[labnetwork] Strange "sample memory" with LOR 5B
Martin, Michael
michael.martin at louisville.edu
Mon Mar 27 10:35:32 EDT 2023
Hi Gustavo,
Like others, we've seen such affects as well. I'm guessing that when you remove the sample from Piranha that a DI water rinse shows uniform sheeting and low contact angle over the entire wafer? It might be worth trying some oxygen plasma after NMP/ stripper. We do this often as our NMP bath is reused for quite some time and ends up with several resists solvated in it. Please share anything else you learn about this!
Regards,
Michael
________________________________
From: labnetwork <labnetwork-bounces at mtl.mit.edu> on behalf of Gustavo de Oliveira Luiz <deolivei at ualberta.ca>
Sent: Friday, March 24, 2023 1:37 PM
To: labnetwork at mtl.mit.edu <labnetwork at mtl.mit.edu>
Subject: [labnetwork] Strange "sample memory" with LOR 5B
You don't often get email from deolivei at ualberta.ca. Learn why this is important<https://aka.ms/LearnAboutSenderIdentification>
CAUTION: This email originated from outside of our organization. Do not click links, open attachments, or respond unless you recognize the sender's email address and know the contents are safe.
Hello everyone,
While working on a recipe for LOR 5B/AZ 1512 in our automatic development system, I encountered some intriguing effects when reusing wafers for my tests. This could be a problem for our users when developing their own process, so we'd appreciate it if anyone could help us to understand what is going on.
Below is a picture of a sample right before exposure, taken using our MLA150. The dark/bright features you see are NOT etched on the wafer (these wafers were never etched). The marks are from a previous lithography test. They become apparent after coating the sample with LOR 5B and even more after adding AZ 1512. And I don't see them when coating only with AZ 1512 (I reused wafers for that process development without any issues).
[Si_w_Oxide_LOR5b-AZ1512_AsCoated.png]
And what is more intriguing is that these features affect exposure/development of my test mask. For instance, on a virgin sample I can expose and auto-develop with the same recipe (dose and development time) I use for the manual process. On a reused sample, the reisst stack behaves as if it were underexposed (a dose test made this very obvious).
Here are the steps during my tests:
1. Piranha clean
2. HMDS prime on a YES oven
3. Spin-coat with LOR 5B/AZ 1512 (marks show up on a reused sample)
4. Expose using either a mask aligner or DWL
5. Auto-develop in our Laurell EDC-650 (resist seems underexposed over the marks)
* AZ Developer 1:1 – 90 s
* Rinse (DI water) and dry (N2+spin) – 60-120 s
* MF-319 – 5 s
* Rinse (DI water) and dry (N2+spin) – 60-120 s
6. Strip resist with Remover PG
7. Repeat all steps for every iteration
At first I thought that this could actually be some etching of my Si wafers by MF-319, even though unlikely given the low TMAH concentration (and I'm not sure why that would affect exposure/development). But the sample in the image above has 2 μm thermal oxide, so practically impervious to TMAH. Not to mention that the brightest crossing marks come from testing a recipe where TMAH was not used at all. This must be some strange interaction between LOR 5B and the sample surface, which I'd expect to be practically reset after piranha and HMDS priming.
My search for more information regarding LOR 5B and it's sensitivity to surface conditions has proven fruitless so far. And requiring a brand new sample for every iteration can get expensive quite quickly. We'd appreciate it if you could point us to some references where this was discussed in any form, or if you know of a method to avoid this from happening.
I'm sorry for the long email, and thank you in advance for any comments.
Best regards,
--
Gustavo de Oliveira Luiz, PhD
Applications/Research Specialist
nanoFAB, University of Alberta
+1 (780) 619-1463
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20230327/f22d117a/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: Si_w_Oxide_LOR5b-AZ1512_AsCoated.png
Type: image/png
Size: 1204197 bytes
Desc: Si_w_Oxide_LOR5b-AZ1512_AsCoated.png
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20230327/f22d117a/attachment.png>
More information about the labnetwork
mailing list