[labnetwork] Temperature specification for facility process chilled water
Anteney I.M.
I.M.Anteney at soton.ac.uk
Tue Sep 15 02:16:24 EDT 2015
Denis,
I'm particularly interested in your comment about polishing the loops to get 1 mega-ohm resistance as we have struggled to do this especially when we dose the system with biocide and corrosion inhibitors. What are you using to polish the system, what inhibitors/biocides do you use and how do these affect the conductivity.
Regards
Iain
Cleanroom Manager
University of Southampton
Sent from my iPhone
On 15 Sep 2015, at 00:36, Dennis Schweiger <schweig at umich.edu<mailto:schweig at umich.edu>> wrote:
Vito,
here at UofMichigan, one of our PCW systems runs about 62.5F (16.9C) with a daily deviation of about 1F. The second runs about 59-60F (15C). It too has a daily deviation of about 1F. For both systems we may see a seasonal deviation of 3-4F as the main chiller plant transitions from "free cooling" to the use of chillers to create cooling water for the campus wide loop. We also "polish" both loops to 1 meg-ohm so that we have a "high resistance" in the RF cooling circuits.
Dennis Schweiger
University of Mighigan/LNF
734.647.2055 Ofc
On Fri, Sep 11, 2015 at 3:47 PM, Vito Logiudice <vito.logiudice at uwaterloo.ca<mailto:vito.logiudice at uwaterloo.ca>> wrote:
Dear Colleagues,
The process chilled water loop in our cleanroom has consistently oscillated between ~15.4C and ~17.2C over a span of roughly 10 minutes. Our Plant Operations group is looking into the possibility of tightening this up for us by tuning the control sequence. They have asked for guidance on an acceptable specification.
In light of this I would appreciate hearing from you as to your facility’s PCW temperature target and tolerance. Thank you much.
Best regards,
Vito
--
Vito Logiudice P.Eng.
Director of Operations, Quantum NanoFab
University of Waterloo
Lazaridis QNC 1207
200 University Avenue West
Waterloo, ON Canada N2L 3G1
Tel.: (519) 888-4567 ext. 38703<tel:%28519%29%20888-4567%20%C2%A0ext.%2038703>
Email: vito.logiudice at uwaterloo.ca<mailto:vito.logiudice at uwaterloo.ca>
Website: https://fab.qnc.uwaterloo.ca
_______________________________________________
labnetwork mailing list
labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>
https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork
_______________________________________________
labnetwork mailing list
labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>
https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork
More information about the labnetwork
mailing list