[labnetwork] Two SEM/EBLs set up in the same room : comments

Mark K Mondol mondol at mit.edu
Tue Dec 19 16:41:36 EST 2017


I think that you would want to have some sort of Spicer or ADE field 
cancellation system on both systems. Large changes in lens current might 
couple to the SEM.

However I defer to Mike Rooks opinion on this as he actually operated a 
lab with both.

I have an Elionix tool about 10 meters from a Raith (Zeiss Gemini 
column) and they seem to do alright, with the Spicer cancellation 
system. There is also a Zeiss Orion He Ion about another 10 meters away 
and a JEOL 2010 TEM nearby too.

Regards,

Mark K Mondol