Fri Jun 13 08:29:44 EDT 2014
Richard,
do you have access to LN2? We'll use a small amount of LN2 in a tray, and
add water. It generates quite a bit of "smoke", is inert, and provides a
good visual for "seeing" the air flow.
Dennis Schweiger
University of Michigan/LNF
Facilities Manager
734.647.2055 Ofc
On Fri, Jun 13, 2014 at 6:19 AM, Morrison, Richard H., Jr. <
rmorrison at draper.com> wrote:
> Hi Everyone,
>
>
>
> I am trying to fix some intermittent issues with air flow around some
> tools in a Class 10 room. I tried to use a smoke stick but they are very
> messy (not Class 10 material), is there a cleanroom smoke generator that I
> can use to look at air flow.
>
>
>
> We have built a physical model in ANSYS and have good results, but I want
> to visualize the flow.
>
>
>
> Thanks and see you at UGIM.
>
>
> Rick
>
>
>
>
>
> Draper Laboratory
>
> Principal Member of the Technical Staff
>
> Group Leader Microfabrication Operations
>