[labnetwork] cleanroom air handling equipment maintenance
Morrison, Richard H., Jr.
rmorrison at draper.com
Thu Jun 21 06:10:42 EDT 2012
HI,
At Draper the central facilities group is responsible for all building equipment. When they have to enter the cleanroom envelope we train them on gowning and tool cleaning procedures. The HVAC equipment vendors trains the facilities technicians during the installation of the equipment. All fans are on a 6 month PM schedule, this entails a published scheduled down time to have this done, usually on a weekend.
All process equipment is maintained by a dedicated staff off 3 equipment technicians and we do not hesitate to bring in the manufacturers service people.
Rick
Rick Morrison
Senior Member Technical Staff
Group Leader Microfabrication Operations
Draper Laboratory
555 Technology Square
Cambridge, MA 02139
W 617-258-3420
C 508-930-3461
-----Original Message-----
From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Iulian Codreanu
Sent: Tuesday, June 19, 2012 4:57 PM
To: Fab Network
Subject: [labnetwork] cleanroom air handling equipment maintenance
Good Afternoon.
We are trying to determine who will maintain the systems that will serve
Penn's new cleanroom and hope that you can share some of your experience
with us.
- Who maintains your cleanroom air handling equipment (e.g. external
contractor, Central Facilities)?
- If the maintenance is done by Facilities, what kind of additional
training/certification do you require?
Thank you very much for your help.
Iulian
--
iulian Codreanu, Ph.D.
Director, Penn NanoFab
200 South 33rd Street
Room 376 GRW Bldg
Philadelphia, PA 19104-6314
P: 215-898-9308
F: 215-573-2068
www.seas.upenn.edu/~nanofab
_______________________________________________
labnetwork mailing list
labnetwork at mtl.mit.edu
https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork
More information about the labnetwork
mailing list