[labnetwork] ISO 5 Cleanroom RH
Morrison, Richard H., Jr
rmorrison at draper.com
Tue Oct 12 08:06:01 EDT 2021
The Draper clean room is 7700sqft of mixed use cleanrooms . Our makeup air unit provides 22000 cfm or 68F +-2F and 45% RHM +-3%. It uses a steam generator with a dispersion coil in the main duct to add steam in the winter and a cooling coil to reduce RH in the summer. When we designed the cleanroom we asked this forum what our RH spec should be and it was clear that 45% was the preferred RH.
Rick
Richard Morrison
PMTS
Draper Laboratory
555 Technology Square
Cambridge Ma 02139
Office: 617-258-3420
Cell: 508-930-3461
From: labnetwork <labnetwork-bounces at mtl.mit.edu> On Behalf Of Aju Jugessur
Sent: Monday, October 11, 2021 3:50 PM
To: labnetwork at mtl.mit.edu
Cc: labnetwork <labnetwork-bounces at mtl.mit.edu>
Subject: [labnetwork] ISO 5 Cleanroom RH
Hello all,
I am reaching out to obtain some inputs on your cleanroom relative humidity (RH).
We have designed our ISO 5 cleanroom with RH (40 +/- 5 %). Among other cleanroom fabrication processes, the cleanroom will support EBeam nanolithography work.
The system to control RH within the required specs is quite expensive and I would like to know if anyone in the network is able to relax the RH specs, without compromising processes OR have alternative ways of locally controlling the RH, say in the actual litho. bay areas or hoods?
Your inputs are very much appreciated.
Thanks
Aju
Aju Jugessur Ph.D. IEEE Sr. Member
Director, Colorado Shared Instrumentation
in Nanofabrication and Characterization (COSINC) Facility
Member of Inclusive Culture Council (ICC)
Member of Engineering Staff Council (ESC)
University of Colorado Boulder | College of Engineering & Applied Science
4001 Discovery Drive, N360G SEEC | Boulder, CO 80303| P: 303.735.5019
E-mail: aju.jugessur at colorado.edu<mailto:aju.jugessur at colorado.edu>
Personal Zoom link: https://cuboulder.zoom.us/my/ajugessur
www.colorado.edu/facility/cosinc<http://www.colorado.edu/facility/cosinc>
[signature_1686830665]
Signature-Strengths: Focus, Activator, Futuristic, Strategic, Achiever
(CliftonStrengths)
________________________________
Notice: This email and any attachments may contain proprietary (Draper non-public) and/or export-controlled information of Draper. If you are not the intended recipient of this email, please immediately notify the sender by replying to this email and immediately destroy all copies of this email.
________________________________
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20211012/ffd71de5/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image001.png
Type: image/png
Size: 6296 bytes
Desc: image001.png
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20211012/ffd71de5/attachment.png>
More information about the labnetwork
mailing list