[labnetwork] LPCVD film anomaly
mark cooper
markc at exper-tech.com
Wed Aug 4 10:11:57 EDT 2021
The soft pump, (sometimes called the half gate) should be non interlocked and allow you to open while you are pulling the load. We did custom tools with a slow pump valve with standard 1/2 inch nupro, a medium pump with a NW25 gate, and a main vacuum valve with ISO 80 valve, gave more options in the pump down and backfill.
The ammonium chloride will build up on Nitride, so blanket heaters are always important.
The N2 backfill should be able to flow while the process door is open, but make sure the flow it attached to the load area either through a connection to the front flange, or through an rear connected injector.
We had built some CVD polysilicon configurations with a blanket or curtain purge in the scavenger box so the wafers would pass through N2.
Mark Cooper
Expertech General Manager
10 Victor Square, Suite 100
Scotts Valley, CA. 95066
Office: (831) 439-9300
Direct:(831) 440-4422
Mobile:(831) 332-9396
markc at exper-tech.com
________________________________
From: labnetwork <labnetwork-bounces at mtl.mit.edu> on behalf of Mary Tang <mtang at stanford.edu>
Sent: 03 August 2021 17:03
To: Tony Whipple <whipp003 at umn.edu>; Terre Briggs <terreb at umich.edu>
Cc: Labnetwork <labnetwork at mtl.mit.edu>
Subject: Re: [labnetwork] LPCVD film anomaly
Thanks Tony!
Yes, I was trying to remember what Ted Berg, our furnace guy at the time, did about the "champagne nitride". I don't recall the plumbing solution exactly, but he took a cue from Berkeley and found a way to ensure a constant, low draw during unload.
Mary
--
Mary X. Tang, Ph.D.
Managing Director
Stanford Nanofabrication Facility
Paul G. Allen Building
420 Via Palou Mall
Stanford, CA 94305
https://snf.stanford.edu
mtang at stanford.edu<mailto:mtang at stanford.edu>
On 8/3/2021 2:29 PM, Tony Whipple wrote:
Hello Terre;
I thought I saw this type of image before and we had a similar problem with our LSN process, see image below.
It turned out that with our setup we were having a reaction with residual that would react when we vented the tube. This
reaction would cause particles to fall ( rain ) on the wafers when the tube was vented and being unloaded. We had talked
with Bob Hamilton at Berekley while visiting there and noticed his LPCVD setup and his solution before we noticed the
same problem with our LPCVD. After installing the fix we did not have this issue anymore.
The solution is to provide a very limited small low flowing path around the gate valve so that when the gate valve is closed
( such as during tube vent ) the tube still has a small draw on it pulling any reactions toward the pump and away from the
wafers. If you have this type of bypass already, the only thing would be to confirm that it is still working I guess. We had a
pressure gauge installed next to this to confirm that it was working and not plugged. We would see a pressure change
that indicated it needed cleaning soon.
Regards, Tony W.
[image.png]
On Tue, Aug 3, 2021 at 11:20 AM Terre Briggs <terreb at umich.edu<mailto:terreb at umich.edu>> wrote:
Hello All,
We are currently having film quality issues with our LPCVD Low Stress Nitride, HTO and Nitride films. Without getting into all of the details just yet I am sharing with you some SEM images. The anomaly, as our furnace manufacturer describes it, are what look like spheres, or bubbles on the surface of the films. We are pretty confident that the spheres are only on the surface and not down into the film itself.
It is very random across the wafers and across the wafer load. After further processing, the spheres come off and leave behind a divot, or crater on the surface. This can also be seen in the SEM images.
My question is, has anyone ever seen anything like this with any of your LPCVD films? I have been working with Tempress for the last several months to resolve the issue, but they have never seen anything like this, and so far we have not made any progress in determining the source. Any help would be appreciated.
Thanks.....Terre
Terre Briggs
Engineer in Research Sr.
Lurie Nanofabrication Laboratory
University of Michigan
1301 Beal Ave.
Ann Arbor, MI 48109
734-320-4470
_______________________________________________
labnetwork mailing list
labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>
https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
_______________________________________________
labnetwork mailing list
labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>
https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
Expert Semiconductor Technology, Inc.
Phone: (831) 439-9300
Fax: (831) 439-8139
This email message is for the sole use of the addressee(s) and may contain Expertech confidential and legally privileged information. Access, review, use, disclosure or distribution of this email by anyone other than the intended addressee is unauthorized and prohibited. If you are not the intended recipient, please notify the sender immediately and destroy all paper and electronic copies of the original message and any attachments.
mark cooper
Expert Semiconductor Technology, Inc.
Phone: (831) 439-9300
Fax: (831) 439-8139
This email message is for the sole use of the addressee(s) and may contain Expertech confidential and legally privileged information. Access, review, use, disclosure or distribution of this email by anyone other than the intended addressee is unauthorized and prohibited. If you are not the intended recipient, please notify the sender immediately and destroy all paper and electronic copies of the original message and any attachments.
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20210804/e6a40a37/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image.png
Type: image/png
Size: 64268 bytes
Desc: image.png
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20210804/e6a40a37/attachment.png>
More information about the labnetwork
mailing list