From hollingshead.19 at osu.edu Mon May 2 10:04:42 2022 From: hollingshead.19 at osu.edu (Hollingshead, Dave) Date: Mon, 2 May 2022 14:04:42 +0000 Subject: [labnetwork] Direct write + plasma etch residue In-Reply-To: <1937977434.1689956.1651277765137@mail.yahoo.com> References: <1937977434.1689956.1651277765137@mail.yahoo.com> Message-ID: Hi Joon, Thank you for the feedback and thank you to everyone else that has responded. I responded to many of you directly, but this topic has come up a few times so I wanted to address it with the whole group. We historically have not noticed this phenomenon on the contact aligner, although we have not run that SPR220 process with an etch in a few years. That is another experiment we are planning but have not done so yet. We will have some SEMs of these particular sidewalls this week, but we do have SEM x-sections of other recent samples and I can confirm that the side walls are directionally angled usually with a slope differential of ~5?. We have worked with Heidelberg a lot on this issue and have it mostly solved, but it is still somewhat visible on thicker resists. The problem was actually much, much worse with negative resists (nLOF2020 or nLOF2070) where the slope variation was so extreme that one side of the pattern would be undercut, while the other positive. Attached is an example of the "poor" profile prior to the fix, along with the more recent "improved" profile. Heidelberg was able to fix the alignment well enough that we were happy with our negative resist profiles but we have not been able to get it perfectly orthogonal. So we know that the light is slightly angled coming in, but I am still unsure why this would lead to the etch issues that we are seeing. Thanks, -Dave -----Original Message----- From: Joonhyeong Park Sent: Friday, April 29, 2022 20:16 To: labnetwork at mtl.mit.edu; Hollingshead, Dave Subject: Re: [labnetwork] Direct write + plasma etch residue Hello Mr. Hollingshead, When you did optical lithography with contact mask aligner, SPR220 and S1827, did you get the similar result such as only one side residue? Have you checked the cross-section in SEM after develop before etching? If there is a directional problem in MLA150, you may observed the difference in side angle in SEM. You may also check sidewall corner residue under darkfield microscope, too. I think that you had better break down and narrow down the cause and result step by step.? Regards, Joon Park Research Engineer Birck Nanotechnology Center Purdue University On Friday, April 29, 2022, 03:40:02 PM EDT, Hollingshead, Dave wrote: ?? Good afternoon everyone, ? We have recently run into some very ?interesting? (completely perplexing) issues while trying to etch some III-V samples in our ICP. Essentially we are seeing some sort of residue on the protected regions of the devices after stripping the resist. Whatever it is, the material seems to also be affecting our etch profiles. The odd part is that the residue has a clear directionality to it and in the present case only appears on two edges. This has led us to believe that is must in some way be a result of some non-uniform / ?angled? exposure from our MLA150. While I can completely understand differing etched sidewall slopes or skewed features if the resist sidewalls are not uniform on all sides, I am having difficulty understanding how this might cause the residue we are seeing (especially in a positive resist). ? Any thoughts or possible explanations would be greatly appreciated. ? Attached are some microscope and SEM images of what we are seeing. I have plenty more I?d be happy to share with those that may be interested! ? A few relevant processing notes: ???? ????* All etches were done using a BCl3/Ar chemistry in our Plasmatherm 770 system ????* ~18-20 minutes of etch time, total depth of ~2.5-3.0?m ????* We have repeated these issues on InGaAs device samples as well as bulk GaAs test samples ????* Issue repeated with both SPR220 and S1827 resist masks (both were un-hardbaked, but this is also something we are going to look into) ????* Exposures seem to be properly dosed, but we are planning an etch test on a dose/defocus array to see if exposure has any effect ????* All resist removal was done in 80C NMP ????* We have attempted some follow-on O?2 plasma etches, that seem to remove the material, but it is very slow ? Thanks, -Dave ? Dave Hollingshead?Manager of Research Operations ? Nanotech West LabThe Ohio State University Suite 100, 1381 Kinnear Road, Columbus, OH 43212 614.292.1355 Office hollingshead.19 at osu.edu?osu.edu ? _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://urldefense.com/v3/__https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork__;!!KGKeukY!yFwm1jA0dCy91O08GbuxMvJ7GBlqrnZAfaSSI8DZK6kuxnPkKlZqVdm1wql5PRdxOygw6jds5P9gHbUrej2bdHBn$ -------------- next part -------------- A non-text attachment was scrubbed... Name: OSU_MLA_asymmetry_comparison.png Type: image/png Size: 574923 bytes Desc: OSU_MLA_asymmetry_comparison.png URL: From rreger at purdue.edu Mon May 2 11:34:15 2022 From: rreger at purdue.edu (Reger, Ronald K) Date: Mon, 2 May 2022 15:34:15 +0000 Subject: [labnetwork] Opportunity at the Birck Nanotechnology Center Message-ID: Dear Colleagues, Please see this link for a Senior Research Associate in our Birck Nanotechnology Center: https://www.talent.com/view?id=264f4934aa8d Please pass this along to anyone who may have an interest. Thanks much, Ron Ron Reger Engineering Manager Birck Nanotechnology Center | Room 2289 | Office: 765.494.6667 | Email: rreger at purdue.edu Wiki: https://wiki.itap.purdue.edu/display/BNCWiki iLabs: https://purdue.ilabsolutions.com/homepage/ -------------- next part -------------- An HTML attachment was scrubbed... URL: From rlc12 at cornell.edu Wed May 4 14:05:26 2022 From: rlc12 at cornell.edu (Rebecca Lee Vliet) Date: Wed, 4 May 2022 18:05:26 +0000 Subject: [labnetwork] June 2022 CNF TCN Short Course - REGISTRATION OPEN! Message-ID: Please post & share the announcement below with other members of the scientific communities. Thank you! CNF Technology & Characterization at the Nanoscale (CNF TCN) IN-PERSON Short Course June 7, 2022 - June 10, 2022 The CNF TCN will be in-person for June 2022! The course will be held Tuesday, June 7, 2022 - Friday, June 10, 2022. Each day offers lectures and laboratory demonstrations designed to impart a broad understanding of the science and technology required to undertake research in nanoscience. TCN is an ideal way for faculty, students, post docs and staff members to rapidly come up to speed in many of the technologies that users of the CNF need to employ. Members of the high tech business community will also find it an effective way to learn best practices for success in a nanofab environment. Attendance is open to the general scientific community. Note: The short course does not replace the three part training required to become a user of our facility. To become a CNF user, please visit the "Getting Started" link (https:cnf.cornell.edu/howto) on the CNF website. For more information and to register visit: https://cnf.cornell.edu/education/tcn [Text Description automatically generated] -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image002.jpg Type: image/jpeg Size: 488127 bytes Desc: image002.jpg URL: From dcchrist at wisc.edu Fri May 6 11:33:24 2022 From: dcchrist at wisc.edu (Daniel Christensen) Date: Fri, 6 May 2022 15:33:24 +0000 Subject: [labnetwork] Preliminary Program for UGIM 2022 Message-ID: Hi all, The Preliminary Program for UGIM 2022 is now online at; https://ugim2020.wisc.edu/program/ We hope to see you all in June! Jerry Hunter and Dan Christensen UGIM 2022 Co-Chairs Daniel C. Christensen Laboratory Manager Nanoscale Fabrication Center University of Wisconsin-Madison -------------- next part -------------- An HTML attachment was scrubbed... URL: From dsbarth at seas.upenn.edu Fri May 6 13:24:17 2022 From: dsbarth at seas.upenn.edu (David Barth) Date: Fri, 6 May 2022 13:24:17 -0400 Subject: [labnetwork] Ion Mill/Ion Beam Etch Tools Message-ID: Hi all, We're looking into buying an ion mill, and I'm hoping to hear some opinions and experiences about your ion mill tools. Any thoughts on the usefulness of optional features (SIMS, OES, reactive gases, etc.) would also be great. I'd appreciate any feedback on these issues. Thanks, David Barth Sr. Manager Quattrone Nanofabrication Facility University of Pennsylvania dsbarth at seas.upenn.edu -------------- next part -------------- An HTML attachment was scrubbed... URL: From alexw at caltech.edu Fri May 6 18:52:24 2022 From: alexw at caltech.edu (Wertheim, Alex R.) Date: Fri, 6 May 2022 22:52:24 +0000 Subject: [labnetwork] Seeking refurbished CTI Cryotorr 8F (Conflat flange) leads Message-ID: Hey All, I'm looking to find leads on some good deals on a CryoTorr 8F conflat cryopump. Rebuild/refurbished preferred. Anyone have any shops you highly recommend? Thanks! ________________________ Alex Wertheim (He/Him) Materials & Process Engineer The Kavli Nanoscience Institute at Caltech Office: 1-626-395-3371 Email: alexw at caltech.edu -------------- next part -------------- An HTML attachment was scrubbed... URL: From James_Goodman at uml.edu Sat May 7 06:27:21 2022 From: James_Goodman at uml.edu (Goodman, James R) Date: Sat, 7 May 2022 10:27:21 +0000 Subject: [labnetwork] Seeking refurbished CTI Cryotorr 8F (Conflat flange) leads In-Reply-To: References: Message-ID: Good morning, I just worked with a company out of Texas called ?Get Spares ?. They have a pretty good selection of cryo pumps and were pretty easy to deal with. Jay. Get Outlook for iOS ________________________________ From: labnetwork on behalf of Wertheim, Alex R. Sent: Friday, May 6, 2022 6:52:24 PM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Seeking refurbished CTI Cryotorr 8F (Conflat flange) leads This e-mail originated from outside the UMass Lowell network. ________________________________ Hey All, I'm looking to find leads on some good deals on a CryoTorr 8F conflat cryopump. Rebuild/refurbished preferred. Anyone have any shops you highly recommend? Thanks! ________________________ Alex Wertheim (He/Him) Materials & Process Engineer The Kavli Nanoscience Institute at Caltech Office: 1-626-395-3371 Email: alexw at caltech.edu -------------- next part -------------- An HTML attachment was scrubbed... URL: From wjkiethe at ncsu.edu Mon May 9 10:28:54 2022 From: wjkiethe at ncsu.edu (William Kiether) Date: Mon, 9 May 2022 10:28:54 -0400 Subject: [labnetwork] Seeking refurbished CTI Cryotorr 8F (Conflat flange) leads In-Reply-To: References: Message-ID: We have had a good experience with Texas Capital Semiconductor with CTI cryos. TCsemi.com Bill Kiether NCSU Nanofabrication Facility On Sat, May 7, 2022 at 8:06 AM Goodman, James R wrote: > Good morning, > > I just worked with a company out of Texas called ?Get Spares ?. They have > a pretty good selection of cryo pumps and were pretty easy to deal with. > > Jay. > > Get Outlook for iOS > ------------------------------ > *From:* labnetwork on behalf of > Wertheim, Alex R. > *Sent:* Friday, May 6, 2022 6:52:24 PM > *To:* labnetwork at mtl.mit.edu > *Subject:* [labnetwork] Seeking refurbished CTI Cryotorr 8F (Conflat > flange) leads > > *This e-mail originated from outside the UMass Lowell network.* > ------------------------------ > Hey All, > > I'm looking to find leads on some good deals on a CryoTorr 8F conflat > cryopump. Rebuild/refurbished preferred. Anyone have any shops you highly > recommend? > > Thanks! > > ________________________ > Alex Wertheim (He/Him) > Materials & Process Engineer > The Kavli Nanoscience Institute at Caltech > Office: 1-626-395-3371 > Email: alexw at caltech.edu > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork > -------------- next part -------------- An HTML attachment was scrubbed... URL: From jmaduzi2 at illinois.edu Mon May 9 10:36:39 2022 From: jmaduzi2 at illinois.edu (Maduzia, Joseph Walter) Date: Mon, 9 May 2022 14:36:39 +0000 Subject: [labnetwork] Seeking refurbished CTI Cryotorr 8F (Conflat flange) leads In-Reply-To: References: Message-ID: Hello Alex, I have had 2 cryo pumps refurbished and purchased 1 refurbished helium compressor and 1 refurbished cryo from Bay Technologies. Tim Valdivia is incredibly responsive and very helpful. He is also very knowledgeable and helps troubleshoot cryos and compressors over the phone. I highly recommend them. All three cryos and the compressor are still running 24/7 since 2017. Tim Valdivia 16540 Railroad Ave Suite 150 Morgan Hill, CA 95037 Tim at bay-technologies.net www.bay-technologies.net 408-313-1100 I also get spares from here https://austincryogenics.com/Home_Page.html Thank you, JOE MADUZIA Senior Research Engineer, Grainger College of Engineering, MechSE, Univ of IL 2239 LuMEB | (P) 217.244.6302 | https://cleanroom.mechse.illinois.edu/ From: labnetwork On Behalf Of Wertheim, Alex R. Sent: Friday, May 6, 2022 5:52 PM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Seeking refurbished CTI Cryotorr 8F (Conflat flange) leads Hey All, I'm looking to find leads on some good deals on a CryoTorr 8F conflat cryopump. Rebuild/refurbished preferred. Anyone have any shops you highly recommend? Thanks! ________________________ Alex Wertheim (He/Him) Materials & Process Engineer The Kavli Nanoscience Institute at Caltech Office: 1-626-395-3371 Email: alexw at caltech.edu -------------- next part -------------- An HTML attachment was scrubbed... URL: From tafuller at eng.ucsd.edu Mon May 9 13:05:04 2022 From: tafuller at eng.ucsd.edu (Timothy Fuller) Date: Mon, 9 May 2022 10:05:04 -0700 Subject: [labnetwork] Seeking refurbished CTI Cryotorr 8F (Conflat flange) leads In-Reply-To: References: Message-ID: I have always relied on Austin Scientific (now part of Trillium) for aftermarket and their own brand of cryo pumps and service. They were having supply chain issues on some items, but hopefully that has eased. Tim Fuller Equipment Engineer UCSD Calit2 Nano3 Facility On Mon, May 9, 2022 at 8:54 AM William Kiether wrote: > We have had a good experience with Texas Capital Semiconductor with CTI > cryos. TCsemi.com > > Bill Kiether > NCSU Nanofabrication Facility > > On Sat, May 7, 2022 at 8:06 AM Goodman, James R > wrote: > >> Good morning, >> >> I just worked with a company out of Texas called ?Get Spares ?. They have >> a pretty good selection of cryo pumps and were pretty easy to deal with. >> >> Jay. >> >> Get Outlook for iOS >> ------------------------------ >> *From:* labnetwork on behalf of >> Wertheim, Alex R. >> *Sent:* Friday, May 6, 2022 6:52:24 PM >> *To:* labnetwork at mtl.mit.edu >> *Subject:* [labnetwork] Seeking refurbished CTI Cryotorr 8F (Conflat >> flange) leads >> >> *This e-mail originated from outside the UMass Lowell network.* >> ------------------------------ >> Hey All, >> >> I'm looking to find leads on some good deals on a CryoTorr 8F conflat >> cryopump. Rebuild/refurbished preferred. Anyone have any shops you highly >> recommend? >> >> Thanks! >> >> ________________________ >> Alex Wertheim (He/Him) >> Materials & Process Engineer >> The Kavli Nanoscience Institute at Caltech >> Office: 1-626-395-3371 >> Email: alexw at caltech.edu >> _______________________________________________ >> labnetwork mailing list >> labnetwork at mtl.mit.edu >> https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork >> > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork > -------------- next part -------------- An HTML attachment was scrubbed... URL: From lvchang at Central.UH.EDU Mon May 9 14:13:31 2022 From: lvchang at Central.UH.EDU (Chang, Long) Date: Mon, 9 May 2022 18:13:31 +0000 Subject: [labnetwork] Lamp for Backside alignment Message-ID: <5F9D5321-B020-4C9D-B74B-D85D7302C9F4@cougarnet.uh.edu> Hello, Anyone know what the lamp is for backside alignment? There are no markings on the lamp with gold reflector to help me find a replacement. We have a stock of EKE 12V150W V5KD with silver reflector that we use for frontside alignment, see photo. I tried to replace it with our stock lamp anyways, but I couldn?t see the alignment mark. Thanks, Long [cid:D5ACBA77-8A73-4C1F-AC6A-6E3152B30E53 at serc.dhcp.e.uh.edu] -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: Screen Shot 2022-05-09 at 1.06.44 PM.png Type: image/png Size: 515694 bytes Desc: Screen Shot 2022-05-09 at 1.06.44 PM.png URL: From gregmullins at berkeley.edu Mon May 9 17:54:05 2022 From: gregmullins at berkeley.edu (Greg Mullins) Date: Mon, 9 May 2022 14:54:05 -0700 Subject: [labnetwork] Lamp for Backside alignment In-Reply-To: <5F9D5321-B020-4C9D-B74B-D85D7302C9F4@cougarnet.uh.edu> References: <5F9D5321-B020-4C9D-B74B-D85D7302C9F4@cougarnet.uh.edu> Message-ID: Hi Long, What system is this for? On my Quintel mask aligner, I use this one! On Mon, May 9, 2022 at 1:48 PM Chang, Long wrote: > Hello, > > Anyone know what the lamp is for backside alignment? There are no markings > on the lamp with gold reflector to help me find a replacement. We have a > stock of EKE 12V150W V5KD with silver reflector that we use for frontside > alignment, see photo. I tried to replace it with our stock lamp anyways, > but I couldn?t see the alignment mark. > > Thanks, > Long > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork > -- Greg Mullins R&D Engineer 4 Marvell Nanofabrication Laboratory University of California at Berkeley 520 Sutardja Dai Hall, MS 1754 Berkeley, CA 94720-1754 Office: (510) 809-8625 Email: gregmullins at berkeley.edu -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: Screen Shot 2022-05-09 at 1.06.44 PM.png Type: image/png Size: 515694 bytes Desc: not available URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: IMG_9982.jpg Type: image/jpg Size: 2566026 bytes Desc: not available URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: IMG_9983.jpg Type: image/jpg Size: 1782057 bytes Desc: not available URL: From sandra.malhotra at tamu.edu Tue May 10 12:46:50 2022 From: sandra.malhotra at tamu.edu (Malhotra, Sandra Guy) Date: Tue, 10 May 2022 16:46:50 +0000 Subject: [labnetwork] Ta etch recipe Message-ID: Howdy All, We have an Oxford Plasmalab 100 ICP RIE in AggieFab, and a user is attempting to develop a Ta etch recipe with straight sidewalls. If anyone has experience developing such a process and can share pointers, we would be very appreciative! Thanks for your help. Sandra G. Malhotra, Ph.D. | Technical Lab Manager AggieFab Nanofabrication Facility Department of Electrical & Computer Engineering, College of Engineering | Texas A&M University 3253 TAMU | College Station, TX 77843 ph: 979.845.3199 | sandra.malhotra at tamu.edu - - - - - - - - - - - - - - - - - - - - - - - - TEXAS A&M UNIVERSITY | FEARLESS on Every Front -------------- next part -------------- An HTML attachment was scrubbed... URL: From rlc12 at cornell.edu Wed May 11 14:39:44 2022 From: rlc12 at cornell.edu (Rebecca Lee Vliet) Date: Wed, 11 May 2022 18:39:44 +0000 Subject: [labnetwork] REMINDER: June 2022 CNF TCN Short Course - REGISTRATION OPEN! Message-ID: Please post & share the announcement below with other members of the scientific communities. Thank you! CNF Technology & Characterization at the Nanoscale (CNF TCN) IN-PERSON Short Course June 7, 2022 - June 10, 2022 The CNF TCN will be in-person for June 2022! The course will be held Tuesday, June 7, 2022 - Friday, June 10, 2022. Each day offers lectures and laboratory demonstrations designed to impart a broad understanding of the science and technology required to undertake research in nanoscience. TCN is an ideal way for faculty, students, post docs and staff members to rapidly come up to speed in many of the technologies that users of the CNF need to employ. Members of the high tech business community will also find it an effective way to learn best practices for success in a nanofab environment. Attendance is open to the general scientific community. Seating is limited. Note: The short course does not replace the three part training required to become a user of our facility. To become a CNF user, please visit the "Getting Started" link (https:cnf.cornell.edu/howto) on the CNF website. Cornell COVID-19 Response guidelines can be found here: https://covid.cornell.edu/visitors/. Masks are required in all CNF lab spaces. For more information and to register visit: https://cnf.cornell.edu/education/tcn [Text Description automatically generated] -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image002.jpg Type: image/jpeg Size: 488127 bytes Desc: image002.jpg URL: From mokhodada at gmail.com Thu May 12 01:10:45 2022 From: mokhodada at gmail.com (Mo Khodadadi) Date: Thu, 12 May 2022 01:10:45 -0400 Subject: [labnetwork] Open Position - Microfabrication Process Engineer Message-ID: Hello everyone, We're looking for qualified applicants to consider our open position of Microfabrication Process Engineer at EnaChip Inc. Link to the position: Lithography & Microfabrication Process Engineer - Rocky Hill, NJ 08553 - Indeed.com Best Regards, Mo Khodadadi -------------- next part -------------- An HTML attachment was scrubbed... URL: From mtang at stanford.edu Thu May 12 10:19:19 2022 From: mtang at stanford.edu (Mary Tang) Date: Thu, 12 May 2022 07:19:19 -0700 Subject: [labnetwork] Process Integration/Curriculum Dev position at the Stanford Nanofab Message-ID: Hi everyone! If you love teaching and working with a diverse community of researchers who want to make the world a better place -consider Stanford. We have an opening at the Nanofab for a Process Integration/Curriculum Development Manager - a long title for fellow staffer who enjoys device technologies and is excited about sharing knowledge. More details can be found at Stanford Jobs (position #17012 ) or contacting me. Cheers! Mary -- Mary X. Tang, Ph.D. Managing Director Stanford Nanofabrication Facility Paul G. Allen Building 420 Via Palou Mall Stanford, CA 94305 https://snf.stanford.edu mtang at stanford.edu -------------- next part -------------- An HTML attachment was scrubbed... URL: From cschnitzer at stonehill.edu Fri May 13 11:13:04 2022 From: cschnitzer at stonehill.edu (Schnitzer, Cheryl) Date: Fri, 13 May 2022 15:13:04 +0000 Subject: [labnetwork] Instructors Needed: Electronics, Manufacturing, and Applied Optics Message-ID: Instructors Needed: Electronics, Manufacturing, and Applied Optics Stonehill College is looking for instructors for its Photonics Certificate Program. This program, which will begin its third year in the fall is intended to train technicians for work in the photonics industry. The program consists of a sequence of general manufacturing and electronics courses, followed by several optics and photonics courses, with emphasis on hands-on work and on skills relevant to the photonics industry. These 12-14 week-long courses are in-person, weekday evenings, and there is no expectation of prior background in math, electronics, or industry for the students in the program. The ideal instructors would have both industry experience and some background in worker training or higher education. The courses place heavy emphasis on hands-on experience (use of machine tools, design and fabrication procedures, etc.), but some of the courses are also expected to cover background in areas such as basic math and software skills. Safety protocols will be strongly emphasized in all courses, and online OSHA-10 and laser safety training is required of the students as part of the program. Classes will be capped at 20 students, and students range in ages from late teens to early 50's with different levels of education and experience. The courses that we expect to need instructors for in the coming year are: * Electricity and Electronics - Fall semester 2022 * Tools and Testing - Fall semester 2022 (Covers basic measurements with oscilloscopes, calipers, microscopes, etc., as well as practice in welding and other basic electronics and manufacturing skills.) * Digital Fundamentals - Spring semester 2023 * Statistical Process Control - Spring semester 2023 * Tools and Materials - Spring semester 2023 (Introduction to manufacturing systems, design of engineering processes, and CNC manufacturing) * Applied Optics - Summer semester 2023 * Fiber Optics - Summer semester 2023 The 12-month program includes an internship with a local company during the final semester and was constructed with extensive input from local companies. The goal is to prepare the students for immediate employment once the certificate is completed, and the program achieved 100% success in placing students from the first graduating class. Stonehill College is a private, liberal arts college in Easton, MA, about 20 miles south of Boston, and is committed to diversity in its student body and in its instructors. More information about the program is available at: www.stonehill.edu/photonics. For further information about the program or the available positions, please contact Profs. Cheryl Schnitzer, Ruby Gu, or David Simon at cschnitzer at stonehill.edu, ggu at stonehill.edu, dsimon at stonehill.edu. _________________________________ Cheryl Schnitzer, Ph.D. (she/her/hers) * Professor, Department of Chemistry, Stonehill College * Director, Photonics Certificate Program * Chair, Environmental Stewardship Council Good chemistry requires all of our elements. The Department of Chemistry and the Biochemistry Program value the perspectives, experiences, and identities of each and every individual. -------------- next part -------------- An HTML attachment was scrubbed... URL: From rlc12 at cornell.edu Mon May 16 11:12:13 2022 From: rlc12 at cornell.edu (Rebecca Lee Vliet) Date: Mon, 16 May 2022 15:12:13 +0000 Subject: [labnetwork] REMINDER: June 2022 CNF TCN Short Course - Seats still available! Message-ID: REMINDER: June 2022 CNF TCN Short Course - Seats still available! Please post & share the announcement below with other members of the scientific communities. Thank you! CNF Technology & Characterization at the Nanoscale (CNF TCN) IN-PERSON Short Course June 7, 2022 - June 10, 2022 The CNF TCN will be in-person for June 2022! The course will be held Tuesday, June 7, 2022 - Friday, June 10, 2022. Each day offers lectures and laboratory demonstrations designed to impart a broad understanding of the science and technology required to undertake research in nanoscience. TCN is an ideal way for faculty, students, post docs and staff members to rapidly come up to speed in many of the technologies that users of the CNF need to employ. Members of the high tech business community will also find it an effective way to learn best practices for success in a nanofab environment. Attendance is open to the general scientific community. Seating is limited. Note: The short course does not replace the three part training required to become a user of our facility. To become a CNF user, please visit the "Getting Started" link (https:cnf.cornell.edu/howto) on the CNF website. Cornell COVID-19 Response guidelines can be found here: https://covid.cornell.edu/visitors/. Masks are required in all CNF lab spaces. For more information and to register visit: https://cnf.cornell.edu/education/tcn [Text Description automatically generated] -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image002.jpg Type: image/jpeg Size: 488127 bytes Desc: image002.jpg URL: From sb86922 at usc.edu Mon May 16 12:55:51 2022 From: sb86922 at usc.edu (Shivakumar Bhaskaran) Date: Mon, 16 May 2022 16:55:51 +0000 Subject: [labnetwork] Job Opportunity at USC Message-ID: I am super excited to announce that John O' Brien Nanofab at USC is fully operational! Our user base is growing, and we are adding new tools. We are looking for an energetic person to work on our team. The USC Viterbi School of Engineering in Los Angeles is looking for Nanofabrication Lab Operation Engineer to work at our cleanroom, John O Brien Nanofabrication Laboratory. Please forward this to anyone interested. Job Description: We are looking for a highly motivated Lab Operation Engineer. Reporting to the Associate Director, this individual will support the cleanroom users with training, develop a nanofabrication process, and collaborate with facility staff and research users to identify tool operation issues and promptly address those issues with proper maintenance and repair and troubleshooting. The ideal candidate should be customer service-oriented and contribute to improving the facility's operation. https://usccareers.usc.edu/job/los-angeles/nanofabrication-lab-operation-engineer/1209/23295731312 Requisition ID: REQ20112629 Job Title: Nanofabrication Lab Operation Engineer Shivakumar Bhaskaran, Ph.D., Associate Director, John D. O'Brien Nanofabrication Laboratory Michelson Hall, 1002 Childs Way, MCB LL121, Los Angeles, California 90089, 213 821 2374 [Sign] -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.jpg Type: image/jpeg Size: 2931 bytes Desc: image001.jpg URL: From sb86922 at usc.edu Mon May 16 18:27:22 2022 From: sb86922 at usc.edu (Shivakumar Bhaskaran) Date: Mon, 16 May 2022 22:27:22 +0000 Subject: [labnetwork] Perchloric Acid Message-ID: Hi All, Do any you use 70% Perchloric Acid in your Acid bench. Is there any additional retrofit needed to use Perchloric acid (Internal wash-down for heating percholoric acid or scrubber recirculated tank). Thanks Shiva Shivakumar Bhaskaran, Ph.D., Associate Director, John D. O'Brien Nanofabrication Laboratory Michelson Hall, 1002 Childs Way, MCB LL121, Los Angeles, California 90089, 213 821 2374 [Sign] -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.jpg Type: image/jpeg Size: 2931 bytes Desc: image001.jpg URL: From deonc69 at illinois.edu Tue May 17 09:56:24 2022 From: deonc69 at illinois.edu (Collins, Deon) Date: Tue, 17 May 2022 13:56:24 +0000 Subject: [labnetwork] Perchloric Acid In-Reply-To: References: Message-ID: Perchloric acid requires a specific fume hood with certain precautions. Glennys could you elaborate a bit on the subject? Deon D. Collins FACILITY MANAGER Holonyak Lab University of Illinois at Urbana-Champaign 208 N Wright St Rm. 248 | 2250 Urbana, IL 61801 217-300-7531 | deonc69 at illinois.edu [https://webtools.illinois.edu/webservices/js/ds/signature_logo.png] Under the Illinois Freedom of Information Act any written communication to or from university employees regarding university business is a public record and may be subject to public disclosure. Life is not about watching other people live it. It's about you living your own! From: labnetwork On Behalf Of Shivakumar Bhaskaran Sent: Monday, May 16, 2022 5:27 PM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Perchloric Acid Hi All, Do any you use 70% Perchloric Acid in your Acid bench. Is there any additional retrofit needed to use Perchloric acid (Internal wash-down for heating percholoric acid or scrubber recirculated tank). Thanks Shiva Shivakumar Bhaskaran, Ph.D., Associate Director, John D. O'Brien Nanofabrication Laboratory Michelson Hall, 1002 Childs Way, MCB LL121, Los Angeles, California 90089, 213 821 2374 [Sign] -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image002.png Type: image/png Size: 2602 bytes Desc: image002.png URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.jpg Type: image/jpeg Size: 2116 bytes Desc: image001.jpg URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: Mr_ Deon D_ Collins.vcf Type: text/x-vcard Size: 2025 bytes Desc: Mr_ Deon D_ Collins.vcf URL: From chris at malocsay.com Wed May 18 16:57:36 2022 From: chris at malocsay.com (Chris Malocsay) Date: Wed, 18 May 2022 13:57:36 -0700 Subject: [labnetwork] Customer needs Cr/Au coating on slides Message-ID: Greetings: I have a customer in Illinois who needs a Sequentially coated 1 x 4 inch substrate with Cr and Au. Is there a University Nano fab in the area they can contract for coating services? Best Regards Chris Chris Malocsay 510-506-5894 -------------- next part -------------- An HTML attachment was scrubbed... URL: From sguo18 at yorku.ca Thu May 19 11:05:06 2022 From: sguo18 at yorku.ca (Xin (Shane) Guo) Date: Thu, 19 May 2022 15:05:06 +0000 Subject: [labnetwork] KLA Tencor D600 profilometer stage cover Message-ID: Hello everyone, The laser shielding cover of our KLA Tencor D600 profilometer was cracked. Has anyone bought a replacement cover from a 3rd party or made one before? The official replacement is way too expensive for us to afford. The laser in the profilometer is Class IIIa, 635nm and 5mW, but it doesn?t seem to be easy to find shield plastic plaques for this wavelength. The original cover looks like dark brown colour. Any idea of what material I can choose or which vendor that I can contact? Thank you! Shane Guo -------------- next part -------------- An HTML attachment was scrubbed... URL: From julia.aebersold at louisville.edu Thu May 19 11:16:43 2022 From: julia.aebersold at louisville.edu (Aebersold,Julia W.) Date: Thu, 19 May 2022 15:16:43 +0000 Subject: [labnetwork] Customer needs Cr/Au coating on slides In-Reply-To: References: Message-ID: Hi Chris! We can perform such services, but can you give more detail as to how much chrome and gold your client needs deposited? Cheers! Julia Aebersold, Ph.D. Manager, Micro/Nano Technology Center University of Louisville Shumaker Research Building, Room 233 2210 South Brook Street Louisville, KY 40292 (502) 852-1572 http://louisville.edu/micronano/ From: labnetwork On Behalf Of Chris Malocsay Sent: Wednesday, May 18, 2022 4:58 PM To: Labnetwork Mailing List Subject: [labnetwork] Customer needs Cr/Au coating on slides Some people who received this message don't often get email from chris at malocsay.com. Learn why this is important CAUTION: This email originated from outside of our organization. Do not click links, open attachments, or respond unless you recognize the sender's email address and know the contents are safe. Greetings: I have a customer in Illinois who needs a Sequentially coated 1 x 4 inch substrate with Cr and Au. Is there a University Nano fab in the area they can contract for coating services? Best Regards Chris Chris Malocsay 510-506-5894 -------------- next part -------------- An HTML attachment was scrubbed... URL: From chris at malocsay.com Fri May 20 11:32:57 2022 From: chris at malocsay.com (Chris Malocsay) Date: Fri, 20 May 2022 08:32:57 -0700 Subject: [labnetwork] Substrate Coating for Cr/Au slides Message-ID: I want to thank the group for all of the great responses. Our client was able to connect with a suitable fab for the project. Thanks again. Chris Chris Malocsay 510-506-5894 -------------- next part -------------- An HTML attachment was scrubbed... URL: From tina at lightandhologram.com Mon May 23 17:53:11 2022 From: tina at lightandhologram.com (Tina Qin) Date: Mon, 23 May 2022 14:53:11 -0700 Subject: [labnetwork] Nanofabrication Engineering Position at Pacific Light & Hologram Message-ID: Hello everyone, We are seeking a self-starting nanofabrication process engineer who can help us imagine, and build, the future of display technology. We are an early to mid-stage startup (about 25 people) in stealth mode located in Pasadena, CA. We have the deep funding and tooling matched to the magnitude of the problems we are tackling. You will find with us an innovative application of semiconductor processing to bring new content and new capabilities to displays. We seek individuals who are comfortable in our very diverse multidisciplinary environment, willing to take technical risks, and experimentally oriented. We practice rapid-prototyping to iterate and push the boundary of technology in our field. In particular, we are looking for people with cleanroom experience to work with patterning and process engineering of nanometer and micrometer features, deposition, and etching of thin films, metrology, and process documentation. Experience with the Raith e-beam systems and resists is a plus. As this is a very hands-on culture, we are currently looking for on-site full-time engineers to work in/with our new cleanroom being built near Pasadena, CA. You are an even better fit if you have experience with python and other programming languages to leave a trail of automation. Successful candidates will work with cross-functional teams of IC designers, PCB, and embedded designers, optical engineers, physicists, and chemists. As we solve each engineering hurdle in front of us, we move on to the next, and with that, there is a lot of room for growth. The evolution of job roles at PL&H is as open as your interests and capabilities. Preferred candidates should have an MS or PhD in electrical engineering, material science, applied physics, or related fields. We offer the usual benefits: health insurance, dental and vision, 401k with matching, relocation expenses, alongside a competitive salary and equity stakes in the company, plus daily lunches with an engaging discussion. Extraneous interests in plants (carnivorous or not), DIY, jet engines, telescopes, and baking bread are a bonus! Please Email your cover letter and CV to iwanttowork at lightandhologram.com ? -------------- next part -------------- An HTML attachment was scrubbed... URL: From terreb at umich.edu Tue May 24 13:56:27 2022 From: terreb at umich.edu (Terre Briggs) Date: Tue, 24 May 2022 13:56:27 -0400 Subject: [labnetwork] High temperature profile thermocouples Message-ID: Hi all, I have found myself in search of a new vendor for our high temperature profile thermocouples, used in our Tempress furnaces systems. I would like to find one that has the capability to reclaim the metals from our non functioning thermocouples to be used in manufacturing new ones. Tempress can supply us with new ones but they can't reclaim the old ones. Any leads would be greatly appreciated. Thanks....Terre -- Terre Briggs Engineer in Research Sr. Lurie Nanofabrication Laboratory University of Michigan 1301 Beal Ave. Ann Arbor, MI 48109 734-320-4470 -------------- next part -------------- An HTML attachment was scrubbed... URL: From kbarnes at semitorrdist.com Tue May 24 16:54:38 2022 From: kbarnes at semitorrdist.com (Keith Barnes) Date: Tue, 24 May 2022 20:54:38 +0000 Subject: [labnetwork] High temperature profile thermocouples In-Reply-To: References: Message-ID: Vulcan Electric! Thank you, Keith F. Barnes Sales Representative SEMITORR GROUP, INC. High Purity, Sanitary, Industrial Divisions 1951 Rutan Drive Livermore, CA 94551 tel: 925.667.3607 cell: 925.337.2221 email: kbarnes at semitorrinc.com web: www.semitorrinc.com ______________________________________________________________________________ ?This email and any files transmitted with it are confidential and intended solely for the use of the individual or entity to whom they are addressed. If you have received this email in error please notify the system manager. This message contains confidential information and is intended only for the individual named. If you are not the named addressee you should not disseminate, distribute or copy this e-mail. Please notify the sender immediately by e-mail if you have received this e-mail by mistake and delete this e-mail from your system. If you are not the intended recipient you are notified that disclosing, copying, distributing or taking any action in reliance on the contents of this information is strictly prohibited.? From: labnetwork On Behalf Of Terre Briggs Sent: Tuesday, May 24, 2022 10:56 AM To: Labnetwork Subject: [labnetwork] High temperature profile thermocouples Hi all, I have found myself in search of a new vendor for our high temperature profile thermocouples, used in our Tempress furnaces systems. I would like to find one that has the capability to reclaim the metals from our non functioning thermocouples to be used in manufacturing new ones. Tempress can supply us with new ones but they can't reclaim the old ones. Any leads would be greatly appreciated. Thanks....Terre -- Terre Briggs Engineer in Research Sr. Lurie Nanofabrication Laboratory University of Michigan 1301 Beal Ave. Ann Arbor, MI 48109 734-320-4470 -------------- next part -------------- An HTML attachment was scrubbed... URL: From terreb at umich.edu Wed May 25 08:02:08 2022 From: terreb at umich.edu (Terre Briggs) Date: Wed, 25 May 2022 08:02:08 -0400 Subject: [labnetwork] High temperature profile thermocouples In-Reply-To: References: Message-ID: Thanks Philip, and Keith for the suggested vendors. I will add them to my contact list and will give them a call when I have a TC to send in. Thanks again......Terre On Wed, May 25, 2022 at 12:31 AM Keith Barnes wrote: > Vulcan Electric! > > > > Thank you, > > > > Keith F. Barnes > > Sales Representative > > > > *SEMITORR GROUP, INC.* > > *High Purity, Sanitary, Industrial Divisions* > > > > 1951 Rutan Drive > > Livermore, CA 94551 > > tel: 925.667.3607 > > cell: 925.337.2221 > > email: kbarnes at semitorrinc.com > > web: www.semitorrinc.com > > > ______________________________________________________________________________ > > ?This email and any files transmitted with it are confidential and > intended solely for the use of the individual or entity to whom they are > addressed. If you have received this email in error please notify the > system manager. This message contains confidential information and is > intended only for the individual named. If you are not the named addressee > you should not disseminate, distribute or copy this e-mail. Please notify > the sender immediately by e-mail if you have received this e-mail by > mistake and delete this e-mail from your system. If you are not the > intended recipient you are notified that disclosing, copying, distributing > or taking any action in reliance on the contents of this information is > strictly prohibited.? > > > > *From:* labnetwork * On Behalf Of *Terre > Briggs > *Sent:* Tuesday, May 24, 2022 10:56 AM > *To:* Labnetwork > *Subject:* [labnetwork] High temperature profile thermocouples > > > > Hi all, > > I have found myself in search of a new vendor for our high temperature > profile thermocouples, used in our Tempress furnaces systems. I would like > to find one that has the capability to reclaim the metals from our non > functioning thermocouples to be used in manufacturing new ones. Tempress > can supply us with new ones but they can't reclaim the old ones. Any leads > would be greatly appreciated. > > > > Thanks....Terre > > > > -- > > Terre Briggs > > Engineer in Research Sr. > > Lurie Nanofabrication Laboratory > > University of Michigan > > 1301 Beal Ave. > > Ann Arbor, MI 48109 > > 734-320-4470 > > > -- Terre Briggs Engineer in Research Sr. Lurie Nanofabrication Laboratory University of Michigan 1301 Beal Ave. Ann Arbor, MI 48109 734-320-4470 -------------- next part -------------- An HTML attachment was scrubbed... URL: From James.Vlahakis at tufts.edu Wed May 25 09:44:47 2022 From: James.Vlahakis at tufts.edu (Vlahakis, James) Date: Wed, 25 May 2022 13:44:47 +0000 Subject: [labnetwork] Startup Job Opportunity Message-ID: Hi all, hope your spring semester has come to a successful close and congrats to any new graduates. Job opportunity at a Boston startup, please see attached for details. They hope to hire someone soon jim -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: 2022_NDT_Microfabrication Engineer.pdf Type: application/pdf Size: 35221 bytes Desc: 2022_NDT_Microfabrication Engineer.pdf URL: From rharlan at purdue.edu Wed May 25 16:49:26 2022 From: rharlan at purdue.edu (Harlan, Richard D) Date: Wed, 25 May 2022 20:49:26 +0000 Subject: [labnetwork] Is Nitrogen Generator a Viable Option Message-ID: Greetings to All, I am writing because we are being challenged, like everyone, to reduce our cost for research. We are considering other options for our Nitrogen gas source including a Nitrogen generator due to the recent cost increase of LN2. We are currently using a 2-tank system which supplies both the liquid and gaseous nitrogen to our facility. We are averaging 80 to 90 SCFM of utility nitrogen and 6 -7 SCFM of High Purity Nitrogen. We would continue to maintain our liquid nitrogen supply. Please provide your positive and negative experiences with Nitrogen generators as well as the manufacturer of the unit. All comments/suggestions are welcomed. Thanks in advance for your input. Best Regards, Rich Harlan Research Engineering Birck Nanotechnology Center | Room 2287C Office: 765.494.6697 | Email: rharlan at purdue.edu [cid:image001.png at 01D87054.ED3CD3E0] -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.png Type: image/png Size: 14241 bytes Desc: image001.png URL: From Justin_Moreau at uml.edu Wed May 25 18:21:05 2022 From: Justin_Moreau at uml.edu (Moreau, Justin) Date: Wed, 25 May 2022 22:21:05 +0000 Subject: [labnetwork] Nanonex NX-2600 NanoImprint tool Message-ID: Hi All I'm looking for a set of electrical schematics for a Nanonex NX-2600 NanoImprint tool. The drawer sensors seem to be working, the chamber moves into the closed position and the lid closed switches are working. The software waits for the lid to close but eventually aborts. I chased the wiring to a control card on the rear panel. Without the schematics I don't know what signals I'm expecting to see through the lid switches. Any help would be appreciated :) [University of Massachusetts Lowell] Justin Moreau Sr. Lab Manager, Nanofabrication Core CORE RESEARCH FACILITIES ETIC E: Justin_Moreau at uml.edu T: 978-934-3615 -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.png Type: image/png Size: 2650 bytes Desc: image001.png URL: From deonc69 at illinois.edu Thu May 26 09:09:10 2022 From: deonc69 at illinois.edu (Collins, Deon) Date: Thu, 26 May 2022 13:09:10 +0000 Subject: [labnetwork] Is Nitrogen Generator a Viable Option In-Reply-To: References: Message-ID: We have N2 supply via a generator. We supply both High purity and Ultra High purity N2 to our labs. The installation of an N2 generator will greatly reduce you LN2 cost but will have a large upfront cost as we have 2 N2 delivery(hard piping) structures in place. I would have to go through some historical data that is decades old to get you numbers but a it all depends on use. Best guess is a couple years out on ROI but it is a no brainer if you are a heavy N2 user. Deon D. Collins FACILITY MANAGER Holonyak Lab University of Illinois at Urbana-Champaign 208 N Wright St Rm. 248 | 2250 Urbana, IL 61801 217-300-7531 | deonc69 at illinois.edu [https://webtools.illinois.edu/webservices/js/ds/signature_logo.png] Under the Illinois Freedom of Information Act any written communication to or from university employees regarding university business is a public record and may be subject to public disclosure. Life is not about watching other people live it. It's about you living your own! From: labnetwork On Behalf Of Harlan, Richard D Sent: Wednesday, May 25, 2022 3:49 PM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Is Nitrogen Generator a Viable Option Greetings to All, I am writing because we are being challenged, like everyone, to reduce our cost for research. We are considering other options for our Nitrogen gas source including a Nitrogen generator due to the recent cost increase of LN2. We are currently using a 2-tank system which supplies both the liquid and gaseous nitrogen to our facility. We are averaging 80 to 90 SCFM of utility nitrogen and 6 -7 SCFM of High Purity Nitrogen. We would continue to maintain our liquid nitrogen supply. Please provide your positive and negative experiences with Nitrogen generators as well as the manufacturer of the unit. All comments/suggestions are welcomed. Thanks in advance for your input. Best Regards, Rich Harlan Research Engineering Birck Nanotechnology Center | Room 2287C Office: 765.494.6697 | Email: rharlan at purdue.edu [cid:image003.png at 01D870D6.E1832450] -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image002.png Type: image/png Size: 2602 bytes Desc: image002.png URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image003.png Type: image/png Size: 14241 bytes Desc: image003.png URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: Mr_ Deon D_ Collins.vcf Type: text/x-vcard Size: 2025 bytes Desc: Mr_ Deon D_ Collins.vcf URL: From Howard.Northfield at uottawa.ca Thu May 26 10:38:13 2022 From: Howard.Northfield at uottawa.ca (Howard Northfield) Date: Thu, 26 May 2022 14:38:13 +0000 Subject: [labnetwork] SAMCO 10NR Message-ID: Our SAMCO 10NR RIE has a vacuum leak. Changed all the easy O-rings. Anyone have an exploded diagram of the Turbo Gate Valve (TGV) or instructions for the replacement of the TGV O-rings. Thanks Howard Northfield Research Associate Center for Research in Photonics University of Ottawa 25 Templeton Street, Ottawa, Ontario Canada, K1N 6N5 -------------- next part -------------- An HTML attachment was scrubbed... URL: From johnboyle at fast.net Thu May 26 13:04:34 2022 From: johnboyle at fast.net (John Boyle) Date: Thu, 26 May 2022 13:04:34 -0400 Subject: [labnetwork] Is Nitrogen Generator a Viable Option In-Reply-To: References: Message-ID: <02df01d87122$ad093c40$071bb4c0$@net> I worked in the industrial gas industry for 25 years and Nitrogen generators are a reliable source of nitrogen for many customers. Usually Nitrogen generators can provide a range of purity suitable for most industrial applications..the purity usually is higher when the unit is operated at lower flow and purity is lower at higher flow rates. For UHP applications you would likely need an add on purifier to remove O2, H20. If your UHP process is sensitive to other contaminants (CO,CO2, THC) there are some purifiers to remove these as well. There are two main types to consider----membrane based units and adsorbent bed units. A Nitrogen generator will not provide Liquid N2 and so any operations that require cryogenic temperatures that LIN provides would not be amenable to generated N2. John Boyle John Boyle Consulting, LLC 4848 Canterbury Drive Emmaus, PA 18049 484-432-7596 Cell 610-965-3208 Office From: labnetwork [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Collins, Deon Sent: Thursday, May 26, 2022 9:09 AM To: Harlan, Richard D; labnetwork at mtl.mit.edu Subject: Re: [labnetwork] Is Nitrogen Generator a Viable Option We have N2 supply via a generator. We supply both High purity and Ultra High purity N2 to our labs. The installation of an N2 generator will greatly reduce you LN2 cost but will have a large upfront cost as we have 2 N2 delivery(hard piping) structures in place. I would have to go through some historical data that is decades old to get you numbers but a it all depends on use. Best guess is a couple years out on ROI but it is a no brainer if you are a heavy N2 user. Deon D. Collins FACILITY MANAGER Holonyak Lab University of Illinois at Urbana-Champaign 208 N Wright St Rm. 248 | 2250 Urbana, IL 61801 217-300-7531 | deonc69 at illinois.edu https://webtools.illinois.edu/webservices/js/ds/signature_logo.png Under the Illinois Freedom of Information Act any written communication to or from university employees regarding university business is a public record and may be subject to public disclosure. Life is not about watching other people live it. It's about you living your own! From: labnetwork On Behalf Of Harlan, Richard D Sent: Wednesday, May 25, 2022 3:49 PM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Is Nitrogen Generator a Viable Option Greetings to All, I am writing because we are being challenged, like everyone, to reduce our cost for research. We are considering other options for our Nitrogen gas source including a Nitrogen generator due to the recent cost increase of LN2. We are currently using a 2-tank system which supplies both the liquid and gaseous nitrogen to our facility. We are averaging 80 to 90 SCFM of utility nitrogen and 6 -7 SCFM of High Purity Nitrogen. We would continue to maintain our liquid nitrogen supply. Please provide your positive and negative experiences with Nitrogen generators as well as the manufacturer of the unit. All comments/suggestions are welcomed. Thanks in advance for your input. Best Regards, Rich Harlan Research Engineering Birck Nanotechnology Center | Room 2287C Office: 765.494.6697 | Email: rharlan at purdue.edu -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.png Type: image/png Size: 2602 bytes Desc: not available URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image002.png Type: image/png Size: 14241 bytes Desc: not available URL: From jmaduzi2 at illinois.edu Thu May 26 13:55:44 2022 From: jmaduzi2 at illinois.edu (Maduzia, Joseph Walter) Date: Thu, 26 May 2022 17:55:44 +0000 Subject: [labnetwork] Is Nitrogen Generator a Viable Option In-Reply-To: References: Message-ID: Hello Richard, At MNMS Cleanroom in MechSE at UIUC, we used www.onsitegas.com to purchase a 99.9995% n2 Gen. They were cheaper for a better purity and flow rate than Parker. It's overkill for most of our N2 needs, but it's cheaper to supply 99.9995% to everything than to have both low and high purity N2 lines running around the facility. OnSiteGas sales team was great. We did have one issue during install. We used a SS flex hose to supply compressed air to the N2 Gen. This caused the system to "ring" every time it generated N2. I recommend using the really simple rubber hose onsitegas provides to connect from CDA to N2Gen to minimize vibration and resulting noise. The maintenance is super simple. Just a few compressed air filters on a regular schedule. Approx $500 a year. As long as the N2Gen has compressed air, it pumps out nitrogen. This was our sales rep. I'm not sure if he's still there. Michael Montesi Sales Manager-Commercial Products On Site Gas Systems, Inc. 35 Budney Rd. Newington, CT 06111 Phone: 860-667-8888 Ext: 230 http://www.onsitegas.com/ Thank You, JOE MADUZIA Senior Research Engineer, Grainger College of Engineering, MechSE, Univ of IL Lab Manager, ResearchPark & EnterpriseWorks, Univ of IL 2239 LuMEB | (P) 217.244.6302 https://cleanroom.mechse.illinois.edu/ | researchpark.illinois.edu From: labnetwork On Behalf Of Harlan, Richard D Sent: Wednesday, May 25, 2022 3:49 PM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Is Nitrogen Generator a Viable Option Greetings to All, I am writing because we are being challenged, like everyone, to reduce our cost for research. We are considering other options for our Nitrogen gas source including a Nitrogen generator due to the recent cost increase of LN2. We are currently using a 2-tank system which supplies both the liquid and gaseous nitrogen to our facility. We are averaging 80 to 90 SCFM of utility nitrogen and 6 -7 SCFM of High Purity Nitrogen. We would continue to maintain our liquid nitrogen supply. Please provide your positive and negative experiences with Nitrogen generators as well as the manufacturer of the unit. All comments/suggestions are welcomed. Thanks in advance for your input. Best Regards, Rich Harlan Research Engineering Birck Nanotechnology Center | Room 2287C Office: 765.494.6697 | Email: rharlan at purdue.edu [cid:image001.png at 01D870FF.E8592F80] -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.png Type: image/png Size: 14241 bytes Desc: image001.png URL: From rlc12 at cornell.edu Thu May 26 12:45:10 2022 From: rlc12 at cornell.edu (Rebecca Lee Vliet) Date: Thu, 26 May 2022 16:45:10 +0000 Subject: [labnetwork] REGISTRATION EXTENDED UNTIL 6/1 - CNF TCN Short Course Message-ID: REGISTRATION EXTENDED UNTIL 6/1 - SEATS STILL AVAILABLE! Please post & share the announcement below. Thank you! CNF Technology & Characterization at the Nanoscale (CNF TCN) IN-PERSON Short Course June 7, 2022 - June 10, 2022 The CNF TCN will be in-person for June 2022! The course will be held Tuesday, June 7, 2022 - Friday, June 10, 2022. Each day offers lectures and laboratory demonstrations designed to impart a broad understanding of the science and technology required to undertake research in nanoscience. TCN is an ideal way for faculty, students, post docs and staff members to rapidly come up to speed in many of the technologies that users of the CNF need to employ. Members of the high tech business community will also find it an effective way to learn best practices for success in a nanofab environment. Attendance is open to the general scientific community. Seating is limited. Note: The short course does not replace the three part training required to become a user of our facility. To become a CNF user, please visit the "Getting Started" link (https:cnf.cornell.edu/howto) on the CNF website. Cornell COVID-19 Response guidelines can be found here: https://covid.cornell.edu/visitors/. Masks are required in all CNF lab spaces. For more information and to register visit: https://cnf.cornell.edu/education/tcn [Text Description automatically generated] -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image002.jpg Type: image/jpeg Size: 488127 bytes Desc: image002.jpg URL: From fm2469 at nyu.edu Thu May 26 17:38:04 2022 From: fm2469 at nyu.edu (Felicia Mcguire) Date: Thu, 26 May 2022 17:38:04 -0400 Subject: [labnetwork] Cleanroom HEPA Filter Replacement Message-ID: Hi All! We (NYU) are working on renovating our cleanroom and had a question: How much does it cost to replace the HEPA filters in the lab (and how big is the cleanroom you are replacing them in)? Thank you! Felicia -------------- next part -------------- An HTML attachment was scrubbed... URL: From johnboyle at fast.net Thu May 26 19:50:04 2022 From: johnboyle at fast.net (John Boyle) Date: Thu, 26 May 2022 19:50:04 -0400 Subject: [labnetwork] Is Nitrogen Generator a Viable Option In-Reply-To: References: Message-ID: <035501d8715b$52b02db0$f8108910$@net> I have used OnSite N2 generators and they have performed well and reliably. I agree with Joe's comments below. If you need >99.9995% you can add a purifier. John Boyle John Boyle Consulting, LLC 4848 Canterbury Drive Emmaus, PA 18049 484-432-7596 Cell 610-965-3208 Office From: labnetwork [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Maduzia, Joseph Walter Sent: Thursday, May 26, 2022 1:56 PM To: Harlan, Richard D; labnetwork at mtl.mit.edu Subject: Re: [labnetwork] Is Nitrogen Generator a Viable Option Hello Richard, At MNMS Cleanroom in MechSE at UIUC, we used www.onsitegas.com to purchase a 99.9995% n2 Gen. They were cheaper for a better purity and flow rate than Parker. It's overkill for most of our N2 needs, but it's cheaper to supply 99.9995% to everything than to have both low and high purity N2 lines running around the facility. OnSiteGas sales team was great. We did have one issue during install. We used a SS flex hose to supply compressed air to the N2 Gen. This caused the system to "ring" every time it generated N2. I recommend using the really simple rubber hose onsitegas provides to connect from CDA to N2Gen to minimize vibration and resulting noise. The maintenance is super simple. Just a few compressed air filters on a regular schedule. Approx $500 a year. As long as the N2Gen has compressed air, it pumps out nitrogen. This was our sales rep. I'm not sure if he's still there. Michael Montesi Sales Manager-Commercial Products On Site Gas Systems, Inc. 35 Budney Rd. Newington, CT 06111 Phone: 860-667-8888 Ext: 230 http://www.onsitegas.com/ Thank You, JOE MADUZIA Senior Research Engineer, Grainger College of Engineering, MechSE, Univ of IL Lab Manager, ResearchPark & EnterpriseWorks, Univ of IL 2239 LuMEB | (P) 217.244.6302 https://cleanroom.mechse.illinois.edu/ | researchpark.illinois.edu From: labnetwork On Behalf Of Harlan, Richard D Sent: Wednesday, May 25, 2022 3:49 PM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Is Nitrogen Generator a Viable Option Greetings to All, I am writing because we are being challenged, like everyone, to reduce our cost for research. We are considering other options for our Nitrogen gas source including a Nitrogen generator due to the recent cost increase of LN2. We are currently using a 2-tank system which supplies both the liquid and gaseous nitrogen to our facility. We are averaging 80 to 90 SCFM of utility nitrogen and 6 -7 SCFM of High Purity Nitrogen. We would continue to maintain our liquid nitrogen supply. Please provide your positive and negative experiences with Nitrogen generators as well as the manufacturer of the unit. All comments/suggestions are welcomed. Thanks in advance for your input. Best Regards, Rich Harlan Research Engineering Birck Nanotechnology Center | Room 2287C Office: 765.494.6697 | Email: rharlan at purdue.edu -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.png Type: image/png Size: 14241 bytes Desc: not available URL: From Kalvin.Huoth at ll.mit.edu Fri May 27 11:10:03 2022 From: Kalvin.Huoth at ll.mit.edu (Huoth, Kalvin - 0835 - MITLL) Date: Fri, 27 May 2022 15:10:03 +0000 Subject: [labnetwork] SAMCO 10NR In-Reply-To: References: Message-ID: Hi Howard, I ran into a similar problem with our Samco 10NR a few years back, the leak came from large flakes on the TGV O-ring seal. Your Kalrez-ring maybe fine to reuse. See attached Vatvalve Schematic. Give me a call at Cell 978-930-3264, I can walk you through it over the phone. Thanks, Kalvin Huoth From: labnetwork > On Behalf Of Howard Northfield Sent: Thursday, May 26, 2022 10:38 AM To: Subject: [labnetwork] SAMCO 10NR Our SAMCO 10NR RIE has a vacuum leak. Changed all the easy O-rings. Anyone have an exploded diagram of the Turbo Gate Valve (TGV) or instructions for the replacement of the TGV O-rings. Thanks Howard Northfield Research Associate Center for Research in Photonics University of Ottawa 25 Templeton Street, Ottawa, Ontario Canada, K1N 6N5 -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: Discontinuance S642 with PM-4_PM-5.pdf Type: application/pdf Size: 144964 bytes Desc: not available URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: 64040-PE52-0008 IOMI 219035EE.pdf Type: application/pdf Size: 525393 bytes Desc: not available URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: 64040-PE52-0008 Drawing 85854EA.pdf Type: application/pdf Size: 104551 bytes Desc: not available URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: smime.p7s Type: application/pkcs7-signature Size: 5645 bytes Desc: not available URL: From beaudoin at physics.ubc.ca Fri May 27 18:43:13 2022 From: beaudoin at physics.ubc.ca (Beaudoin, Mario) Date: Fri, 27 May 2022 15:43:13 -0700 Subject: [labnetwork] Looking for small footprint LPCVD system Message-ID: I'm looking for a recommended small footprint LPCVD system.? My cleanroom "chase" is above the room so the entire furnace system needs to be inside the clean area.? I'm looking for maximum 4" size wafers. -- -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: Mario%20Beaudoin%20SBQMI%20sig%202.jpg Type: image/jpeg Size: 21446 bytes Desc: not available URL: From erica.alvarezconde at kaust.edu.sa Sun May 29 09:54:00 2022 From: erica.alvarezconde at kaust.edu.sa (Erica Alvarez Conde) Date: Sun, 29 May 2022 13:54:00 +0000 Subject: [labnetwork] Help with SSEC system In-Reply-To: References: Message-ID: Hello everyone; In our cleanroom we have an old SSEC mask processor, model M3300ML. The tool has been off for a long time and now we are trying to bring the tool to life again, but we are facing some problems and we are stuck. If someone knows any engineer that can provide help or maintenance service for this equipment, please contact me. I couldn't find anyone and the company doesn't exist anymore. erica.alvarezconde at kaust.edu.sa Thank you. Regards; Erica Alvarez Conde Senior Technical Specialist, Nanofabrication Core Lab KAUST King Abdullah University of Science and Technology Building 3, Level 0, Office 0250-WS05 Thuwal 23955-6900 Kingdom of Saudi Arabia Office: +966 12 808 0752 Mobile: +966 (0) 544700505 Email: erica.alvarezconde at kaust.edu.sa Website: www.kaust.edu.sa , https://corelabs.kaust.edu.sa/ NOTE: Remember that the weekends in Saudi Arabia are Fridays and Saturdays -------------- next part -------------- An HTML attachment was scrubbed... URL: