[labnetwork] Regarding : Aluminium E-Beam Evaporation
Ted Wangensteen
twangens at mail.usf.edu
Tue Aug 30 09:05:27 EDT 2022
Ryan has hit on the key thermal explanation, and the best way to help this
is the Carbon spacer he mentioned.
We used this on Lesker tools.
Kurt Lesker sells the spacer, but I'm not sure what tool you are using.
-Ted Wangensteen
On Tue, Aug 30, 2022 at 8:07 AM Ryan Rivers <rrivers at berkeley.edu> wrote:
> The best way around the aluminum crucible shattering problem is to use a
> FABMATE crucible, but also to use a carbon spacer under your liner. A 1/8"
> thick carbon disk works in a pinch, you can get fancier with some machine
> shop time. We use them in all our ebeam tools the Marvell NanoLab and I've
> only seen a handful of crucible liners shatter from thermal load in the
> last decade.
>
> Primary cause of crucible liner breaks are incomplete thermal contact on
> one side of the crucible. When you're pouring too much heat in to make up
> for that contact, one side goes red hot, the other never gets there.
> Thermal expansion drives the rest. The spacer prevents your liner from
> cooling through the sides and only cools at the bottom. You heat with less
> joules in, so less shock. Lets your entire crucible heat more evenly and
> more readily stabilize through the problematic regions of melt formation.
>
> -Ryan
>
> On Mon, Aug 29, 2022 at 4:47 PM Michael Yakimov <yakimom at sunypoly.edu>
> wrote:
>
>> One of things which may hinder Al deposition from carbon crucible is
>> Aluminum carbide formation on the surface. It will show up as yellow stuff
>> on the surface after venting, eventually decomposing in the air.
>> It is a part of tribal knowledge for me; I couldn't find a good reference
>> in 5 minutes. there is mention of the effect here:
>> https://link.springer.com/article/10.1007/BF00792405
>> <https://link.springer.com/article/10.1007/BF00792405>
>> Industrial operation of vacuum aluminum evaporators made in refractory
>> compound alloys - SpringerLink
>> <https://link.springer.com/article/10.1007/BF00792405>
>> Vaporizing elements of various types and sizes are being produced from a
>> TiB2-TiC alloy in the Special Design and Technology Department of the
>> Institute of Materials Science, Academy of Sciences of the Ukrainian SSR,
>> and widely used in vacuum deposition plants for the application of aluminum
>> coatings to glass, plastics, fabrics, paper, metals, and film and other
>> coiling materials. The ...
>> link.springer.com
>>
>>
>>
>> ------------------------------
>> *From:* labnetwork <labnetwork-bounces at mtl.mit.edu> on behalf of Zhichao
>> Wang <zhichaow at udel.edu>
>> *Sent:* Monday, August 29, 2022 12:14 PM
>> *To:* Chandan H B <chandanachar95 at gmail.com>
>> *Cc:* labnetwork at mtl.mit.edu <labnetwork at mtl.mit.edu>
>> *Subject:* Re: [labnetwork] Regarding : Aluminium E-Beam Evaporation
>>
>> Chandan,
>>
>> We use a pattern beam with the pattern sweeping/rotating constantly
>> during the deposition, and a fabmate crucible. The rest of the parameters
>> are pretty similar. The highest deposition rate we've tried is 15
>> Angstrom/s.
>>
>> --
>> Chao
>>
>> On Mon, Aug 29, 2022 at 8:07 AM Chandan H B <chandanachar95 at gmail.com>
>> wrote:
>>
>> Dear Lab network community,
>>
>> We are currently facing an issue in the deposition of Aluminium in one of
>> our Electron beam evaporation tools, we tried intermetallic, fab mate,
>> copper, glassy coated graphite, and graphite crucibles for the evaporation.
>> It seems that none of them are working out.
>> We see crucible gets broken at the initial deposition quite often.
>>
>> Are we missing out on any parameters unchecked?
>> Kindly recommend us a few parameters or solutions for the same.
>>
>> Any suggestions/Inputs are highly appreciated.
>> Thanks in advance!
>>
>> Regards,
>> CHANDAN H B
>> THIN FILM ENGINEER
>>
>> P.S: Here are a few parameters that are provided to the tool.
>> Power: 10KW
>> Voltage: 10KV constant
>> Current: Variable
>> Rise1 & Soak1: 5min & 1min /8min & 2min
>> Rise1 Power: 4% (for intermetallic crucible)
>> Rise2 & Soak2: 5min & 1min /8min & 2min
>> Rise2 Power: 8% (for intermetallic crucible)
>> Ramp down: 5min
>> Rate of deposition: 0.1nm/sec
>> Beam Pattern: Spot Beam at the center of the crucible
>> Crucible Volume: 20cc
>> Material fill %: As recommended 67-75%
>>
>>
>> _______________________________________________
>> labnetwork mailing list
>> labnetwork at mtl.mit.edu
>> https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
>>
>> _______________________________________________
>> labnetwork mailing list
>> labnetwork at mtl.mit.edu
>> https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
>>
> _______________________________________________
> labnetwork mailing list
> labnetwork at mtl.mit.edu
> https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20220830/544cf599/attachment.html>
More information about the labnetwork
mailing list