[labnetwork] LPCVD film anomaly
Mary Tang
mtang at stanford.edu
Tue Aug 3 20:03:26 EDT 2021
Thanks Tony!
Yes, I was trying to remember what Ted Berg, our furnace guy at the
time, did about the "champagne nitride". I don't recall the plumbing
solution exactly, but he took a cue from Berkeley and found a way to
ensure a constant, low draw during unload.
Mary
--
Mary X. Tang, Ph.D.
Managing Director
Stanford Nanofabrication Facility
Paul G. Allen Building
420 Via Palou Mall
Stanford, CA 94305
https://snf.stanford.edu
mtang at stanford.edu
On 8/3/2021 2:29 PM, Tony Whipple wrote:
> Hello Terre;
> I thought I saw this type of image before and we had a similar problem
> with our LSN process, see image below.
> It turned out that with our setup we were having a reaction with
> residual that would react when we vented the tube. This
> reaction would cause particles to fall ( rain ) on the wafers when the
> tube was vented and being unloaded. We had talked
> with Bob Hamilton at Berekley while visiting there and noticed his
> LPCVD setup and his solution before we noticed the
> same problem with our LPCVD. After installing the fix we did not
> have this issue anymore.
>
> The solution is to provide a very limited small low flowing path
> around the gate valve so that when the gate valve is closed
> ( such as during tube vent ) the tube still has a small draw on it
> pulling any reactions toward the pump and away from the
> wafers. If you have this type of bypass already, the only thing would
> be to confirm that it is still working I guess. We had a
> pressure gauge installed next to this to confirm that it was working
> and not plugged. We would see a pressure change
> that indicated it needed cleaning soon.
>
> Regards, Tony W.
>
>
> image.png
>
>
>
>
> On Tue, Aug 3, 2021 at 11:20 AM Terre Briggs <terreb at umich.edu
> <mailto:terreb at umich.edu>> wrote:
>
> Hello All,
> We are currently having film quality issues with our LPCVD Low
> Stress Nitride, HTO and Nitride films. Without getting into all of
> the details just yet I am sharing with you some SEM images. The
> anomaly, as our furnace manufacturer describes it, are what look
> like spheres, or bubbles on the surface of the films. We are
> pretty confident that the spheres are only on the surface and not
> down into the film itself.
> It is very random across the wafers and across the wafer load.
> After further processing, the spheres come off and leave behind a
> divot, or crater on the surface. This can also be seen in the SEM
> images.
> My question is, has anyone ever seen anything like this with
> any of your LPCVD films? I have been working with Tempress for the
> last several months to resolve the issue, but they have never seen
> anything like this, and so far we have not made any progress
> in determining the source. Any help would be appreciated.
>
> Thanks.....Terre
>
> Terre Briggs
> Engineer in Research Sr.
> Lurie Nanofabrication Laboratory
> University of Michigan
> 1301 Beal Ave.
> Ann Arbor, MI 48109
> 734-320-4470
>
> _______________________________________________
> labnetwork mailing list
> labnetwork at mtl.mit.edu <mailto:labnetwork at mtl.mit.edu>
> https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
> <https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork>
>
>
> _______________________________________________
> labnetwork mailing list
> labnetwork at mtl.mit.edu
> https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20210803/eeafebea/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image.png
Type: image/png
Size: 64268 bytes
Desc: not available
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20210803/eeafebea/attachment.png>
More information about the labnetwork
mailing list