We are in the process if installing a new PECVD system which will use 100%
SiH4. In reviewing CGA-G13, it states:
>
>
> *7.2 Indoor storage and use*Buildings, rooms, or areas used for silane
> storage shall be constructed in accordance with the requirements of the
> local AHJ. *Silane shall not be stored in locations below grade* and
> silane bulk sources shall not be located indoors. For transfilling
> exceptions, see Section 18.
In our case, an installation below grade may be preferred. I am wondering
if anyone can provide any insight into the reasoning behind this. Does
anybody store/use Silane below grade?