[labnetwork] Sputtering CuS and Y2O3
Brent Gila
bgila at ufl.edu
Thu Aug 24 09:32:49 EDT 2017
Hi Fouad,
We avoid S containing materials in our multi user tools, you will get
contamination. Sulfur has a high vapor pressure and if there is any
heating on the sputter chimney or if you have a substrate heater, you
will redeposit/redistribute S in the system and into films. I have
always used 1E-4 Torr as the "line in the sand" for an elements vapor
pressure vs temperature. Below this pressure, there is a very small
amount of flux from the element source. Above this pressure, the
element can produce a high enough flux for MBE growth. On a vapor
pressure/temperature curve, S has a vapor pressure of 1E-4 Torr at the
50C-60C range and I would imaging that the chimney and shutter will get
hotter than that during a sputter run.
We have a Y2O3 target and we have no known issues with this material.
We do store our targets in a desiccator and Y2O3 is hygroscopic, so you
will want to keep it in a dry ambient when not in the sputter tool.
Best Regards,
Brent
On 8/23/2017 9:48 PM, Fouad Karouta wrote:
>
> Hi all,
>
> I am looking for feedback about the use of CuS and/or Y2O3 targets in
> a multi users facility.
>
> 1-Is there any risk that CuS may decompose during sputtering and
> contaminate the system with sulphur?
>
> 2-The SDS I consulted for Y2O3 mentioned some hazards and I wonder if
> there are experiences with this material in the network especially if
> there are any health concerns.
>
> Thanks,
>
> Fouad Karouta
>
> *************************************
>
> Manager ANFF ACT Node
>
> Australian National Fabrication Facility
>
> Research School of Physics and Engineering
>
> L. Huxley Building (#56), Mills Road, Room 4.02
>
> Australian National University
>
> ACT 0200, Canberra, Australia
>
> Tel: + 61 2 6125 7174
>
> Mob: + 61 451 046 412
>
> Email: fouad.karouta at anu.edu.au <mailto:fouad.karouta at anu.edu.au>
>
> http://anff-act.anu.edu.au/
>
>
>
> _______________________________________________
> labnetwork mailing list
> labnetwork at mtl.mit.edu
> https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork