From schweig at umich.edu Fri Mar 1 05:36:01 2024 From: schweig at umich.edu (Dennis Schweiger) Date: Fri, 1 Mar 2024 05:36:01 -0500 Subject: [labnetwork] Forming gas: 5% or 10% hydrogen In-Reply-To: References: Message-ID: Mario, good morning. Here at the University of Michigan/LNF we use a 5%H2/Nitrogen mix wherever we need forming gas. We had one application that was using 2%, however that tool has gone off-line and that gas mix was removed. Our EHS group has capped us at 5% for not needing any type of H2 detection in case there is a leak. So far, i've heard of no complaints that we need either more, or less, H2 in the mix, or that we need to switch to another carrier gas. Thanks, Dennis 734.647.2055 Ofc "People can be divided into 3 groups - those that make things happen, those that watch things happen, and those that wonder what happened." Within which group do you belong? On Tue, Feb 27, 2024 at 3:55?PM Beaudoin, Mario wrote: > Dear Network, > > We're setting up a new RTA and will run it with Ar, N2 and forming gas. > We're uncertain about the % level of hydrogen we want in our forming gas. > Can someone recomment the best hydrogen % to use for MOS capacitor > annealing? > > Regards, > > Mario > -- > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork > -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: Mario Beaudoin SBQMI sig 3.jpg Type: image/jpeg Size: 22413 bytes Desc: not available URL: From johnboyle at fast.net Fri Mar 1 09:06:28 2024 From: johnboyle at fast.net (John Boyle) Date: Fri, 1 Mar 2024 09:06:28 -0500 Subject: [labnetwork] Forming gas: 5% or 10% hydrogen In-Reply-To: References: Message-ID: <034f01da6be1$a7e86590$f7b930b0$@net> Most industrial gas suppliers are reluctant to go above 5% H2 in N2 as this is the threshold where the mix can become flammable. If the process can tolerate <=5% H2 in N2 then that is a preferred supply option. Now, the same gas suppliers have no problem supplying 100% H2 but then 100% H2 is shipped, stored and used under a set of codes and standards that are specific to the flammability issues with 100% H2. John John Boyle John Boyle Consulting, LLC 4848 Canterbury Drive Emmaus, PA 18049 484-432-7596 Cell 610-965-3208 Office From: labnetwork [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Dennis Schweiger Sent: Friday, March 01, 2024 5:36 AM To: Beaudoin, Mario Cc: labnetwork at mtl.mit.edu Subject: Re: [labnetwork] Forming gas: 5% or 10% hydrogen Mario, good morning. Here at the University of Michigan/LNF we use a 5%H2/Nitrogen mix wherever we need forming gas. We had one application that was using 2%, however that tool has gone off-line and that gas mix was removed. Our EHS group has capped us at 5% for not needing any type of H2 detection in case there is a leak. So far, i've heard of no complaints that we need either more, or less, H2 in the mix, or that we need to switch to another carrier gas. Thanks, Dennis 734.647.2055 Ofc "People can be divided into 3 groups - those that make things happen, those that watch things happen, and those that wonder what happened." Within which group do you belong? On Tue, Feb 27, 2024 at 3:55?PM Beaudoin, Mario wrote: Dear Network, We're setting up a new RTA and will run it with Ar, N2 and forming gas. We're uncertain about the % level of hydrogen we want in our forming gas. Can someone recomment the best hydrogen % to use for MOS capacitor annealing? Regards, Mario -- _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.jpg Type: image/jpeg Size: 22413 bytes Desc: not available URL: From peng.yang2 at mcgill.ca Fri Mar 1 11:30:25 2024 From: peng.yang2 at mcgill.ca (Peng Yang, Mr) Date: Fri, 1 Mar 2024 16:30:25 +0000 Subject: [labnetwork] Supplier of Clorine gas for RIE Message-ID: Dear network, We are looking for a replacement Clorine gas cylinder for our RIE system. The old cylinder we have is 99.99%. Our regular supplier Linde Canada and Air Liquid Canada doesn't have this purity available. The only thing I can get is 99.9%. I am wondering if any member knows a supplier that can supply high purity Clorine gas cylinder to Canada? Thanks, Peng McGill University Canada -------------- next part -------------- An HTML attachment was scrubbed... URL: From johnboyle at fast.net Fri Mar 1 16:13:37 2024 From: johnboyle at fast.net (John Boyle) Date: Fri, 1 Mar 2024 16:13:37 -0500 Subject: [labnetwork] Supplier of Clorine gas for RIE In-Reply-To: References: Message-ID: <044301da6c1d$54fc6750$fef535f0$@net> Matheson provides >99.99% pure chlorine. I think they have a location in/near Erie PA but may not have a plant in Canada MATHESON Erie 1313 Chestnut Street Erie PA 16501 USA Phone: 814-453-5637 Fax: 814-456-5914 John Boyle John Boyle Consulting, LLC 4848 Canterbury Drive Emmaus, PA 18049 484-432-7596 Cell 610-965-3208 Office From: labnetwork [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Peng Yang, Mr Sent: Friday, March 01, 2024 11:30 AM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Supplier of Clorine gas for RIE Dear network, We are looking for a replacement Clorine gas cylinder for our RIE system. The old cylinder we have is 99.99%. Our regular supplier Linde Canada and Air Liquid Canada doesn't have this purity available. The only thing I can get is 99.9%. I am wondering if any member knows a supplier that can supply high purity Clorine gas cylinder to Canada? Thanks, Peng McGill University Canada -------------- next part -------------- An HTML attachment was scrubbed... URL: From rrenzas at unr.edu Fri Mar 1 16:50:18 2024 From: rrenzas at unr.edu (James Renzas) Date: Fri, 1 Mar 2024 21:50:18 +0000 Subject: [labnetwork] New fab questions: Bottle washer, QDR w/o SRD, blanket expose Message-ID: Hi everybody, Three questions for smaller fabs: 1. Do you find a bottle washer to be necessary? We have a bio area with (probably) an autoclave...is that adequate? Not something I've dealt with before. 2. QDR: I have an opportunity to upgrade one hood to have a QDR (acid hood, I'm thinking), but likely can't afford SRD presently. Is it too silly to have people using an N2 gun after the QDR? 3. Blanket exposure tools: any recommendations, are these helpful? We have an old used aligner and a uMLA on order. Doubt we'll be throughput limited. I have another 100 questions too, all the exciting little things like gas cylinder pricing and when I can get away with filtered CDA vs. house N2 vs process (cylinder local) N2, and thoughts on how to set up glassware cleaning station, and recommendations for plastic dishes for HF for 4" wafers...UGIM can't come soon enough! Russ Russ Renzas, Ph.D. Director, Davidson Foundation Cleanroom College of Engineering University of Nevada, Reno -------------- next part -------------- An HTML attachment was scrubbed... URL: From schweig at umich.edu Fri Mar 1 18:04:52 2024 From: schweig at umich.edu (Dennis Schweiger) Date: Fri, 1 Mar 2024 18:04:52 -0500 Subject: [labnetwork] Supplier of Clorine gas for RIE In-Reply-To: References: Message-ID: Peng, When we were looking for Cl2 gas about a year ago, our vendor told us that the material was hard to get, however they could supply it in lecture bottle sizes. It required an adapter to fit into the DISS cylinder connection of our gas cabinet, but that's what we've been running for a couple of months now. It was a 1# fill, purchased through Doug Reid Account Manager CALL MAIN FOR ORDERS MAIN: (313)835-5513 CELL: (313)539-2716 dougreid at metrowelding.com *www.cryogenicgas.com * *www.metrowelding.com * *I do not know if they would be able to ship to Canada though.* Dennis Schweiger University of Michigan/LNF 734.647.2055 Ofc On Fri, Mar 1, 2024 at 3:19?PM Peng Yang, Mr wrote: > Dear network, > > > > We are looking for a replacement Clorine gas cylinder for our RIE system. > The old cylinder we have is 99.99%. Our regular supplier Linde Canada and > Air Liquid Canada doesn?t have this purity available. The only thing I can > get is 99.9%. > > > > I am wondering if any member knows a supplier that can supply high purity > Clorine gas cylinder to Canada? > > > > > > > > Thanks, > > Peng > > > > McGill University > > Canada > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork > -------------- next part -------------- An HTML attachment was scrubbed... URL: From lvchang at Central.UH.EDU Mon Mar 4 13:04:04 2024 From: lvchang at Central.UH.EDU (Chang, Long) Date: Mon, 4 Mar 2024 18:04:04 +0000 Subject: [labnetwork] Precious Metal Reclaim Service Message-ID: Hi All, Who do you use for precious metal reclaiming from spent targets and chamber/shield scraping? Thanks, Long Chang UH Nanofabrication Facility Houston, TX lvchang at central.uh.edu -------------- next part -------------- An HTML attachment was scrubbed... URL: From yb2471 at columbia.edu Mon Mar 4 19:18:15 2024 From: yb2471 at columbia.edu (Youry Borisenkov) Date: Mon, 4 Mar 2024 16:18:15 -0800 Subject: [labnetwork] Precious Metal Reclaim Service In-Reply-To: References: Message-ID: Hi, Here at Columbia CNI we use Kurt J Lesker Best, Youry On Mon, Mar 4, 2024 at 1:13?PM Chang, Long wrote: > Hi All, > > Who do you use for precious metal reclaiming from spent targets and > chamber/shield scraping? > > Thanks, > Long Chang > UH Nanofabrication Facility > Houston, TX > lvchang at central.uh.edu > > > > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork > -------------- next part -------------- An HTML attachment was scrubbed... URL: From cveith at seas.upenn.edu Mon Mar 4 16:26:15 2024 From: cveith at seas.upenn.edu (Charles Veith) Date: Mon, 4 Mar 2024 16:26:15 -0500 Subject: [labnetwork] Precious Metal Reclaim Service In-Reply-To: References: Message-ID: Hi Long: I have been using ACI Alloys. Charles Albert 408-259-7337, Regards, On Mon, Mar 4, 2024 at 4:13?PM Chang, Long wrote: > Hi All, > > Who do you use for precious metal reclaiming from spent targets and > chamber/shield scraping? > > Thanks, > Long Chang > UH Nanofabrication Facility > Houston, TX > lvchang at central.uh.edu > > > > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > > https://urldefense.com/v3/__https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork__;!!IBzWLUs!SSw4GJA29lXK12fs9-Q9OW6d9mWcb8MH-bWglbWwnewGyFLKZB4sCyo9SC3JU6oYgq2pUdwqFdkB6Uvoc9KN5KDxTw$ > -------------- next part -------------- An HTML attachment was scrubbed... URL: From rrand at ucdavis.edu Tue Mar 5 09:07:17 2024 From: rrand at ucdavis.edu (Ryan R Anderson) Date: Tue, 5 Mar 2024 14:07:17 +0000 Subject: [labnetwork] Precious Metal Reclaim Service In-Reply-To: References: Message-ID: Hi all, We use David H Fell & Company Ryan Anderson Manager, Center for Nano-Micro Manufacturing 1209 Kemper Hall University of California Davis One Shields Avenue Davis, CA 95616 Phone: 530-601-3943 Email: rrand at ucdavis.edu https://cnm2.ucdavis.edu/ ________________________________ From: labnetwork on behalf of Charles Veith Sent: Monday, March 4, 2024 1:26 PM To: Chang, Long Cc: Fab Network Subject: Re: [labnetwork] Precious Metal Reclaim Service Hi Long: I have been using ACI Alloys. Charles Albert 408-259-7337, Regards, On Mon, Mar 4, 2024 at 4:13?PM Chang, Long > wrote: Hi All, Who do you use for precious metal reclaiming from spent targets and chamber/shield scraping? Thanks, Long Chang UH Nanofabrication Facility Houston, TX lvchang at central.uh.edu _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://urldefense.com/v3/__https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork__;!!IBzWLUs!SSw4GJA29lXK12fs9-Q9OW6d9mWcb8MH-bWglbWwnewGyFLKZB4sCyo9SC3JU6oYgq2pUdwqFdkB6Uvoc9KN5KDxTw$ -------------- next part -------------- An HTML attachment was scrubbed... URL: From lvchang at Central.UH.EDU Tue Mar 5 16:32:24 2024 From: lvchang at Central.UH.EDU (Chang, Long) Date: Tue, 5 Mar 2024 21:32:24 +0000 Subject: [labnetwork] Precious Metal Reclaim Service In-Reply-To: References: Message-ID: Thanks for the recommendations everyone. Long Chang UH Nanofabrication Facility Houston, TX lvchang at central.uh.edu On Mar 5, 2024, at 8:07 AM, Ryan R Anderson > wrote: Hi all, We use David H Fell & Company Ryan Anderson Manager, Center for Nano-Micro Manufacturing 1209 Kemper Hall University of California Davis One Shields Avenue Davis, CA 95616 Phone: 530-601-3943 Email: rrand at ucdavis.edu https://cnm2.ucdavis.edu/ ________________________________ From: labnetwork > on behalf of Charles Veith > Sent: Monday, March 4, 2024 1:26 PM To: Chang, Long > Cc: Fab Network > Subject: Re: [labnetwork] Precious Metal Reclaim Service Hi Long: I have been using ACI Alloys. Charles Albert 408-259-7337, Regards, On Mon, Mar 4, 2024 at 4:13?PM Chang, Long > wrote: Hi All, Who do you use for precious metal reclaiming from spent targets and chamber/shield scraping? Thanks, Long Chang UH Nanofabrication Facility Houston, TX lvchang at central.uh.edu _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://urldefense.com/v3/__https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork__;!!IBzWLUs!SSw4GJA29lXK12fs9-Q9OW6d9mWcb8MH-bWglbWwnewGyFLKZB4sCyo9SC3JU6oYgq2pUdwqFdkB6Uvoc9KN5KDxTw$ -------------- next part -------------- An HTML attachment was scrubbed... URL: From ben.w.schmidt at Vanderbilt.Edu Tue Mar 5 16:54:37 2024 From: ben.w.schmidt at Vanderbilt.Edu (Schmidt, Benjamin Willis) Date: Tue, 5 Mar 2024 21:54:37 +0000 Subject: [labnetwork] Precious Metal Reclaim Service In-Reply-To: References: Message-ID: Hi Long, We have used Reldan the last couple years. https://reldan.com/ Ben Schmidt, Ph.D. Cleanroom Manager Vanderbilt Institute of Nanoscale Science and Engineering Vanderbilt University Nashville, TN Ben.schmidt at vanderbilt.edu 615.875.9294 From: labnetwork On Behalf Of Ryan R Anderson Sent: Tuesday, March 5, 2024 8:07 AM To: Charles Veith ; Chang, Long Cc: Fab Network Subject: Re: [labnetwork] Precious Metal Reclaim Service You don't often get email from rrand at ucdavis.edu. Learn why this is important Hi all, We use David H Fell & Company Ryan Anderson Manager, Center for Nano-Micro Manufacturing 1209 Kemper Hall University of California Davis One Shields Avenue Davis, CA 95616 Phone: 530-601-3943 Email: rrand at ucdavis.edu https://cnm2.ucdavis.edu/ ________________________________ From: labnetwork > on behalf of Charles Veith > Sent: Monday, March 4, 2024 1:26 PM To: Chang, Long > Cc: Fab Network > Subject: Re: [labnetwork] Precious Metal Reclaim Service Hi Long: I have been using ACI Alloys. Charles Albert 408-259-7337, Regards, On Mon, Mar 4, 2024 at 4:13?PM Chang, Long > wrote: Hi All, Who do you use for precious metal reclaiming from spent targets and chamber/shield scraping? Thanks, Long Chang UH Nanofabrication Facility Houston, TX lvchang at central.uh.edu _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://urldefense.com/v3/__https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork__;!!IBzWLUs!SSw4GJA29lXK12fs9-Q9OW6d9mWcb8MH-bWglbWwnewGyFLKZB4sCyo9SC3JU6oYgq2pUdwqFdkB6Uvoc9KN5KDxTw$ -------------- next part -------------- An HTML attachment was scrubbed... URL: From jerry.hunter at wisc.edu Wed Mar 6 10:18:38 2024 From: jerry.hunter at wisc.edu (Jerry Hunter) Date: Wed, 6 Mar 2024 15:18:38 +0000 Subject: [labnetwork] University of Wisconsin-Madison Nanoscale Fabrication Center Laboratory Manager Position Opening In-Reply-To: References: Message-ID: We have reopened this position listing to a worldwide search. The University of Wisconsin-Madison, Wisconsin Centers for Nanoscale Technology-Nanoscale Fabrication center has an immediate opening for a cleanroom laboratory manager. The center is a 10,000 sf cleanroom facility for nanoscale fabrication that supports a broad range of research by students, staff, faculty and industrial clients with interest in micro and nano fabrication and device development. The position will involve all aspects of Nano Fabrication operations from user training, safety, process development, equipment maintenance & repair, and cleanroom facility management, to budgeting and user and personnel management. The position requires a broad knowledge and experience in nanofabrication cleanrooms, processes, equipment, and process integration. Special consideration will be given to persons with deep understanding and experience in equipment maintenance and repair. This position is now live on the UW Jobs website: https://jobs.wisc.edu/jobs/lab-manager-nfc-madison-wisconsin-united-states-c17cd70f-a88e-4cee-8932-fa0dbc6dcc3b Deadline is March 13, 2024, at 11:55pm CST. -------------- next part -------------- An HTML attachment was scrubbed... URL: From codreanu at udel.edu Wed Mar 6 16:22:29 2024 From: codreanu at udel.edu (Iulian Codreanu) Date: Wed, 6 Mar 2024 16:22:29 -0500 Subject: [labnetwork] Photomask blanks free to go to a new home In-Reply-To: <080B7D9A-2C17-4045-83DB-616F91ECBE8B@udel.edu> References: <080B7D9A-2C17-4045-83DB-616F91ECBE8B@udel.edu> Message-ID: Hello, We bought some mask blanks a long time ago and never used them; pictures attached.? The boxes are still inside the sealed plastic bags. The resist is most likely no good by now but the substrates should be still usable. You can have them if you send us a shipping label. Cheers, Iulian -------------- next part -------------- A non-text attachment was scrubbed... Name: IMG_5776.jpg Type: image/jpeg Size: 1156269 bytes Desc: not available URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: IMG_5777.jpg Type: image/jpeg Size: 2105690 bytes Desc: not available URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: IMG_5778.jpg Type: image/jpeg Size: 1380274 bytes Desc: not available URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: IMG_5779.jpg Type: image/jpeg Size: 1544638 bytes Desc: not available URL: From scholvin at mit.edu Thu Mar 7 12:47:05 2024 From: scholvin at mit.edu (Jorg Scholvin) Date: Thu, 7 Mar 2024 17:47:05 +0000 Subject: [labnetwork] UGIM 2024 - Announcing Working Groups Message-ID: Dear Labnetwork and UGIM Enthusiasts: We're excited that UGIM 2024 is quickly approaching (June 23-26). A few reminders, and two exciting new opportunities to participate: 1. Registration is open (https://mitnano.mit.edu/events/UGIM2024/registration). Early registrations help us plan better, and we offer reduced rates until April 15. 2. Call for papers are still open (https://mitnano.mit.edu/ugim-abstract), deadline is April 15. 3. Sponsorship opportunities for vendors: (https://mitnano.mit.edu/form/vendor-interest-ugim-2024) New this UGIM: Working-Groups We will prototype two working-groups, to explore this concept for future UGIMs. The goal of a working group is (1) to pre-discuss a key topic of broad interest and (2), in doing so, pool in-depth data and information from across many different participant organizations. The resulting group-consensus (or lack thereof) will be presented at UGIM in a 40 min talk, followed by 20 mins of panel discussion. The working group therefore creates a reference point, or even an "industry standard", which lab managers can point towards in their decision making (or when convincing their administration). Working Group #1 - Staffing & Continuity - Led by Mary Tang (Stanford) Abstract: Do you have a succession plan for the critical roles in your organization? Experimental research facilities have a unique set of operational challenges; this brings us together every two years to share experiences, commiserate, and find inspiration. One unique challenge is professional staffing from recruiting and career growth to retention. For UGIM 2024, we propose a working group to frame professional development for technical staff in research fab facilities. We aim to identify key skills and training that would help prepare a recruit or current staffer for a variety of technical or administrative professional roles. With national attention to workforce development, we will also explore opportunities and resources that align with these efforts. Interested in participating? Reach out to Mary at mtang at stanford.edu Working Group #2 - Process Control in Shared Facilities - Led by Brian Thibeault (UCSB) Abstract: The overarching goal of open access fabrication facilities is to provide the systems and expertise to enable success for the science and engineering communities we serve. In order to improve the success rate and outcomes of our users, process controls of various kinds need to be implemented and refined over time. The controls and feedback take various forms including system operating procedures, preventative maintenance, SPC charts, process centering, in-situ measurements, system log file analysis, efficient and high quality issue reporting and remedy, materials restrictions, etc. In our university facilities, we are asked to support a diverse range of materials and processes, using limited sets of equipment that can most often not be dedicated to a restricted set of materials and processes. We are often understaffed and also have to accommodate a wide variety of user skill levels. Improving process controls within this environment in service to our user communities can be very challenging. The process control workgroup will seek to explore these process control practices, effectiveness, and improvements within our greater community in order to help us all better reach our fabrication service goals. Interested in participating? Reach out to Brian at thibeault at ece.ucsb.edu -------------- next part -------------- An HTML attachment was scrubbed... URL: From massey21 at llnl.gov Thu Mar 7 18:22:16 2024 From: massey21 at llnl.gov (Massey, Travis) Date: Thu, 7 Mar 2024 23:22:16 +0000 Subject: [labnetwork] Process engineer position at LLNL Message-ID: Hi Lab Network, Lawrence Livermore National Laboratory has an opening for a staff process engineer in our multi-user nanofabrication facility. More about the position here: https://www.llnl.gov/join-our-team/careers/find-your-job/all/process%20engineer/3743990003439076 Feel free to reach out if you have questions. Best, Travis Massey -------------- next part -------------- An HTML attachment was scrubbed... URL: From furao at ucsd.edu Fri Mar 8 14:42:00 2024 From: furao at ucsd.edu (Fubo Rao) Date: Fri, 8 Mar 2024 11:42:00 -0800 Subject: [labnetwork] Job Posting - Equipment Engineer @ UC San Diego Nanofab Message-ID: Dear Colleagues, The Nano3 Research Facility at UC San Diego has an immediate position opening for an experienced Equipment Engineer to support the maintenance of our Cleanroom equipment. Job Link: https://employment.ucsd.edu/equipment-engineer-3-127916/job/27858123 Besides its premier location, UCSD offers many good benefits, including 13 paid holidays, *Pension retirement plan*, tuition benefits, and many others... Thanks, Fubo Rao, Ph.D., Technical Director, Nano3 Research Facility, Qualcomm Institute/Atkinson Hall, UC San Diego 9500 Gilman Drive #0436 La Jolla, CA 92093-0436 Tel. (858)534-4518 E-mail: furao at ucsd.edu -------------- next part -------------- An HTML attachment was scrubbed... URL: From rrenzas at unr.edu Tue Mar 12 21:35:11 2024 From: rrenzas at unr.edu (Russ Renzas) Date: Wed, 13 Mar 2024 01:35:11 +0000 Subject: [labnetwork] Basic shelf-type plasma etchers? Message-ID: Does anyone have experience with shelf-type etchers? Specifically like Anatech SCE-1414RF? Rectangular Al chamber, can power a shelf for some light anisotropy. Looking for a basic etch/ash solution to complement my fancy ICP RIE ALE. Anatech is one of the few local equipment manufacturers in Nevada. At minimum I'd get their quartz barrel asher but be nice to get just a little more capability. Russ Russ Renzas University of Nevada, Reno -------------- next part -------------- An HTML attachment was scrubbed... URL: From jks7 at buffalo.edu Wed Mar 13 08:28:18 2024 From: jks7 at buffalo.edu (Jeffrey Salzmann) Date: Wed, 13 Mar 2024 12:28:18 +0000 Subject: [labnetwork] Chemical PPE aprons with snap-clips? Message-ID: Good morning! I am looking to get some new PPE in my facility and I'm trying to locate a supplier with chemical PPE aprons that have snap-clips instead of 'behind the back' ties. The usual approach to search engines doesn't seem to work. Thank you in advance!!! Regards, Jeff Jeff Salzmann Assistant Professor of Research Cleanroom Manager, Shared Instrumentation Laboratories School of Engineering and Applied Sciences University at Buffalo 114A Davis Hall Buffalo, NY 14260 -------------- next part -------------- An HTML attachment was scrubbed... URL: From daniel.woodie at princeton.edu Wed Mar 13 16:33:17 2024 From: daniel.woodie at princeton.edu (Dan P. Woodie) Date: Wed, 13 Mar 2024 20:33:17 +0000 Subject: [labnetwork] Chemical PPE aprons with snap-clips? In-Reply-To: References: Message-ID: Jeff, Many years back at another university, I created the buckles for our aprons using basic hand tools. I purchased webbing, buckles, and the tools to put in grommets. Once I created a batch of the needed straps, they could be swapped to new nitrile aprons as they wore out and lasted a long time. Webbing https://www.mcmaster.com/products/webbing/webbing-2~/ Grommets https://www.mcmaster.com/products/grommets/grommets~/?s=metal+grommets Dan Daniel Woodie Director, Micro/Nano Fabrication Center, Princeton Materials Institute From: labnetwork On Behalf Of Jeffrey Salzmann Sent: Wednesday, March 13, 2024 8:28 AM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Chemical PPE aprons with snap-clips? Good morning! I am looking to get some new PPE in my facility and I'm trying to locate a supplier with chemical PPE aprons that have snap-clips instead of 'behind the back' ties. The usual approach to search engines doesn't seem to work. Thank you in advance!!! Regards, Jeff Jeff Salzmann Assistant Professor of Research Cleanroom Manager, Shared Instrumentation Laboratories School of Engineering and Applied Sciences University at Buffalo 114A Davis Hall Buffalo, NY 14260 -------------- next part -------------- An HTML attachment was scrubbed... URL: From pi12 at cornell.edu Thu Mar 14 11:08:35 2024 From: pi12 at cornell.edu (Philip Infante) Date: Thu, 14 Mar 2024 15:08:35 +0000 Subject: [labnetwork] Chemical PPE aprons with snap-clips? In-Reply-To: References: Message-ID: Warning: This message has had one or more attachments removed Warning: (favicon.ico). Warning: Please read the "MTL-Attachment-Warning.txt" attachment(s) for more information. Hi Jeff, I switched to some custom fabricated straps from Strapworks, the squeeze release/slip on buckles we had would constantly break with frequent use. They make custom strap assemblies, I used a Fidlock magnetic release buckle with clip on hooks at the apron end and have not had any issues in over 5 yrs. We can easily swap out old/torn aprons in a couple of minutes. No changing of clasps, hammering grommets ,etc. Cost in ~2019 was about $15/ strap assembly - very reasonable and well worth it in my opinion. Straps, Webbing, Tie-Downs, Buckles & More strapworks.com [favicon.ico] Phil Infante Cornell Nanoscale Facility Cornell University 250 Duffield Hall Ithaca, NY 14853 607-254-4926 On Mar 13, 2024, at 4:33?PM, Dan P. Woodie wrote: Jeff, Many years back at another university, I created the buckles for our aprons using basic hand tools. I purchased webbing, buckles, and the tools to put in grommets. Once I created a batch of the needed straps, they could be swapped to new nitrile aprons as they wore out and lasted a long time. Webbing https://www.mcmaster.com/products/webbing/webbing-2~/ Grommets https://www.mcmaster.com/products/grommets/grommets~/?s=metal+grommets Dan Daniel Woodie Director, Micro/Nano Fabrication Center, Princeton Materials Institute From: labnetwork > On Behalf Of Jeffrey Salzmann Sent: Wednesday, March 13, 2024 8:28 AM To: labnetwork at mtl.mit.edu Subject: [labnetwork] Chemical PPE aprons with snap-clips? Good morning! I am looking to get some new PPE in my facility and I?m trying to locate a supplier with chemical PPE aprons that have snap-clips instead of ?behind the back? ties. The usual approach to search engines doesn?t seem to work. Thank you in advance!!! Regards, Jeff Jeff Salzmann Assistant Professor of Research Cleanroom Manager, Shared Instrumentation Laboratories School of Engineering and Applied Sciences University at Buffalo 114A Davis Hall Buffalo, NY 14260 _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- This is a message from the Microsystems Technology Laboratories e-mail server. The original e-mail attachment "favicon.ico" is on the list of unacceptable attachments for this site and has been replaced by this warning message. If you wish to receive a copy of the original email, please e-mail mail-problems at mtl.mit.edu and include the whole of this message in your request. Below is a detailed summary: Windows Icon files are dangerous in email (favicon.ico) Note to Compute Team: Look on mtl.mit.edu in /usr/local/MailScanner/var/spool/quarantine/20240314 (message 1rkmhB-0003G9-S0). -- Questions/Problems? Send e-mail to mail-problems at mtl.mit.edu From codreanu at udel.edu Thu Mar 14 12:06:47 2024 From: codreanu at udel.edu (Iulian Codreanu) Date: Thu, 14 Mar 2024 12:06:47 -0400 Subject: [labnetwork] Chemical PPE aprons with snap-clips? In-Reply-To: References: Message-ID: Hi Jeff, We replace the ties with snap-clips in a manner similar to the one described by Dan Woodie. Cheers, Iulian iulian Codreanu, Ph.D. Director, Nanofabrication Facility University of Delaware Harker ISE Lab, Room 163 221 Academy Street Newark, DE 19716 302-831-2784 https://udnf.udel.edu On 3/13/2024 8:28 AM, Jeffrey Salzmann wrote: > > Good morning! > > I am looking to get some new PPE in my facility and I?m trying to > locate a supplier with chemical PPE aprons that have snap-clips > instead of ?behind the back? ties. > > The usual approach to search engines doesn?t seem to work. Thank you > in advance!!! > > Regards, > > Jeff > > Jeff Salzmann > > Assistant Professor of Research > > Cleanroom Manager, Shared Instrumentation Laboratories > > School of Engineering and Applied Sciences > > University at Buffalo > > 114A Davis Hall > > Buffalo, NY 14260 > > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: From joseph.losby at ucalgary.ca Thu Mar 14 15:11:58 2024 From: joseph.losby at ucalgary.ca (Joseph Losby) Date: Thu, 14 Mar 2024 19:11:58 +0000 Subject: [labnetwork] Job posting (Calgary, Alberta, Canada) : Cleanroom Technical Specialist Message-ID: Hello, please apply if interested. https://internal.careers.ucalgary.ca/jobs/14104660-cleanroom-technical-specialist-vp-research Cheers, Joe Joseph Losby, PhD Manager, qLab Operations joseph.losby at ucalgary.ca [cid:efb5d824-1aba-4096-9ad0-7b349e1469f5] -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: Outlook-t0qrfotf.jpg Type: image/jpeg Size: 13901 bytes Desc: Outlook-t0qrfotf.jpg URL: From tobi at stanford.edu Thu Mar 14 15:24:05 2024 From: tobi at stanford.edu (Tobi Beetz) Date: Thu, 14 Mar 2024 19:24:05 +0000 Subject: [labnetwork] Job opportunity: Lab Operations Engineer, Stanford Nano Shared Facilities (SNSF) Message-ID: Dear All - The Stanford Nano Shared Facilities (SNSF) is seeking a Lab Operations Engineer to support highly specialized technical research operations within the SNSF's Nanopatterning and Flexible Cleanrooms. A successful candidate will manage day-to-day operations, maintenance of a number of fabrication and characterization instruments within the cleanrooms, and foster a sense of community between users from various technical backgrounds. S(he) will provide training and support to researchers, develop and implement advanced techniques, and also be involved in planning for upgrades and new equipment. https://careersearch.stanford.edu/jobs/research-and-development-scientist-and-engineer-1-24739 Please help us fill this position. Thanks, Tobi Tobi Beetz, Ph.D. Associate Director, Stanford Nano Shared Facilities, Stanford University, http://snsf.stanford.edu -------------- next part -------------- An HTML attachment was scrubbed... URL: From christophe.clement at polymtl.ca Tue Mar 19 09:03:21 2024 From: christophe.clement at polymtl.ca (=?iso-8859-1?Q?Christophe_Cl=E9ment?=) Date: Tue, 19 Mar 2024 09:03:21 -0400 (EDT) Subject: [labnetwork] OXford Plasmalab System 100 (ICP-RIE) - PCB BOARD ISSUE Message-ID: <08dc01da79fd$d1f804b0$75e80e10$@polymtl.ca> Hello community We recently had a water leak on our ICP-RIE Plasmalab System100 (an ICP-RIE). Water went on the main PCB board without touching the X20 PLC by chance. We could not have +24 +15 and -15VDC on screen to restart all pumps. After some work on the PCB board we could find a faulty resistor and could finally restart the pumping, but still some issues were still present (no CM Gauge reading on the software, and some heaters at max temperature). After some step forwards, we did step backwards because now we have lost the communication between PC and PLC. We have then decided to ask Oxford for a PCB replacement (94-PC91S21083 REV 3), but what we received is not what we were expecting (I did not know if I had to laugh or cry when opening the box ) Left: what we need, right: what we received from Oxford. If some Oxford tool owners, or any refurbished equipment?s company, or any third party company that works with Oxford tools read this message and know where we can buy a refurbished 94-PC91S21083 REV 3 PCB board or offer any advice, that will be greatly appreciated! Thank you Christophe Christophe Cl?ment Professionnel de recherche Laboratoire de microfabrication (LMF) Groupe des Couches Minces (GCM) www.gcmlab.ca Ecole Polytechnique de Montr?al www.polymtl.ca D?partement de g?nie physique * 2900 Boulevard Edouard Monpetit Pavillon JAB Campus de l'Universit? de Montr?al Montr?al (Qu?bec) H3T 1J4 8 christophe.clement at polymtl.ca ( 514 340 4711 #2417 Fax : 514 340 5195 -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.jpg Type: image/jpeg Size: 18882 bytes Desc: not available URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image002.jpg Type: image/jpeg Size: 16132 bytes Desc: not available URL: From Justin_Moreau at uml.edu Tue Mar 19 11:55:36 2024 From: Justin_Moreau at uml.edu (Moreau, Justin) Date: Tue, 19 Mar 2024 15:55:36 +0000 Subject: [labnetwork] OXford Plasmalab System 100 (ICP-RIE) - PCB BOARD ISSUE In-Reply-To: <08dc01da79fd$d1f804b0$75e80e10$@polymtl.ca> References: <08dc01da79fd$d1f804b0$75e80e10$@polymtl.ca> Message-ID: Hi Christophe 94-PC91S21083 is the PN for the unpopulated card. You need the assembled card. The PN looks like it should be 94-MA91Z21084 I suggest calling Oxford they may have one. [University of Massachusetts Lowell] Justin Moreau Director, Nanofabrication Core CORE RESEARCH FACILITIES ETIC E: Justin_Moreau at uml.edu T: 978-934-3615 M: 978-337-8148 [cid:image004.jpg at 01DA79F4.591E4CB0] From: labnetwork On Behalf Of Christophe Cl?ment Sent: Tuesday, March 19, 2024 9:03 AM To: labnetwork at mtl.mit.edu Subject: [labnetwork] OXford Plasmalab System 100 (ICP-RIE) - PCB BOARD ISSUE Hello community We recently had a water leak on our ICP-RIE Plasmalab System100 (an ICP-RIE). Water went on the main PCB board without touching the X20 PLC by chance. We could not have +24 +15 and -15VDC on screen to restart all pumps. After ZjQcmQRYFpfptBannerStart This Message Is From an External Sender This message came from outside your organization. Report Suspicious ? ZjQcmQRYFpfptBannerEnd Hello community We recently had a water leak on our ICP-RIE Plasmalab System100 (an ICP-RIE). Water went on the main PCB board without touching the X20 PLC by chance. We could not have +24 +15 and -15VDC on screen to restart all pumps. After some work on the PCB board we could find a faulty resistor and could finally restart the pumping, but still some issues were still present (no CM Gauge reading on the software, and some heaters at max temperature). After some step forwards, we did step backwards because now we have lost the communication between PC and PLC. We have then decided to ask Oxford for a PCB replacement (94-PC91S21083 REV 3), but what we received is not what we were expecting (I did not know if I had to laugh or cry when opening the box?) Left: what we need, right: what we received from Oxford. [cid:image005.jpg at 01DA79F4.591E4CB0][cid:image006.jpg at 01DA79F4.591E4CB0] If some Oxford tool owners, or any refurbished equipment?s company, or any third party company that works with Oxford tools read this message and know where we can buy a refurbished 94-PC91S21083 REV 3 PCB board or offer any advice, that will be greatly appreciated! Thank you Christophe Christophe Cl?ment Professionnel de recherche Laboratoire de microfabrication (LMF) Groupe des Couches Minces (GCM) www.gcmlab.ca Ecole Polytechnique de Montr?al www.polymtl.ca D?partement de g?nie physique ? 2900 Boulevard Edouard Monpetit Pavillon JAB Campus de l'Universit? de Montr?al Montr?al (Qu?bec) H3T 1J4 ? christophe.clement at polymtl.ca ? 514 340 4711 #2417 Fax : 514 340 5195 -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image003.png Type: image/png Size: 2650 bytes Desc: image003.png URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image004.jpg Type: image/jpeg Size: 3088 bytes Desc: image004.jpg URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image005.jpg Type: image/jpeg Size: 18882 bytes Desc: image005.jpg URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image006.jpg Type: image/jpeg Size: 16132 bytes Desc: image006.jpg URL: From christophe.clement at polymtl.ca Tue Mar 19 16:56:08 2024 From: christophe.clement at polymtl.ca (=?iso-8859-1?Q?Christophe_Cl=E9ment?=) Date: Tue, 19 Mar 2024 16:56:08 -0400 (EDT) Subject: [labnetwork] OXford Plasmalab System 100 (ICP-RIE) - PCB BOARD ISSUE In-Reply-To: <08dc01da79fd$d1f804b0$75e80e10$@polymtl.ca> References: <08dc01da79fd$d1f804b0$75e80e10$@polymtl.ca> Message-ID: <009701da7a3f$ddd43050$997c90f0$@polymtl.ca> Small edit of my previous message: Oxford has solved today our issue. As our tool is old (bought in 2002), the tool is not anymore listed in their inventory. Bill of material was hard to find for them for such old tool. Finally, we will receive in few days a fully completed PCB assembly! Thanks you for all your answer, Phillip, Justin, Denis, Tobi, Julia. I appreciate you help. And thank you Labnetwork. Best Chris Christophe Cl?ment Technicien laboratoire Laboratoire de microfabrication (LMF) Groupe des Couches Minces (GCM) www.gcmlab.ca Ecole Polytechnique de Montr?al www.polymtl.ca D?partement de g?nie physique * 2900 Boulevard Edouard Monpetit Pavillon JAB Campus de l'Universit? de Montr?al Montr?al (Qu?bec) H3T 1J4 * christophe.clement at polymtl.ca * 514 340 4711 #2417 Fax : 514 340 5195 De : labnetwork De la part de Christophe Cl?ment Envoy? : 19 mars 2024 09:03 ? : labnetwork at mtl.mit.edu Objet : [labnetwork] OXford Plasmalab System 100 (ICP-RIE) - PCB BOARD ISSUE Hello community We recently had a water leak on our ICP-RIE Plasmalab System100 (an ICP-RIE). Water went on the main PCB board without touching the X20 PLC by chance. We could not have +24 +15 and -15VDC on screen to restart all pumps. After some work on the PCB board we could find a faulty resistor and could finally restart the pumping, but still some issues were still present (no CM Gauge reading on the software, and some heaters at max temperature). After some step forwards, we did step backwards because now we have lost the communication between PC and PLC. We have then decided to ask Oxford for a PCB replacement (94-PC91S21083 REV 3), but what we received is not what we were expecting (I did not know if I had to laugh or cry when opening the box ) Left: what we need, right: what we received from Oxford. If some Oxford tool owners, or any refurbished equipment?s company, or any third party company that works with Oxford tools read this message and know where we can buy a refurbished 94-PC91S21083 REV 3 PCB board or offer any advice, that will be greatly appreciated! Thank you Christophe Christophe Cl?ment Professionnel de recherche Laboratoire de microfabrication (LMF) Groupe des Couches Minces (GCM) www.gcmlab.ca Ecole Polytechnique de Montr?al www.polymtl.ca D?partement de g?nie physique * 2900 Boulevard Edouard Monpetit Pavillon JAB Campus de l'Universit? de Montr?al Montr?al (Qu?bec) H3T 1J4 * christophe.clement at polymtl.ca * 514 340 4711 #2417 Fax : 514 340 5195 -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.jpg Type: image/jpeg Size: 18882 bytes Desc: not available URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image002.jpg Type: image/jpeg Size: 16132 bytes Desc: not available URL: From codreanu at udel.edu Wed Mar 20 12:17:01 2024 From: codreanu at udel.edu (Iulian Codreanu) Date: Wed, 20 Mar 2024 12:17:01 -0400 Subject: [labnetwork] Cleanroom career opportunities at UCF Message-ID: <14417adb-73ab-4f69-a595-cd9a15bd15d0@udel.edu> Dear Colleagues, My old school has a couple of cleanroom career opportunities; the links to the jobs are below. To learn more, please contact Prof. Abdolvand (Reza.Abdolvand at ucf.edu). https://ucf.wd1.myworkdayjobs.com/en-US/careers/job/Cleanroom-Engineer_R107438-1 https://ucf.wd1.myworkdayjobs.com/en-US/careers/job/Associate-Director--Cleanroom_R107407 Cheers, Iulian -- iulian Codreanu, Ph.D. Director, Nanofabrication Facility University of Delaware Harker ISE Lab, Room 163 221 Academy Street Newark, DE 19716 302-831-2784 https://udnf.udel.edu From patricns at uw.edu Wed Mar 20 16:41:11 2024 From: patricns at uw.edu (N Shane Patrick) Date: Wed, 20 Mar 2024 13:41:11 -0700 Subject: [labnetwork] Gas Cabinet Service Providers Message-ID: Hello All, Does anyone know of any third-party vendors who will provide maintenance service to Praxair made (now owned by Linde) gas cabinets? Our efforts to obtain service or parts through Linde have been disappointing to say the least. Alternatively, who would you consider in terms of new gas cabinets that you feel has a reliable reputation for service? N. Shane Patrick Manager, Lab Operations and Safety Electron Beam Lithography Washington Nanofabrication Facility (WNF) National Nanotechnology Coordinated Infrastructure (NNCI) University of Washington - NanoES Fluke Hall 129, Box 352143 (206) 221-1045 patricns at uw.edu http://www.wnf.washington.edu/ -------------- next part -------------- An HTML attachment was scrubbed... URL: From dwolff at nelhydrogen.com Wed Mar 20 19:55:16 2024 From: dwolff at nelhydrogen.com (David Wolff) Date: Wed, 20 Mar 2024 23:55:16 +0000 Subject: [labnetwork] Gas Cabinet Service Providers In-Reply-To: References: Message-ID: Try Mark Dye at SilPac: Mark Dye SilPac International markdye at silpac.net (408) 768-1362 (Mobile) David Wolff Industrial Sales Director - North America [cid:image001.png at 01DA7B00.550F3F50] Nel Hydrogen 10 Technology Drive, Wallingford, CT 06492 United States Office: M +1 860-604-3282 Fax: +1 203-949-8016 dwolff at nelhydrogen.com From: labnetwork On Behalf Of N Shane Patrick Sent: Wednesday, March 20, 2024 16:41 To: Fab Network Subject: [labnetwork] Gas Cabinet Service Providers Hello All, Does anyone know of any third-party vendors who will provide maintenance service to Praxair made (now owned by Linde) gas cabinets? Our efforts to obtain service or parts through Linde have been disappointing to say the least. Alternatively, who would you consider in terms of new gas cabinets that you feel has a reliable reputation for service? N. Shane Patrick Manager, Lab Operations and Safety Electron Beam Lithography Washington Nanofabrication Facility (WNF) National Nanotechnology Coordinated Infrastructure (NNCI) University of Washington - NanoES Fluke Hall 129, Box 352143 (206) 221-1045 patricns at uw.edu http://www.wnf.washington.edu/ -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.png Type: image/png Size: 36237 bytes Desc: image001.png URL: From schweig at umich.edu Thu Mar 21 04:48:14 2024 From: schweig at umich.edu (Dennis Schweiger) Date: Thu, 21 Mar 2024 04:48:14 -0400 Subject: [labnetwork] Gas Cabinet Service Providers In-Reply-To: References: Message-ID: Shane, Good morning. I would try Critical Systems ( https://www.criticalsystemsinc.com/gas-delivery-systems) for the support of your existing cabinets. They may be able to help you out. For the new cabinets, I would recommend Applied Energy Systems ( https://www.appliedenergysystems.com/) out of Pennsylvania. I've installed a lot of their automatic cabinets over the years, and have nothing but good things to say about their build, and support, of their equipment. As a side note, we have quite a few older Air Products cabinets in our facility, and when the controllers for those finally give out, we'll retrofit them with AES controllers. Feel free to give me a call if you want to discuss further. Thanks, Dennis Schweiger Facilities Manager University of Michigan/LNF 734.647.2055 Ofc On Wed, Mar 20, 2024 at 7:53?PM N Shane Patrick wrote: > Hello All, > > Does anyone know of any third-party vendors who will provide maintenance > service to Praxair made (now owned by Linde) gas cabinets? Our efforts to > obtain service or parts through Linde have been disappointing to say the > least. > > Alternatively, who would you consider in terms of new gas cabinets that > you feel has a reliable reputation for service? > > > N. Shane Patrick > Manager, Lab Operations and Safety > Electron Beam Lithography > Washington Nanofabrication Facility (WNF) > National Nanotechnology Coordinated Infrastructure (NNCI) > University of Washington - NanoES > Fluke Hall 129, Box 352143 > (206) 221-1045 > patricns at uw.edu > http://www.wnf.washington.edu/ > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork > -------------- next part -------------- An HTML attachment was scrubbed... URL: From codreanu at udel.edu Thu Mar 21 08:45:32 2024 From: codreanu at udel.edu (Iulian Codreanu) Date: Thu, 21 Mar 2024 08:45:32 -0400 Subject: [labnetwork] Gas Cabinet Service Providers In-Reply-To: References: Message-ID: <91f83057-7941-4b07-98b3-77cc4000e864@udel.edu> I would also recommend?Applied Energy Systems. iulian Codreanu, Ph.D. Director, Nanofabrication Facility University of Delaware Harker ISE Lab, Room 163 221 Academy Street Newark, DE 19716 302-831-2784 https://udnf.udel.edu On 3/20/2024 4:41 PM, N Shane Patrick wrote: > Hello All, > > Does anyone know of any third-party vendors who will provide > maintenance service to Praxair made (now owned by Linde) gas cabinets? > Our efforts to obtain service or parts through Linde have been > disappointing to say the least. > > Alternatively, who would you consider in terms of new gas cabinets > that you feel has a reliable reputation for service? > > > N. Shane Patrick > Manager, Lab Operations and Safety > Electron Beam Lithography > Washington Nanofabrication Facility?(WNF) > National Nanotechnology Coordinated?Infrastructure (NNCI) > University of Washington - NanoES > Fluke Hall 129, Box 352143 > (206) 221-1045 > patricns at uw.edu > http://www.wnf.washington.edu/ > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: From Jackson.Anderson at uvm.edu Thu Mar 21 09:07:31 2024 From: Jackson.Anderson at uvm.edu (Jackson Anderson (he/him)) Date: Thu, 21 Mar 2024 13:07:31 +0000 Subject: [labnetwork] Reichert-Jung Polyvar microscope controller schematics Message-ID: Hello all, I have a Reichert-Jung 6526-04 Polyvar microscope controller that is acting up (no 12V coming out for the light source) and am wondering if anyone has any schematics or documentation for troubleshooting these units. Thanks, Jackson Jackson Anderson Research Assistant Professor Electrical and Biomedical Engineering University of Vermont Votey Hall, Room 301A 33 Colchester Ave. Burlington, VT 05405 -------------- next part -------------- An HTML attachment was scrubbed... URL: From deonc69 at illinois.edu Thu Mar 21 11:07:19 2024 From: deonc69 at illinois.edu (Collins, Deon) Date: Thu, 21 Mar 2024 15:07:19 +0000 Subject: [labnetwork] Gas Cabinet Service Providers In-Reply-To: References: Message-ID: Getting service from Linde on almost anything has been very poor lately. We went to CSI(Critical Systems Inc) for gas cabinets and service. Deon D. Collins Director of Facilities The Grainger College Of Engineering UIUC Holonyak Micro & Nanotechnology Lab 208 N Wright St Rm. 1114 | MC-249 Urbana, IL 61801 217-300-7531 | deonc69 at illinois.edu [cid:image001.png at 01DA7B77.8E7492A0] Under the Illinois Freedom of Information Act any written communication to or from university employees regarding university business is a public record and may be subject to public disclosure. Life begins at the end of your comfort zone. Propeck Amanda From: labnetwork On Behalf Of N Shane Patrick Sent: Wednesday, March 20, 2024 3:41 PM To: Fab Network Subject: [labnetwork] Gas Cabinet Service Providers Hello All, Does anyone know of any third-party vendors who will provide maintenance service to Praxair made (now owned by Linde) gas cabinets? Our efforts to obtain service or parts through Linde have been disappointing to say the least. Alternatively, who would you consider in terms of new gas cabinets that you feel has a reliable reputation for service? N. Shane Patrick Manager, Lab Operations and Safety Electron Beam Lithography Washington Nanofabrication Facility (WNF) National Nanotechnology Coordinated Infrastructure (NNCI) University of Washington - NanoES Fluke Hall 129, Box 352143 (206) 221-1045 patricns at uw.edu http://www.wnf.washington.edu/ -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.png Type: image/png Size: 2602 bytes Desc: image001.png URL: From julia.aebersold at louisville.edu Thu Mar 21 11:12:12 2024 From: julia.aebersold at louisville.edu (Aebersold, Julia) Date: Thu, 21 Mar 2024 15:12:12 +0000 Subject: [labnetwork] Gas Cabinet Service Providers In-Reply-To: References: Message-ID: We also enjoy good support from AES and recommend them. Cheers! Julia Aebersold, Ph.D. Director of Operations Micro/Nano Technology Center University of Louisville 2210 South Brook Street Shumaker Research Building, Room 233 Louisville, KY 40292 (502) 852-1572 http://louisville.edu/micronano/ From: labnetwork On Behalf Of Dennis Schweiger Sent: Thursday, March 21, 2024 4:48 AM To: N Shane Patrick Cc: Fab Network Subject: Re: [labnetwork] Gas Cabinet Service Providers CAUTION: This email originated from outside of our organization. Do not click links, open attachments, or respond unless you recognize the sender's email address and know the contents are safe. Shane, Good morning. I would try Critical Systems (https://www.criticalsystemsinc.com/gas-delivery-systems) for the support of your existing cabinets. They may be able to help you out. For the new cabinets, I would recommend Applied Energy Systems (https://www.appliedenergysystems.com/) out of Pennsylvania. I've installed a lot of their automatic cabinets over the years, and have nothing but good things to say about their build, and support, of their equipment. As a side note, we have quite a few older Air Products cabinets in our facility, and when the controllers for those finally give out, we'll retrofit them with AES controllers. Feel free to give me a call if you want to discuss further. Thanks, Dennis Schweiger Facilities Manager University of Michigan/LNF 734.647.2055 Ofc On Wed, Mar 20, 2024 at 7:53?PM N Shane Patrick > wrote: Hello All, Does anyone know of any third-party vendors who will provide maintenance service to Praxair made (now owned by Linde) gas cabinets? Our efforts to obtain service or parts through Linde have been disappointing to say the least. Alternatively, who would you consider in terms of new gas cabinets that you feel has a reliable reputation for service? N. Shane Patrick Manager, Lab Operations and Safety Electron Beam Lithography Washington Nanofabrication Facility (WNF) National Nanotechnology Coordinated Infrastructure (NNCI) University of Washington - NanoES Fluke Hall 129, Box 352143 (206) 221-1045 patricns at uw.edu http://www.wnf.washington.edu/ _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: From jeanne.guo at rice.edu Fri Mar 22 09:29:54 2024 From: jeanne.guo at rice.edu (Jing Guo) Date: Fri, 22 Mar 2024 08:29:54 -0500 Subject: [labnetwork] Gas Cabinet Service Providers In-Reply-To: <91f83057-7941-4b07-98b3-77cc4000e864@udel.edu> References: <91f83057-7941-4b07-98b3-77cc4000e864@udel.edu> Message-ID: <24FFB815-8E26-4D31-B8FE-C633168EF885@rice.edu> One more ?yes? to AES --------------------------------------- Jing Guo Ph.D. Research Scientist SEA Cleanroom (SST 017) Rice University Houston, TX jeanne.guo at rice.edu 713-348-8227 > On Mar 21, 2024, at 7:45 AM, Iulian Codreanu wrote: > > I would also recommend Applied Energy Systems. > iulian Codreanu, Ph.D. > Director, Nanofabrication Facility > University of Delaware > Harker ISE Lab, Room 163 > 221 Academy Street > Newark, DE 19716 > 302-831-2784 > https://udnf.udel.edu > On 3/20/2024 4:41 PM, N Shane Patrick wrote: >> Hello All, >> >> Does anyone know of any third-party vendors who will provide maintenance service to Praxair made (now owned by Linde) gas cabinets? Our efforts to obtain service or parts through Linde have been disappointing to say the least. >> >> Alternatively, who would you consider in terms of new gas cabinets that you feel has a reliable reputation for service? >> >> >> N. Shane Patrick >> Manager, Lab Operations and Safety >> Electron Beam Lithography >> Washington Nanofabrication Facility (WNF) >> National Nanotechnology Coordinated Infrastructure (NNCI) >> University of Washington - NanoES >> Fluke Hall 129, Box 352143 >> (206) 221-1045 >> patricns at uw.edu >> http://www.wnf.washington.edu/ >> >> >> _______________________________________________ >> labnetwork mailing list >> labnetwork at mtl.mit.edu >> https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: From mjohnso6 at tulane.edu Fri Mar 22 09:45:07 2024 From: mjohnso6 at tulane.edu (Johnson, Michael W) Date: Fri, 22 Mar 2024 13:45:07 +0000 Subject: [labnetwork] PI funding gaps Message-ID: How do other shared user facilites deal with a situation when a regular user research group PI has a funding shortfall and can't pay user fees? Is it typical to have some type of temporary waiver mechanism enshrined in the shared user structure or does that PI appeal to their department head to bail them out to bridge a gap? A waiver feels problematic because it would seem to violate basic fairness in that one group could be getting "free stuff" and everybody else has to pay. I would be interested to hear how other places deal with a sticky situation like this. -------------- next part -------------- An HTML attachment was scrubbed... URL: From johnboyle at fast.net Fri Mar 22 10:07:04 2024 From: johnboyle at fast.net (johnboyle) Date: Fri, 22 Mar 2024 10:07:04 -0400 Subject: [labnetwork] Gas Cabinet Service Providers In-Reply-To: Message-ID: <1A.E2.25880.8809DF56@smtp04.aqua.email-ash1.sync.lan> SDC in Saugerties NY. They may have a worked with Prax/Linde on cabinetsJohn BoyleSent from my Verizon, Samsung Galaxy smartphone -------- Original message --------From: N Shane Patrick Date: 3/20/24 7:52 PM (GMT-05:00) To: Fab Network Subject: [labnetwork] Gas Cabinet Service Providers Hello All,Does anyone know of any third-party vendors who will provide maintenance service to Praxair made (now owned by Linde) gas cabinets? Our efforts to obtain service or parts through Linde have been disappointing to say the least.Alternatively, who would you consider in terms of new gas cabinets that you feel has a reliable reputation for service? N. Shane PatrickManager, Lab Operations and SafetyElectron Beam LithographyWashington Nanofabrication Facility?(WNF)?National Nanotechnology Coordinated?Infrastructure (NNCI)University of Washington - NanoESFluke Hall 129, Box 352143(206) 221-1045patricns at uw.eduhttp://www.wnf.washington.edu/ -------------- next part -------------- An HTML attachment was scrubbed... URL: From rgc5 at psu.edu Fri Mar 22 10:38:01 2024 From: rgc5 at psu.edu (Cornwall, Bob) Date: Fri, 22 Mar 2024 14:38:01 +0000 Subject: [labnetwork] PI funding gaps In-Reply-To: References: Message-ID: Within our cores, faculty research groups cant do the work unless they provide a valid fund number. WE pull that out of the financial system nightly, and if the fund is not allowed to be charged against, it is removed from the users allowable charge accounts in our reservation and billing system which is based on NEMO. They cannot do the work on that account and faculty are made aware that they are not allowed to charge other Award/contracts and move charges later. They can charge internal university funds, but may not be allowed to move those charges to an award after the fact. That is handled at the department level. Bob Robert Cornwall Managing Director - Penn State - Materials Research Institute rgc5 at psu.edu C 814-571-3438 W 814-863-8735 Availability Connect with MRI: https://www.mri.psu.edu/opt ________________________________ From: labnetwork on behalf of Johnson, Michael W Sent: Friday, March 22, 2024 9:45 AM To: Fab Network Subject: [labnetwork] PI funding gaps You don't often get email from mjohnso6 at tulane.edu. Learn why this is important How do other shared user facilites deal with a situation when a regular user research group PI has a funding shortfall and can't pay user fees? Is it typical to have some type of temporary waiver mechanism enshrined in the shared user structure or does that PI appeal to their department head to bail them out to bridge a gap? A waiver feels problematic because it would seem to violate basic fairness in that one group could be getting "free stuff" and everybody else has to pay. I would be interested to hear how other places deal with a sticky situation like this. -------------- next part -------------- An HTML attachment was scrubbed... URL: From codreanu at udel.edu Fri Mar 22 10:53:34 2024 From: codreanu at udel.edu (Iulian Codreanu) Date: Fri, 22 Mar 2024 10:53:34 -0400 Subject: [labnetwork] PI funding gaps In-Reply-To: References: Message-ID: <7bde8e51-d1b5-4951-bd09-139897aaa3f1@udel.edu> Hi Michael, I have not encountered the scenario you mentioned but I am providing some comments you may find useful. Our reservation system "talks" to the University's financial system every time a user makes a reservation or enables a tool for usage. If the financial account the user has on file is out of money, expired, etc. the reservation system presents the user with the ability to proceed with using a "suspense code" (I am not sure why I agreed to that name). If the user makes that choice, an email is generated to provide a valid account at the earliest convenience. We have not had any situations where we did not get a valid account number by the time we completed our monthly billing cycle. The bailing mechanism you mentioned would probably be a good way to handle this. I believe that charging some users but not charging others (for whatever reason) runs afoul of the federal service center rules our labs have to adhere to. Cheers, Iulian iulian Codreanu, Ph.D. Director, Nanofabrication Facility University of Delaware Harker ISE Lab, Room 163 221 Academy Street Newark, DE 19716 302-831-2784 https://udnf.udel.edu On 3/22/2024 9:45 AM, Johnson, Michael W wrote: > How do other shared user facilites deal with a situation when a > regular user research group PI has a funding shortfall and can't pay > user fees? > > Is it typical to have some type of temporary waiver mechanism > enshrined in the shared user structure or does that PI appeal to their > department head to bail them out to bridge a gap? > > A waiver feels problematic because it would seem to violate basic > fairness in that one group could be getting "free stuff" and everybody > else has to pay. > > I would be interested to hear how other places deal with a sticky > situation like this. > > > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork From bgila at ufl.edu Fri Mar 22 12:40:43 2024 From: bgila at ufl.edu (Gila,Brent P) Date: Fri, 22 Mar 2024 12:40:43 -0400 Subject: [labnetwork] PI funding gaps In-Reply-To: References: Message-ID: <537aeed0-434f-4af6-84b1-f43d4961e6ed@ufl.edu> At UF we require the faculty to have their Returned Overhead (ROH) accounts set up in our system as a fall back in these instances.? If they have no ROH then we ask their home department for an unrestricted department account to cover any future transactions/invoices until the PI's account is back in the positive.? This way the department chair is aware and can deal with it internally when the bill comes due. Our fiscal system does read the account details from the university system similar to other university core facilities so we can monitor if needed. Best Regards, Brent -- Brent P. Gila, PhD. Associate Director, Research Service Centers University of Florida 1041 Center Drive Gainesville, Florida 32611 Tel:352-273-2245,? Fax:352-846-2877 email: bgila at ufl.edu On 3/22/2024 9:45 AM, Johnson, Michael W wrote: > *[External Email]* > > How do other shared user facilites deal with a situation when a > regular user research group PI has a funding shortfall and can't pay > user fees? > > Is it typical to have some type of temporary waiver mechanism > enshrined in the shared user structure or does that PI appeal to their > department head to bail them out to bridge a gap? > > A waiver feels problematic because it would seem to violate basic > fairness in that one group could be getting "free stuff" and everybody > else has to pay. > > I would be interested to hear how other places deal with a sticky > situation like this. > > > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: From odc1n08 at soton.ac.uk Fri Mar 22 12:43:50 2024 From: odc1n08 at soton.ac.uk (Owain Clark) Date: Fri, 22 Mar 2024 16:43:50 +0000 Subject: [labnetwork] PI funding gaps In-Reply-To: References: Message-ID: For us when a project is set up, a maximum spend per research facility is assigned. Bookings made in that facility are then subtracted from this figure as they are created and once it hits zero then no more bookings are possible on that code. Users are assigned access to project codes in the booking system during their account setup, or when they are added to a project. Our admin team work with our finance and RIS departments to administrate this. We currently use The Living Database (Phoenix) as a booking system but will move to PPMS/Stratocore later this year. Our financial integration is separate (but driven from reports generated from the booking system) as a lot of our research projects are not pay as you go, and are instead charged a fixed amount per month regardless of actual facility usage. So the financial spend figure held in our booking system is only protection against budget over runs. Owain. From: labnetwork On Behalf Of Cornwall, Bob Sent: Friday, March 22, 2024 2:38 PM To: Johnson, Michael W ; Fab Network Subject: Re: [labnetwork] PI funding gaps CAUTION: This e-mail originated outside the University of Southampton. Within our cores, faculty research groups cant do the work unless they provide a valid fund number. WE pull that out of the financial system nightly, and if the fund is not allowed to be charged against, it is removed from the users allowable charge accounts in our reservation and billing system which is based on NEMO. They cannot do the work on that account and faculty are made aware that they are not allowed to charge other Award/contracts and move charges later. They can charge internal university funds, but may not be allowed to move those charges to an award after the fact. That is handled at the department level. Bob Robert Cornwall Managing Director - Penn State - Materials Research Institute rgc5 at psu.edu C 814-571-3438 W 814-863-8735 Availability Connect with MRI: https://www.mri.psu.edu/opt ________________________________ From: labnetwork > on behalf of Johnson, Michael W > Sent: Friday, March 22, 2024 9:45 AM To: Fab Network > Subject: [labnetwork] PI funding gaps You don't often get email from mjohnso6 at tulane.edu. Learn why this is important How do other shared user facilites deal with a situation when a regular user research group PI has a funding shortfall and can't pay user fees? Is it typical to have some type of temporary waiver mechanism enshrined in the shared user structure or does that PI appeal to their department head to bail them out to bridge a gap? A waiver feels problematic because it would seem to violate basic fairness in that one group could be getting "free stuff" and everybody else has to pay. I would be interested to hear how other places deal with a sticky situation like this. -------------- next part -------------- An HTML attachment was scrubbed... URL: From mtang at stanford.edu Fri Mar 22 12:48:26 2024 From: mtang at stanford.edu (Mary Tang) Date: Fri, 22 Mar 2024 16:48:26 +0000 Subject: [labnetwork] PI funding gaps In-Reply-To: References: Message-ID: <05b202fa-720e-45d1-91ba-d173ea64d04a@stanford.edu> Hi Michael - Excellent question for which there is no "right" answer. You are correct, if you operate as a Service Center, you can't waive debt. The first priority is the service center, making sure it balances at the end of the fiscal year. The next priority is to fix this problem and make sure it doesn't happen again. I know, easier said than done. We are fortunate to have access to an unrestricted account -- not everyone has this luxury. But perhaps this is something you can lobby for. Depending on how much the amount owed and how your compliance office works, bad debt might have to be carried over into the next fiscal year (so your rates would need to be increased accordingly), or would need to be reconciled in the current fiscal year (so your fees would have to be retroactively increased accordingly for everyone) -- neither option will lead to happy faculty. If your administration is like mine, someone higher up would rather find a way to cover this debt (with strings attached) and deal with one, rather than a boatload of unhappy faculty. We had a professor who routinely couldn't pay. A well-loved, charming, otherwise highly ethical person who was simply terrible with money. His department said, "everyone knows he has no money. Why did you let him use your facility?" Like we would know and have the ability to say "no?" In the end, we now have an account underwritten by the School of Engineering to cover debt. This allows us to close the fiscal year and buy us time to address the situation. (While we no longer accept his accounts, it doesn't affect his research -- he happily collaborates with other PI's with better financial sense.) A shared facility is like a business in that we need to be customer-service oriented. For compliance reasons, we don't have as many adjustable knobs at our disposal. But if your administration is amenable, there are probably many different ways they can work with you to address these needs while remaining compliant. Mary -- Mary X. Tang, Ph.D. Managing Director Stanford Nanofabrication Facility Paul G. Allen Building 420 Via Palou Mall Stanford, CA 94305 https://snf.stanford.edu mtang at stanford.edu On 3/22/2024 6:45 AM, Johnson, Michael W wrote: How do other shared user facilites deal with a situation when a regular user research group PI has a funding shortfall and can't pay user fees? Is it typical to have some type of temporary waiver mechanism enshrined in the shared user structure or does that PI appeal to their department head to bail them out to bridge a gap? A waiver feels problematic because it would seem to violate basic fairness in that one group could be getting "free stuff" and everybody else has to pay. I would be interested to hear how other places deal with a sticky situation like this. _______________________________________________ labnetwork mailing list labnetwork at mtl.mit.edu https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: From carl.chow at hpe.com Fri Mar 22 13:31:12 2024 From: carl.chow at hpe.com (Chow, Carl) Date: Fri, 22 Mar 2024 17:31:12 +0000 Subject: [labnetwork] Gas Cabinet Service Providers In-Reply-To: <1A.E2.25880.8809DF56@smtp04.aqua.email-ash1.sync.lan> References: <1A.E2.25880.8809DF56@smtp04.aqua.email-ash1.sync.lan> Message-ID: We have SDC CiphercoN 1500 gas cabinets. They are only 4 years old, so we have not had any issues or service calls. SDC was willing to come out and show us how to change toxic gas cylinders for $8K. They would not do the work, but showed us what to do and we made our own video for documentation. Another vendor wanted $27K to change 6 cylinders, so now we are self-sufficient. We had a good experience with SDC. Regards, Carl Chow Hewlett Packard Enterprise Labs From: labnetwork On Behalf Of johnboyle Sent: Friday, March 22, 2024 7:07 AM To: N Shane Patrick ; Fab Network ; johnboyle at fast.net Subject: Re: [labnetwork] Gas Cabinet Service Providers SDC in Saugerties NY. They may have a worked with Prax/Linde on cabinets John Boyle Sent from my Verizon, Samsung Galaxy smartphone -------- Original message -------- From: N Shane Patrick > Date: 3/20/24 7:52 PM (GMT-05:00) To: Fab Network > Subject: [labnetwork] Gas Cabinet Service Providers Hello All, Does anyone know of any third-party vendors who will provide maintenance service to Praxair made (now owned by Linde) gas cabinets? Our efforts to obtain service or parts through Linde have been disappointing to say the least. Alternatively, who would you consider in terms of new gas cabinets that you feel has a reliable reputation for service? N. Shane Patrick Manager, Lab Operations and Safety Electron Beam Lithography Washington Nanofabrication Facility (WNF) National Nanotechnology Coordinated Infrastructure (NNCI) University of Washington - NanoES Fluke Hall 129, Box 352143 (206) 221-1045 patricns at uw.edu http://www.wnf.washington.edu/ -------------- next part -------------- An HTML attachment was scrubbed... URL: From sb86922 at usc.edu Mon Mar 25 15:10:16 2024 From: sb86922 at usc.edu (Shivakumar Bhaskaran) Date: Mon, 25 Mar 2024 19:10:16 +0000 Subject: [labnetwork] USC Job Opening - Nanofabrication Lab Operation Engineer In-Reply-To: References: Message-ID: USC Nanofab is looking for Nanofabrication Lab Operation Engineer. REQ20146594 Nanofabrication Lab Operation Engineer https://usccareers.usc.edu/job/los-angeles/nanofabrication-lab-operation-engineer/1209/62638875264 Job Description: We are looking for a highly-motivated Lab Operation Engineer. Reporting to the Associate Director, this individual will support the cleanroom users with training, developing nanofabrication process, create new processes, and improve yield and device performance. Collaborate with facility staff and research users to identify tool operation, process issues and address those issues promptly with proper maintenance, repair, and troubleshooting. The ideal candidate should be customer service oriented and contribute towards improving the operation of the facility. Shiva Shivakumar Bhaskaran, Ph.D., Associate Director, John D. O'Brien Nanofabrication Laboratory Michelson Hall, 1002 Childs Way, MCB LL121, Los Angeles, California 90089, 213 821 2374 [Sign] -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: image001.jpg Type: image/jpeg Size: 2931 bytes Desc: image001.jpg URL: From codreanu at udel.edu Mon Mar 25 16:58:07 2024 From: codreanu at udel.edu (Iulian Codreanu) Date: Mon, 25 Mar 2024 16:58:07 -0400 Subject: [labnetwork] Precious Metal Reclaim Service In-Reply-To: References: Message-ID: <252b76d4-9036-488f-b74f-ebe15d25192c@udel.edu> I use Glines and Rhodes. iulian Codreanu, Ph.D. Director, Nanofabrication Facility University of Delaware Harker ISE Lab, Room 163 221 Academy Street Newark, DE 19716 302-831-2784 https://udnf.udel.edu On 3/4/2024 1:04 PM, Chang, Long wrote: > Hi All, > > Who do you use for precious metal reclaiming from spent targets and > chamber/shield scraping? > > Thanks, > Long Chang > UH Nanofabrication Facility > Houston, TX > lvchang at central.uh.edu > > > > > > _______________________________________________ > labnetwork mailing list > labnetwork at mtl.mit.edu > https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork -------------- next part -------------- An HTML attachment was scrubbed... URL: From kckeenan at seas.upenn.edu Wed Mar 27 16:37:40 2024 From: kckeenan at seas.upenn.edu (Kyle Keenan) Date: Wed, 27 Mar 2024 16:37:40 -0400 Subject: [labnetwork] service/repair help for an ABM 300HR Message-ID: Hello Lab Network, We have an old ABM 300HR aligner (HTG rebuild) that is experiencing an electrical failure that has been difficult to diagnose. Given the age of the system, we are finding it very difficult to get anyone who is willing or able to give us technical support and/or repair service. If you know of someone who might be able to help us and can share their contact information, it would be greatly appreciated. This repair is needed somewhat urgently, as the system has flexibility not found in more modern aligners that some of our researchers need to do their work. Thanks for your help. -- Kyle Keenan Senior Manager - Laboratory Operations Quattrone Nanofabrication Facility University of Pennsylvania P: 215-898-7560 F: 215-573-4925 -------------- next part -------------- An HTML attachment was scrubbed... URL: