From swagone at gmu.edu Wed Feb 1 07:22:52 2023 From: swagone at gmu.edu (Shawn Wagoner) Date: Wed, 1 Feb 2023 12:22:52 +0000 Subject: [labnetwork] Question about DC sputtering of zinc In-Reply-To: References: Message-ID: I would not allow Zinc in any system I ran at Binghamton University. I would not even allow it in the cleanroom. It represents a significant contamination risk for shared use facilities. If I were fortunate enough to have two systems, one clean and one "dirty", I might consider it in the dirty system. Its vapor pressure is so high I am not sure there is much you can do to cover it up with Ti. Shawn From: labnetwork On Behalf Of Malhotra, Sandra Guy Sent: Monday, January 30, 2023 4:40 PM To: Fab Network Subject: [labnetwork] Question about DC sputtering of zinc Howdy All from AggieFab, We have a user who requested permission to sputter zinc in our shared Lesker PVD75 DC sputtering tool. On the Lesker website, there is the following note: "This material has a relatively high vapor pressure at low temperatures which can cause ppm level contamination in the chamber and subsequent films. Careful consideration should be taken before this material is used in any deposition system. For this reason, some users prefer to use only dedicated vacuum chambers for deposition." I'd like to ask the group if they allow zinc depositions in their shared tools if they follow it by a thorough paste with another material, like Ti? Or if this material is prohibited because it is problematic. Thanks in advance for any inputs you may have. Best, Sandra G. Malhotra, Ph.D. | Technical Lab Manager AggieFab Nanofabrication Facility https://aggiefab.tamu.edu/ Department of Electrical & Computer Engineering, College of Engineering | Texas A&M University 3253 TAMU | College Station, TX 77843 ph: 979.845.3199 | sandra.malhotra at tamu.edu - - - - - - - - - - - - - - - - - - - - - - - - TEXAS A&M UNIVERSITY | FEARLESS on Every Front -------------- next part -------------- An HTML attachment was scrubbed... URL: From yakimom at sunypoly.edu Wed Feb 1 10:20:51 2023 From: yakimom at sunypoly.edu (Michael Yakimov) Date: Wed, 1 Feb 2023 15:20:51 +0000 Subject: [labnetwork] Question about DC sputtering of zinc In-Reply-To: References: Message-ID: I would also ask if process goal can be achieved with zinc compound, like ZnO. Thanks Mike ________________________________ From: labnetwork on behalf of Shawn Wagoner Sent: Wednesday, February 1, 2023 7:22 AM To: Malhotra, Sandra Guy ; Fab Network Subject: Re: [labnetwork] Question about DC sputtering of zinc I would not allow Zinc in any system I ran at Binghamton University. I would not even allow it in the cleanroom. It represents a significant contamination risk for shared use facilities. If I were fortunate enough to have two systems, one clean and one ?dirty?, I might consider it in the dirty system. Its vapor pressure is so high I am not sure there is much you can do to cover it up with Ti. Shawn From: labnetwork On Behalf Of Malhotra, Sandra Guy Sent: Monday, January 30, 2023 4:40 PM To: Fab Network Subject: [labnetwork] Question about DC sputtering of zinc Howdy All from AggieFab, We have a user who requested permission to sputter zinc in our shared Lesker PVD75 DC sputtering tool. On the Lesker website, there is the following note: "This material has a relatively high vapor pressure at low temperatures which can cause ppm level contamination in the chamber and subsequent films. Careful consideration should be taken before this material is used in any deposition system. For this reason, some users prefer to use only dedicated vacuum chambers for deposition." I'd like to ask the group if they allow zinc depositions in their shared tools if they follow it by a thorough paste with another material, like Ti? Or if this material is prohibited because it is problematic. Thanks in advance for any inputs you may have. Best, Sandra G. Malhotra, Ph.D. | Technical Lab Manager AggieFab Nanofabrication Facility https://aggiefab.tamu.edu/ Department of Electrical & Computer Engineering, College of Engineering | Texas A&M University 3253 TAMU | College Station, TX 77843 ph: 979.845.3199 | sandra.malhotra at tamu.edu - - - - - - - - - - - - - - - - - - - - - - - - TEXAS A&M UNIVERSITY | FEARLESS on Every Front -------------- next part -------------- An HTML attachment was scrubbed... URL: From James.Vlahakis at tufts.edu Wed Feb 1 10:43:02 2023 From: James.Vlahakis at tufts.edu (Vlahakis, James) Date: Wed, 1 Feb 2023 15:43:02 +0000 Subject: [labnetwork] [External] Question about DC sputtering of zinc In-Reply-To: References: Message-ID: Hi Sandra, a few years ago we allowed Zn deposition in our "dirty" sputter tool without a heated platen. Consider the attached vapor pressure v Temp chart - at the temps seen in our tool the vapor pressure is below the system base pressure. Since that time we've noted no significant changes in base pressure and other users haven't complained about their films being lower quality. Please take this with a grain of salt - Zn dep is not very common here, just a few times per year. In addition, our users are not fabricating CMOS devices, almost all are biomed focused these days jim From: labnetwork On Behalf Of Malhotra, Sandra Guy Sent: Monday, January 30, 2023 4:40 PM To: Fab Network Subject: [External] [labnetwork] Question about DC sputtering of zinc Howdy All from AggieFab, We have a user who requested permission to sputter zinc in our shared Lesker PVD75 DC sputtering tool. On the Lesker website, there is the following note: "This material has a relatively high vapor pressure at low temperatures which can cause ppm level contamination in the chamber and subsequent films. Careful consideration should be taken before this material is used in any deposition system. For this reason, some users prefer to use only dedicated vacuum chambers for deposition." I'd like to ask the group if they allow zinc depositions in their shared tools if they follow it by a thorough paste with another material, like Ti? Or if this material is prohibited because it is problematic. Thanks in advance for any inputs you may have. Best, Sandra G. Malhotra, Ph.D. | Technical Lab Manager AggieFab Nanofabrication Facility https://aggiefab.tamu.edu/ Department of Electrical & Computer Engineering, College of Engineering | Texas A&M University 3253 TAMU | College Station, TX 77843 ph: 979.845.3199 | sandra.malhotra at tamu.edu - - - - - - - - - - - - - - - - - - - - - - - - TEXAS A&M UNIVERSITY | FEARLESS on Every Front Caution: This message originated from outside of the Tufts University organization. Please exercise caution when clicking links or opening attachments. When in doubt, email the TTS Service Desk at it at tufts.edu or call them directly at 617-627-3376. -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: vapor pressure 2.pdf Type: application/pdf Size: 1038255 bytes Desc: vapor pressure 2.pdf URL: From odc1n08 at soton.ac.uk Wed Feb 1 11:49:01 2023 From: odc1n08 at soton.ac.uk (Owain Clark) Date: Wed, 1 Feb 2023 16:49:01 +0000 Subject: [labnetwork] Question about DC sputtering of zinc In-Reply-To: References: Message-ID: Agreed, tried it once in a dirty evaporator and it makes a difficult to remove low density coating. Did evaporate ok with e-beam though. Another cleanroom here less concerned about such things sputters it regularly for Zn doped Li Niobate applications. All depends what you want out of your shared tools. From: labnetwork On Behalf Of Shawn Wagoner Sent: 01 February 2023 12:23 To: Malhotra, Sandra Guy ; Fab Network Subject: Re: [labnetwork] Question about DC sputtering of zinc CAUTION: This e-mail originated outside the University of Southampton. I would not allow Zinc in any system I ran at Binghamton University. I would not even allow it in the cleanroom. It represents a significant contamination risk for shared use facilities. If I were fortunate enough to have two systems, one clean and one "dirty", I might consider it in the dirty system. Its vapor pressure is so high I am not sure there is much you can do to cover it up with Ti. Shawn From: labnetwork > On Behalf Of Malhotra, Sandra Guy Sent: Monday, January 30, 2023 4:40 PM To: Fab Network > Subject: [labnetwork] Question about DC sputtering of zinc Howdy All from AggieFab, We have a user who requested permission to sputter zinc in our shared Lesker PVD75 DC sputtering tool. On the Lesker website, there is the following note: "This material has a relatively high vapor pressure at low temperatures which can cause ppm level contamination in the chamber and subsequent films. Careful consideration should be taken before this material is used in any deposition system. For this reason, some users prefer to use only dedicated vacuum chambers for deposition." I'd like to ask the group if they allow zinc depositions in their shared tools if they follow it by a thorough paste with another material, like Ti? Or if this material is prohibited because it is problematic. Thanks in advance for any inputs you may have. Best, Sandra G. Malhotra, Ph.D. | Technical Lab Manager AggieFab Nanofabrication Facility https://aggiefab.tamu.edu/ Department of Electrical & Computer Engineering, College of Engineering | Texas A&M University 3253 TAMU | College Station, TX 77843 ph: 979.845.3199 | sandra.malhotra at tamu.edu - - - - - - - - - - - - - - - - - - - - - - - - TEXAS A&M UNIVERSITY | FEARLESS on Every Front -------------- next part -------------- An HTML attachment was scrubbed... URL: From price.798 at osu.edu Wed Feb 1 12:36:57 2023 From: price.798 at osu.edu (Price, Aimee) Date: Wed, 1 Feb 2023 17:36:57 +0000 Subject: [labnetwork] Signatures with affiliation Message-ID: Hi Labnetwork administrators, Would it be possible to kindly ask those who post and who respond to include their affiliation in a signature block? It would be very helpful and instructive to know where and what type of institution was responding as that often matters, I.e. industry and academia are different, different regions have different needs, etc. You can't always tell the from tne email address or name alone. Thank you, Aimee Aimee Bross Price Manager, Nanofabrication Nanotech West Lab The Ohio State University -------------- next part -------------- An HTML attachment was scrubbed... URL: From C.Heidelberger at ll.mit.edu Wed Feb 1 12:59:57 2023 From: C.Heidelberger at ll.mit.edu (Heidelberger, Christopher - 0889 - MITLL) Date: Wed, 1 Feb 2023 17:59:57 +0000 Subject: [labnetwork] Contract SiO2 Deposition on 300 mm Wafers Message-ID: Hello, Does anyone know of a business or other entity that does contract deposition of SiO2 films on 300 mm wafers? I would like to have 50 nm SiO2 deposited by PECVD on just two 300 mm Si wafers. It would need to be in a Au-free system. Best, Chris - Christopher Heidelberger, PhD Technical Staff, Quantum Information and Integrated Nanosystems Group MIT Lincoln Laboratory, 244 Wood Street, Lexington, MA 02421-6246 Phone: 781-981-9052 | Email: c.heidelberger at ll.mit.edu | he/him -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: smime.p7s Type: application/pkcs7-signature Size: 5669 bytes Desc: not available URL: From lvchang at Central.UH.EDU Wed Feb 1 14:02:00 2023 From: lvchang at Central.UH.EDU (Chang, Long) Date: Wed, 1 Feb 2023 19:02:00 +0000 Subject: [labnetwork] Question about DC sputtering of zinc In-Reply-To: References: Message-ID: <8F0F4660-12C7-497E-AF80-8853B535CB87@central.uh.edu> Hi All, Does anyone have a list of banned materials for sputtering they can share? I made a list over the years through discussions on labnetwork whenever a new material gets requested, see attached photo. It?ll be nice to have a publicly searchable database, similar to this google sheet (https://docs.google.com/spreadsheets/d/1z0iyCsDUTpi6nkhtIxSp_rSMKGF4jeiqGbVRfnF5CRk). Thanks, Long [cid:DDECC2FB-B7D7-4A97-BBFD-B712E4A2427F] On Feb 1, 2023, at 9:20 AM, Michael Yakimov > wrote: I would also ask if process goal can be achieved with zinc compound, like ZnO. Thanks Mike ________________________________ From: labnetwork > on behalf of Shawn Wagoner > Sent: Wednesday, February 1, 2023 7:22 AM To: Malhotra, Sandra Guy >; Fab Network > Subject: Re: [labnetwork] Question about DC sputtering of zinc I would not allow Zinc in any system I ran at Binghamton University. I would not even allow it in the cleanroom. It represents a significant contamination risk for shared use facilities. If I were fortunate enough to have two systems, one clean and one ?dirty?, I might consider it in the dirty system. Its vapor pressure is so high I am not sure there is much you can do to cover it up with Ti. Shawn From: labnetwork > On Behalf Of Malhotra, Sandra Guy Sent: Monday, January 30, 2023 4:40 PM To: Fab Network > Subject: [labnetwork] Question about DC sputtering of zinc Howdy All from AggieFab, We have a user who requested permission to sputter zinc in our shared Lesker PVD75 DC sputtering tool. On the Lesker website, there is the following note: "This material has a relatively high vapor pressure at low temperatures which can cause ppm level contamination in the chamber and subsequent films. Careful consideration should be taken before this material is used in any deposition system. For this reason, some users prefer to use only dedicated vacuum chambers for deposition." I'd like to ask the group if they allow zinc depositions in their shared tools if they follow it by a thorough paste with another material, like Ti? Or if this material is prohibited because it is problematic. Thanks in advance for any inputs you may have. Best, Sandra G. Malhotra, Ph.D. | Technical Lab Manager AggieFab Nanofabrication Facility https://aggiefab.tamu.edu/ Department of Electrical & Computer Engineering, College of Engineering | Texas A&M University 3253 TAMU | College Station, TX 77843 ph: 979.845.3199 | sandra.malhotra at tamu.edu - - - - - - - - - - - - - - - - - - - - - - - - TEXAS A&M UNIVERSITY | FEARLESS on Every Front _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://urldefense.com/v3/__https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork__;!!LkSTlj0I!Aesqckjzk4BJyH4BTgx7gpd2zSbcvoD0aNb_4EiUCkoLdUYfIvVH6gEOcN2-6xdSIojkXl2Q3CtgiyRTh1O4DIsD2Q$ -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: Screen Shot 2023-02-01 at 12.58.13 PM.png Type: image/png Size: 306491 bytes Desc: Screen Shot 2023-02-01 at 12.58.13 PM.png URL: From beaudoin at physics.ubc.ca Wed Feb 1 14:07:43 2023 From: beaudoin at physics.ubc.ca (Beaudoin, Mario) Date: Wed, 1 Feb 2023 11:07:43 -0800 Subject: [labnetwork] Question: adhesion of metal to BPDA/PPD polyimide Message-ID: <3d0e9163-6746-d694-7209-b84abc4598be@physics.ubc.ca> Dear Networkers, I'm new to polyimide processing.? We have a student project that requires deposition of metal on polyimide (BPDA/PPD PI-2600 series from HD microsystems).? We tried liftoff with depositon of Ti/Pt by Ebeam.? Despite a 1min O2 plasma, albeit at low power, the metal was peeling off.? Moreover, the photoresist was also peeling off from the polyimide.? Does anyone have a good working recipe for metalization of polyimide? Thanks, Mario -- -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: Mario Beaudoin SBQMI sig 2.jpg Type: image/jpeg Size: 21446 bytes Desc: not available URL: From skooi at mit.edu Wed Feb 1 15:30:48 2023 From: skooi at mit.edu (Steven Earl Kooi) Date: Wed, 1 Feb 2023 20:30:48 +0000 Subject: [labnetwork] Job posting - Instrumentation Specialist Message-ID: Hello, The Institute for Soldier Nanotechnologies (ISN) at MIT has an open position for an "Instrumentation Specialist?. We are looking for someone interested in hands-on daily responsibility for user training, operation, and maintenance for a range of instruments (e.g., microscopy; spectroscopy; optical, thermal and mechanical characterization) in our research facility. Anyone interested can find out more about the position and apply at: [reach-5c9f84a028.jpeg] Instrumentation Specialist careers.peopleclick.com Best, Steve. Steven Kooi, Ph.D. Principal Research Scientist Institute for Soldier Nanotechnologies Massachusetts Institute of Technology Phone: 617-324-6416 Email: skooi at mit.edu -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: reach-5c9f84a028.jpeg Type: image/jpeg Size: 20930 bytes Desc: reach-5c9f84a028.jpeg URL: From bill_flounders at berkeley.edu Wed Feb 1 16:51:11 2023 From: bill_flounders at berkeley.edu (Albert William (Bill) Flounders) Date: Wed, 1 Feb 2023 13:51:11 -0800 Subject: [labnetwork] Signatures with affiliation In-Reply-To: References: Message-ID: Aimee, Thank you for this excellent etiquette recommendation. I have also requested that the exceptional moderator of LabNetwork consider adding this guideline to the Lab Network sign up page. Sincerely, Bill Bill Flounders UC Berkeley NanoLab https://nanolab.berkeley.edu/ On Wed, Feb 1, 2023 at 12:48 PM Price, Aimee wrote: > Hi Labnetwork administrators, > > Would it be possible to kindly ask those who post and who respond to > include their affiliation in a signature block? It would be very helpful > and instructive to know where and what type of institution was responding > as that often matters, I.e. industry and academia are different, different > regions have different needs, etc. You can't always tell the from tne > email address or name alone. > > Thank you, > > Aimee > > Aimee Bross Price > Manager, Nanofabrication > Nanotech West Lab > The Ohio State University > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork > -------------- next part -------------- An HTML attachment was scrubbed... URL: From Khaled.Mnaymneh at nrc-cnrc.gc.ca Wed Feb 1 20:59:15 2023 From: Khaled.Mnaymneh at nrc-cnrc.gc.ca (Mnaymneh, Khaled) Date: Thu, 2 Feb 2023 01:59:15 +0000 Subject: [labnetwork] Postdoctoral Fellowship, Silicon Nitride Platform Message-ID: Hi All, Please distribute to interested parties and networks. Cheers, Khaled. Postdoctoral Fellowship, A Canadian-Based Silicon Nitride Platform (nrc-cnrc.gc.ca) This project will establish a silicon-nitride platform for optical devices targeting classical and quantum applications. Silicon nitride is set to be the new "silicon" for the world's next-generation photonic technologies. Because we are in early days of adaption, there is an excellent opportunity to shape Canada's nitride platform in a unique way and to be an international trendsetter. Continuing to build upon current material and device research, the successful applicant will be expected to lead design, fabrication and characterization efforts in silicon nitride photonics and phononics. Specifically, this will entail a Quantum-Optics-Lab-On-A-Chip approach; design, fabricate and characterize chip components that perform and process optical information. Working with our university partner, you will perform a careful study of film-material parameters; more specifically, the refractive index of highly stoichiometric Si3N4, and how it governs the dispersion engineering and nonlinear optical qualities needed to achieve on-chip all-optical performance. Education PhD in Physics, Electrical Engineering, Engineering Physics or a relevant Engineering field. For information on certificates and diplomas issued abroad, please see Degree equivalency Experience * Hands-on nanofabrication and thin-film experience * Knowledge of integrated optics and chip design, designing and constructing experimental apparatus * Experience with lasers, non-linear optics, and detectors * Experience with fiber optics * Experience with cryostats, vacuum cryogenics would be considered as an asset Language Requirements English, Information on language requirements and self-assessment tests Condition of Employment Reliability Status Who is eligible? * Fellowships will be for two years contingent on satisfactory progress achieved during the first year. * Candidates should have obtained a PhD (or equivalent) within the past three years (PhD received on or after July 1, 2020) or expect to complete their PhD within 6 months of appointment. * Fellows will work in a lab under the direct supervision of an NRC researcher. Application Requirements In order to be considered for the program please include the following in your application, please note that you will need to attach the required documents as per the list below when submitting your application. Failure to do so will result in your application being excluded from searches. * Resume * Statement of Interest in the project (maximum one page in length) * PhD Transcript - an electronic copy is sufficient, it does not have to be an official version. * List of Publications When submitting your application you can include the required documents in any attachment field such as ?Second language evaluation results? or ?Other attachments? In addition, applicants who best meet the requirements of the position will be asked to provide three letters of recommendation at a later stage of the competition process. Relocation Relocation assistance will be determined in accordance with the NRC's directives. Compensation The intent of this hiring action is to staff through the Postdoctoral Fellowships Program at the RCO-2/AsRO level, which is an early-career level position with a salary range of $74,230 to $103,093. NOTE: Salary determination will be based on a review of the candidate's expertise, outcomes and impacts of their previous work experience relative to the requirements of the level. As a guide, the current annual PhD recruiting rate is $74,230. NRC employees enjoy a wide-range of competitive benefits including comprehensive health and dental plans, pension and insurance plans, vacation and other leave entitlements. Notes * Fellowships are open to nationals of all countries, although preference will be given to Canadian Citizens and Permanent Residents of Canada. Please include citizenship information in your application. * The incumbent must adhere to safe workplace practices at all times. * We thank all those who apply, however only those selected for further consideration will be contacted. Please direct your questions, with the requisition number (18898) to: E-mail: NRC.NRCHiring-EmbaucheCNRC.CNRC at nrc-cnrc.gc.ca Telephone: 343-548-5849 Closing Date: 27 March 2023 - 23:59 Eastern Time -- Dr Khaled Mnaymneh, PEng Senior Staff Scientist, National Research Council Canada Adjunct Research Professor, Carleton University This message and any attachments may contain confidential and privileged material classified as "Protected A" and is for the sole use of the intended recipient. Any review, use, distribution or disclosure by (or to) others is strictly prohibited. If you are not the intended recipient (or authorized to receive for the recipient), please contact the sender by reply email and delete all copies of this message. Thank You. Ce message et toute pi?ce jointe connexe pourraient contenir de l'information confidentielle ou privil?gi?e d?sign?e ? Prot?g? A ?. Ils s'adressent exclusivement ? leur destinataire. L'examen, l'utilisation, la distribution et la communication de ce message par ou ? d'autres personnes sont strictement interdits. Si ce message ne vous est pas destin? (ou si vous n'avez pas l'autorisation expresse de le recevoir au nom du destinataire), veuillez contacter l'exp?diteur en r?pondant ? ce courriel et d?truire toute copie du message. Merci. -------------- next part -------------- An HTML attachment was scrubbed... URL: From reet at dtu.dk Thu Feb 2 05:35:31 2023 From: reet at dtu.dk (Rebecca Bolt Ettlinger) Date: Thu, 2 Feb 2023 10:35:31 +0000 Subject: [labnetwork] [External] Question about DC sputtering of zinc In-Reply-To: References: Message-ID: <4166ef25910444038e5ec4b00007263d@dtu.dk> Hi Sandra, We have allowed Zn in a thermal evaporator here at DTU Nanolab (a multi-user university facility). We didn't allow it in our sputter systems as we were concerned about the cryo pumps on those systems being overloaded. Zn was not so easy to control in thermal evaporation and we concluded that the optimal solution would be sputtering in a dedicated system with a turbo pump. After evaporation of course Zn was all over the chamber so it required thorough cleaning and also several "bakeout" evaporation runs of Al to get rid of Zn in the deposited layers (we usually use the evaporator for Al and Ag and the chamber gets heated most from the Al evaporation). We checked the deposited Al for traces of Zn by XPS before and after allowing Zn in the chamber and eventually got back to a state where the Zn was not detectable by XPS. We have had no user complaints afterwards but are wary of letting future users deposit Zn again because of the work involved to clean it out and check whether it is gone. All the best, Rebecca Process specialist at DTU Nanolab, Technical University of Denmark From: labnetwork On Behalf Of Vlahakis, James Sent: 1. februar 2023 16:43 To: Malhotra, Sandra Guy ; Fab Network Subject: Re: [labnetwork] [External] Question about DC sputtering of zinc Hi Sandra, a few years ago we allowed Zn deposition in our "dirty" sputter tool without a heated platen. Consider the attached vapor pressure v Temp chart - at the temps seen in our tool the vapor pressure is below the system base pressure. Since that time we've noted no significant changes in base pressure and other users haven't complained about their films being lower quality. Please take this with a grain of salt - Zn dep is not very common here, just a few times per year. In addition, our users are not fabricating CMOS devices, almost all are biomed focused these days jim From: labnetwork > On Behalf Of Malhotra, Sandra Guy Sent: Monday, January 30, 2023 4:40 PM To: Fab Network > Subject: [External] [labnetwork] Question about DC sputtering of zinc Howdy All from AggieFab, We have a user who requested permission to sputter zinc in our shared Lesker PVD75 DC sputtering tool. On the Lesker website, there is the following note: "This material has a relatively high vapor pressure at low temperatures which can cause ppm level contamination in the chamber and subsequent films. Careful consideration should be taken before this material is used in any deposition system. For this reason, some users prefer to use only dedicated vacuum chambers for deposition." I'd like to ask the group if they allow zinc depositions in their shared tools if they follow it by a thorough paste with another material, like Ti? Or if this material is prohibited because it is problematic. Thanks in advance for any inputs you may have. Best, Sandra G. Malhotra, Ph.D. | Technical Lab Manager AggieFab Nanofabrication Facility https://aggiefab.tamu.edu/ Department of Electrical & Computer Engineering, College of Engineering | Texas A&M University 3253 TAMU | College Station, TX 77843 ph: 979.845.3199 | sandra.malhotra at tamu.edu - - - - - - - - - - - - - - - - - - - - - - - - TEXAS A&M UNIVERSITY | FEARLESS on Every Front Caution: This message originated from outside of the Tufts University organization. Please exercise caution when clicking links or opening attachments. When in doubt, email the TTS Service Desk at it at tufts.edu or call them directly at 617-627-3376. -------------- next part -------------- An HTML attachment was scrubbed... URL: From renato.beraldo at hotmail.com Thu Feb 2 11:44:49 2023 From: renato.beraldo at hotmail.com (Renato Massaroto Beraldo) Date: Thu, 2 Feb 2023 16:44:49 +0000 Subject: [labnetwork] Help with technical manual Message-ID: Hello, can you have a technical manual from JETFIRST 150 ? I saw your messagen in a internet page https://mtl.mit.edu/pipermail/labnetwork/2021-November/005093.html And I'm trying to get this manual. Best regards. Ph.D. Canditade - Renato M. Beraldo - Metal Thin film Dep. Division UNICAMP University - Brazil -------------- next part -------------- An HTML attachment was scrubbed... URL: From grallion at ncsu.edu Fri Feb 3 11:04:59 2023 From: grallion at ncsu.edu (Greg Allion) Date: Fri, 3 Feb 2023 11:04:59 -0500 Subject: [labnetwork] Semiconductor Research Engineer positions opening Message-ID: Greetings Colleagues, Please forward this to any potential candidates that may come to mind. I?m pleased to announce that the North Carolina State University Nanofabrication Facility (NNF) is expanding. We will be filling two positions for process engineers. The ideal candidate will have a strong background with either plasma/dry etch or Lithography and be familiar with other semiconductor fab processing capabilities. Processing experience with wide band gap materials is also highly desirable. NNF is located in Raleigh at the eastern tip of the Research Triangle which is defined by the top 3 research Universities in the state. The triangle area has become one of the largest research parks in the country and is one of the fastest growing regions. The Raleigh area offers an east coast lifestyle, sunny ?winters?, and is still very close to a midwest cost of living. Raleigh is only a 2-3 hour drive from either mountains and beaches. Please see the link to the job posting below and feel free to contact me directly if you have any additional questions or issues accessing the job posting. A flyer is also attached with some additional details. https://jobs.ncsu.edu/postings/178184 Best Regards, Greg -- Greg Allion NC State University Nanofabrication Facility (NNF) Process Integration Engineering Manager Monteith Research Center 2410 Campus Shore Drive rm.243E Raleigh, NC 27606 -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: NC State Nanofabrication Facility Positions February 2023.pdf Type: application/pdf Size: 390164 bytes Desc: not available URL: From ana.n.cohen.ctr at army.mil Fri Feb 3 14:59:49 2023 From: ana.n.cohen.ctr at army.mil (Cohen, Ana N CTR USARMY DEVCOM ARL (USA)) Date: Fri, 3 Feb 2023 19:59:49 +0000 Subject: [labnetwork] OSRAM 1000W Bulbs Message-ID: Good afternoon Travis, No spares here. Unfortunately, our lab has had a purchase of these bulbs on back order for multiple years now. There's been some conversation on here about others stuck in similar predicaments. Depending on how soon you need 1000W bulbs again, Ushio usually has quickly available stock of their USH1000KS. They also recently have started supplying USH1001KS bulbs that are supposed to more match the 1500hr lifetime of the Osram bulbs. If you're interested in using this opportunity to retrofit your tool, there also are LED UV sources available. (Also, if anyone has had any luck receiving orders from Osram again, please let me know! I've taken a break from bothering the distributor for more details, but I'll take it up again if other orders are coming in.) Kindly, Ana -- Ana N. Cohen [she/her/hers] Photolithography Cleanroom Technician Contractor | General Technical Services, LLC US Army Research Laboratory -- [labnetwork] Suss MA06 Mask Aligner Travis Venables venables at seas.upenn.edu Mon Jan 30 13:06:13 EST 2023 Hello, I am reaching out from UPENN's Quattrone Nanofabrication Facility. We have recently ordered some spare bulbs for our MA06 mask aligner that seem to have quite a long lead time. I was wondering if anyone had some spare units that we could borrow and replace upon the receipt of our newly ordered bulbs. Attached is a link to the specific bulb we are interested in. Thanks! https://rstvisions.com/products/osram-69200-hbo-1000w-d?_pos=1&_sid=c03e23b8 2&_ss=r -- Travis Venables Lead Cleanroom Equipment Engineer Quattrone Nanofabrication Facility University of Pennsylvania P: 215-898-1787 -------------- next part -------------- A non-text attachment was scrubbed... Name: smime.p7s Type: application/pkcs7-signature Size: 6587 bytes Desc: not available URL: From James.Grant at glasgow.ac.uk Mon Feb 6 09:37:33 2023 From: James.Grant at glasgow.ac.uk (James Grant) Date: Mon, 6 Feb 2023 14:37:33 +0000 Subject: [labnetwork] SPTS Rapier Module - Laser/White Light Interferometry Message-ID: Hello, Hoping the community can feedback on their experiences. In 2019 we purchased an SPTS Rapier DSiE tool. As well as doing Bosch processing we also require the tool to do mixed process etching. Quite a few of our users, on our other Si etch tools (OIPT Estrelas and STS Multiplex), use mixed process to etch SOI materials or a-Si on fused silica substrates. These systems have Intellemetrics LEP 410/500 red laser interferometer units for the purpose of end-point detection. When we purchased the Rapier, SPTS did tell us that a laser end-point system could not be used. They encouraged us instead to use the Claritas OES system on the tool, however when we pointed out our typical use scenario (small substrate size, low % open area) they admitted that the Claritas would struggle to differentiate when to end-point. Our preferred laser interferometer vendor, Intellemetrics, say they have sold LEP 500 red laser interferometer units to SPTS but cannot say on what systems they have been used. We have found that EPFL use white light interferometry end-point detection on their Rapier module - have contacted Phillipe Fluckiger separately. My question to the community is has anyone used a laser interferometry system on an SPTS Rapier? If so, could they feedback with their experiences. Apologies I'm asking such an open question! Cheers, James Dr. James Paul Grant Research Engineer in Plasma Processing Plasma Processing Group james.grant at glasgow.ac.uk [cid:image001.png at 01D93A36.DE509BD0] www.JWNC.gla.ac.uk [cid:image002.png at 01D93A36.DE509BD0] LinkedIn.com/company/JWNC [cid:image003.png at 01D93A36.DE509BD0] @UofG_JWNC -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.png Type: image/png Size: 26666 bytes Desc: image001.png URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image002.png Type: image/png Size: 1538 bytes Desc: image002.png URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image003.png Type: image/png Size: 1717 bytes Desc: image003.png URL: From massey21 at llnl.gov Mon Feb 6 13:58:47 2023 From: massey21 at llnl.gov (Massey, Travis) Date: Mon, 6 Feb 2023 18:58:47 +0000 Subject: [labnetwork] Question: adhesion of metal to BPDA/PPD polyimide In-Reply-To: <3d0e9163-6746-d694-7209-b84abc4598be@physics.ubc.ca> References: <3d0e9163-6746-d694-7209-b84abc4598be@physics.ubc.ca> Message-ID: Hi Mario, We use PI-2600 (2610 and 2611) extensively with Ti-Pt and Ti-Au metallization (sputtered). Films are 2 um and 5 um, respectively. Recipes and procedures are often tailored to specific toolsets and projects, so let?s start higher level with some general thoughts: 1. I suggest a vacuum dehydration bake before each film deposition ? so, before the polyimide and again before the metal. We use our polyimide oven for this. The dehydration bake before metal dep has had a marked impact on our metal adhesion. 2. Are you imizing your polyimide at >350 C under an inert (N2) atmosphere? We do 375C because nothing in our stack is that temperature sensitive. Occasionally our oven crashes before reaching 375C and I?ve seen bubbles beneath the metal and other issues down the line. I?ve heard of some people curing the polyimide at lower temperatures, and while you can drive off all the solvent and get a mechanically stable film, the imidization may be incomplete. 3. Our process deviates from the datasheet on the softbake. We softbake for 10 minutes at 95C (so, lower temp for longer time). This is unusually cool for NMP drive-off, but the process has worked for our team since long before I arrived. If your films are thicker, you may need to increase one parameter or the other. 4. Does your polyimide curing oven get used for organics other than polyimide? 5. I avoid adhesion promoters between Si and PI unless it?s strictly necessary (and it?s generally not, for my processes). I view it as another variable and possible contaminant down the line ? this contamination possibility is why I bring it up here. That said, others in my facility have used adhesion promoters for polyimide on glass without issue? but it?s good to reduce variables where you can. I?m happy to discuss further if you have additional questions/issues or want a detailed procedure. Best, Travis Massey Lawrence Livermore National Laboratory 7000 East Avenue, Livermore, CA 94550 925-422-9509 (desk) 925-495-7103 (work cell) massey21 at llnl.gov From: labnetwork On Behalf Of Beaudoin, Mario Sent: Wednesday, February 1, 2023 11:08 AM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Question: adhesion of metal to BPDA/PPD polyimide Dear Networkers, I'm new to polyimide processing. We have a student project that requires deposition of metal on polyimide (BPDA/PPD PI-2600 series from HD microsystems). We tried liftoff with depositon of Ti/Pt by Ebeam. Despite a 1min O2 plasma, albeit at low power, the metal was peeling off. Moreover, the photoresist was also peeling off from the polyimide. Does anyone have a good working recipe for metalization of polyimide? Thanks, Mario -- [cid:image001.jpg at 01D93A19.FC28DC70] -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.jpg Type: image/jpeg Size: 21446 bytes Desc: image001.jpg URL: From michael.rooks at yale.edu Mon Feb 6 15:47:26 2023 From: michael.rooks at yale.edu (Michael Rooks) Date: Mon, 6 Feb 2023 15:47:26 -0500 Subject: [labnetwork] job opening - yale nano Message-ID: Job opening at Yale University: e-beam lithography, electron microscopy, and AFM in a shared core facility. New Haven CT (USA), BS-level or higher. Check out the job posting here. -------------------------------- Michael Rooks Yale University nano.yale.edu -------------- next part -------------- An HTML attachment was scrubbed... URL: From daniel.woodie at cornell.edu Mon Feb 6 15:57:14 2023 From: daniel.woodie at cornell.edu (Dan P. Woodie) Date: Mon, 6 Feb 2023 20:57:14 +0000 Subject: [labnetwork] Question: adhesion of metal to BPDA/PPD polyimide In-Reply-To: References: <3d0e9163-6746-d694-7209-b84abc4598be@physics.ubc.ca> Message-ID: A solid cure of the polyimide is critical, as the imidization reaction releases water as a by-product, which will cause bubbling and delamination issues for any films over it. The challenge is that you never reach 100% curing as it is an asymptotic approach. So, you need to ensure that the cure is done at a sufficiently long enough time and high enough temperature that no further cross-linking will occur during any subsequent steps. You may want to also consider using Cyclotene from Dow instead of polyimide. It has many properties like polyimide but does not have any volatile byproducts from the cross linking, so issues like this mostly go away. I used it for my thesis electroplating copper into some thick structures and is an improvement on polyimide in many aspects. Dan Daniel Woodie Interim Asst. Director of Facilities / Safety Manager / Facilities Engineer - (formerly an engineer at the Cornell NanoScale Facility) College of Engineering From: labnetwork On Behalf Of Massey, Travis Sent: Monday, February 6, 2023 1:59 PM To: Beaudoin, Mario ; labnetwork at mtl.mit.edu Subject: Re: [labnetwork] Question: adhesion of metal to BPDA/PPD polyimide Hi Mario, We use PI-2600 (2610 and 2611) extensively with Ti-Pt and Ti-Au metallization (sputtered). Films are 2 um and 5 um, respectively. Recipes and procedures are often tailored to specific toolsets and projects, so let?s start higher level with some general thoughts: 1. I suggest a vacuum dehydration bake before each film deposition ? so, before the polyimide and again before the metal. We use our polyimide oven for this. The dehydration bake before metal dep has had a marked impact on our metal adhesion. 2. Are you imizing your polyimide at >350 C under an inert (N2) atmosphere? We do 375C because nothing in our stack is that temperature sensitive. Occasionally our oven crashes before reaching 375C and I?ve seen bubbles beneath the metal and other issues down the line. I?ve heard of some people curing the polyimide at lower temperatures, and while you can drive off all the solvent and get a mechanically stable film, the imidization may be incomplete. 3. Our process deviates from the datasheet on the softbake. We softbake for 10 minutes at 95C (so, lower temp for longer time). This is unusually cool for NMP drive-off, but the process has worked for our team since long before I arrived. If your films are thicker, you may need to increase one parameter or the other. 4. Does your polyimide curing oven get used for organics other than polyimide? 5. I avoid adhesion promoters between Si and PI unless it?s strictly necessary (and it?s generally not, for my processes). I view it as another variable and possible contaminant down the line ? this contamination possibility is why I bring it up here. That said, others in my facility have used adhesion promoters for polyimide on glass without issue? but it?s good to reduce variables where you can. I?m happy to discuss further if you have additional questions/issues or want a detailed procedure. Best, Travis Massey Lawrence Livermore National Laboratory 7000 East Avenue, Livermore, CA 94550 925-422-9509 (desk) 925-495-7103 (work cell) massey21 at llnl.gov From: labnetwork > On Behalf Of Beaudoin, Mario Sent: Wednesday, February 1, 2023 11:08 AM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Question: adhesion of metal to BPDA/PPD polyimide Dear Networkers, I'm new to polyimide processing. We have a student project that requires deposition of metal on polyimide (BPDA/PPD PI-2600 series from HD microsystems). We tried liftoff with depositon of Ti/Pt by Ebeam. Despite a 1min O2 plasma, albeit at low power, the metal was peeling off. Moreover, the photoresist was also peeling off from the polyimide. Does anyone have a good working recipe for metalization of polyimide? Thanks, Mario -- [cid:image001.jpg at 01D93A43.AD93E170] -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.jpg Type: image/jpeg Size: 21446 bytes Desc: image001.jpg URL: From odc1n08 at soton.ac.uk Tue Feb 7 04:31:20 2023 From: odc1n08 at soton.ac.uk (Owain Clark) Date: Tue, 7 Feb 2023 09:31:20 +0000 Subject: [labnetwork] SPTS Rapier Module - Laser/White Light Interferometry In-Reply-To: References: Message-ID: Hi James, we don't have a rapier but we have an intellemetrics system on an OIPT chamber here. If there is a suitable optical KF/NW standard 40mm+ port located normally above the wafer then I expect it would work fine. We use IR wavelength for better compatibility with Si+Si dielectrics. I integrated the system without an assistance from OIPT, it does not interact with the tool control s/w so you need to manually observe one and hit stop on the other. Downsides of our system are tricky to manually focus and get the back reflection in the detector plus you have to create your masks with an opening in the correct spot, with limited XY translation possible. Can't tell you anything about etch uniformity either. I agree with SPTS in general, OES is a better and more versatile method of etch detection. Particularly if you have a full spectrum based system, also can be useful for plasma diagnostics and chamber cleaning information so it is more than just etch process. An inferometer gives you none of that additional info. Make us an offer if you like, I can't remember the last time our system got used. It's easier/quicker to take wafers out part etch and use the ellipsometer to characterise... Very easy to detach from tool and reinstall elsewhere. BR, Owain From: labnetwork On Behalf Of James Grant Sent: 06 February 2023 14:38 To: labnetwork at mtl.mit.edu Subject: [labnetwork] SPTS Rapier Module - Laser/White Light Interferometry CAUTION: This e-mail originated outside the University of Southampton. Hello, Hoping the community can feedback on their experiences. In 2019 we purchased an SPTS Rapier DSiE tool. As well as doing Bosch processing we also require the tool to do mixed process etching. Quite a few of our users, on our other Si etch tools (OIPT Estrelas and STS Multiplex), use mixed process to etch SOI materials or a-Si on fused silica substrates. These systems have Intellemetrics LEP 410/500 red laser interferometer units for the purpose of end-point detection. When we purchased the Rapier, SPTS did tell us that a laser end-point system could not be used. They encouraged us instead to use the Claritas OES system on the tool, however when we pointed out our typical use scenario (small substrate size, low % open area) they admitted that the Claritas would struggle to differentiate when to end-point. Our preferred laser interferometer vendor, Intellemetrics, say they have sold LEP 500 red laser interferometer units to SPTS but cannot say on what systems they have been used. We have found that EPFL use white light interferometry end-point detection on their Rapier module - have contacted Phillipe Fluckiger separately. My question to the community is has anyone used a laser interferometry system on an SPTS Rapier? If so, could they feedback with their experiences. Apologies I'm asking such an open question! Cheers, James Dr. James Paul Grant Research Engineer in Plasma Processing Plasma Processing Group james.grant at glasgow.ac.uk [cid:image001.png at 01D93AD6.EF4C07D0] www.JWNC.gla.ac.uk [cid:image004.png at 01D93AD6.EF4C07D0] LinkedIn.com/company/JWNC [cid:image005.png at 01D93AD6.EF4C07D0] @UofG_JWNC -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.png Type: image/png Size: 26666 bytes Desc: image001.png URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image004.png Type: image/png Size: 465 bytes Desc: image004.png URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image005.png Type: image/png Size: 476 bytes Desc: image005.png URL: From kmcpeak at lsu.edu Tue Feb 7 18:01:17 2023 From: kmcpeak at lsu.edu (Kevin M McPeak) Date: Tue, 7 Feb 2023 23:01:17 +0000 Subject: [labnetwork] Budgeting a cleanroom build Message-ID: Dear Colleagues, At LSU, we are putting together a proposal to build a cleanroom on campus (our current one is 15 min away). Our VP of research has asked me to determine a rough budget for constructing a microelectronics cleanroom. Below are the itemized details: - Size: 2000 sqft of class 100 space and 2000 sqft of grey room space. Our current off-campus cleanroom is 2000 sqft. - Location: 5th floor (top floor) of chemistry building (total space of 5th floor is 12,000 sqft) - Current condition of space: 5th floor is unfinished, i.e., concrete floors with pipes coming from the ground and 10ft ceilings. I want to factor in the following costs: air handlers, toxic gas handling for PECVD, LPCVD, and ICP-RIE machines, and getting utilities to each tool, i.e., power, water, and N2. $8 million is the current number that people are tossing around. I would greatly appreciate the feedback on that value. I certainly don't want to underfund the project. Thanks! Regards, Kevin From beaudoin at physics.ubc.ca Tue Feb 7 20:05:53 2023 From: beaudoin at physics.ubc.ca (Beaudoin, Mario) Date: Tue, 7 Feb 2023 17:05:53 -0800 Subject: [labnetwork] Budgeting a cleanroom build In-Reply-To: References: Message-ID: <1b544477-05b4-7e93-2203-4acff0961e5d@physics.ubc.ca> Kevin, We did a 1500 sq ft cleanroom: * Yellow room, class 100, 350 sq ft * Yellow room, class 1000, 350 sq ft * white room, class 1000, 900 sq ft * 9ft high inside clearance The cost to purchase the cleanroom and have it installed in an old building where all the services are in a mezzanine above the room cost us 5M$ CAD (roughly 4M$ USD).? Hope this helps. Mario On 2023-02-07 3:01 p.m., Kevin M McPeak wrote: > [CAUTION: Non-UBC Email] > > Dear Colleagues, > > At LSU, we are putting together a proposal to build a cleanroom on campus (our current one is 15 min away). Our VP of research has asked me to determine a rough budget for constructing a microelectronics cleanroom. Below are the itemized details: > > - Size: 2000 sqft of class 100 space and 2000 sqft of grey room space. Our current off-campus cleanroom is 2000 sqft. > - Location: 5th floor (top floor) of chemistry building (total space of 5th floor is 12,000 sqft) > - Current condition of space: 5th floor is unfinished, i.e., concrete floors with pipes coming from the ground and 10ft ceilings. > > I want to factor in the following costs: air handlers, toxic gas handling for PECVD, LPCVD, and ICP-RIE machines, and getting utilities to each tool, i.e., power, water, and N2. > > $8 million is the current number that people are tossing around. I would greatly appreciate the feedback on that value. I certainly don't want to underfund the project. Thanks! > > Regards, > Kevin > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork > -- -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: Mario Beaudoin SBQMI sig 2.jpg Type: image/jpeg Size: 21446 bytes Desc: not available URL: From julia.aebersold at louisville.edu Wed Feb 8 12:36:22 2023 From: julia.aebersold at louisville.edu (Aebersold, Julia) Date: Wed, 8 Feb 2023 17:36:22 +0000 Subject: [labnetwork] Budgeting a cleanroom build In-Reply-To: References: Message-ID: Kevin, Dennis Grimard at MITnano would be a great resource to ask this question. Essentially, its going to cost more than you expect. Also, make sure you lock in rock solid commitment of long term support for infrastructure and equipment. Cheers! Julia Aebersold, Ph.D. Manager, Micro/Nano Technology Center University of Louisville Shumaker Research Building, Room 233 2210 South Brook Street Louisville, KY? 40292 (502) 852-1572 ? http://louisville.edu/micronano/ -----Original Message----- From: labnetwork On Behalf Of Kevin M McPeak Sent: Tuesday, February 7, 2023 6:01 PM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Budgeting a cleanroom build CAUTION: This email originated from outside of our organization. Do not click links, open attachments, or respond unless you recognize the sender's email address and know the contents are safe. Dear Colleagues, At LSU, we are putting together a proposal to build a cleanroom on campus (our current one is 15 min away). Our VP of research has asked me to determine a rough budget for constructing a microelectronics cleanroom. Below are the itemized details: - Size: 2000 sqft of class 100 space and 2000 sqft of grey room space. Our current off-campus cleanroom is 2000 sqft. - Location: 5th floor (top floor) of chemistry building (total space of 5th floor is 12,000 sqft) - Current condition of space: 5th floor is unfinished, i.e., concrete floors with pipes coming from the ground and 10ft ceilings. I want to factor in the following costs: air handlers, toxic gas handling for PECVD, LPCVD, and ICP-RIE machines, and getting utilities to each tool, i.e., power, water, and N2. $8 million is the current number that people are tossing around. I would greatly appreciate the feedback on that value. I certainly don't want to underfund the project. Thanks! Regards, Kevin _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://nam11.safelinks.protection.outlook.com/?url=https%3A%2F%2Fmtl.mit.edu%2Fmailman%2Flistinfo.cgi%2Flabnetwork&data=05%7C01%7Cjulia.aebersold%40louisville.EDU%7Cbfc2b5d1a59a47aa0bf008db096f5484%7Cdd246e4a54344e158ae391ad9797b209%7C0%7C0%7C638114145992578330%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000%7C%7C%7C&sdata=61ESPgIoWI8eABK0px1EK8Ro2pTyinSJOJUnYGTvm4I%3D&reserved=0 From lvchang at Central.UH.EDU Thu Feb 9 10:14:36 2023 From: lvchang at Central.UH.EDU (Chang, Long) Date: Thu, 9 Feb 2023 15:14:36 +0000 Subject: [labnetwork] HDD Best Practice Message-ID: <590390B9-BEB6-4FB7-AA4F-EC7FE04A1270@central.uh.edu> Hi All, Yesterday we experienced a HDD failure for our AFM. This event should have been prevented by the Raid Mirror with 2 HDD. Our setup where we just have a single working drive and a spare backup clone has been more successful. We are planning to move away from Raid Mirror. Anyone has expertise/experience here they can share? Thanks, Long Chang Technical Director UH Nanofabrication Facility Houston, TX lvchang at central.uh.edu From demis at ucsb.edu Thu Feb 9 12:08:47 2023 From: demis at ucsb.edu (Demis D. John) Date: Thu, 9 Feb 2023 09:08:47 -0800 Subject: [labnetwork] HDD Best Practice In-Reply-To: <590390B9-BEB6-4FB7-AA4F-EC7FE04A1270@central.uh.edu> References: <590390B9-BEB6-4FB7-AA4F-EC7FE04A1270@central.uh.edu> Message-ID: <5AA84C0C-1CB7-4560-8632-A077E1E41BB6@ucsb.edu> We have also found that keeping a cloned HDD *outside* the computer chassis is very important. Once a tool has been set up satisfactorily, we then take the tool down, clone the HDD onto a similar one (using external HDD cloners), and then keep the clone in a separate cabinet (anti-static etc). In addition, we have all user recipes, software folders backed up from the live tool computer to our network storage. So for restore - we can swap the cloned HDD back into the computer, And then copy the backed up network files to the computer. We had a power glitch result in damaging Both HDD?s (including raid backup) within our ASML computer)! -- Demis > On Feb 9, 2023, at 08:42, Chang, Long wrote: > > ?Hi All, > > Yesterday we experienced a HDD failure for our AFM. This event should have been prevented by the Raid Mirror with 2 HDD. Our setup where we just have a single working drive and a spare backup clone has been more successful. We are planning to move away from Raid Mirror. Anyone has expertise/experience here they can share? > > Thanks, > Long Chang > Technical Director > UH Nanofabrication Facility > Houston, TX > lvchang at central.uh.edu > > > > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork From patricns at uw.edu Thu Feb 9 12:17:45 2023 From: patricns at uw.edu (N Shane Patrick) Date: Thu, 9 Feb 2023 09:17:45 -0800 Subject: [labnetwork] HDD Best Practice In-Reply-To: <590390B9-BEB6-4FB7-AA4F-EC7FE04A1270@central.uh.edu> References: <590390B9-BEB6-4FB7-AA4F-EC7FE04A1270@central.uh.edu> Message-ID: RAID mirrors are incredibly useful tools when the expectations and management are properly handled, but they are not a backup and don?t replace the need for a separate backup. RAID mirrors are a downtime mitigation tool. They allow the computer to continue operating in the event a single disk fails (assuming RAID 1, there are multi-mirror RAID setups as well) and are based on it being statistically unlikely that all disks in the array fail simultaneously. This gives the maintainer time to swap out the bad disk and allow the array to re-silver, all while keeping the system online in most cases. It isn?t uncommon to treat a raid array as a ?built-in backup?, but that?s not the true purpose of it. For instance, having a RAID mirror array will fail if there aren?t policies requiring regular checks on the status of the array or automated monitoring and alert tasks to notify the maintainer of a disk problem. It will also fail if nothing is done about any signs of failure or alerts. It will also not save you if the controller (software or hardware) malfunctions or fails and starts writing corrupted data to both mirrors, or if something in an operating system or on the data bus goes screwy and causes data streams to be garbled before or during transmission. Bad data into the array, bad array. You still need a full backup solution, or at least a backup solution of critical data, like an automated incremental backup solution or a regular disk clone solution like you mention in your post. So. RAID mirror - minimize downtime if managed properly, not a backup. Backups allow for recovery when, not if, a failure and downtime occurs. Backups should be segregated copies of data ideally held somewhere other than on the system being backed up. N. Shane Patrick Manager, Lab Operations and Safety Electron Beam Lithography Washington Nanofabrication Facility (WNF) National Nanotechnology Coordinated Infrastructure (NNCI) University of Washington - NanoES Fluke Hall 129, Box 352143 (206) 221-1045 patricns at uw.edu http://www.wnf.washington.edu/ > On Feb 9, 2023, at 7:14 AM, Chang, Long wrote: > > Hi All, > > Yesterday we experienced a HDD failure for our AFM. This event should have been prevented by the Raid Mirror with 2 HDD. Our setup where we just have a single working drive and a spare backup clone has been more successful. We are planning to move away from Raid Mirror. Anyone has expertise/experience here they can share? > > Thanks, > Long Chang > Technical Director > UH Nanofabrication Facility > Houston, TX > lvchang at central.uh.edu > > > > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://urldefense.com/v3/__https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork__;!!K-Hz7m0Vt54!kD3GvIsu3cdHCobhN6JZoubyakmhXr0GbvPzLpQkuvlVc4J3kVmQo00o47eLqZ8G7rDSz02L0KnseCkmbfg2$ -------------- next part -------------- An HTML attachment was scrubbed... URL: From michael.martin at louisville.edu Thu Feb 9 12:52:13 2023 From: michael.martin at louisville.edu (Martin, Michael) Date: Thu, 9 Feb 2023 17:52:13 +0000 Subject: [labnetwork] HDD Best Practice In-Reply-To: <590390B9-BEB6-4FB7-AA4F-EC7FE04A1270@central.uh.edu> References: <590390B9-BEB6-4FB7-AA4F-EC7FE04A1270@central.uh.edu> Message-ID: Hi Long, Yes, we use the drive clone method instead of RAID. We clone the drive and then put the clone in the computers as sometimes there are clone glitches. Regards, Michael ________________________________ From: labnetwork on behalf of Chang, Long Sent: Thursday, February 9, 2023 10:14 AM To: Fab Network Subject: [labnetwork] HDD Best Practice CAUTION: This email originated from outside of our organization. Do not click links, open attachments, or respond unless you recognize the sender's email address and know the contents are safe. Hi All, Yesterday we experienced a HDD failure for our AFM. This event should have been prevented by the Raid Mirror with 2 HDD. Our setup where we just have a single working drive and a spare backup clone has been more successful. We are planning to move away from Raid Mirror. Anyone has expertise/experience here they can share? Thanks, Long Chang Technical Director UH Nanofabrication Facility Houston, TX lvchang at central.uh.edu _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://nam11.safelinks.protection.outlook.com/?url=https%3A%2F%2Fmtl.mit.edu%2Fmailman%2Flistinfo.cgi%2Flabnetwork&data=05%7C01%7Cmichael.martin%40louisville.EDU%7Ca4e33133feba45a1a05108db0abc9127%7Cdd246e4a54344e158ae391ad9797b209%7C0%7C0%7C638115577231692696%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C3000%7C%7C%7C&sdata=pCsDHDd1QY%2BTCypSbcJRVDBa0ZjW960e1M9JO16Mprw%3D&reserved=0 -------------- next part -------------- An HTML attachment was scrubbed... URL: From sieb at 4dlabs.ca Thu Feb 9 12:57:57 2023 From: sieb at 4dlabs.ca (Nathanael Sieb) Date: Thu, 9 Feb 2023 09:57:57 -0800 Subject: [labnetwork] HDD Best Practice In-Reply-To: <590390B9-BEB6-4FB7-AA4F-EC7FE04A1270@central.uh.edu> References: <590390B9-BEB6-4FB7-AA4F-EC7FE04A1270@central.uh.edu> Message-ID: Hi Long, We use Macrium Reflect to save an image of our drives on an external server.? That way when a hard-drive dies, we just need to restore the image.? We set it on a schedule for some systems where frequent backups are useful.? Other systems are a single backup and then we disconnect the tool from the network connection.? It has saved us a lot of times. Thanks, Nathanael On 2023-02-09 7:14 a.m., Chang, Long wrote: > Hi All, > > Yesterday we experienced a HDD failure for our AFM. This event should have been prevented by the Raid Mirror with 2 HDD. Our setup where we just have a single working drive and a spare backup clone has been more successful. We are planning to move away from Raid Mirror. Anyone has expertise/experience here they can share? > > Thanks, > Long Chang > Technical Director > UH Nanofabrication Facility > Houston, TX > lvchang at central.uh.edu > > > > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork *Nathanael Sieb * Director of Operations and Administration ?| 4D LABS Simon Fraser University 8888 University Dr., Burnaby, B.C. V5A 1S6 T: 778.782.8084?| F: 778.782.3765 | www.4dlabs.ca Facebook ?| Twitter ?| LinkedIn -------------- next part -------------- An HTML attachment was scrubbed... URL: From massey21 at llnl.gov Thu Feb 9 16:01:33 2023 From: massey21 at llnl.gov (Massey, Travis) Date: Thu, 9 Feb 2023 21:01:33 +0000 Subject: [labnetwork] HDD Best Practice In-Reply-To: <590390B9-BEB6-4FB7-AA4F-EC7FE04A1270@central.uh.edu> References: <590390B9-BEB6-4FB7-AA4F-EC7FE04A1270@central.uh.edu> Message-ID: Hi Long, Very few of our tools are networked (or could be networked) for backups to a server, so I recently started using Clonezilla to create clones of all of the tools. Clonezilla is a lightweight bootable Linux build whose sole purpose is to clone drives. Just put it on a flash drive (or CD, for tools older than flash drive support -- first-gen Windows XP and earlier), stick it in the tool, and boot to it. It'll guide you through the steps. Once I have the clone, I keep the dedicated clone drive in a safe place *and* manually put the clone image on a server in case something happens to the physical copy of the clone. Best, Travis Massey Center for Micro and Nanotechnology | Center for Bioengineering Lawrence Livermore National Laboratory -----Original Message----- From: labnetwork On Behalf Of Chang, Long Sent: Thursday, February 9, 2023 7:15 AM To: Fab Network Subject: [labnetwork] HDD Best Practice Hi All, Yesterday we experienced a HDD failure for our AFM. This event should have been prevented by the Raid Mirror with 2 HDD. Our setup where we just have a single working drive and a spare backup clone has been more successful. We are planning to move away from Raid Mirror. Anyone has expertise/experience here they can share? Thanks, Long Chang Technical Director UH Nanofabrication Facility Houston, TX lvchang at central.uh.edu _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://urldefense.us/v3/__https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork__;!!G2kpM7uM-TzIFchu!iZO-N1yGztn99AIvATf_riH_OinnPkqjv6GW8GmSVVfpaTXYv-PKEVu_asDHO1px0Q$ From whipp003 at umn.edu Thu Feb 9 16:40:56 2023 From: whipp003 at umn.edu (Tony Whipple) Date: Thu, 9 Feb 2023 15:40:56 -0600 Subject: [labnetwork] HDD Best Practice In-Reply-To: <590390B9-BEB6-4FB7-AA4F-EC7FE04A1270@central.uh.edu> References: <590390B9-BEB6-4FB7-AA4F-EC7FE04A1270@central.uh.edu> Message-ID: Hello Long; We have some systems that are older but still work great and have a legacy computer. The computer sometimes can not be upgraded if it also uses older cards such as ISA slots. The limited life time of spinning hard drives caused us to switch to using SSD for the older IDE interface drives. Making a duplicate on a second SSD and keeping the backup in a static bag nearby is a low cost backup. We have needed to use this a few times and it was quick and easy to swap it out. Regards, Tony W. On Thu, Feb 9, 2023 at 10:38 AM Chang, Long wrote: > Hi All, > > Yesterday we experienced a HDD failure for our AFM. This event should have > been prevented by the Raid Mirror with 2 HDD. Our setup where we just have > a single working drive and a spare backup clone has been more successful. > We are planning to move away from Raid Mirror. Anyone has > expertise/experience here they can share? > > Thanks, > Long Chang > Technical Director > UH Nanofabrication Facility > Houston, TX > lvchang at central.uh.edu > > > > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork > -------------- next part -------------- An HTML attachment was scrubbed... URL: From lino.eugene at uwaterloo.ca Fri Feb 10 13:20:13 2023 From: lino.eugene at uwaterloo.ca (Lino Eugene) Date: Fri, 10 Feb 2023 18:20:13 +0000 Subject: [labnetwork] HDD Best Practice In-Reply-To: <590390B9-BEB6-4FB7-AA4F-EC7FE04A1270@central.uh.edu> References: <590390B9-BEB6-4FB7-AA4F-EC7FE04A1270@central.uh.edu> Message-ID: Hi Long, Most of our computers are connected to a Synology NAS with RAID 1 configuration in a private network. The Synology software Active to Backup Business does back up the Windows 7 and 10 computers and Macrium Reflect is used for Windows XP. Backups are run weekly and several versions are kept for few months. We also have an offline backup of the NAS. When a hard drive fails, a new one is installed and the computer can be booted from a recovery media and connects to the NAS to recover a backup. This has saved us from long downtimes few times. Best, Lino Eugene, P.Eng., Ph.D., Micro/nanofabrication process engineer Quantum-Nano Fabrication and Characterization Facility Office of Research QNC 1611 University of Waterloo 200 University Avenue West Waterloo, ON, Canada N2L 3G1 Ph: +1 519-888-4567 #37788 Cell: +1 226-929-1685 Websites: https://uwaterloo.ca/quantum-nano-fabrication-and-characterization-facility/ https://fab.qnc.uwaterloo.ca/ -----Message d'origine----- De : labnetwork De la part de Chang, Long Envoy? : 9 f?vrier 2023 10:15 ? : Fab Network Objet : [labnetwork] HDD Best Practice Hi All, Yesterday we experienced a HDD failure for our AFM. This event should have been prevented by the Raid Mirror with 2 HDD. Our setup where we just have a single working drive and a spare backup clone has been more successful. We are planning to move away from Raid Mirror. Anyone has expertise/experience here they can share? Thanks, Long Chang Technical Director UH Nanofabrication Facility Houston, TX lvchang at central.uh.edu _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: From jtamelier at eng.ucsd.edu Mon Feb 13 12:02:45 2023 From: jtamelier at eng.ucsd.edu (John Tamelier) Date: Mon, 13 Feb 2023 09:02:45 -0800 Subject: [labnetwork] O2 8%/CF4 gas cylinder Message-ID: <6443ece4-9b28-096b-4a78-473408048db3@eng.ucsd.edu> Hello all, We have been having difficulty getting a cylinder of O2 8%/CF4. We placed an order in early November 2022 with Linde and have yet to receive a ship date for the material. We have three orders out, but need the material urgently. The shortest lead time we have found is 9 weeks. Does anyone have any contacts for purchasing this gas or other options for obtaining a cylinder of this gas? Is anyone else experiencing this type of lead time for gas cylinders? Thank you, John ----------------------------------- John Tamelier, Ph.D. Interim Director - Nano3 University of California, San Diego Atkinson Hall, 5th Floor Front Desk, M/C 0436 9500 Gilman Drive La Jolla, California 92093-0436 Phone: (858) 246-2735 Fax:?? (858) 534-9092 E-mail: jtamelier at ucsd.edu http://nano3.calit2.net ---------------------------------- From olms0025 at umn.edu Mon Feb 13 15:37:18 2023 From: olms0025 at umn.edu (Brian K. Olmsted) Date: Mon, 13 Feb 2023 14:37:18 -0600 Subject: [labnetwork] group buy: Entegris 5" mask holder? Message-ID: Hi, We are looking at adding a 5" mask holder from Entegris (A94-0215), but the minimum order is 8. Is anyone interested in going in together to share a minimum quantity order? We only need one. The price per unit is $643.26. Thanks, Brian K. Olmsted Associate Director of Laboratory Operations University of Minnesota | MNC cse.umn.edu/mnc 612.626.3287 olms0025 at umn.edu [image: image.png] -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image.png Type: image/png Size: 544842 bytes Desc: not available URL: From massey21 at llnl.gov Mon Feb 13 17:39:56 2023 From: massey21 at llnl.gov (Massey, Travis) Date: Mon, 13 Feb 2023 22:39:56 +0000 Subject: [labnetwork] Light filtering in photolithography bays Message-ID: Hi all, LLNL recently changed out its cleanroom lighting to LEDs from fluorescent tubes. A few months later, we found we were having lithography issues - our resist is getting exposed under ambient light within the litho bays within 5-15 minutes. The lights in our litho bays all have ~30 year old plastic diffusers that snap on over the tubes, and these diffusers have a filter film inside. We speculate that the new LEDs may have photobleached the aging photosensitive pigments. Measurements of the light intensity in the cleanroom at 365, 405, and 435 nm indicate that our litho bays have just as much light at these wavelengths than our non-litho bays! -- As an alternative to filtering the light, has anyone found LED cleanroom lighting that does not emit in the >450 nm range? For example, form-compatible LED 'tubes' that only emit at ~580 nm (yellow)? -- How do all of you solve this problem? What product do you use to filter your lights (e.g., tube covers, panel covers, applied films, solid dyed plastics, laser safety products...) -- Have you found that your filters have a finite lifespan? For example, I've heard in previous discussions that window films are only good for 5 years or so before the wavelength-absorbent pigment decomposes and is ineffective. -- To what extent have you found computer monitors' light emissions to be a point of concern? -- Other thoughts/suggestions that we should explore or consider? Thanks, Travis Massey Center for Micro and Nanotechnology (CMNT) Lawrence Livermore National Laboratory -------------- next part -------------- An HTML attachment was scrubbed... URL: From jtmitch5 at ncsu.edu Tue Feb 14 01:05:41 2023 From: jtmitch5 at ncsu.edu (James Mitchell) Date: Tue, 14 Feb 2023 01:05:41 -0500 Subject: [labnetwork] O2 8%/CF4 gas cylinder In-Reply-To: <6443ece4-9b28-096b-4a78-473408048db3@eng.ucsd.edu> References: <6443ece4-9b28-096b-4a78-473408048db3@eng.ucsd.edu> Message-ID: Have you tried Airgas? Jim On Mon, Feb 13, 2023, 5:37 PM John Tamelier wrote: > Hello all, > > We have been having difficulty getting a cylinder of O2 8%/CF4. We > placed an order in early November 2022 with Linde and have yet to > receive a ship date for the material. We have three orders out, but need > the material urgently. The shortest lead time we have found is 9 weeks. > > Does anyone have any contacts for purchasing this gas or other options > for obtaining a cylinder of this gas? Is anyone else experiencing this > type of lead time for gas cylinders? > > Thank you, > > John > > ----------------------------------- > John Tamelier, Ph.D. > Interim Director - Nano3 > > University of California, San Diego > Atkinson Hall, 5th Floor Front Desk, M/C 0436 > 9500 Gilman Drive > La Jolla, California 92093-0436 > Phone: (858) 246-2735 > Fax: (858) 534-9092 > E-mail: jtamelier at ucsd.edu > http://nano3.calit2.net > ---------------------------------- > > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork > -------------- next part -------------- An HTML attachment was scrubbed... URL: From christian.pies at heidelberg-instruments.com Tue Feb 14 03:48:48 2023 From: christian.pies at heidelberg-instruments.com (Christian Pies) Date: Tue, 14 Feb 2023 08:48:48 +0000 Subject: [labnetwork] Light filtering in photolithography bays In-Reply-To: References: Message-ID: Hi Travis, when we planned our new cleanroom facilities, we wanted to switch to LED lighting, but were also quite concerned about accidentally exposing the photoresists. After some discussions with one of our resist suppliers, we learned that wavelengths below 500 nm must be avoided. Unfortunately, most yellow LEDs show a sidepeak around 420 - 450 nm, which is often not visible in the datasheets unless the emission spectrum is shown in a logarithmic representation. The LEDs we originally looked at had a side peak of 0.4% intensity, which was still too much. Pure yellow light LEDs have a lower efficiency, but covering the lamps with a filter film probably reduces the output in a similar way or even more. We finally decided for the following LEDs: https://dammedia.osram.info/media/resource/hires/osram-dam-6665600/LY%20CKBP_EN.pdf They are in use for almost two years now and we haven't experienced any problems with our resists so far. They are a bit more orange than the typical lighting, which feels a bit unfamiliar at the beginning. Computer monitors can be an issue. If we leave our coated substrates on the desk in front of the monitor, the resist is typically exposed within a few hours. A short time next to the monitor while setting up the machine is usually uncritical (depending on the sensitivity of the resist). "Light contamination" of the entire room from the monitors has not been a problem in our case. But of course this depends on the local conditions. Turning down the brightness and switching off the monitors when not in use is for sure not a mistake. Best regards, Christian Dr. Christian Pies | Head of the Process & Application Lab T | +49 6221 728899-0 E | christian.pies at heidelberg-instruments.com Heidelberg Instruments Mikrotechnik GmbH Mittelgewannweg 27 69123 Heidelberg [Heidelberg Instruments] Handelsregister: Mannheim HRB 333834 Gesch?ftsf?hrer: Konrad Roessler, Steffen Diez Diese E-Mail enth?lt vertrauliche und/oder rechtlich gesch?tzte Informationen. Wenn Sie nicht der richtige Adressat sind oder diese E-Mail irrt?mlich erhalten haben, informieren Sie bitte sofort den Absender und vernichten Sie diese Mail. Das unerlaubte Kopieren sowie die unbefugte Weitergabe dieser Mail sind nicht gestattet. This e-mail may contain confidential and/or privileged information. If you are not the intended recipient (or have received this e-mail in error) please notify the sender immediately and destroy this e-mail. Any unauthorized copying, disclosure, or distribution of the material in this e-mail is strictly forbidden. www.heidelberg-instruments.com Von: labnetwork Im Auftrag von Massey, Travis Gesendet: Montag, 13. Februar 2023 23:40 An: labnetwork at mtl.mit.edu Betreff: [labnetwork] Light filtering in photolithography bays Hi all, LLNL recently changed out its cleanroom lighting to LEDs from fluorescent tubes. A few months later, we found we were having lithography issues - our resist is getting exposed under ambient light within the litho bays within 5-15 minutes. The lights in our litho bays all have ~30 year old plastic diffusers that snap on over the tubes, and these diffusers have a filter film inside. We speculate that the new LEDs may have photobleached the aging photosensitive pigments. Measurements of the light intensity in the cleanroom at 365, 405, and 435 nm indicate that our litho bays have just as much light at these wavelengths than our non-litho bays! -- As an alternative to filtering the light, has anyone found LED cleanroom lighting that does not emit in the >450 nm range? For example, form-compatible LED 'tubes' that only emit at ~580 nm (yellow)? -- How do all of you solve this problem? What product do you use to filter your lights (e.g., tube covers, panel covers, applied films, solid dyed plastics, laser safety products...) -- Have you found that your filters have a finite lifespan? For example, I've heard in previous discussions that window films are only good for 5 years or so before the wavelength-absorbent pigment decomposes and is ineffective. -- To what extent have you found computer monitors' light emissions to be a point of concern? -- Other thoughts/suggestions that we should explore or consider? Thanks, Travis Massey Center for Micro and Nanotechnology (CMNT) Lawrence Livermore National Laboratory -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.png Type: image/png Size: 4947 bytes Desc: image001.png URL: From johnboyle at fast.net Tue Feb 14 11:38:08 2023 From: johnboyle at fast.net (John Boyle) Date: Tue, 14 Feb 2023 11:38:08 -0500 Subject: [labnetwork] O2 8%/CF4 gas cylinder In-Reply-To: References: <6443ece4-9b28-096b-4a78-473408048db3@eng.ucsd.edu> Message-ID: <019201d94092$b871f820$2955e860$@net> Matheson Gas. They have 2 locations in Southern California and numerous locations in US. I think they carry CF4 and they do blending. John Boyle John Boyle Consulting, LLC 4848 Canterbury Drive Emmaus, PA 18049 484-432-7596 Cell 610-965-3208 Office From: labnetwork [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of James Mitchell Sent: Tuesday, February 14, 2023 1:06 AM To: John Tamelier Cc: Fab Network Subject: Re: [labnetwork] O2 8%/CF4 gas cylinder Have you tried Airgas? Jim On Mon, Feb 13, 2023, 5:37 PM John Tamelier wrote: Hello all, We have been having difficulty getting a cylinder of O2 8%/CF4. We placed an order in early November 2022 with Linde and have yet to receive a ship date for the material. We have three orders out, but need the material urgently. The shortest lead time we have found is 9 weeks. Does anyone have any contacts for purchasing this gas or other options for obtaining a cylinder of this gas? Is anyone else experiencing this type of lead time for gas cylinders? Thank you, John ----------------------------------- John Tamelier, Ph.D. Interim Director - Nano3 University of California, San Diego Atkinson Hall, 5th Floor Front Desk, M/C 0436 9500 Gilman Drive La Jolla, California 92093-0436 Phone: (858) 246-2735 Fax: (858) 534-9092 E-mail: jtamelier at ucsd.edu http://nano3.calit2.net ---------------------------------- _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: From Pavel at efgases.com Tue Feb 14 18:37:00 2023 From: Pavel at efgases.com (Pavel Perlov) Date: Tue, 14 Feb 2023 23:37:00 +0000 Subject: [labnetwork] O2 8%/CF4 gas cylinder In-Reply-To: <019201d94092$b871f820$2955e860$@net> References: <6443ece4-9b28-096b-4a78-473408048db3@eng.ucsd.edu> <019201d94092$b871f820$2955e860$@net> Message-ID: Jim, We can help you out with this mix. I'm with Electronic Fluorocarbons www.efgases.com, I cc'd our sales team to assist with your request below. Regards Pavel Sent via the Samsung Galaxy S22+ 5G, an AT&T 5G smartphone -------- Original message -------- From: John Boyle Date: 2/14/23 6:24 PM (GMT-05:00) To: 'James Mitchell' , 'John Tamelier' Cc: 'Fab Network' Subject: Re: [labnetwork] O2 8%/CF4 gas cylinder CAUTION: This email originated from outside of the organization. Do not click links or open attachments unless you recognize the sender and know the content is safe. Matheson Gas. They have 2 locations in Southern California and numerous locations in US. I think they carry CF4 and they do blending. John Boyle John Boyle Consulting, LLC 4848 Canterbury Drive Emmaus, PA 18049 484-432-7596 Cell 610-965-3208 Office From: labnetwork [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of James Mitchell Sent: Tuesday, February 14, 2023 1:06 AM To: John Tamelier Cc: Fab Network Subject: Re: [labnetwork] O2 8%/CF4 gas cylinder Have you tried Airgas? Jim On Mon, Feb 13, 2023, 5:37 PM John Tamelier > wrote: Hello all, We have been having difficulty getting a cylinder of O2 8%/CF4. We placed an order in early November 2022 with Linde and have yet to receive a ship date for the material. We have three orders out, but need the material urgently. The shortest lead time we have found is 9 weeks. Does anyone have any contacts for purchasing this gas or other options for obtaining a cylinder of this gas? Is anyone else experiencing this type of lead time for gas cylinders? Thank you, John ----------------------------------- John Tamelier, Ph.D. Interim Director - Nano3 University of California, San Diego Atkinson Hall, 5th Floor Front Desk, M/C 0436 9500 Gilman Drive La Jolla, California 92093-0436 Phone: (858) 246-2735 Fax: (858) 534-9092 E-mail: jtamelier at ucsd.edu http://nano3.calit2.net ---------------------------------- _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: From savithap at iisc.ac.in Wed Feb 15 07:19:03 2023 From: savithap at iisc.ac.in (Savitha P) Date: Wed, 15 Feb 2023 12:19:03 +0000 Subject: [labnetwork] HF storage Message-ID: Hi! We recently had a couple of occasions where users had mistakenly used Hydrofluoric acid in place of Hydrochloric acid during RCA cleaning. To avoid this, one of the suggestions we have received from our OLSEH is to store HF in a different coloured bottle (currently all acids are stored in white translucent polypropylene bottles which are properly labelled). Accordingly, we had considered storing HF in opaque, brown colour HDPE bottles. However, that was not deemed acceptable by a section of users saying bottles should be translucent so that acid level inside can be judged. Could you please let me know how small quantities of HF are stored in your respective fabs. Is there any regulation governing storage of HF, so that we could implement the same. Thanks and regards, Savitha Dr. Savitha P Chief Operating Officer National Nanofabrication Centre Centre for Nanoscience and Engineering Indian Institute of Science Bangalore - 560012 India. Ph. +91 80 2293 3319 www.cense.iisc.ac.in -------------- next part -------------- An HTML attachment was scrubbed... URL: From hosler0 at purdue.edu Wed Feb 15 10:03:43 2023 From: hosler0 at purdue.edu (Hosler, Richard S) Date: Wed, 15 Feb 2023 15:03:43 +0000 Subject: [labnetwork] HF storage In-Reply-To: References: Message-ID: For our RCA process at Purdue/BNC, we have a small (6ft.) dedicated Kinetics RCA cleaning hood with heated below deck level baths for SC1/SC2 and a smaller unheated bath for HF. I have installed a separate cabinet for the HF and HCl adjacent to the hood. This is marked as for RCA only to a) separate the RCA supplies from the main supply and b) to shorten the distance between the cabinet and hood. The HF is in a smaller semi-opaque white bottle that is easy to determine the level. The HCl bottles are double the size and glass. Each has their own shelf to visually separate them as well. H2O2 goes into a small fridge on the other side of the hood, again to separate the supply and provide quick access. Access to this hood is restricted by requiring a 2 hour training session in which the user sets up, uses, and cleans up the process from start to finish. Prior HF procedures that are outlined originally in the cleanroom access training are reinforced here and the nearest HF exposure kit (6ft behind the user) is pointed out. Basically the strategy is a sort of defense-in-depth of individual mistake-proofing features. To my knowledge we haven't had any exposure incidents stemming from this hood in the ~4 years it's been operational. My catchphrase when working with RCA and Piranha is as follows: "If you ever stop being scared and respectful of these chemicals, you need to stop working with them." Rich Hosler Research Engineer (Thermal/PECVD) Purdue University - Birck Nanotechnology Center hosler0 at purdue.edu From: labnetwork On Behalf Of Savitha P Sent: Wednesday, February 15, 2023 7:19 AM To: labnetwork at mtl.mit.edu Subject: Re: [labnetwork] HF storage ---- External Email: Use caution with attachments, links, or sharing data ---- Hi! We recently had a couple of occasions where users had mistakenly used Hydrofluoric acid in place of Hydrochloric acid during RCA cleaning. To avoid this, one of the suggestions we have received from our OLSEH is to store HF in a different coloured bottle (currently all acids are stored in white translucent polypropylene bottles which are properly labelled). Accordingly, we had considered storing HF in opaque, brown colour HDPE bottles. However, that was not deemed acceptable by a section of users saying bottles should be translucent so that acid level inside can be judged. Could you please let me know how small quantities of HF are stored in your respective fabs. Is there any regulation governing storage of HF, so that we could implement the same. Thanks and regards, Savitha Dr. Savitha P Chief Operating Officer National Nanofabrication Centre Centre for Nanoscience and Engineering Indian Institute of Science Bangalore - 560012 India. Ph. +91 80 2293 3319 www.cense.iisc.ac.in -------------- next part -------------- An HTML attachment was scrubbed... URL: From deolivei at ualberta.ca Wed Feb 15 11:25:19 2023 From: deolivei at ualberta.ca (Gustavo de Oliveira Luiz) Date: Wed, 15 Feb 2023 09:25:19 -0700 Subject: [labnetwork] HF storage In-Reply-To: References: Message-ID: Hi Savitha, Here at the University of Alberta's nanoFAB, we store HF and HCl in completely separate cabinets. In fact, the HF cabinet only contains 49%-HF, BOE, Borofloat etch (HNO, HF mixture) and CaCl2 (for neutralization of these solutions after usage). HCl is kept in a cabinet where other acids and metal etchants may also be found, and it is the responsibility of the user to read labels before pouring chemicals into beakers or other containers. It is understandable that we want to have our labs basically fool-proof, but space always comes at a premium and it may not be an option to have separate cabinets for every single chemical we stock ? again, we still have different acids in the same cabinets, only HF and sulfuric have their unique cabinets. The only option left then is taking administrative action to make sure users follow basic rules. We require all users to be trained and signed off by a staff member before they can process by themselves. Whenever something this serious happens, we remove their user status and set them back to trainee, and they are not allowed to process alone until they regain a staff memeber's confidence that they can do so safely. And, of course, reincidence is followed by more drastic consequences affecting their access to the process or even the clean room. I echo Rich's motto, everyone must be a little bit afraid of these chemicals, just enough so we remember to be careful around them. Best, -- Gustavo de Oliveira Luiz, PhD Applications/Research Specialist nanoFAB, University of Alberta On Wed, Feb 15, 2023 at 9:14 AM Hosler, Richard S wrote: > For our RCA process at Purdue/BNC, we have a small (6ft.) dedicated > Kinetics RCA cleaning hood with heated below deck level baths for SC1/SC2 > and a smaller unheated bath for HF. I have installed a separate cabinet for > the HF and HCl adjacent to the hood. This is marked as for RCA only to a) > separate the RCA supplies from the main supply and b) to shorten the > distance between the cabinet and hood. > > > > The HF is in a smaller semi-opaque white bottle that is easy to determine > the level. The HCl bottles are double the size and glass. Each has their > own shelf to visually separate them as well. H2O2 goes into a small fridge > on the other side of the hood, again to separate the supply and provide > quick access. > > > > Access to this hood is restricted by requiring a 2 hour training session > in which the user sets up, uses, and cleans up the process from start to > finish. Prior HF procedures that are outlined originally in the cleanroom > access training are reinforced here and the nearest HF exposure kit (6ft > behind the user) is pointed out. > > > > Basically the strategy is a sort of defense-in-depth of individual > mistake-proofing features. To my knowledge we haven?t had any exposure > incidents stemming from this hood in the ~4 years it?s been operational. > > > > My catchphrase when working with RCA and Piranha is as follows: ?If you > ever stop being scared and respectful of these chemicals, you need to stop > working with them.? > > > > Rich Hosler > > Research Engineer (Thermal/PECVD) > > Purdue University ? Birck Nanotechnology Center > > hosler0 at purdue.edu > > > > *From:* labnetwork * On Behalf Of *Savitha > P > *Sent:* Wednesday, February 15, 2023 7:19 AM > *To:* labnetwork at mtl.mit.edu > *Subject:* Re: [labnetwork] HF storage > > > > ---- *External Email*: Use caution with attachments, links, or sharing > data ---- > > > > Hi! > > > > We recently had a couple of occasions where users had mistakenly used > Hydrofluoric acid in place of Hydrochloric acid during RCA cleaning. To > avoid this, one of the suggestions we have received from our OLSEH is to > store HF in a different coloured bottle (currently all acids are stored in > white translucent polypropylene bottles which are properly labelled). > Accordingly, we had considered storing HF in opaque, brown colour HDPE > bottles. However, that was not deemed acceptable by a section of users > saying bottles should be translucent so that acid level inside can be > judged. > > > > Could you please let me know how small quantities of HF are stored in your > respective fabs. Is there any regulation governing storage of HF, so that > we could implement the same. > > > > Thanks and regards, > > Savitha > > > > Dr. Savitha P > Chief Operating Officer > > National Nanofabrication Centre > Centre for Nanoscience and Engineering > Indian Institute of Science > Bangalore - 560012 > India. > Ph. +91 80 2293 3319 > www.cense.iisc.ac.in > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork > -------------- next part -------------- An HTML attachment was scrubbed... URL: From mmoneck at andrew.cmu.edu Wed Feb 15 12:14:57 2023 From: mmoneck at andrew.cmu.edu (Matt Moneck) Date: Wed, 15 Feb 2023 12:14:57 -0500 Subject: [labnetwork] HSQ Availability Message-ID: An HTML attachment was scrubbed... URL: From yakimom at sunypoly.edu Wed Feb 15 12:14:56 2023 From: yakimom at sunypoly.edu (Michael Yakimov) Date: Wed, 15 Feb 2023 17:14:56 +0000 Subject: [labnetwork] HF storage In-Reply-To: References: Message-ID: I wonder if your bottles come from the same supplier, look the same, and have chemical name buried into huge DANGER, HAZARD, WARNING symbols. I definitely had some bottles (not pointing fingers, but I have some specific companies in mind) where finding chemical name required literally 2-3 minutes of reading. Can you share bottle pictures to see if that is the issue? If I had a penny every time I was told "check the bottle to make sure..." - so that is what I would emphasize. Instantly readable labeling is a must. Low cost ideas": Secondary "HF" or "HCl" stickers on the bottle in large letters, maybe in different colors. Highlighting chemical name on the label, possibly with different color markers. Switching one of chemicals to a different brand so bottle has different shape and label style... Mike _______________________________________________ Michael Yakimov Research scientist College of Nanoscale Science and Engineering SUNY Polytechnic Institute 253 Fuller rd. Albany NY 12203 Phone: 518-437-8609 lab e-mail: yakimom at sunypoly.edu ________________________________ From: labnetwork on behalf of Savitha P Sent: Wednesday, February 15, 2023 7:19 AM To: labnetwork at mtl.mit.edu Subject: Re: [labnetwork] HF storage Hi! We recently had a couple of occasions where users had mistakenly used Hydrofluoric acid in place of Hydrochloric acid during RCA cleaning. To avoid this, one of the suggestions we have received from our OLSEH is to store HF in a different coloured bottle (currently all acids are stored in white translucent polypropylene bottles which are properly labelled). Accordingly, we had considered storing HF in opaque, brown colour HDPE bottles. However, that was not deemed acceptable by a section of users saying bottles should be translucent so that acid level inside can be judged. Could you please let me know how small quantities of HF are stored in your respective fabs. Is there any regulation governing storage of HF, so that we could implement the same. Thanks and regards, Savitha Dr. Savitha P Chief Operating Officer National Nanofabrication Centre Centre for Nanoscience and Engineering Indian Institute of Science Bangalore - 560012 India. Ph. +91 80 2293 3319 www.cense.iisc.ac.in -------------- next part -------------- An HTML attachment was scrubbed... URL: From hathaway at cns.fas.harvard.edu Wed Feb 15 13:12:34 2023 From: hathaway at cns.fas.harvard.edu (Hathaway, Malcolm R) Date: Wed, 15 Feb 2023 18:12:34 +0000 Subject: [labnetwork] HF storage In-Reply-To: References: Message-ID: Hi Savitha, This is Mac Hathaway, the Nanofabrication Safety officer at Harvard CNS. This is an interesting question from a general perspective, of what and how much to do about the occasional lapse in user behavior which could have bad consequences. In our lab, we do not repackage hazardous chemicals, and rely on proper reading of labels for proper use of chemicals. Occasionally, we have considered repackaging for various purposes (smaller bottle size, usually) but have concluded that the additional risk to staff from extra handling of hazardous materials would not be worth it. (it turns out pouring a full gallon bottle of toxic chemical from one jug to another through a funnel is not as trivial as it may sound.) Different cap colors might help (our acids come with different cap colors, but the particular colors seem relatively random. For reference, HF cap is white). Marking HF bottles in large bold letters with a permanent marker might also help. The most important element in this kind of situation is to teach the users the importance of the concept of Situation Awareness, the idea that every user must Pay Attention to what they are doing ALL THE TIME. I don't want this to sound facetious, as it is literally the most important concept that we teach in our chemical safety training. We hammer away at this concept all the time, for exactly the situation you describe. Quite seriously, no amount of fancy labeling, signage, color codes, etc. will prevent the user lacking situation awareness from making the most monumental errors. It would be interesting to know what the circumstances were surrounding these incidents. Was this late at night? Did they mistake the bottles, or did they mistake the recipe? Where they rookies, or in a hurry? Often, attention to these other aspects of the incidents can pay greater dividends. Specifically with regard to HF, I find that scaring the dickens out of the users during their introduction to the chemicals enhances their retention of the important points. Mac Hathaway Process and Systems Engineer Harvard CNS ________________________________ From: labnetwork on behalf of Savitha P Sent: Wednesday, February 15, 2023 7:19 AM To: labnetwork at mtl.mit.edu Subject: Re: [labnetwork] HF storage Hi! We recently had a couple of occasions where users had mistakenly used Hydrofluoric acid in place of Hydrochloric acid during RCA cleaning. To avoid this, one of the suggestions we have received from our OLSEH is to store HF in a different coloured bottle (currently all acids are stored in white translucent polypropylene bottles which are properly labelled). Accordingly, we had considered storing HF in opaque, brown colour HDPE bottles. However, that was not deemed acceptable by a section of users saying bottles should be translucent so that acid level inside can be judged. Could you please let me know how small quantities of HF are stored in your respective fabs. Is there any regulation governing storage of HF, so that we could implement the same. Thanks and regards, Savitha Dr. Savitha P Chief Operating Officer National Nanofabrication Centre Centre for Nanoscience and Engineering Indian Institute of Science Bangalore - 560012 India. Ph. +91 80 2293 3319 www.cense.iisc.ac.in -------------- next part -------------- An HTML attachment was scrubbed... URL: From price.798 at osu.edu Wed Feb 15 16:03:38 2023 From: price.798 at osu.edu (Price, Aimee) Date: Wed, 15 Feb 2023 21:03:38 +0000 Subject: [labnetwork] HSQ Availability In-Reply-To: References: Message-ID: Hi Matthew, The MAEBL community has dealt with this for many years. We reached out to Dischem (Andy Thompson) and he is providing many of us Ebeamers in the US with our HSQ. He is actually in your backyard in western Pa. He packages up HSQ into the same/similar thicknesses and solids content that Dow did. He also will make small quantities and ships fast. I just bought a 20mL bottle from him which works great for us. Happy to share more if you are interested. It?s not cheap, but it?s inline with the trajectory of the HSQ from Dow before it skyrocketed. You could compare directly with a quote from him. Andy?s contact info is Andy Thompson andy at discheminc.com I do know that some people are still able to get it from Dow distributors, but it can be quite painful. Feel free to call me if you have any questions. Aimee Manager, Nanofabrication The Ohio State University Nanotech West Lab Institute for Materials Research 1381 Kinnear Road Suite 100 Columbus, OH 43212 614-292-2753 Price.798 at osu.edu Nanotech.osu.edu Pronouns: she/her/hers From: labnetwork On Behalf Of Matt Moneck Sent: Wednesday, February 15, 2023 12:15 PM To: labnetwork Subject: [labnetwork] HSQ Availability Hi All, I would like to solicit your feedback on HSQ options for ebeam litho. Our group has traditionally worked with Dow Corning?s XR-1541-002 HSQ, and it has fit our needs for many years. However, as our stock is now reaching a reorder point, Hi All, I would like to solicit your feedback on HSQ options for ebeam litho. Our group has traditionally worked with Dow Corning?s XR-1541-002 HSQ, and it has fit our needs for many years. However, as our stock is now reaching a reorder point, we have reached out to vendors for quotes only to find that pricing has increased to a level we could have never imagined (not to mention lead times on the order of 80 business days). With that in mind, we have been looking at alternative suppliers, including EM Resist (distributed through Integrated Micro Materials) and Allresist (Medusa 82). If anyone has experience working with the EM Resist and Allresist products, we would appreciate any feedback you might you might have on how they compare to the XR-1541. For example, do you see any major differences in resolution, adhesion, dosing, etch resistance, etc.? If you have other alternatives, we would love to hear about them as well, although, we do prefer to buy pre-mixed solutions as opposed to mixing dry HSQ with MIBK ourselves. Best Regards, Matt -- Matthew T. Moneck, Ph.D Executive Director, Claire & John Bertucci Nanotechnology Laboratory Electrical & Computer Engineering | Carnegie Mellon University 5000 Forbes Avenue, Pittsburgh, PA 15213-3890 Phone: 412-268-5430 ece.cmu.edu nanofab.ece.cmu.edu -------------- next part -------------- An HTML attachment was scrubbed... URL: From kevlister at mac.com Wed Feb 15 16:14:51 2023 From: kevlister at mac.com (Kevin Lister) Date: Wed, 15 Feb 2023 13:14:51 -0800 Subject: [labnetwork] HSQ Availability In-Reply-To: References: Message-ID: <87FBFC02-CCC3-432F-BA2A-FFC16F8CAEC0@mac.com> Hi Matt, I would suggest you look at DisChem (https://discheminc.com/ ) as they have a good HSQ substitute with a short lead time. I have no affliction with DisChem. Allresists? Medusa 82 is not really a direct substitute for HSQ. Best Kevin > On Feb 15, 2023, at 9:14 AM, Matt Moneck wrote: > > Hi All, > > I would like to solicit your feedback on HSQ options for ebeam litho. Our group has traditionally worked with Dow Corning?s XR-1541-002 HSQ, and it has fit our needs for many years. However, as our stock is now reaching a reorder point, we have reached out to vendors for quotes only to find that pricing has increased to a level we could have never imagined (not to mention lead times on the order of 80 business days). With that in mind, we have been looking at alternative suppliers, including EM Resist (distributed through Integrated Micro Materials) and Allresist (Medusa 82). > > If anyone has experience working with the EM Resist and Allresist products, we would appreciate any feedback you might you might have on how they compare to the XR-1541. For example, do you see any major differences in resolution, adhesion, dosing, etch resistance, etc.? > > If you have other alternatives, we would love to hear about them as well, although, we do prefer to buy pre-mixed solutions as opposed to mixing dry HSQ with MIBK ourselves. > > Best Regards, > > Matt > > -- > Matthew T. Moneck, Ph.D > Executive Director, Claire & John Bertucci Nanotechnology Laboratory > Electrical & Computer Engineering | Carnegie Mellon University > 5000 Forbes Avenue, Pittsburgh, PA 15213-3890 > Phone: 412-268-5430 > ece.cmu.edu > nanofab.ece.cmu.edu > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: smime.p7s Type: application/pkcs7-signature Size: 1376 bytes Desc: not available URL: From michael.rooks at yale.edu Wed Feb 15 16:34:26 2023 From: michael.rooks at yale.edu (Michael Rooks) Date: Wed, 15 Feb 2023 16:34:26 -0500 Subject: [labnetwork] HSQ Availability In-Reply-To: References: Message-ID: We are now buying HSQ from Dischem , in Pennsylvania. Dischem and EMresist buy HSQ powder from AQM in Canada, and then mix it in MIBK. The resolution and sensitivity are similar to that of HSQ from Dow, but we found that the Dischem HSQ sticks better if we bake it at 120C after spinning. Prices for thin HSQ are high, but manageable. Prices for thick HSQ (like Dow's Fox-16) are prohibitively expensive. We will have to rework our processes when we run out. We store HSQ in liquid nitrogen, so shelf life is not an issue. You can also buy HSQ powder directly from AQM. -------------------------------- Michael Rooks nano.yale.edu On Wed, Feb 15, 2023 at 3:59 PM Matt Moneck wrote: > Hi All, > > > > I would like to solicit your feedback on HSQ options for ebeam litho. Our > group has traditionally worked with Dow Corning?s XR-1541-002 HSQ, and it > has fit our needs for many years. However, as our stock is now reaching a > reorder point, we have reached out to vendors for quotes only to find that > pricing has increased to a level we could have never imagined (not to > mention lead times on the order of 80 business days). With that in mind, > we have been looking at alternative suppliers, including EM Resist > (distributed through Integrated Micro Materials) and Allresist (Medusa > 82). > > > > If anyone has experience working with the EM Resist and Allresist > products, we would appreciate any feedback you might you might have on how > they compare to the XR-1541. For example, do you see any major differences > in resolution, adhesion, dosing, etch resistance, etc.? > > > > If you have other alternatives, we would love to hear about them as well, > although, we do prefer to buy pre-mixed solutions as opposed to mixing dry > HSQ with MIBK ourselves. > > > > Best Regards, > > > Matt > > > > -- > Matthew T. Moneck, Ph.D > Executive Director, Claire & John Bertucci Nanotechnology Laboratory > Electrical & Computer Engineering | Carnegie Mellon University > 5000 Forbes Avenue, Pittsburgh, PA 15213-3890 > Phone: 412-268-5430 > ece.cmu.edu > nanofab.ece.cmu.edu > > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork > -------------- next part -------------- An HTML attachment was scrubbed... URL: From lvchang at Central.UH.EDU Wed Feb 15 17:03:32 2023 From: lvchang at Central.UH.EDU (Chang, Long) Date: Wed, 15 Feb 2023 22:03:32 +0000 Subject: [labnetwork] HDD Best Practice In-Reply-To: References: <590390B9-BEB6-4FB7-AA4F-EC7FE04A1270@central.uh.edu> Message-ID: Hi All, Thanks for the excellent feedback. I?ve learned that: 1) Raid Mirror can fail and requires proper management 2) Clone the drive and use the clone 3) Cloning software works, but not on all systems Spoke with our IT guys and they did not recommend cloning over Raid Mirror because some security stuff doesn?t get cloned. The most likely scenario for our incident is that Drive 0 (master) got corrupted, Drive 1 mirrored and got corrupted, then Drive 0 failed. To keep our IT guys happy, we will do Raid Mirror and clone both drives. Use the cloned drives and store the original. Thanks, Long Chang Technical Director UH Nanofabrication Facility Houston, TX lvchang at central.uh.edu On Feb 10, 2023, at 12:20 PM, Lino Eugene > wrote: Hi Long, Most of our computers are connected to a Synology NAS with RAID 1 configuration in a private network. The Synology software Active to Backup Business does back up the Windows 7 and 10 computers and Macrium Reflect is used for Windows XP. Backups are run weekly and several versions are kept for few months. We also have an offline backup of the NAS. When a hard drive fails, a new one is installed and the computer can be booted from a recovery media and connects to the NAS to recover a backup. This has saved us from long downtimes few times. Best, Lino Eugene, P.Eng., Ph.D., Micro/nanofabrication process engineer Quantum-Nano Fabrication and Characterization Facility Office of Research QNC 1611 University of Waterloo 200 University Avenue West Waterloo, ON, Canada N2L 3G1 Ph: +1 519-888-4567 #37788 Cell: +1 226-929-1685 Websites: https://uwaterloo.ca/quantum-nano-fabrication-and-characterization-facility/ https://fab.qnc.uwaterloo.ca/ -----Message d'origine----- De : labnetwork > De la part de Chang, Long Envoy? : 9 f?vrier 2023 10:15 ? : Fab Network > Objet : [labnetwork] HDD Best Practice Hi All, Yesterday we experienced a HDD failure for our AFM. This event should have been prevented by the Raid Mirror with 2 HDD. Our setup where we just have a single working drive and a spare backup clone has been more successful. We are planning to move away from Raid Mirror. Anyone has expertise/experience here they can share? Thanks, Long Chang Technical Director UH Nanofabrication Facility Houston, TX lvchang at central.uh.edu _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: From tony.olsen at utah.edu Wed Feb 15 19:12:49 2023 From: tony.olsen at utah.edu (Tony L Olsen) Date: Thu, 16 Feb 2023 00:12:49 +0000 Subject: [labnetwork] HF storage In-Reply-To: References: Message-ID: Savitha Unfortunately, I learned a long time ago that no matter how foolproof your system is, there is always a bigger fool. I have also found that complex systems or exceptions tend to make things worse. The simpler the system, the easier to follow and manage. With that said, we developed a system that has worked well for us. We have created several chemical categories. Each category has a two-letter code and description (e.g., GA - General Acid, PR - Photoresist, HF - Hydrofluoric Acid, OA - Oxidizing Acid, SA - Sulfuric Acid, etc.). We apply a chemical code sticker to EVERY chemical container that enters our facility - both chemicals supplied by Staff and those purchased separately by lab members. The sticker is usually placed near the neck of the bottle. The stickers are 1.25 X 1.625" in size. The two-letter code is a 60pt font, while the description is only 11pt. We place a similar, but larger sticker at every location where that chemical can be stored or used. Those stickers are 4 x 3.25", with the code at 200pt and the description at 30pt. For storage and usage, lab members simply have to match the code on the bottle to the code on the wet bench, fume hood, storage cabinet, or storage drawer. If I find a chemical that needs some special handling or segregation I don't create an exception, I create a new category. We decided not to try any color coding for two primary reasons: 1) yellow lights distort colors and 2) color blindness. Prior to this system, I was frequently finding chemicals where they shouldn't be. Since this system was implemented, I have rarely seen a violation (refer to my first statement). This approach may not necessarily address your real concern, but it has eased my worries considerably. In my opinion, using a different color bottle would only be effective if it is the ONLY chemical in the facility of that or similar color bottle. And, you may run in to supply chain issues for non-standard bottles once in a while. We provide both 49% HF and 6:1 BOE in gallon bottles. Tony Olsen Nanofab Cleanroom Supervisor/Process Engineer University of Utah 36 S Wasatch Drive, Suite 2500 Salt Lake City, UT 84112 801-587-0651 www.nanofab.utah.edu From: Savitha P Sent: Wednesday, February 15, 2023 05:19 To: labnetwork at mtl.mit.edu Subject: Re: [labnetwork] HF storage Hi! We recently had a couple of occasions where users had mistakenly used Hydrofluoric acid in place of Hydrochloric acid during RCA cleaning. To avoid this, one of the suggestions we have received from our OLSEH is to store HF in a different coloured bottle (currently all acids are stored in white translucent polypropylene bottles which are properly labelled). Accordingly, we had considered storing HF in opaque, brown colour HDPE bottles. However, that was not deemed acceptable by a section of users saying bottles should be translucent so that acid level inside can be judged. Could you please let me know how small quantities of HF are stored in your respective fabs. Is there any regulation governing storage of HF, so that we could implement the same. Thanks and regards, Savitha Dr. Savitha P Chief Operating Officer National Nanofabrication Centre Centre for Nanoscience and Engineering Indian Institute of Science Bangalore - 560012 India. Ph. +91 80 2293 3319 www.cense.iisc.ac.in -------------- next part -------------- An HTML attachment was scrubbed... URL: From jeanne.guo at rice.edu Thu Feb 16 11:11:52 2023 From: jeanne.guo at rice.edu (Jing Guo) Date: Thu, 16 Feb 2023 10:11:52 -0600 Subject: [labnetwork] HF storage In-Reply-To: References: Message-ID: <25713BEA-37FD-4DC2-A57E-37A08B3A338D@rice.edu> Hi Savitha, This is Jing from Rice U Nanofab. Fortunately we have enough space to store HF or BOE in a separate cabinet. HCl is stored with the other acids in the ?General Acids? cabinet. Also all HF related processes will be performed in the separate HF wet bench which is a independent bench from other benches. For all of our chemicals, we are using the ?Two-letter? labelling system which I learned from Utah NanoFab a long time ago. All chemical bottles, storage cabinets and wet benches have a ?two letter? sticker on. All HF bottles have the ?HF? stickers. HCl or other Acids have the ?GA? stickers (general acid). The stickers have to match to the cabinets and benches when you need to store or use them. For very limited lab or storage space, the simple solution I can think about will be separate secondary containers with different colors, and put different color stickers on the bottle caps. Also put a big color chart beside the storage cabinets and benches to clarify which is which. I totally agree with Tony, no matter how great foolproof your system is, there will still be someone making mistakes. Wish our system can provide some idea to help. Best, Jing --------------------------------------- Jing Guo Ph.D. Research Scientist SEA Cleanroom (SST 017) Rice University Houston, TX jeanne.guo at rice.edu 713-348-8227 > On Feb 15, 2023, at 6:19 AM, Savitha P wrote: > > Hi! > > We recently had a couple of occasions where users had mistakenly used Hydrofluoric acid in place of Hydrochloric acid during RCA cleaning. To avoid this, one of the suggestions we have received from our OLSEH is to store HF in a different coloured bottle (currently all acids are stored in white translucent polypropylene bottles which are properly labelled). Accordingly, we had considered storing HF in opaque, brown colour HDPE bottles. However, that was not deemed acceptable by a section of users saying bottles should be translucent so that acid level inside can be judged. > > Could you please let me know how small quantities of HF are stored in your respective fabs. Is there any regulation governing storage of HF, so that we could implement the same. > > Thanks and regards, > Savitha > > Dr. Savitha P > Chief Operating Officer > National Nanofabrication Centre > Centre for Nanoscience and Engineering > Indian Institute of Science > Bangalore - 560012 > India. > Ph. +91 80 2293 3319 > www.cense.iisc.ac.in > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://urldefense.com/v3/__https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork__;!!BuQPrrmRaQ!jwbzV_ZvSyB1Q5acEoqn2bdk1V1F959zmmHlnmL0yfvnJJ7KjK-9pQMpfGdp8hSSHKhA-Mmu9UAZSVxRNrPjWiD0FvT2dA$ -------------- next part -------------- An HTML attachment was scrubbed... URL: From venables at seas.upenn.edu Thu Feb 16 14:47:38 2023 From: venables at seas.upenn.edu (Travis Venables) Date: Thu, 16 Feb 2023 14:47:38 -0500 Subject: [labnetwork] Explorer 14 gate valve adjustment. Message-ID: Hello, We are currently trying to adjust the cryogate valve intermediate position on our Explorer 14 sputtering chamber. The gate valve in question is a "VAT 64046-PE48-0008/0053 A-1142418" Photo attached. Any information/guidance on how to do this would be greatly appreciated. Tech service is giving us a hard time with getting someone to talk to. Thanks! -- Travis Venables Lead Cleanroom Equipment Engineer Quattrone Nanofabrication Facility University of Pennsylvania P: 215-898-1787 -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: IMG_2315.HEIC Type: image/heif Size: 1447626 bytes Desc: not available URL: From venables at seas.upenn.edu Thu Feb 16 14:48:23 2023 From: venables at seas.upenn.edu (Travis Venables) Date: Thu, 16 Feb 2023 14:48:23 -0500 Subject: [labnetwork] Explorer 14 gate valve adjustment. In-Reply-To: References: Message-ID: A better image On Thu, Feb 16, 2023 at 2:47 PM Travis Venables wrote: > Hello, > > We are currently trying to adjust the cryogate valve intermediate > position on our Explorer 14 sputtering chamber. The gate valve in question > is a "VAT 64046-PE48-0008/0053 A-1142418" Photo attached. > > Any information/guidance on how to do this would be greatly appreciated. > Tech service is giving us a hard time with getting someone to talk to. > Thanks! > -- > Travis Venables > > Lead Cleanroom Equipment Engineer > Quattrone Nanofabrication Facility > University of Pennsylvania > P: 215-898-1787 > -- Travis Venables Lead Cleanroom Equipment Engineer Quattrone Nanofabrication Facility University of Pennsylvania P: 215-898-1787 -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: IMG_2313.HEIC Type: image/heif Size: 1370647 bytes Desc: not available URL: From Paul.Reynolds at glasgow.ac.uk Fri Feb 17 07:46:27 2023 From: Paul.Reynolds at glasgow.ac.uk (Paul Reynolds) Date: Fri, 17 Feb 2023 12:46:27 +0000 Subject: [labnetwork] HSQ Availability In-Reply-To: References: Message-ID: Hi all, I've just bought a final 1L bottle of Fox-16, price has doubled since my last bottle in 2020 (unsure how much of that is a Brexit surcharge...). As we all know, actually communicating with Dow / DuPont is basically impossible. I've bought through Univar and Ellsworth in the past. This time, Ellsworth confirmed what we heard at MAEBL 22: 1 litre is the only available size and the product will be obsolete: The effective date of the product discontinuation will be July 1, 2023, with final orders being accepted through May 1, 2023. So it seems like the end of the road for 16% HSQ solutions we can freeze with ease. It's been a happy 3 years since I followed Mike's instructions on storage in LN2[1][2]. A big dewar cost me ?2.5k or so and it's paid for itself. We freeze in 8ml bottles and put a new one out every week/two in the 4C fridge. This bottle should last us another couple of years. ***update???*** While writing this message I emailed Ellsworth again and they sent the following memo from DuPont: UPDATE 11/11/22 To support and hearing our customers concerns DuPont APT intends to maintain FOx and XR-1541 series supply beyond July 1st, 2023, until 2025.However, potential supply risk still remains. Production hardware safety plus capability assessment to maintain high quality resin manufacturing and the final product is still in progress. The cliff edge on Fox/XR-1541 has been pushed back for now, expect the prices to continue increasing though! Best wishes Paul -- Dr Paul Reynolds Electron Beam Lithography Engineer James Watt Nanofabrication Centre University of Glasgow Working pattern: Mon,Wed,Thu,Fri 0800-1745 [cid:c657e835-5222-4c9e-b925-1a609a1ad6a9] ________________________________ From: labnetwork on behalf of Michael Rooks Sent: 15 February 2023 21:34 To: Matt Moneck Cc: labnetwork Subject: Re: [labnetwork] HSQ Availability We are now buying HSQ from Dischem, in Pennsylvania. Dischem and EMresist buy HSQ powder from AQM in Canada, and then mix it in MIBK. The resolution and sensitivity are similar to that of HSQ from Dow, but we found that the Dischem HSQ sticks better if we bake it at 120C after spinning. Prices for thin HSQ are high, but manageable. Prices for thick HSQ (like Dow's Fox-16) are prohibitively expensive. We will have to rework our processes when we run out. We store HSQ in liquid nitrogen, so shelf life is not an issue. You can also buy HSQ powder directly from AQM. -------------------------------- Michael Rooks nano.yale.edu On Wed, Feb 15, 2023 at 3:59 PM Matt Moneck > wrote: Hi All, I would like to solicit your feedback on HSQ options for ebeam litho. Our group has traditionally worked with Dow Corning?s XR-1541-002 HSQ, and it has fit our needs for many years. However, as our stock is now reaching a reorder point, we have reached out to vendors for quotes only to find that pricing has increased to a level we could have never imagined (not to mention lead times on the order of 80 business days). With that in mind, we have been looking at alternative suppliers, including EM Resist (distributed through Integrated Micro Materials) and Allresist (Medusa 82). If anyone has experience working with the EM Resist and Allresist products, we would appreciate any feedback you might you might have on how they compare to the XR-1541. For example, do you see any major differences in resolution, adhesion, dosing, etch resistance, etc.? If you have other alternatives, we would love to hear about them as well, although, we do prefer to buy pre-mixed solutions as opposed to mixing dry HSQ with MIBK ourselves. Best Regards, Matt -- Matthew T. Moneck, Ph.D Executive Director, Claire & John Bertucci Nanotechnology Laboratory Electrical & Computer Engineering | Carnegie Mellon University 5000 Forbes Avenue, Pittsburgh, PA 15213-3890 Phone: 412-268-5430 ece.cmu.edu nanofab.ece.cmu.edu _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: Outlook-a15f0thd.png Type: image/png Size: 108365 bytes Desc: Outlook-a15f0thd.png URL: From rosendo.bindoy at kaust.edu.sa Fri Feb 17 06:54:23 2023 From: rosendo.bindoy at kaust.edu.sa (Rosendo Bindoy) Date: Fri, 17 Feb 2023 11:54:23 +0000 Subject: [labnetwork] Explorer 14 gate valve adjustment. In-Reply-To: References: Message-ID: Hello Travis, Please see attached manual and refer to page 18-19. Hope this helps. Regards, Rosendo Rosendo Bindoy Equipment Specialist KAUST Nanofabrication Core Lab ________________________________ From: labnetwork on behalf of Travis Venables Sent: Friday, February 17, 2023 12:08 AM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Explorer 14 gate valve adjustment. Hello, We are currently trying to adjust the cryogate valve intermediate position on our Explorer 14 sputtering chamber. The gate valve in question is a "VAT 64046-PE48-0008/0053 A-1142418" Photo attached. Any information/guidance on how to do this would be greatly appreciated. Tech service is giving us a hard time with getting someone to talk to. Thanks! -- Travis Venables Lead Cleanroom Equipment Engineer Quattrone Nanofabrication Facility University of Pennsylvania P: 215-898-1787 -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: 875067ea-0.pdf Type: application/pdf Size: 2514637 bytes Desc: 875067ea-0.pdf URL: From mondol at mit.edu Fri Feb 17 09:16:47 2023 From: mondol at mit.edu (Mark K Mondol) Date: Fri, 17 Feb 2023 14:16:47 +0000 Subject: [labnetwork] HSQ Availability In-Reply-To: References: Message-ID: Thanks for the update on the Dow supply. I have been very happy with both the DisChem and the EM Resist products and they appear to be virtually the same as the Dow product. I would just add that I store my decanted vials in a -60C fridge. I have been using a Stirling UltraCold NEU25NEU. Regards, Mark K Mondol ---------------------------------------------------------------------------------- Mark K Mondol Assistant Director NanoStructures Laboratory And Facility Manager Scanning Electron Beam Lithography Facility Bldg 36 Room 229 mondol at mit.edu office - 617-253-9617 cell - 617-224-8756 ________________________________ From: labnetwork on behalf of Paul Reynolds Sent: Friday, February 17, 2023 7:46 AM To: Michael Rooks ; Matt Moneck Cc: labnetwork Subject: Re: [labnetwork] HSQ Availability Hi all, I've just bought a final 1L bottle of Fox-16, price has doubled since my last bottle in 2020 (unsure how much of that is a Brexit surcharge...). As we all know, actually communicating with Dow / DuPont is basically impossible. I've bought through Univar and Ellsworth in the past. This time, Ellsworth confirmed what we heard at MAEBL 22: 1 litre is the only available size and the product will be obsolete: The effective date of the product discontinuation will be July 1, 2023, with final orders being accepted through May 1, 2023. So it seems like the end of the road for 16% HSQ solutions we can freeze with ease. It's been a happy 3 years since I followed Mike's instructions on storage in LN2[1][2]. A big dewar cost me ?2.5k or so and it's paid for itself. We freeze in 8ml bottles and put a new one out every week/two in the 4C fridge. This bottle should last us another couple of years. ***update???*** While writing this message I emailed Ellsworth again and they sent the following memo from DuPont: UPDATE 11/11/22 To support and hearing our customers concerns DuPont APT intends to maintain FOx and XR-1541 series supply beyond July 1st, 2023, until 2025.However, potential supply risk still remains. Production hardware safety plus capability assessment to maintain high quality resin manufacturing and the final product is still in progress. The cliff edge on Fox/XR-1541 has been pushed back for now, expect the prices to continue increasing though! Best wishes Paul -- Dr Paul Reynolds Electron Beam Lithography Engineer James Watt Nanofabrication Centre University of Glasgow Working pattern: Mon,Wed,Thu,Fri 0800-1745 [cid:c657e835-5222-4c9e-b925-1a609a1ad6a9] ________________________________ From: labnetwork on behalf of Michael Rooks Sent: 15 February 2023 21:34 To: Matt Moneck Cc: labnetwork Subject: Re: [labnetwork] HSQ Availability We are now buying HSQ from Dischem, in Pennsylvania. Dischem and EMresist buy HSQ powder from AQM in Canada, and then mix it in MIBK. The resolution and sensitivity are similar to that of HSQ from Dow, but we found that the Dischem HSQ sticks better if we bake it at 120C after spinning. Prices for thin HSQ are high, but manageable. Prices for thick HSQ (like Dow's Fox-16) are prohibitively expensive. We will have to rework our processes when we run out. We store HSQ in liquid nitrogen, so shelf life is not an issue. You can also buy HSQ powder directly from AQM. -------------------------------- Michael Rooks nano.yale.edu On Wed, Feb 15, 2023 at 3:59 PM Matt Moneck > wrote: Hi All, I would like to solicit your feedback on HSQ options for ebeam litho. Our group has traditionally worked with Dow Corning?s XR-1541-002 HSQ, and it has fit our needs for many years. However, as our stock is now reaching a reorder point, we have reached out to vendors for quotes only to find that pricing has increased to a level we could have never imagined (not to mention lead times on the order of 80 business days). With that in mind, we have been looking at alternative suppliers, including EM Resist (distributed through Integrated Micro Materials) and Allresist (Medusa 82). If anyone has experience working with the EM Resist and Allresist products, we would appreciate any feedback you might you might have on how they compare to the XR-1541. For example, do you see any major differences in resolution, adhesion, dosing, etch resistance, etc.? If you have other alternatives, we would love to hear about them as well, although, we do prefer to buy pre-mixed solutions as opposed to mixing dry HSQ with MIBK ourselves. Best Regards, Matt -- Matthew T. Moneck, Ph.D Executive Director, Claire & John Bertucci Nanotechnology Laboratory Electrical & Computer Engineering | Carnegie Mellon University 5000 Forbes Avenue, Pittsburgh, PA 15213-3890 Phone: 412-268-5430 ece.cmu.edu nanofab.ece.cmu.edu _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: Outlook-a15f0thd.png Type: image/png Size: 108365 bytes Desc: Outlook-a15f0thd.png URL: From ana.n.cohen.ctr at army.mil Fri Feb 17 10:55:44 2023 From: ana.n.cohen.ctr at army.mil (Cohen, Ana N CTR USARMY DEVCOM ARL (USA)) Date: Fri, 17 Feb 2023 15:55:44 +0000 Subject: [labnetwork] Light filtering in photolithography bays Message-ID: Hi Travis, A couple years ago we switched our lithography bay fluorescents (T8) to amber tube LEDS from Super Bright LEDs. The light is more orange than yellow so took a little adjusting (many users first thought it was darker than usual), but visibility has actually been much better since we these bulbs are more stable and we don't end up with dark patches from dead fluorescents. https://www.superbrightleds.com/t8-uv-blocking-led-tube-clean-room-18w-1800- lumens-single-end-dual-end-ballast-bypass-type-b-32w-equivalent-1800k+packam t-30-Pack Our fume hoods are still fluorescent bulbs with filters. I've definitely noticed a degredation over time, but our facilities also have purchased different filters over the years, so I can't provide much comparison to your situation. However, it may be that your LEDS are emitting higher wavelengths than the fluorescents. UV Process has UV filters along with the color tubes which could be useful https://www.uvprocess.com/c3/1790-gamtube-and-supertube-filters.html We don't filter our computer screens because don't notice much light pollution and we don't keep anything photosensitive nearby long-term, but you can definitely turn down the brightness and blue saturation on the monitors. We also have not needed to replace the darker pigment filters on our doors' windows, but similar to the monitors they are placed sufficiently distant from the photoresists/coated substrates. Good luck! Ana -- Ana N. Cohen [she/her/hers] Photolithography Cleanroom Technician Contractor | General Technical Services, LLC US Army Research Laboratory 2800 Powder Mill Road, Adelphi, MD 20783 Tel: 301-394-1527 Email: ana.n.cohen.ctr at army.mil -------------- next part -------------- A non-text attachment was scrubbed... Name: smime.p7s Type: application/pkcs7-signature Size: 6587 bytes Desc: not available URL: From robert.ilic at nist.gov Fri Feb 17 19:00:09 2023 From: robert.ilic at nist.gov (Ilic, Robert (Fed)) Date: Sat, 18 Feb 2023 00:00:09 +0000 Subject: [labnetwork] Process engineer job opening Message-ID: The CNST NanoFab at NIST in Gaithersburg Maryland has a Nanofabrication Process Engineer job opening. Please see the job posting at: https://www.usajobs.gov/job/707487400 Regards, Rob B. Robert Ilic, PhD National Institute of Standards and Technology, Physical Measurement Laboratory Physicist ? Microsystems and Nanotechnology Division Group Leader and Manager ? Center for Nanoscale Science and Technology NanoFab 100 Bureau Drive, Mail Stop 6201, Building 216 Gaithersburg, MD 20899 (301) 975-3712 robert.ilic at nist.gov https://www.nist.gov/cnst http://scholar.google.com/citations?user=FxzEsBQAAAAJ&hl=en -------------- next part -------------- An HTML attachment was scrubbed... URL: From James.Grant at glasgow.ac.uk Sun Feb 19 05:49:15 2023 From: James.Grant at glasgow.ac.uk (James Grant) Date: Sun, 19 Feb 2023 10:49:15 +0000 Subject: [labnetwork] SPTS Rapier Module - Laser/White Light Interferometry In-Reply-To: References: Message-ID: Hello all, Been almost two weeks since I posted this and believe we have come to some conclusions. Many thanks to all those who contributed. Tyndall and EPFL people were especially helpful. 1. We had a lot of back and forth with SPTS. They said that while they could see benefit of laser interferometry for our application, they currently did not offer the hardware to mount laser interferometry endpoint systems to the Rapier. We asked them to quote to so do, but they were unwilling. Fair enough. 2. I don't think the folks at Tyndall will mind me mentioning the challenges they have faced in getting laser interferometry on their SPTS Synapse module working. I understand the issue is spot stability and wobble, thought to be cased by heat from chamber walls - they Synapse walls are heated to 130oC. It's thought this won't be a problem on the Rapier as chamber walls are only heated to 55oC. Let's hope so. 3. We sent pictures and some dimensions of the pre-drilled holes at the top of our Rapier to Intellemetrics, our preferred laser endpoint system vendor. They believe it possible to mount a LEP 500 unit to the tool with not too much trouble. Hope summary information is useful. Cheers, James Dr. James Paul Grant Research Engineer in Plasma Processing Plasma Etch Group james.grant at glasgow.ac.uk [cid:image001.png at 01D9444F.CE818480] www.JWNC.gla.ac.uk [cid:image002.png at 01D9444F.CE818480] LinkedIn.com/company/JWNC [cid:image003.png at 01D9444F.CE818480] @UofG_JWNC From: James Grant Sent: 06 February 2023 14:38 To: labnetwork at mtl.mit.edu Subject: SPTS Rapier Module - Laser/White Light Interferometry Hello, Hoping the community can feedback on their experiences. In 2019 we purchased an SPTS Rapier DSiE tool. As well as doing Bosch processing we also require the tool to do mixed process etching. Quite a few of our users, on our other Si etch tools (OIPT Estrelas and STS Multiplex), use mixed process to etch SOI materials or a-Si on fused silica substrates. These systems have Intellemetrics LEP 410/500 red laser interferometer units for the purpose of end-point detection. When we purchased the Rapier, SPTS did tell us that a laser end-point system could not be used. They encouraged us instead to use the Claritas OES system on the tool, however when we pointed out our typical use scenario (small substrate size, low % open area) they admitted that the Claritas would struggle to differentiate when to end-point. Our preferred laser interferometer vendor, Intellemetrics, say they have sold LEP 500 red laser interferometer units to SPTS but cannot say on what systems they have been used. We have found that EPFL use white light interferometry end-point detection on their Rapier module - have contacted Phillipe Fluckiger separately. My question to the community is has anyone used a laser interferometry system on an SPTS Rapier? If so, could they feedback with their experiences. Apologies I'm asking such an open question! Cheers, James Dr. James Paul Grant Research Engineer in Plasma Processing Plasma Processing Group james.grant at glasgow.ac.uk [cid:image001.png at 01D9444F.CE818480] www.JWNC.gla.ac.uk [cid:image002.png at 01D9444F.CE818480] LinkedIn.com/company/JWNC [cid:image003.png at 01D9444F.CE818480] @UofG_JWNC -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.png Type: image/png Size: 26666 bytes Desc: image001.png URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image002.png Type: image/png Size: 1538 bytes Desc: image002.png URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image003.png Type: image/png Size: 1717 bytes Desc: image003.png URL: From mmoneck at andrew.cmu.edu Mon Feb 20 18:06:16 2023 From: mmoneck at andrew.cmu.edu (Matt Moneck) Date: Mon, 20 Feb 2023 18:06:16 -0500 Subject: [labnetwork] HSQ Availability In-Reply-To: References: , Message-ID: <1F76742D-3596-4A7C-9100-5B826994C7C6@hxcore.ol> An HTML attachment was scrubbed... URL: From gruis at cst.rwth-aachen.de Wed Feb 22 03:43:37 2023 From: gruis at cst.rwth-aachen.de (Jan Gruis) Date: Wed, 22 Feb 2023 09:43:37 +0100 Subject: [labnetwork] Repair of Lehighton 1510C Sheet Resistance Measurement System Message-ID: Hello all! Our Lehighton 1510C sheet resistance measurement system PC (from the 1990's) broke, we lost the hard drive and we need to reinstall and reconfigure the software. Semilab LEI (the former Lehighton Electronics team that manufactured our model) is not very helpful. Even after a few emails back and forth, it is not even clear (to us) if they still have the software for this tool. Is there anyone else or maybe a company that can still repair this tool? The main software Lei1500 for the device is stored on 5 diskettes, but the fifth is corrupt. Does anyone still have these installation disks? But even with these, it seems like additional setup steps are needed to set up the two ISA cards and other hardware. I managed to recover data from the broken hard drive, but I couldn't get the software to control the hardware. Apparently, the PCs were delivered pre-configured, so simply copying the software is not enough. Thank you and greetings from Germany Jan -- Jan Gruis Reinraumingenieur CST - Compound Semiconductor Technology RWTH Aachen University Sommerfeldstra?e 18 52074 Aachen +49 241 80-27748 gruis at cst.rwth-aachen.de www.cst.rwth-aachen.de -------------- next part -------------- A non-text attachment was scrubbed... Name: smime.p7s Type: application/pkcs7-signature Size: 5345 bytes Desc: S/MIME Cryptographic Signature URL: From joseph.losby at ucalgary.ca Wed Feb 22 13:04:14 2023 From: joseph.losby at ucalgary.ca (Joseph Losby) Date: Wed, 22 Feb 2023 18:04:14 +0000 Subject: [labnetwork] Thoughts about sunlight entering cleanroom Message-ID: Hello, in the planning of our new facility the architects have placed windows along a side of the cleanroom (which is fine). Outside of the windows is a hallway which is lined with large windows to the outside. This would allow for sunlight to pass through two sets of windows and into the cleanroom. Although standard windows are adequate for filtering a majority of shorter wavelength UV, some of the longer wavelength UV (UV-A, ~300-400 nm) will still transmit. It is of my belief that it's standard practice not to have sunlight entering the cleanroom (especially a litho area). Are any of your cleanrooms exposed to sunlight (I haven't seen any) and, if so, what type of window filtering do you use? Cheers, Joseph Losby, PhD. Operations Manager, qLab Quantum City, University of Calgary -------------- next part -------------- An HTML attachment was scrubbed... URL: From alfred.renaldo at sjsu.edu Wed Feb 22 16:46:21 2023 From: alfred.renaldo at sjsu.edu (alfred renaldo) Date: Wed, 22 Feb 2023 13:46:21 -0800 Subject: [labnetwork] Thoughts about sunlight entering cleanroom In-Reply-To: References: Message-ID: <7EED2E21-64E4-4EC3-99FB-14A35CC53BC7@hxcore.ol> An HTML attachment was scrubbed... URL: From mpleil at unm.edu Wed Feb 22 17:32:00 2023 From: mpleil at unm.edu (Matthias Pleil) Date: Wed, 22 Feb 2023 22:32:00 +0000 Subject: [labnetwork] Thoughts about sunlight entering cleanroom In-Reply-To: References: Message-ID: You can apply yellow film to the windows to filter the outside light. Having windows does a lot for your folks. Kind Regards, Matthias Pleil, Ph.D. Director Manufacturing Engineering Program UNM MTTC Cleanroom Manager Research Professor & Senior Lecturer III of Mech. Eng - UNM PI - Support Center for Microsystems Education Find great curriculum at: www.scme-support.org (505)272-7157 Join the MNTeSIG Team at www.mntesig.net [cid:4a36f1a5-3462-41d8-8310-2da0d1a30599] ________________________________ From: labnetwork on behalf of Joseph Losby Sent: Wednesday, February 22, 2023 11:04 AM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Thoughts about sunlight entering cleanroom You don't often get email from joseph.losby at ucalgary.ca. Learn why this is important [EXTERNAL] Hello, in the planning of our new facility the architects have placed windows along a side of the cleanroom (which is fine). Outside of the windows is a hallway which is lined with large windows to the outside. This would allow for sunlight to pass through two sets of windows and into the cleanroom. Although standard windows are adequate for filtering a majority of shorter wavelength UV, some of the longer wavelength UV (UV-A, ~300-400 nm) will still transmit. It is of my belief that it's standard practice not to have sunlight entering the cleanroom (especially a litho area). Are any of your cleanrooms exposed to sunlight (I haven't seen any) and, if so, what type of window filtering do you use? Cheers, Joseph Losby, PhD. Operations Manager, qLab Quantum City, University of Calgary -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: Outlook-tdoec0lk.png Type: image/png Size: 13711 bytes Desc: Outlook-tdoec0lk.png URL: From rmorrison at draper.com Wed Feb 22 17:51:12 2023 From: rmorrison at draper.com (Morrison, Richard H., Jr) Date: Wed, 22 Feb 2023 22:51:12 +0000 Subject: [labnetwork] [EXTERNAL] Thoughts about sunlight entering cleanroom In-Reply-To: References: Message-ID: This is Rick Morrison from Draper, when we built our fab we have windows that let in sunlight to a clean corridor, on the Litho doors we used a yellow filter film attenuated the Sunlight. Check this site out, good product Rick Richard Morrison DMTS Draper Laboratory 555 Technology Square Cambridge Ma 02139 Office: 617-258-3420 Cell: 508-930-3461 From: labnetwork On Behalf Of Joseph Losby Sent: Wednesday, February 22, 2023 1:04 PM To: labnetwork at mtl.mit.edu Subject: [EXTERNAL] [labnetwork] Thoughts about sunlight entering cleanroom CAUTION: This email originated from outside of the organization. Do not click links or open attachments unless you recognize the sender and know the content is safe. Hello, in the planning of our new facility the architects have placed windows along a side of the cleanroom (which is fine). Outside of the windows is a hallway which is lined with large windows to the outside. This would allow for sunlight to pass through two sets of windows and into the cleanroom. Although standard windows are adequate for filtering a majority of shorter wavelength UV, some of the longer wavelength UV (UV-A, ~300-400 nm) will still transmit. It is of my belief that it's standard practice not to have sunlight entering the cleanroom (especially a litho area). Are any of your cleanrooms exposed to sunlight (I haven't seen any) and, if so, what type of window filtering do you use? Cheers, Joseph Losby, PhD. Operations Manager, qLab Quantum City, University of Calgary -------------- next part -------------- An HTML attachment was scrubbed... URL: From savithap at iisc.ac.in Thu Feb 23 03:32:01 2023 From: savithap at iisc.ac.in (Savitha P) Date: Thu, 23 Feb 2023 08:32:01 +0000 Subject: [labnetwork] HF storage In-Reply-To: <25713BEA-37FD-4DC2-A57E-37A08B3A338D@rice.edu> References: <25713BEA-37FD-4DC2-A57E-37A08B3A338D@rice.edu> Message-ID: Dear All: Thank you very much for all the suggestions. We have currently incorporated the following as per our space availability. HF has been segregated and kept in a secondary container so that users will not mistake the same with HCl. Size of Labels on the bottles have been increased for HF only, so that users have another mark for identification. As per our protocol, all safety incidents at wet etch leads to an automatic removal of access to the cleanroom. The user will be allowed only after a retraining and retest and will be provided access only after a satisfactory drive in. Regards, Savitha ________________________________ From: Jing Guo Sent: 16 February 2023 21:41 To: Savitha P Cc: labnetwork at mtl.mit.edu Subject: Re: [labnetwork] HF storage External Email Hi Savitha, This is Jing from Rice U Nanofab. Fortunately we have enough space to store HF or BOE in a separate cabinet. HCl is stored with the other acids in the ?General Acids? cabinet. Also all HF related processes will be performed in the separate HF wet bench which is a independent bench from other benches. For all of our chemicals, we are using the ?Two-letter? labelling system which I learned from Utah NanoFab a long time ago. All chemical bottles, storage cabinets and wet benches have a ?two letter? sticker on. All HF bottles have the ?HF? stickers. HCl or other Acids have the ?GA? stickers (general acid). The stickers have to match to the cabinets and benches when you need to store or use them. For very limited lab or storage space, the simple solution I can think about will be separate secondary containers with different colors, and put different color stickers on the bottle caps. Also put a big color chart beside the storage cabinets and benches to clarify which is which. I totally agree with Tony, no matter how great foolproof your system is, there will still be someone making mistakes. Wish our system can provide some idea to help. Best, Jing --------------------------------------- Jing Guo Ph.D. Research Scientist SEA Cleanroom (SST 017) Rice University Houston, TX jeanne.guo at rice.edu 713-348-8227 On Feb 15, 2023, at 6:19 AM, Savitha P > wrote: Hi! We recently had a couple of occasions where users had mistakenly used Hydrofluoric acid in place of Hydrochloric acid during RCA cleaning. To avoid this, one of the suggestions we have received from our OLSEH is to store HF in a different coloured bottle (currently all acids are stored in white translucent polypropylene bottles which are properly labelled). Accordingly, we had considered storing HF in opaque, brown colour HDPE bottles. However, that was not deemed acceptable by a section of users saying bottles should be translucent so that acid level inside can be judged. Could you please let me know how small quantities of HF are stored in your respective fabs. Is there any regulation governing storage of HF, so that we could implement the same. Thanks and regards, Savitha Dr. Savitha P Chief Operating Officer National Nanofabrication Centre Centre for Nanoscience and Engineering Indian Institute of Science Bangalore - 560012 India. Ph. +91 80 2293 3319 www.cense.iisc.ac.in _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://urldefense.com/v3/__https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork__;!!BuQPrrmRaQ!jwbzV_ZvSyB1Q5acEoqn2bdk1V1F959zmmHlnmL0yfvnJJ7KjK-9pQMpfGdp8hSSHKhA-Mmu9UAZSVxRNrPjWiD0FvT2dA$ -------------- next part -------------- An HTML attachment was scrubbed... URL: From yakimom at sunypoly.edu Thu Feb 23 09:07:50 2023 From: yakimom at sunypoly.edu (Michael Yakimov) Date: Thu, 23 Feb 2023 14:07:50 +0000 Subject: [labnetwork] HF storage In-Reply-To: References: <25713BEA-37FD-4DC2-A57E-37A08B3A338D@rice.edu> Message-ID: One issue with "automatic removal of access": punitive measures are not the best if safety is involved. Especially for minor things, where user may have a small event that didn't lead to serious consequences - but the user may share the problem, get some advice, and some things may improve after all. If user is discouraged from that by automatic punishment, then things work worse overall. It's a big philosophic question in general but looks like safety actually benefits from a less punishing approach. Of course, such dilemmas go well beyond cleanroom or chemical safety and experience comes from many industries. I heard a lot of such discussions in aviation safety - and I would call that even more safety-critical business than what we're doing... Thanks Mike _______________________________________________ Michael Yakimov Research scientist College of Nanoscale Science and Engineering SUNY Polytechnic Institute 253 Fuller rd. Albany NY 12203 Phone: 518-437-8609 lab e-mail: yakimom at sunypoly.edu ________________________________ From: labnetwork on behalf of Savitha P Sent: Thursday, February 23, 2023 3:32 AM To: labnetwork at mtl.mit.edu Subject: Re: [labnetwork] HF storage Dear All: Thank you very much for all the suggestions. We have currently incorporated the following as per our space availability. HF has been segregated and kept in a secondary container so that users will not mistake the same with HCl. Size of Labels on the bottles have been increased for HF only, so that users have another mark for identification. As per our protocol, all safety incidents at wet etch leads to an automatic removal of access to the cleanroom. The user will be allowed only after a retraining and retest and will be provided access only after a satisfactory drive in. Regards, Savitha ________________________________ From: Jing Guo Sent: 16 February 2023 21:41 To: Savitha P Cc: labnetwork at mtl.mit.edu Subject: Re: [labnetwork] HF storage External Email Hi Savitha, This is Jing from Rice U Nanofab. Fortunately we have enough space to store HF or BOE in a separate cabinet. HCl is stored with the other acids in the ?General Acids? cabinet. Also all HF related processes will be performed in the separate HF wet bench which is a independent bench from other benches. For all of our chemicals, we are using the ?Two-letter? labelling system which I learned from Utah NanoFab a long time ago. All chemical bottles, storage cabinets and wet benches have a ?two letter? sticker on. All HF bottles have the ?HF? stickers. HCl or other Acids have the ?GA? stickers (general acid). The stickers have to match to the cabinets and benches when you need to store or use them. For very limited lab or storage space, the simple solution I can think about will be separate secondary containers with different colors, and put different color stickers on the bottle caps. Also put a big color chart beside the storage cabinets and benches to clarify which is which. I totally agree with Tony, no matter how great foolproof your system is, there will still be someone making mistakes. Wish our system can provide some idea to help. Best, Jing --------------------------------------- Jing Guo Ph.D. Research Scientist SEA Cleanroom (SST 017) Rice University Houston, TX jeanne.guo at rice.edu 713-348-8227 On Feb 15, 2023, at 6:19 AM, Savitha P > wrote: Hi! We recently had a couple of occasions where users had mistakenly used Hydrofluoric acid in place of Hydrochloric acid during RCA cleaning. To avoid this, one of the suggestions we have received from our OLSEH is to store HF in a different coloured bottle (currently all acids are stored in white translucent polypropylene bottles which are properly labelled). Accordingly, we had considered storing HF in opaque, brown colour HDPE bottles. However, that was not deemed acceptable by a section of users saying bottles should be translucent so that acid level inside can be judged. Could you please let me know how small quantities of HF are stored in your respective fabs. Is there any regulation governing storage of HF, so that we could implement the same. Thanks and regards, Savitha Dr. Savitha P Chief Operating Officer National Nanofabrication Centre Centre for Nanoscience and Engineering Indian Institute of Science Bangalore - 560012 India. Ph. +91 80 2293 3319 www.cense.iisc.ac.in _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://urldefense.com/v3/__https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork__;!!BuQPrrmRaQ!jwbzV_ZvSyB1Q5acEoqn2bdk1V1F959zmmHlnmL0yfvnJJ7KjK-9pQMpfGdp8hSSHKhA-Mmu9UAZSVxRNrPjWiD0FvT2dA$ -------------- next part -------------- An HTML attachment was scrubbed... URL: From hollingshead.19 at osu.edu Thu Feb 23 09:09:01 2023 From: hollingshead.19 at osu.edu (Hollingshead, Dave) Date: Thu, 23 Feb 2023 14:09:01 +0000 Subject: [labnetwork] Thoughts about sunlight entering cleanroom In-Reply-To: References: Message-ID: We have a full bank of windows on the outside wall of our hallway/user storage area. The litho bay segment is protected by yellow film. -Dave Dave Hollingshead Manager, Research Operations The Ohio State University Nanotech West Labs Suite 100, 1381 Kinnear Rd, Columbus, OH 43212 614.292.1355 Office hollingshead.19 at osu.edu / nanotech.osu.edu Pronouns: he/him/his From: labnetwork On Behalf Of Matthias Pleil Sent: Wednesday, February 22, 2023 17:32 To: Joseph Losby ; labnetwork at mtl.mit.edu Subject: Re: [labnetwork] Thoughts about sunlight entering cleanroom You can apply yellow film to the windows to filter the outside light. Having windows does a lot for your folks. Kind Regards, Matthias Pleil, Ph.?D. Director Manufacturing Engineering Program UNM MTTC Cleanroom Manager Research Professor & You can apply yellow film to the windows to filter the outside light. Having windows does a lot for your folks. Kind Regards, Matthias Pleil, Ph.D. Director Manufacturing Engineering Program UNM MTTC Cleanroom Manager Research Professor & Senior Lecturer III of Mech. Eng - UNM PI - Support Center for Microsystems Education Find great curriculum at: www.scme-support.org (505)272-7157 Join the MNTeSIG Team at www.mntesig.net [cid:image001.png at 01D94766.779107D0] ________________________________ From: labnetwork > on behalf of Joseph Losby > Sent: Wednesday, February 22, 2023 11:04 AM To: labnetwork at mtl.mit.edu > Subject: [labnetwork] Thoughts about sunlight entering cleanroom You don't often get email from joseph.losby at ucalgary.ca. Learn why this is important [EXTERNAL] Hello, in the planning of our new facility the architects have placed windows along a side of the cleanroom (which is fine). Outside of the windows is a hallway which is lined with large windows to the outside. This would allow for sunlight to pass through two sets of windows and into the cleanroom. Although standard windows are adequate for filtering a majority of shorter wavelength UV, some of the longer wavelength UV (UV-A, ~300-400 nm) will still transmit. It is of my belief that it's standard practice not to have sunlight entering the cleanroom (especially a litho area). Are any of your cleanrooms exposed to sunlight (I haven't seen any) and, if so, what type of window filtering do you use? Cheers, Joseph Losby, PhD. Operations Manager, qLab Quantum City, University of Calgary -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.png Type: image/png Size: 13711 bytes Desc: image001.png URL: From kckeenan at seas.upenn.edu Thu Feb 23 09:16:36 2023 From: kckeenan at seas.upenn.edu (Kyle Keenan) Date: Thu, 23 Feb 2023 09:16:36 -0500 Subject: [labnetwork] Thoughts about sunlight entering cleanroom In-Reply-To: References: Message-ID: We have the same setup as Draper. Whatever filtering we have in place on the sliding glass doors to the litho bays must be working, because we've had no issues, and the afternoon sun shines very brightly on the front of our building, which is primarily glass. Kyle On Wed, Feb 22, 2023 at 4:33 PM Joseph Losby wrote: > Hello, in the planning of our new facility the architects have placed > windows along a side of the cleanroom (which is fine). Outside of the > windows is a hallway which is lined with large windows to the outside. > This would allow for sunlight to pass through two sets of windows and into > the cleanroom. Although standard windows are adequate for filtering a > majority of shorter wavelength UV, some of the longer wavelength UV (UV-A, > ~300-400 nm) will still transmit. It is of my belief that it's standard > practice not to have sunlight entering the cleanroom (especially a litho > area). > > Are any of your cleanrooms exposed to sunlight (I haven't seen any) and, > if so, what type of window filtering do you use? > > Cheers, > > Joseph Losby, PhD. > Operations Manager, qLab > Quantum City, University of Calgary > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > > https://urldefense.com/v3/__https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork__;!!IBzWLUs!WYiY5TVXZ8JhMAXv0Lf5L1HuBBoUET6R62XM6HDnZnawMnAOBGjQSvMnKOIYHAFdrZJlh_bFWVnAM3RpYEo5OQ_dEvD_5Yk$ > -- Kyle Keenan Senior Manager - Laboratory Operations Quattrone Nanofabrication Facility University of Pennsylvania P: 215-898-7560 F: 215-573-4925 -------------- next part -------------- An HTML attachment was scrubbed... URL: From sandra.malhotra at tamu.edu Thu Feb 23 09:30:21 2023 From: sandra.malhotra at tamu.edu (Malhotra, Sandra Guy) Date: Thu, 23 Feb 2023 14:30:21 +0000 Subject: [labnetwork] Thoughts about sunlight entering cleanroom In-Reply-To: References: Message-ID: At AggieFab at Texas A&M, we have windows along one side of the mobile cleanroom (MCR) units that face a hallway that has windows to the outside. The natural light is quite pleasant. We applied yellow film along the MCR windows for the bay where litho processing is done. We have not experienced any issues. Sandra G. Malhotra, Ph.D. | Technical Lab Manager AggieFab Nanofabrication Facility https://aggiefab.tamu.edu/ Department of Electrical & Computer Engineering, College of Engineering | Texas A&M University 3253 TAMU | College Station, TX 77843 ph: 979.845.3199 | sandra.malhotra at tamu.edu - - - - - - - - - - - - - - - - - - - - - - - - TEXAS A&M UNIVERSITY | FEARLESS on Every Front ________________________________ From: labnetwork on behalf of Joseph Losby Sent: Wednesday, February 22, 2023 12:04 PM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Thoughts about sunlight entering cleanroom Hello, in the planning of our new facility the architects have placed windows along a side of the cleanroom (which is fine). Outside of the windows is a hallway which is lined with large windows to the outside. This would allow for sunlight ZjQcmQRYFpfptBannerStart This Message Is From an External Sender This message came from outside your organization. ZjQcmQRYFpfptBannerEnd Hello, in the planning of our new facility the architects have placed windows along a side of the cleanroom (which is fine). Outside of the windows is a hallway which is lined with large windows to the outside. This would allow for sunlight to pass through two sets of windows and into the cleanroom. Although standard windows are adequate for filtering a majority of shorter wavelength UV, some of the longer wavelength UV (UV-A, ~300-400 nm) will still transmit. It is of my belief that it's standard practice not to have sunlight entering the cleanroom (especially a litho area). Are any of your cleanrooms exposed to sunlight (I haven't seen any) and, if so, what type of window filtering do you use? Cheers, Joseph Losby, PhD. Operations Manager, qLab Quantum City, University of Calgary -------------- next part -------------- An HTML attachment was scrubbed... URL: From repemc at rit.edu Thu Feb 23 09:56:50 2023 From: repemc at rit.edu (Robert Pearson) Date: Thu, 23 Feb 2023 14:56:50 +0000 Subject: [labnetwork] HF storage In-Reply-To: References: <25713BEA-37FD-4DC2-A57E-37A08B3A338D@rice.edu> Message-ID: At Rochester Institute of Technology we have designated HF baths of various water to HF ratios (50:1, 10:1 etc.) that are on certain dedicated benches. The baths are only changed on the normal schedule or when the etch rate monitors indicate a change, or there is some contamination suspected. No students (or faculty) can access the 49% HF used to change the baths, it is only done by staff. Rob Pearson, RIT Microelectronic Engineering. From: labnetwork On Behalf Of Michael Yakimov Sent: Thursday, February 23, 2023 9:08 AM To: Savitha P ; labnetwork at mtl.mit.edu Subject: Re: [labnetwork] HF storage One issue with "automatic removal of access": punitive measures are not the best if safety is involved. Especially for minor things, where user may have a small event that didn't lead to serious consequences - but the user may share the problem, get some advice, and some things may improve after all. If user is discouraged from that by automatic punishment, then things work worse overall. It's a big philosophic question in general but looks like safety actually benefits from a less punishing approach. Of course, such dilemmas go well beyond cleanroom or chemical safety and experience comes from many industries. I heard a lot of such discussions in aviation safety - and I would call that even more safety-critical business than what we're doing... Thanks Mike _______________________________________________ Michael Yakimov Research scientist College of Nanoscale Science and Engineering SUNY Polytechnic Institute 253 Fuller rd. Albany NY 12203 Phone: 518-437-8609 lab e-mail: yakimom at sunypoly.edu ________________________________ From: labnetwork > on behalf of Savitha P > Sent: Thursday, February 23, 2023 3:32 AM To: labnetwork at mtl.mit.edu > Subject: Re: [labnetwork] HF storage Dear All: Thank you very much for all the suggestions. We have currently incorporated the following as per our space availability. HF has been segregated and kept in a secondary container so that users will not mistake the same with HCl. Size of Labels on the bottles have been increased for HF only, so that users have another mark for identification. As per our protocol, all safety incidents at wet etch leads to an automatic removal of access to the cleanroom. The user will be allowed only after a retraining and retest and will be provided access only after a satisfactory drive in. Regards, Savitha ________________________________ From: Jing Guo > Sent: 16 February 2023 21:41 To: Savitha P > Cc: labnetwork at mtl.mit.edu > Subject: Re: [labnetwork] HF storage External Email Hi Savitha, This is Jing from Rice U Nanofab. Fortunately we have enough space to store HF or BOE in a separate cabinet. HCl is stored with the other acids in the 'General Acids' cabinet. Also all HF related processes will be performed in the separate HF wet bench which is a independent bench from other benches. For all of our chemicals, we are using the 'Two-letter' labelling system which I learned from Utah NanoFab a long time ago. All chemical bottles, storage cabinets and wet benches have a 'two letter' sticker on. All HF bottles have the 'HF' stickers. HCl or other Acids have the 'GA' stickers (general acid). The stickers have to match to the cabinets and benches when you need to store or use them. For very limited lab or storage space, the simple solution I can think about will be separate secondary containers with different colors, and put different color stickers on the bottle caps. Also put a big color chart beside the storage cabinets and benches to clarify which is which. I totally agree with Tony, no matter how great foolproof your system is, there will still be someone making mistakes. Wish our system can provide some idea to help. Best, Jing --------------------------------------- Jing Guo Ph.D. Research Scientist SEA Cleanroom (SST 017) Rice University Houston, TX jeanne.guo at rice.edu 713-348-8227 On Feb 15, 2023, at 6:19 AM, Savitha P > wrote: Hi! We recently had a couple of occasions where users had mistakenly used Hydrofluoric acid in place of Hydrochloric acid during RCA cleaning. To avoid this, one of the suggestions we have received from our OLSEH is to store HF in a different coloured bottle (currently all acids are stored in white translucent polypropylene bottles which are properly labelled). Accordingly, we had considered storing HF in opaque, brown colour HDPE bottles. However, that was not deemed acceptable by a section of users saying bottles should be translucent so that acid level inside can be judged. Could you please let me know how small quantities of HF are stored in your respective fabs. Is there any regulation governing storage of HF, so that we could implement the same. Thanks and regards, Savitha Dr. Savitha P Chief Operating Officer National Nanofabrication Centre Centre for Nanoscience and Engineering Indian Institute of Science Bangalore - 560012 India. Ph. +91 80 2293 3319 www.cense.iisc.ac.in _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://urldefense.com/v3/__https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork__;!!BuQPrrmRaQ!jwbzV_ZvSyB1Q5acEoqn2bdk1V1F959zmmHlnmL0yfvnJJ7KjK-9pQMpfGdp8hSSHKhA-Mmu9UAZSVxRNrPjWiD0FvT2dA$ -------------- next part -------------- An HTML attachment was scrubbed... URL: From jpalmer at Princeton.EDU Thu Feb 23 11:10:51 2023 From: jpalmer at Princeton.EDU (Joseph E. Palmer) Date: Thu, 23 Feb 2023 11:10:51 -0500 Subject: [labnetwork] Thoughts about sunlight entering cleanroom In-Reply-To: References: Message-ID: A cleanroom should never have windows to the outside.? While using yellow film to reduce UV from sunlight works, this must periodically be replaced.? Also, the window interferes with the temperature stability of the environment. Your situation may be ok, as you will only need filters. Regards, Joe On 2/22/2023 1:04 PM, Joseph Losby wrote: > Hello, in the planning of our new facility the architects have placed > windows along a side of the cleanroom (which is fine). Outside of the > windows is a hallway which is lined with large windows to the > outside.? This would allow for sunlight to pass through two sets of > windows and into the cleanroom. Although standard windows are adequate > for filtering a majority of shorter wavelength UV, some of the longer > wavelength UV (UV-A, ~300-400 nm) will still transmit. It is of my > belief that it's standard practice not to have sunlight entering the > cleanroom (especially a litho area). > > Are any of your cleanrooms exposed to sunlight (I haven't seen any) > and, if so, what type of window filtering do you use? > > Cheers, > > Joseph Losby, PhD. > Operations Manager, qLab > Quantum City, University of Calgary > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork -- Joseph E. Palmer Chief of Operations for the MNFC PIM, Princeton University Contact: Office: 609-258-4706 Cell: 609-731-8962 -------------- next part -------------- An HTML attachment was scrubbed... URL: From ianharvey at weber.edu Thu Feb 23 11:35:34 2023 From: ianharvey at weber.edu (Ian Harvey) Date: Thu, 23 Feb 2023 09:35:34 -0700 Subject: [labnetwork] on how to impose reasonable consequences for infractions / Fwd: HF storage References: Message-ID: <37A8F8E3-DD6F-4B58-A565-79F93575177B@weber.edu> Dear friends, long-time?. But I do enjoy following the feeds. Savitha, I agree with Michael and offer a suggestion that seemed to work well in past lives: Revoke nightime access for the offender. (Contact the faculty member before you do this. They will likely support you and appreciate the approach.) ?I?m sorry you demonstrated that you cannot work unsupervised. Nightime access is revoked so that we can provide additional training and oversight. You may apply again for off-hours access again in This approach has the following benefits: ? it focuses on a key purpose of the educational / research facility: impactful hands-on experiences, learning in a safe place (they may be fired in industry for this, but here we learn in a safe place) ? nonconfrontational. Actions lead directly to reasonable consequences without scolding or bitter feelings involved. ? it avoids having the faculty member jump down your throat, because it allows the research to continue ? it only inconveniences the student, and this very much. The pain of facing the faculty member is quite severe. All the best to you, ?Ian Ian R. Harvey, Ph.D. Director, Miller Advanced Research Solutions College of Engineering, Applied Science & Technology Weber State University Falcon Hill (HAFB West Gate) https://weber.edu/mars https://www.linkedin.com/in/ian-harvey-967422173/ 633 N Falcon Hills Dr. Clearfield , UT 84015 801/726-7648 (c) ? Begin forwarded message: From: Michael Yakimov Subject: Re: [labnetwork] HF storage Date: February 23, 2023 at 7:07:50 AM MST To: Savitha P , "labnetwork at mtl.mit.edu" One issue with "automatic removal of access": punitive measures are not the best if safety is involved. Especially for minor things, where user may have a small event that didn't lead to serious consequences - but the user may share the problem, get some advice, and some things may improve after all. If user is discouraged from that by automatic punishment, then things work worse overall. It's a big philosophic question in general but looks like safety actually benefits from a less punishing approach. Of course, such dilemmas go well beyond cleanroom or chemical safety and experience comes from many industries. I heard a lot of such discussions in aviation safety - and I would call that even more safety-critical business than what we're doing... Thanks Mike _______________________________________________ Michael Yakimov Research scientist College of Nanoscale Science and Engineering SUNY Polytechnic Institute 253 Fuller rd. Albany NY 12203 Phone: 518-437-8609 lab e-mail: yakimom at sunypoly.edu From: labnetwork > on behalf of Savitha P > Sent: Thursday, February 23, 2023 3:32 AM To: labnetwork at mtl.mit.edu > Subject: Re: [labnetwork] HF storage Dear All: Thank you very much for all the suggestions. We have currently incorporated the following as per our space availability. HF has been segregated and kept in a secondary container so that users will not mistake the same with HCl. Size of Labels on the bottles have been increased for HF only, so that users have another mark for identification. As per our protocol, all safety incidents at wet etch leads to an automatic removal of access to the cleanroom. The user will be allowed only after a retraining and retest and will be provided access only after a satisfactory drive in. Regards, Savitha From: Jing Guo > Sent: 16 February 2023 21:41 To: Savitha P > Cc: labnetwork at mtl.mit.edu > Subject: Re: [labnetwork] HF storage External Email Hi Savitha, This is Jing from Rice U Nanofab. Fortunately we have enough space to store HF or BOE in a separate cabinet. HCl is stored with the other acids in the ?General Acids? cabinet. Also all HF related processes will be performed in the separate HF wet bench which is a independent bench from other benches. For all of our chemicals, we are using the ?Two-letter? labelling system which I learned from Utah NanoFab a long time ago. All chemical bottles, storage cabinets and wet benches have a ?two letter? sticker on. All HF bottles have the ?HF? stickers. HCl or other Acids have the ?GA? stickers (general acid). The stickers have to match to the cabinets and benches when you need to store or use them. For very limited lab or storage space, the simple solution I can think about will be separate secondary containers with different colors, and put different color stickers on the bottle caps. Also put a big color chart beside the storage cabinets and benches to clarify which is which. I totally agree with Tony, no matter how great foolproof your system is, there will still be someone making mistakes. Wish our system can provide some idea to help. Best, Jing --------------------------------------- Jing Guo Ph.D. Research Scientist SEA Cleanroom (SST 017) Rice University Houston, TX jeanne.guo at rice.edu 713-348-8227 > On Feb 15, 2023, at 6:19 AM, Savitha P > wrote: > > Hi! > > We recently had a couple of occasions where users had mistakenly used Hydrofluoric acid in place of Hydrochloric acid during RCA cleaning. To avoid this, one of the suggestions we have received from our OLSEH is to store HF in a different coloured bottle (currently all acids are stored in white translucent polypropylene bottles which are properly labelled). Accordingly, we had considered storing HF in opaque, brown colour HDPE bottles. However, that was not deemed acceptable by a section of users saying bottles should be translucent so that acid level inside can be judged. > > Could you please let me know how small quantities of HF are stored in your respective fabs. Is there any regulation governing storage of HF, so that we could implement the same. > > Thanks and regards, > Savitha > > Dr. Savitha P > Chief Operating Officer > National Nanofabrication Centre > Centre for Nanoscience and Engineering > Indian Institute of Science > Bangalore - 560012 > India. > Ph. +91 80 2293 3319 > www.cense.iisc.ac.in > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://urldefense.com/v3/__https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork__;!!BuQPrrmRaQ!jwbzV_ZvSyB1Q5acEoqn2bdk1V1F959zmmHlnmL0yfvnJJ7KjK-9pQMpfGdp8hSSHKhA-Mmu9UAZSVxRNrPjWiD0FvT2dA$ _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://www.google.com/url?q=https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork&source=gmail-imap&ust=1677768487000000&usg=AOvVaw3cxIZlEUR1BOaG6kJ3DLL4 -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: email2016b.png Type: image/png Size: 7866 bytes Desc: not available URL: From delima1 at llnl.gov Thu Feb 23 18:15:26 2023 From: delima1 at llnl.gov (DeLima, Terri L.) Date: Thu, 23 Feb 2023 23:15:26 +0000 Subject: [labnetwork] HF storage In-Reply-To: References: Message-ID: ,[cid:image001.png at 01D94799.A7941B10][cid:image002.png at 01D94799.A7941B10] We have had this problem in the past due to a lot of the chemical bottles look so similar, and the labeling of the product can be very small. As a "best practice" we label all manufacturer bottles of anything containing HF with a "Need A Buddy" sticker as soon as it arrives to our facility. We also have a "need a buddy" rule when you pour HF, the label also acts as a reminder. Hope this helps. Terri DeLima Operations Manager/Supervisor Center for Micro/Nano & Bioengineering Phone: 925-423-4526 Mobile: 925-315-0731 Email: delima1 at llnl.gov 7000 East Ave L-222 Livermore , CA 94550 Lawrence Livermore National Laboratory (LLNL) From: labnetwork On Behalf Of Savitha P Sent: Wednesday, February 15, 2023 4:19 AM To: labnetwork at mtl.mit.edu Subject: Re: [labnetwork] HF storage Hi! We recently had a couple of occasions where users had mistakenly used Hydrofluoric acid in place of Hydrochloric acid during RCA cleaning. To avoid this, one of the suggestions we have received from our OLSEH is to store HF in a different coloured bottle (currently all acids are stored in white translucent polypropylene bottles which are properly labelled). Accordingly, we had considered storing HF in opaque, brown colour HDPE bottles. However, that was not deemed acceptable by a section of users saying bottles should be translucent so that acid level inside can be judged. Could you please let me know how small quantities of HF are stored in your respective fabs. Is there any regulation governing storage of HF, so that we could implement the same. Thanks and regards, Savitha Dr. Savitha P Chief Operating Officer National Nanofabrication Centre Centre for Nanoscience and Engineering Indian Institute of Science Bangalore - 560012 India. Ph. +91 80 2293 3319 www.cense.iisc.ac.in -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.png Type: image/png Size: 454243 bytes Desc: image001.png URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image002.png Type: image/png Size: 393022 bytes Desc: image002.png URL: From hollingshead.19 at osu.edu Fri Feb 24 17:06:06 2023 From: hollingshead.19 at osu.edu (Hollingshead, Dave) Date: Fri, 24 Feb 2023 22:06:06 +0000 Subject: [labnetwork] third party companies for servicing PlasmaTherm DRIE systems In-Reply-To: References: Message-ID: Fubo et al, We have also recently been having some issues with responsiveness and communication from PlasmaTherm. In this case we are dealing with tool upgrades/installation so we don?t necessarily have the option of going to a third party, but wanted to share our difficulties. I would also be interested in hearing if others have been having similar experiences or have any suggestions. Thanks, -Dave Dave Hollingshead Manager, Research Operations The Ohio State University Nanotech West Labs Suite 100, 1381 Kinnear Rd, Columbus, OH 43212 614.292.1355 Office hollingshead.19 at osu.edu / nanotech.osu.edu Pronouns: he/him/his From: labnetwork On Behalf Of Football Sent: Thursday, January 26, 2023 16:45 To: labnetwork at mtl.mit.edu Subject: [labnetwork] third party companies for servicing PlasmaTherm DRIE systems Dear all, We are having trouble getting PlasmaTherm to respond to our questions and service quote. Wanted to ask if you know of any third party companies that work on PlasmaTherm systems, specifically their DRIE tool? Thanks for any suggestions!Fubo Dear all, We are having trouble getting PlasmaTherm to respond to our questions and service quote. Wanted to ask if you know of any third party companies that work on PlasmaTherm systems, specifically their DRIE tool? Thanks for any suggestions! Fubo Rao, Ph.D., Nanofabrication Cleanroom Manager, Center for Nanoscale Materials, Argonne National Laboratory 9700 S. Cass Ave, Lemont, IL 60439 Phone: 630-252-5708 -------------- next part -------------- An HTML attachment was scrubbed... URL: From robert.macdonald at ge.com Mon Feb 27 09:44:33 2023 From: robert.macdonald at ge.com (Macdonald, Robert (GE Research, US)) Date: Mon, 27 Feb 2023 14:44:33 +0000 Subject: [labnetwork] HF storage- chemical training In-Reply-To: References: <25713BEA-37FD-4DC2-A57E-37A08B3A338D@rice.edu> Message-ID: <9e4e892dbcce4d488403797ad86135c7@ge.com> HI Savitha, I was trained to check the label when I pulled the chemical from storage and before pouring it. I have always trained others this way. I would check your protocols for this small detail as it can help. If we pour the wrong chemical it isn't just HF that can be dangerous. Rob GE Research From: labnetwork On Behalf Of Savitha P Sent: Thursday, February 23, 2023 3:32 AM To: labnetwork at mtl.mit.edu Subject: EXT: Re: [labnetwork] HF storage WARNING: This email originated from outside of GE. Please validate the sender's email address before clicking on links or attachments as they may not be safe. Dear All: Thank you very much for all the suggestions. We have currently incorporated the following as per our space availability. HF has been segregated and kept in a secondary container so that users will not mistake the same with HCl. Size of Labels on the bottles have been increased for HF only, so that users have another mark for identification. As per our protocol, all safety incidents at wet etch leads to an automatic removal of access to the cleanroom. The user will be allowed only after a retraining and retest and will be provided access only after a satisfactory drive in. Regards, Savitha ________________________________ From: Jing Guo > Sent: 16 February 2023 21:41 To: Savitha P > Cc: labnetwork at mtl.mit.edu > Subject: Re: [labnetwork] HF storage External Email Hi Savitha, This is Jing from Rice U Nanofab. Fortunately we have enough space to store HF or BOE in a separate cabinet. HCl is stored with the other acids in the 'General Acids' cabinet. Also all HF related processes will be performed in the separate HF wet bench which is a independent bench from other benches. For all of our chemicals, we are using the 'Two-letter' labelling system which I learned from Utah NanoFab a long time ago. All chemical bottles, storage cabinets and wet benches have a 'two letter' sticker on. All HF bottles have the 'HF' stickers. HCl or other Acids have the 'GA' stickers (general acid). The stickers have to match to the cabinets and benches when you need to store or use them. For very limited lab or storage space, the simple solution I can think about will be separate secondary containers with different colors, and put different color stickers on the bottle caps. Also put a big color chart beside the storage cabinets and benches to clarify which is which. I totally agree with Tony, no matter how great foolproof your system is, there will still be someone making mistakes. Wish our system can provide some idea to help. Best, Jing --------------------------------------- Jing Guo Ph.D. Research Scientist SEA Cleanroom (SST 017) Rice University Houston, TX jeanne.guo at rice.edu 713-348-8227 On Feb 15, 2023, at 6:19 AM, Savitha P > wrote: Hi! We recently had a couple of occasions where users had mistakenly used Hydrofluoric acid in place of Hydrochloric acid during RCA cleaning. To avoid this, one of the suggestions we have received from our OLSEH is to store HF in a different coloured bottle (currently all acids are stored in white translucent polypropylene bottles which are properly labelled). Accordingly, we had considered storing HF in opaque, brown colour HDPE bottles. However, that was not deemed acceptable by a section of users saying bottles should be translucent so that acid level inside can be judged. Could you please let me know how small quantities of HF are stored in your respective fabs. Is there any regulation governing storage of HF, so that we could implement the same. Thanks and regards, Savitha Dr. Savitha P Chief Operating Officer National Nanofabrication Centre Centre for Nanoscience and Engineering Indian Institute of Science Bangalore - 560012 India. Ph. +91 80 2293 3319 www.cense.iisc.ac.in _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://urldefense.com/v3/__https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork__;!!BuQPrrmRaQ!jwbzV_ZvSyB1Q5acEoqn2bdk1V1F959zmmHlnmL0yfvnJJ7KjK-9pQMpfGdp8hSSHKhA-Mmu9UAZSVxRNrPjWiD0FvT2dA$ -------------- next part -------------- An HTML attachment was scrubbed... URL: From jeanne.guo at rice.edu Mon Feb 27 14:15:44 2023 From: jeanne.guo at rice.edu (Jing Guo) Date: Mon, 27 Feb 2023 13:15:44 -0600 Subject: [labnetwork] third party companies for servicing PlasmaTherm DRIE systems In-Reply-To: References: Message-ID: Hi Fubo, I am not sure which email or contact you requested the quote. Usually the response time is longer from PlasmaTherm comparing to other vendors. We only have one PECVD system from them, so we don?t have a third party company for the service. David Harris is the business development manager for university and R&D. His email is David.harris at plasmatherm.com and his cell is #727-370-3947. It should be fine to directly call him. I have some other contacts from plasma-therm tech-support or hardware service. Depending on different requests, I usually will send message to all of them. Please let me know whether you need other contacts information. Best, Jing --------------------------------------- Jing Guo Ph.D. Research Scientist SEA Cleanroom (SST 017) Rice University Houston, TX jeanne.guo at rice.edu 713-348-8227 > On Jan 26, 2023, at 3:45 PM, Football wrote: > > Dear all, > > We are having trouble getting PlasmaTherm to respond to our questions and service quote. Wanted to ask if you know of any third party companies that work on PlasmaTherm systems, specifically their DRIE tool? > > Thanks for any suggestions! > > Fubo Rao, Ph.D., > Nanofabrication Cleanroom Manager, > Center for Nanoscale Materials, > Argonne National Laboratory > 9700 S. Cass Ave, Lemont, IL 60439 > Phone: 630-252-5708 > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://urldefense.com/v3/__https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork__;!!BuQPrrmRaQ!nr9gEfF_LOC45c28UD1lJn-F8Rf9xVboLB2Px_P67EiCS6CRWdO99D57RvLBZwBgawSPE0Z1O_p7npg5YM55D-EDfpqDu78$ -------------- next part -------------- An HTML attachment was scrubbed... URL: From yaofootball at gmail.com Mon Feb 27 15:00:33 2023 From: yaofootball at gmail.com (Football) Date: Mon, 27 Feb 2023 14:00:33 -0600 Subject: [labnetwork] third party companies for servicing PlasmaTherm DRIE systems In-Reply-To: References: Message-ID: Thank you all very much for providing the very helpful information!! Thanks, *Fubo Rao, Ph.D.,* *Nanofabrication Cleanroom Manager,* *Center for Nanoscale Materials,* *Argonne National Laboratory* *9700 S. Cass Ave, Lemont, IL 60439* *Email: **frao at anl.gov* *Phone: 630-252-5708* On Mon, Feb 27, 2023 at 1:15 PM Jing Guo wrote: > Hi Fubo, > > I am not sure which email or contact you requested the quote. Usually the > response time is longer from PlasmaTherm comparing to other vendors. We > only have one PECVD system from them, so we don?t have a third party > company for the service. > > David Harris is the business development manager for university and R&D. > His email is David.harris at plasmatherm.com and his cell is #727-370-3947. > It should be fine to directly call him. > I have some other contacts from plasma-therm tech-support or hardware > service. Depending on different requests, I usually will send message to > all of them. > Please let me know whether you need other contacts information. > > > > Best, > Jing > --------------------------------------- > Jing Guo Ph.D. > Research Scientist > SEA Cleanroom (SST 017) > Rice University > Houston, TX > jeanne.guo at rice.edu > 713-348-8227 > > > > > > > > On Jan 26, 2023, at 3:45 PM, Football wrote: > > Dear all, > > We are having trouble getting PlasmaTherm to respond to our questions and > service quote. Wanted to ask if you know of any third party companies that > work on PlasmaTherm systems, specifically their DRIE tool? > > Thanks for any suggestions! > > *Fubo Rao, Ph.D.,* > *Nanofabrication Cleanroom Manager,* > *Center for Nanoscale Materials,* > *Argonne National Laboratory* > *9700 S. Cass Ave, Lemont, IL 60439* > > *Phone: 630-252-5708* > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > > https://urldefense.com/v3/__https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork__;!!BuQPrrmRaQ!nr9gEfF_LOC45c28UD1lJn-F8Rf9xVboLB2Px_P67EiCS6CRWdO99D57RvLBZwBgawSPE0Z1O_p7npg5YM55D-EDfpqDu78$ > > > -------------- next part -------------- An HTML attachment was scrubbed... URL: From david.harris at plasmatherm.com Mon Feb 27 15:24:20 2023 From: david.harris at plasmatherm.com (Harris, David (Plasma-Therm LLC)) Date: Mon, 27 Feb 2023 20:24:20 +0000 Subject: [labnetwork] Labnetwork and Plasma-Therm In-Reply-To: References: Message-ID: Hello Labnetwork Community, Below are a few points we would like to share with this community: * The Semiconductor industry has been badly disrupted due to supply chain issues causing severe shortage of material. This is further exaggerated by the tight labor market, also a result of the Pandemic. * Multiple Plasma-Therm Resources have been shifted to support short term gaps in our field service scheduling and on-site support. * By copying or sending emails directly to TechSupport at Plasmatherm.com our internal tracking system is invoked, ensuring a formal response and Service Ticket is formally generated. * We have multiple Engineering (Field Service and other) openings / requisitions for which we appreciate any help you can provide with recommended candidates http://plasmatherm.com/careers/index.html Thank you! David Harris Business Development Manager ? Universities and R&D [cid:image001.png at 01D94ABC.24C5CC30] phone +1.215.321.1037 mobile +1.727.370.3947 David.Harris at plasmatherm.com | www.plasmatherm.com 10050 16th St. North | St. Petersburg, FL 33716 USA [Text Description automatically generated] From: labnetwork On Behalf Of Hollingshead, Dave Sent: Friday, February 24, 2023 5:06 PM To: Football ; labnetwork at mtl.mit.edu Subject: Re: [labnetwork] third party companies for servicing PlasmaTherm DRIE systems CAUTION: This email originated from outside of the organization. Do not click links or open attachments unless you recognize the sender and know the content is safe. Fubo et al, We have also recently been having some issues with responsiveness and communication from PlasmaTherm. In this case we are dealing with tool upgrades/installation so we don?t necessarily have the option of going to a third party, but wanted to share our difficulties. I would also be interested in hearing if others have been having similar experiences or have any suggestions. Thanks, -Dave Dave Hollingshead Manager, Research Operations The Ohio State University Nanotech West Labs Suite 100, 1381 Kinnear Rd, Columbus, OH 43212 614.292.1355 Office hollingshead.19 at osu.edu / nanotech.osu.edu Pronouns: he/him/his From: labnetwork > On Behalf Of Football Sent: Thursday, January 26, 2023 16:45 To: labnetwork at mtl.mit.edu Subject: [labnetwork] third party companies for servicing PlasmaTherm DRIE systems Dear all, We are having trouble getting PlasmaTherm to respond to our questions and service quote. Wanted to ask if you know of any third party companies that work on PlasmaTherm systems, specifically their DRIE tool? Thanks for any suggestions!Fubo Dear all, We are having trouble getting PlasmaTherm to respond to our questions and service quote. Wanted to ask if you know of any third party companies that work on PlasmaTherm systems, specifically their DRIE tool? Thanks for any suggestions! Fubo Rao, Ph.D., Nanofabrication Cleanroom Manager, Center for Nanoscale Materials, Argonne National Laboratory 9700 S. Cass Ave, Lemont, IL 60439 Phone: 630-252-5708 -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.png Type: image/png Size: 4851 bytes Desc: image001.png URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image002.png Type: image/png Size: 55044 bytes Desc: image002.png URL: From rmorrison at draper.com Mon Feb 27 17:51:54 2023 From: rmorrison at draper.com (Morrison, Richard H., Jr) Date: Mon, 27 Feb 2023 22:51:54 +0000 Subject: [labnetwork] cleanroom deep clean Message-ID: Hi Folks, We are bringing up a closed cleanroom in St. Petersburg Florida, any folks in that area recommend a good company for performing a deep clean of the cleanroom? Thanks Rick Richard Morrison DMTS Draper Laboratory 555 Technology Square Cambridge Ma 02139 Office: 617-258-3420 Cell: 508-930-3461 -------------- next part -------------- An HTML attachment was scrubbed... URL: From meredith.metzler at nist.gov Tue Feb 28 15:08:07 2023 From: meredith.metzler at nist.gov (Metzler, Meredith G. (Fed)) Date: Tue, 28 Feb 2023 20:08:07 +0000 Subject: [labnetwork] Cryo-piping contractors for LN2 piping sought Message-ID: We are hoping the community can assist us in identifying a vendor who can inspect an existing LN2 delivery piping system in the cleanroom for our cryo-etch tools. We've recently started experiencing issues with a phase separator that keeps building up a large block of ice on the outside of it. We have a 5 Gallon phase separator made by Chart Inc. with one outlet. It feeds liquid nitrogen to several tools in the etch bay of the cleanroom. The LN2 system is house regulated to 40 PSI at the inlet. The unit separates the gaseous N2 from the liquid. The liquid is stored in the 5 gallon volume surrounded by a vacuum jacket to prevent icing. The final LN2 inlet connection to the phase separator, the exhaust connection, and the over pressure relief connection seem to be the biggest points of failure. At these three points we have cold air and moisture condensation that forms an ice ball. Although, we have made many at wrapping and insulating these connection points, we cannot quite get the coverage to eliminate this ice ball. Does the community have any preferred cryo-piping contractors they can recommend to assist with solving this issue and any of our future cryo-piping needs? Thanks! Meredith Metzler National Institute of Standards and Technology, Physical Measurement Laboratory Process Engineer - Center for Nanoscale Science and Technology Facility, NanoFab 100 Bureau Drive, Mail Stop 6201, Building 216 Gaithersburg, MD 20899 301-975-8187 meredith.metzler at nist.gov https://www.nist.gov/cnst -------------- next part -------------- An HTML attachment was scrubbed... URL: From yakimom at sunypoly.edu Tue Feb 28 15:46:43 2023 From: yakimom at sunypoly.edu (Michael Yakimov) Date: Tue, 28 Feb 2023 20:46:43 +0000 Subject: [labnetwork] Cryo-piping contractors for LN2 piping sought In-Reply-To: References: Message-ID: you probably know about vacuum barrier corp., https://vacuumbarrier.com/ I adon't know if they would do a retrofit, but wouldn't hurt asking As for handling those ice balls... First - make sure there are no LN2 leaks, even small one can cause problems. And what kind of wrapping do you use? My experience is that packaging bubble wrap with some layers of foil works best on such connections. Mike _______________________________________________ Michael Yakimov Research scientist College of Nanoscale Science and Engineering SUNY Polytechnic Institute 253 Fuller rd. Albany NY 12203 Phone: 518-437-8609 lab e-mail: yakimom at sunypoly.edu ________________________________ From: labnetwork on behalf of Metzler, Meredith G. (Fed) Sent: Tuesday, February 28, 2023 3:08 PM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Cryo-piping contractors for LN2 piping sought We are hoping the community can assist us in identifying a vendor who can inspect an existing LN2 delivery piping system in the cleanroom for our cryo-etch tools. We've recently started experiencing issues with a phase separator that keeps building up a large block of ice on the outside of it. We have a 5 Gallon phase separator made by Chart Inc. with one outlet. It feeds liquid nitrogen to several tools in the etch bay of the cleanroom. The LN2 system is house regulated to 40 PSI at the inlet. The unit separates the gaseous N2 from the liquid. The liquid is stored in the 5 gallon volume surrounded by a vacuum jacket to prevent icing. The final LN2 inlet connection to the phase separator, the exhaust connection, and the over pressure relief connection seem to be the biggest points of failure. At these three points we have cold air and moisture condensation that forms an ice ball. Although, we have made many at wrapping and insulating these connection points, we cannot quite get the coverage to eliminate this ice ball. Does the community have any preferred cryo-piping contractors they can recommend to assist with solving this issue and any of our future cryo-piping needs? Thanks! Meredith Metzler National Institute of Standards and Technology, Physical Measurement Laboratory Process Engineer ? Center for Nanoscale Science and Technology Facility, NanoFab 100 Bureau Drive, Mail Stop 6201, Building 216 Gaithersburg, MD 20899 301-975-8187 meredith.metzler at nist.gov https://www.nist.gov/cnst -------------- next part -------------- An HTML attachment was scrubbed... URL: From schweig at umich.edu Tue Feb 28 16:36:27 2023 From: schweig at umich.edu (Dennis Schweiger) Date: Tue, 28 Feb 2023 16:36:27 -0500 Subject: [labnetwork] Cryo-piping contractors for LN2 piping sought In-Reply-To: References: Message-ID: Meredith, We have five phase separators in our facility, all supplied by Vacuum Barrier Corporation, and whenever we see the icing issue, it's usually attributed to moisture in the vacuum jacket, and we typically will warm the system up, and purge with dry Nitrogen for 24 hours. that corrects the issue, and then we refill the system with cryogenic material after we've pumped the vacuum jacket down. Also, all of our phase separators have a heater on the "vent" port of the phase separator to keep it from icing up at that discharge location. These guys aren't that far from you. *VACUUM BARRIER CORPORATION* 4 Barten Lane Woburn, MA 01801 USA 781.933.3570 www.vacuumbarrier.com Have you tried doing a Helium leak test on your vacuum jacket to make sure you don't have a leak? Are you sure that your high vacuum pump is operating correctly? We've had issues with the heater in the diffusion pump failing, and we weren't getting a low enough vacuum. Facilities Supervisor University of Michigan/LNF Dennis 734.647.2055 Ofc On Tue, Feb 28, 2023 at 3:53?PM Metzler, Meredith G. (Fed) < meredith.metzler at nist.gov> wrote: > We are hoping the community can assist us in identifying a vendor who can > inspect an existing LN2 delivery piping system in the cleanroom for our > cryo-etch tools. We've recently started experiencing issues with a phase > separator that keeps building up a large block of ice on the outside of it. > > > > We have a 5 Gallon phase separator made by Chart Inc. with one outlet. It > feeds liquid nitrogen to several tools in the etch bay of the cleanroom. > The LN2 system is house regulated to 40 PSI at the inlet. The unit > separates the gaseous N2 from the liquid. The liquid is stored in the 5 > gallon volume surrounded by a vacuum jacket to prevent icing. The final LN2 > inlet connection to the phase separator, the exhaust connection, and the > over pressure relief connection seem to be the biggest points of failure. > At these three points we have cold air and moisture condensation that forms > an ice ball. > > > > Although, we have made many at wrapping and insulating these connection > points, we cannot quite get the coverage to eliminate this ice ball. > > > > Does the community have any preferred cryo-piping contractors they can > recommend to assist with solving this issue and any of our future > cryo-piping needs? > > > > Thanks! > > > > Meredith Metzler > > National Institute of Standards and Technology, Physical Measurement > Laboratory > > Process Engineer ? Center for Nanoscale Science and Technology Facility, > NanoFab > > 100 Bureau Drive, Mail Stop 6201, Building 216 > > Gaithersburg, MD 20899 > > 301-975-8187 > > meredith.metzler at nist.gov > > https://www.nist.gov/cnst > > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork > -------------- next part -------------- An HTML attachment was scrubbed... URL: From deonc69 at illinois.edu Tue Feb 28 16:37:17 2023 From: deonc69 at illinois.edu (Collins, Deon) Date: Tue, 28 Feb 2023 21:37:17 +0000 Subject: [labnetwork] Cryo-piping contractors for LN2 piping sought In-Reply-To: References: Message-ID: Your vacuum jacket has failed. We are going through the same situation with our static LN2 delivery lines. Is your phase separator a Static or Dynamic system? From: labnetwork On Behalf Of Metzler, Meredith G. (Fed) Sent: Tuesday, February 28, 2023 2:08 PM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Cryo-piping contractors for LN2 piping sought We are hoping the community can assist us in identifying a vendor who can inspect an existing LN2 delivery piping system in the cleanroom for our cryo-etch tools. We've recently started experiencing issues with a phase separator that keeps building up a large block of ice on the outside of it. We have a 5 Gallon phase separator made by Chart Inc. with one outlet. It feeds liquid nitrogen to several tools in the etch bay of the cleanroom. The LN2 system is house regulated to 40 PSI at the inlet. The unit separates the gaseous N2 from the liquid. The liquid is stored in the 5 gallon volume surrounded by a vacuum jacket to prevent icing. The final LN2 inlet connection to the phase separator, the exhaust connection, and the over pressure relief connection seem to be the biggest points of failure. At these three points we have cold air and moisture condensation that forms an ice ball. Although, we have made many at wrapping and insulating these connection points, we cannot quite get the coverage to eliminate this ice ball. Does the community have any preferred cryo-piping contractors they can recommend to assist with solving this issue and any of our future cryo-piping needs? Thanks! Meredith Metzler National Institute of Standards and Technology, Physical Measurement Laboratory Process Engineer - Center for Nanoscale Science and Technology Facility, NanoFab 100 Bureau Drive, Mail Stop 6201, Building 216 Gaithersburg, MD 20899 301-975-8187 meredith.metzler at nist.gov https://www.nist.gov/cnst -------------- next part -------------- An HTML attachment was scrubbed... URL: From jeffhawks at verizon.net Tue Feb 28 20:14:17 2023 From: jeffhawks at verizon.net (Jeff Hawks) Date: Tue, 28 Feb 2023 20:14:17 -0500 Subject: [labnetwork] Cryo-piping contractors for LN2 piping sought In-Reply-To: References: Message-ID: <002701d94bdb$2578b910$706a2b30$@net> https://www.appliedenergysystems.com/ From: labnetwork [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Collins, Deon Sent: Tuesday, February 28, 2023 4:37 PM To: Metzler, Meredith G. (Fed); labnetwork at mtl.mit.edu Subject: Re: [labnetwork] Cryo-piping contractors for LN2 piping sought Your vacuum jacket has failed. We are going through the same situation with our static LN2 delivery lines. Is your phase separator a Static or Dynamic system? From: labnetwork On Behalf Of Metzler, Meredith G. (Fed) Sent: Tuesday, February 28, 2023 2:08 PM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Cryo-piping contractors for LN2 piping sought We are hoping the community can assist us in identifying a vendor who can inspect an existing LN2 delivery piping system in the cleanroom for our cryo-etch tools. We've recently started experiencing issues with a phase separator that keeps building up a large block of ice on the outside of it. We have a 5 Gallon phase separator made by Chart Inc with one outlet. It feeds liquid nitrogen to several tools in the etch bay of the cleanroom. The LN2 system is house regulated to 40 PSI at the inlet. The unit separates the gaseous N2 from the liquid. The liquid is stored in the 5 gallon volume surrounded by a vacuum jacket to prevent icing. The final LN2 inlet connection to the phase separator, the exhaust connection, and the over pressure relief connection seem to be the biggest points of failure. At these three points we have cold air and moisture condensation that forms an ice ball. Although, we have made many at wrapping and insulating these connection points, we cannot quite get the coverage to eliminate this ice ball. Does the community have any preferred cryo-piping contractors they can recommend to assist with solving this issue and any of our future cryo-piping needs? Thanks! Meredith Metzler National Institute of Standards and Technology, Physical Measurement Laboratory Process Engineer - Center for Nanoscale Science and Technology Facility, NanoFab 100 Bureau Drive, Mail Stop 6201, Building 216 Gaithersburg, MD 20899 301-975-8187 meredith.metzler at nist.gov https://www.nist.gov/cnst -------------- next part -------------- An HTML attachment was scrubbed... URL: From johnboyle at fast.net Tue Feb 28 21:20:30 2023 From: johnboyle at fast.net (johnboyle) Date: Tue, 28 Feb 2023 21:20:30 -0500 Subject: [labnetwork] Cryo-piping contractors for LN2 piping sought In-Reply-To: Message-ID: Check Acme Cryogenics, Allentown, PA. They make VIP, phase separators, do installation and repairs. Ask for Field ServicesJohn BoyleSent from my Verizon, Samsung Galaxy smartphone -------- Original message --------From: "Collins, Deon" Date: 2/28/23 7:46 PM (GMT-05:00) To: "Metzler, Meredith G. (Fed)" , labnetwork at mtl.mit.edu Subject: Re: [labnetwork] Cryo-piping contractors for LN2 piping sought Your vacuum jacket has failed. We are going through the same situation with our static LN2 delivery lines. ? Is your phase separator a Static or Dynamic system? ? From: labnetwork On Behalf Of Metzler, Meredith G. (Fed) Sent: Tuesday, February 28, 2023 2:08 PM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Cryo-piping contractors for LN2 piping sought ? We are hoping the community can assist us in identifying a vendor who can inspect an existing LN2 delivery piping system in the cleanroom for our cryo-etch tools. We've recently started experiencing issues with a phase separator that keeps building up a large block of ice on the outside of it. ? We have a 5 Gallon phase separator made by Chart Inc. with one outlet. It feeds liquid nitrogen to several tools in the etch bay of the cleanroom. The LN2 system is house regulated to 40 PSI at the inlet. The unit separates the gaseous N2 from the liquid. The liquid is stored in the 5 gallon volume surrounded by a vacuum jacket to prevent icing. The final LN2 inlet connection to the phase separator, the exhaust connection, and the over pressure relief connection seem to be the biggest points of failure. At these three points we have cold air and moisture condensation that forms an ice ball. ? Although, we have made many at wrapping and insulating these connection points, we cannot quite get the coverage to eliminate this ice ball. ? Does the community have any preferred cryo-piping contractors they can recommend to assist with solving this issue and any of our future cryo-piping needs? ? Thanks! ? Meredith Metzler National Institute of Standards and Technology, Physical Measurement Laboratory Process Engineer ? Center for Nanoscale Science and Technology Facility, NanoFab 100 Bureau Drive, Mail Stop 6201, Building 216 Gaithersburg, MD 20899 301-975-8187 meredith.metzler at nist.gov https://www.nist.gov/cnst ? -------------- next part -------------- An HTML attachment was scrubbed... URL: