[labnetwork] Two SEM/EBLs set up in the same room : comments
Mark K Mondol
mondol at mit.edu
Tue Dec 19 16:41:36 EST 2017
I think that you would want to have some sort of Spicer or ADE field
cancellation system on both systems. Large changes in lens current might
couple to the SEM.
However I defer to Mike Rooks opinion on this as he actually operated a
lab with both.
I have an Elionix tool about 10 meters from a Raith (Zeiss Gemini
column) and they seem to do alright, with the Spicer cancellation
system. There is also a Zeiss Orion He Ion about another 10 meters away
and a JEOL 2010 TEM nearby too.
Regards,
Mark K Mondol
On 12/19/2017 3:04 PM, Palard, Marylene wrote:
> Dear Lab Managers
> E-beam photolithography/ SEM experts,
>
> At UT Austin we are lucky to have the task of installing a new EBL/SEM in our NNCI cleanroom lab.
>
> The operation team is leaning towards having the new EBL/SEM system in the same room as the 10 years old one.
>
> Assuming that
> - both items are fitting in the room,
> - the manufacturer site survey passed,
> - we agreed with jagged edges pictures due to cleanroom flow
> Could you please share your thoughts on that plan.
>
> Sincerely
> Marylene, an avid reader of the labnetwork messages
--
Mark K Mondol
Assistant Director NanoStructures Laboratory
And
Facility Manager
Scanning Electron Beam Lithography Facility
Bldg 36 Room 229
www.rle.mit.edu/sebl
mondol at mit.edu
office - 617-253-9617
cell - 617-224-8756
More information about the labnetwork
mailing list