Hi Richard,
I'm not sure if this is the right way to reply to your message in the forum but I'll try (for some reason your message isn't appearing in my thread).
I have attached the drawings for the AMC modules (sorry they are not totally clear but they are photocopies of photocopies). There is a block overview along with the capacitor adjust PCB (located on the side of the coil MU) and 4 drawings of the module, the important one is Page 3 module_main. In my experience that is where things go wrong.
You wrote "I watched the voltage going to the motor and it is in fact getting a pulse as opposed to a constant voltage. That being said, Tune Cap 2 looks like it is stepping more than just adjusting slowly. I watched the resistive pot and it "steps" past where it needs to be and then tries to adjust back to the resistive value it was looking for. The problem here is that it always overshoots the needed resistive value so it never actually adjust correctly. "
It could be the gain for the motor control is set too high. That is it accelerates too fast and overshoots the setpoint, turns back goes past again and so on, this kind of looks like its pulsing. Check on the AMC analogue board that jumpers LK1 and 2 are set correctly (also try the opposite to what they are and see what effect it has)
Setting LINK's
LINK1 TUNE CAPACITOR IN for VACUUM TYPE
LINK2 LOAD CAPACITOR OUT for AIR SPACED or SF6 Types
Also take a look at the small POT PCB on the side of the coil MU (drawing attached), this can also be limiting the range of the movement (try comparing both load and tune POT values).
Finally if the motor is always slowly creeping I have also seen a problem with a leaking OP amp. Try changing the components (IC's) in the load circuit for ones in the tune circuit and see if the problem swaps over.
Let me know how you get on. I have some more docs I can dig out if you need more help.
Roy
DTU Danchip
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20150812/38870d65/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: CT8048-002_Page 1 of 4 AMC module.pdf
Type: application/pdf
Size: 538195 bytes
Desc: CT8048-002_Page 1 of 4 AMC module.pdf
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20150812/38870d65/attachment.pdf>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: CT8048-002_Page 2 of 4 AMC module.pdf
Type: application/pdf
Size: 785538 bytes
Desc: CT8048-002_Page 2 of 4 AMC module.pdf
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20150812/38870d65/attachment-0001.pdf>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: CT8048-002_Page 3 of 4 AMC module_main.pdf
Type: application/pdf
Size: 766846 bytes
Desc: CT8048-002_Page 3 of 4 AMC module_main.pdf
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20150812/38870d65/attachment-0002.pdf>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: CT8048-002_Page 4 of 4 AMC module.pdf
Type: application/pdf
Size: 35402 bytes
Desc: CT8048-002_Page 4 of 4 AMC module.pdf
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20150812/38870d65/attachment-0003.pdf>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: ED300374 AMC Module wiring.pdf
Type: application/pdf
Size: 456761 bytes
Desc: ED300374 AMC Module wiring.pdf
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20150812/38870d65/attachment-0004.pdf>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: 8063-003CD Capacitor adjust PCB.pdf
Type: application/pdf
Size: 48774 bytes
Desc: 8063-003CD Capacitor adjust PCB.pdf
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20150812/38870d65/attachment-0005.pdf>
More information about the labnetwork
mailing list