[labnetwork] Chamber pressure during reactive e-beam evaporation
Mark Weiler
mweiler at andrew.cmu.edu
Thu Feb 7 21:57:52 EST 2019
Hi Aamar,
You could also try making a small 1/16” or 1/8” SS tube feed-through that allows a length of tube to enter the chamber and then bent into a position terminating nearby the substrate(s), thus permitting the small amount of O2 you already/currently use to be discharged directly at the substrate. Inexpensive and worth a shot...
You would secure the swage fitting on the ATM side only after the correct length of tube within the chamber is determined. If you do not have a small swage feedthrough, you could use a 1/4” feed-through (or 6mm) and then downsize with an adapter to 1/8 (3mm) nearby the substrate and have a small section of tube affixed there.
Best regards,
Mark
Mark Weiler
Equipment & Facilities Manager
Claire and John Bertucci Nanotechnology Laboratory
Electrical and Computer Engineering | Carnegie Mellon University
5000 Forbes Ave., Pittsburgh, PA 15213-3890<x-apple-data-detectors://3/1>
T: 412.268.2471<tel:412.268.5430>
F: 412.268.3497<tel:412.268.3497>
www.ece.cmu.edu<http://www.ece.cmu.edu/>
nanofab.ece.cmu.edu<http://nanofab.ece.cmu.edu/>
On Feb 7, 2019, at 22:25, Dr. Aamer Mahmood <amahmood at hbku.edu.qa<mailto:amahmood at hbku.edu.qa>> wrote:
Hello all,
We have been doing reactive oxide depositions with e-beam evaporation of metals and a few sccm oxygen flowing through our system. We do not have a variable position gate valve and can only control the oxygen MFC settings. The corresponding chamber pressure has been less than 2e-4 Torr.
I have recently received a request to deposit reactive metal oxide films at higher oxygen flow rates that'll correspond to higher chamber pressure during deposition.
I am hesitant to increase the deposition chamber pressure due to the increased flow rate because the manufacturer mentioned the aforementioned chamber pressure as an acceptable upper limit for the e-gun.
I'd appreciate any insights/experiences on this matter particularly 1) how does increased oxygen pressure in the chamber affect the system hardware, 2) any safety concerns etc.
Thanks in advance.
-Aamer Mahmood
Qatar Environment and Energy Research Institute
Doha, Qatar
This email message, including any attachments is intended solely for the addressee(s), and may contain information that is confidential or legally protected. Any use, disclosure, copying, printing or distribution of the information contained herein by persons other than the designated addressee is unauthorized and may be unlawful. If you believe that you have received this email in error, please notify the sender immediately and delete this email and any attachments from your system and discard any printout thereof. Qatar Foundation makes no warranty that this email is error- or virus-free .The views, opinions, conclusions, and other information expressed in this email are not necessarily endorsed by Qatar Foundation unless otherwise indicated by an authorized representative. As such, Qatar Foundation does not accept any liability for any emails and attachments that may violate the relevant applicable laws.
_______________________________________________
labnetwork mailing list
labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>
https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20190208/4c7c0ef8/attachment.html>
More information about the labnetwork
mailing list