[labnetwork] Sputter problem
Mac Hathaway
Hathaway at cns.fas.harvard.edu
Fri Jun 24 13:19:49 EDT 2011
Apologies, I meant no other responses of any kind!
: )
Mac Hathaway wrote:
> Hi Keith,
>
> Since I see no other better responses yet, I'll give you my two cents
> (not being a sputtering guru, myself):
>
> Sounds like something is getting hot. Things to check:
>
> Loose RF connections in the chamber, and back to the tuning network.
> Also, check inside your tuning network for loose connections.
>
> Does your forward/reflected power change over the day?
>
> Another place to look is your RF power supply.
>
>
> Mac Hathaway
> Harvard CNS
>
>
>
> Keith Bradshaw wrote:
>
>> We are sputtering TiO2 .
>>
>> We begin the day with a 250 angstrom rate....each subsequent run is
>> reduced by 10-15% in rate until we are at a 125 angstrom rate. We are
>> not changing anything.
>>
>> Next day we begin again at 250.
>>
>> Tried argon clean between runs, tried 3 hour wait between runs, we are
>> using a load lock and vacuum looks stable , still rate drops on each run.
>>
>> Any ideas?
>>
>> cordially,
>>
>> Keith Bradshaw
>> Dallas
>>
>