[labnetwork] Spot issue in Unaxis-750 PECVD

April, Mark R. mapril at draper.com
Mon Aug 23 06:48:16 EDT 2021


Good Morning Vibhor,
I concur with Mr. Paolini.  I had a similar issue on a PECVD tool.  After a thorough shower head clean, including the diffuser plate, the shower head pattern was eliminated.  This routinely gets done each year now, during a major PM.  You may need to perform more due to your usage.
Hope this helps.

Mark R. April
Senior  Equipment Engineer
Microfabrication Laboratory
[color_logo_small4]
555 Technology Square
Cambridge, MA  02139
mapril at draper.com<mailto:mapril at draper.com>
O # 617-258-1613 C # 617-455-1596
www.draper.com<http://www.draper.com/>



From: labnetwork <labnetwork-bounces at mtl.mit.edu> On Behalf Of Paolini, Steven
Sent: Friday, August 20, 2021 4:02 PM
To: Vibhor Kumar <vk409 at soe.rutgers.edu>; labnetwork at mtl.mit.edu
Subject: Re: [labnetwork] Spot issue in Unaxis-750 PECVD

Vibhor,
  You are most likely seeing the effect of the showerhead being in need of cleaning, particularly the backside. The Deposition of dielectric films creates a powdery substance just outside of the intended plasma concentrate. This debris, over time, will allow small reactions to happen under the showerhead and show up as spots on your wafer, most likely in the pattern of the holes in the showerhead. I would suggest removing the showerhead and diffuser plate and performing a very thorough clean on all the parts. In my experience, and depending on the usage, this should be performed routinely at an interval that keeps the effect at bay.
  Hope this helps,
    Steve Paolini
   Equipment Dood

Steve Paolini
Principal Equipment Engineer
Harvard University Center for Nanoscale Systems
11 Oxford St.
Cambridge, MA 02138
617- 496- 9816
spaolini at cns.fas.harvard.edu<mailto:spaolini at cns.fas.harvard.edu>
www.cns.fas.harvard.edu<http://www.cns.fas.harvard.edu>

From: labnetwork <labnetwork-bounces at mtl.mit.edu<mailto:labnetwork-bounces at mtl.mit.edu>> On Behalf Of Vibhor Kumar
Sent: Friday, August 20, 2021 1:49 PM
To: labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>
Subject: [labnetwork] Spot issue in Unaxis-750 PECVD

Hello Everyone,

During nitride/oxide deposition with Unaxis-750 PECVD, we are getting a mirror image of shower head on the wafer, i.e., spots. We cleaned the chamber with CF4 for 2 consecutive days, each day 4:30 hrs. The conditions were

CF4 = 100 sccm, Temp = 250 C, chamber pressure = 600 mT, time = 4:30 hrs

We also evacuated the silane gas line, but there is no leakage.

After completing the cleaning, when we conditioned the chamber and deposited nitride, the spots were lightened. But on the second day, the spots look enlarging again.

Can anyone please suggest solution to this issue?

Thanks,
-Vibhor Kumar-
Postdoctoral Associate
Rutgers, The State University of New Jersey.

________________________________
Notice: This email and any attachments may contain proprietary (Draper non-public) and/or export-controlled information of Draper. If you are not the intended recipient of this email, please immediately notify the sender by replying to this email and immediately destroy all copies of this email.
________________________________
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20210823/4ec26379/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image001.png
Type: image/png
Size: 1560 bytes
Desc: image001.png
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20210823/4ec26379/attachment.png>


More information about the labnetwork mailing list