> some advice.
>
> After typing in the password and hitting run the system starts
> initialization. It moves the optical head, mask stage, WEC, and then gives
> a double beep - all of this is normal. Instead of bringing up the recipes
> it has ERROR and look at the log. The log has as the last entry - "Check
> distance measurement !". All menus are greyed out so I can not access any
> other part of the software.
>
> Regards,
>
> Chito Kendrick
>
>
>
> --
> Chito Kendrick Ph.D. <https://sites.google.com/site/chitokendrickphd/>
>
> Managing Director of the Microfabrication Facility
> Research Assistant Professor
> Electrical and Computer Engineering
> Michigan Technological University
> Room 436 M&M Building
> 1400 Townsend Dr.
> Houghton, Michigan 49931-1295
>
> 814-308-4255
> _______________________________________________
> labnetwork mailing list
> labnetwork at mtl.mit.edu
> https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20200131/f81be95f/attachment.html>


More information about the labnetwork mailing list