[labnetwork] Shared Chiller [for equipment cooling] Issues
Mike Young
myoung6 at nd.edu
Mon Sep 22 08:58:09 EDT 2014
Kamal, we use flow limiters ("circuit setters" in the US) in the *return* line of each shared instrument, when they are installed in parallel on one cooling loop. Each flow limiter set to the maximum flow required by the single tool ahead of it. That way, as other tools intermittently increase their coolant flow, they are not able to "starve" other equipment, by virtue of the flow limiter. In order for this scheme to work, the chiller has to be able to supply the total of the maximum flow required of all its commected tools, *at the differential pressure required by the the tool with the highest required differential*. If you chiller can't meet this requirement, you need a larger chiller.
Good luck!
--Mike
On Sep 22, 2014, at 2:22 AM, Kamal Yadav <kamal.yadav at gmail.com> wrote:
> Dear All,
>
> We are facing chiller [for cooling water for equipment] issues, most likely flow issues, probably due to using same chiller for multiple equipment. Though chiller total cooling capacity is able to provide flow/pressure but in an shared equipment scenario, could there be issues. Suddenly some equipment, in use, trips, because of low flow of cooling water probably.
>
> Did anybody faced these issues, are we diagnosing it correctly?
>
> Thanks a lot.
>
> --
> Thanks,
> Kamal Yadav
> Sr. Process Technologist
> IITBNF, EE Department, Annexe,
> IIT Bombay, Powai
> Mumbai 400076
> Internal: 4435
> Cell: 7506144798
> Email: kamal.yadav at gmail.com, kamalyadav at ee.iitb.ac.in
> _______________________________________________
> labnetwork mailing list
> labnetwork at mtl.mit.edu
> https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork
--
Michael P. Young (574) 631-3268 (office)
Nanofabrication Specialist (574) 631-4393 (fax)
Department of Electrical Engineering (765) 412-6728 (cell)
University of Notre Dame mike.young at nd.edu