[labnetwork] Class 10 air flow detection
Paolini, Steven
spaolini at cns.fas.harvard.edu
Fri Jun 13 10:15:50 EDT 2014
Rich,
We used to use a simple drug store vaporizer for visual feedback on air flows. The only caveat is the added moisture put into the air.
Steve Paolini
Equipment dood
Harvard University Center for Nanoscale Systems
From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Morrison, Richard H., Jr.
Sent: Friday, June 13, 2014 6:20 AM
To: labnetwork at mtl.mit.edu
Subject: [labnetwork] Class 10 air flow detection
Hi Everyone,
I am trying to fix some intermittent issues with air flow around some tools in a Class 10 room. I tried to use a smoke stick but they are very messy (not Class 10 material), is there a cleanroom smoke generator that I can use to look at air flow.
We have built a physical model in ANSYS and have good results, but I want to visualize the flow.
Thanks and see you at UGIM.
Rick
Draper Laboratory
Principal Member of the Technical Staff
Group Leader Microfabrication Operations
555 Technology Square
Cambridge Ma, 02139-3563
www.draper.com<http://www.draper.com>
rmorrison at draper.com<mailto:rmorrison at draper.com>
W 617-258-3420
C 508-930-3461