Dear all,
Thank you for your recommendations. They were all very helpful.
Best,
Saba
________________________________________
From: labnetwork-bounces at mtl.mit.edu <labnetwork-bounces at mtl.mit.edu> on behalf of Stephen Howe <info at fabsurplus.com>
Sent: March 1, 2021 5:50:43 AM
To: N Shane Patrick; James Mitchell
Cc: Fab Network
Subject: Re: [labnetwork] cleaning procedure for chillers
Dear All,
To avoid algae growth, it is very important not to use transparent chiller hoses. Also, make sure the lid of the chiller is closed, to stop light from getting in.
It is the presence of light which encourages the growth of organics in your chiller water.
I have encountered people who had to scrap several really expensive RTP tools, due to the cooling circuits were blocked with organic gunk.
-
[SDI Fabsurplus 20 years]
Yours Sincerely,
SDI Fabsurplus LLC
Stephen Howe
Company Owner
email: info at fabsurplus.com<mailto:info at fabsurplus.com>
Mobile: +1 830-388-1071
Skype: stephencshowe
WWW.FABSURPLUS.COM<mail://local/>
Linkedin: https://www.linkedin.com/company/sdi-fabsurplus
Twitter: https://twitter.com/fabsurplus
Instagram: www.instagram.com/fabsurplus<https://www.instagram.com/fabsurplus/>
Facebook:www.facebook.com/fabsurplus<https://www.facebook.com/fabsurplus>
Important Notice
The information contained in this email is confidential. If you are not the intended recipient,you must not disclose or use the information in this email in any way.If you received it in error, please tell us immediately by return email and delete this message.SDI-Fabsurplus, LLC does not guarantee the integrity of any emails or attached files. It is also not responsible for any changes made to them by any other person.
Fraud Alert - The strategically important semiconductor equipment sector is currently under attack by fraudsters.Many false and mis-spelled versions of common websites are being registered for use in"Business E-mail Compromise" schemes. False companies have been set up in all parts of the world, including the USA, to do wire fraud. Avoid e-mail fraud by defining an I.T. security policy and carefully monitoring any bank transfers for false information.Check any e-mails sent to you are legitimate by looking at the e-mail headers, and scan any attachments for viruses before opening them.
On Sat, 2021-02-27 at 09:34 -0800, N Shane Patrick wrote:
We have a haskris chiller on a JEOL SEM and the agreed procedure is to add a bottle of drug store peroxide to tank, let it run normally, no isolation, overnight or so, then drain, flush, and fill and return to service. Algae is usually our biggest issue, and this is more than enough to handle it. I suspect most anything will hold up fine to the dilute peroxide solution. I have other chillers and systems where CLR has been used for other types of clean out.
-Shane
Sent from my iPhone
On Feb 27, 2021, at 9:10 AM, James Mitchell <jtmitch5 at ncsu.edu> wrote:
I have used CLR to clean out cooling lines but you would want to flush it out afterwards. Mix the CLR with water and drain out after cycling. There are biocides you can add to cooling water to keep it clean.
Jim
On Fri, Feb 26, 2021, 3:42 PM Saba Sadeghi <saba.sadeghi at uwaterloo.ca<mailto:saba.sadeghi at uwaterloo.ca>> wrote:
Dear colleagues,
We have chillers from Haskris for cooling our MBE deposition system and also XRD system; I was wondering if you have a procedure that you recommend for cleaning the chillers' water. According to Haskris you can use hydrogen peroxide to clean the water should there be any biological growth in the chiller. For this, do I need to isolate the chiller from the system or can hydrogen peroxide be run through the entire system?
I appreciate your input.
Best,
Saba
--
Saba Sadeghi, PhD
Quantum Devices Fabrication Scientist
Institute for Quantum Computing, University of Waterloo
Email: saba.sadeghi at uwaterloo.ca<mailto:saba.sadeghi at uwaterloo.ca>
_______________________________________________
labnetwork mailing list
labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>
https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
_______________________________________________
labnetwork mailing list
labnetwork at mtl.mit.edu
https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
_______________________________________________
labnetwork mailing list
labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>
https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
More information about the labnetwork
mailing list