[labnetwork] Hold time after fusion bond
JEFFREY MATTHEW NICOLAS MC KAY
jeffmckay at ucla.edu
Thu Oct 14 17:40:30 EDT 2021
Hi Rob,
Is there any reason to hold the pair before the anneal, some other
processing being done? I assume you are doing manual wafer bonding (not
tool assisted)? During graduate work, we made a lot of manual hydrophilic
SiO2-SiO2 bonds, and did not notice any degradation on holding off anneals
in terms of bond strength/bond area for sample weeks+ old (samples kept in
~70 F environment, low humidity).
Kind Regards,
Jeff McKay
On Tue, Oct 12, 2021 at 5:06 PM Macdonald, Robert (GE Research, US) <
robert.macdonald at ge.com> wrote:
> I am looking for information on the limits of hold time after a fusion
> bond.
>
> Whether it be Si to Si or Si to SiO2 or SiO2 to SiO2; Whether hydrophobic
> or hydrophilic, for what amount of time can the bonded pair be held prior
> to anneal?
>
> I have not found a paper with this information. We have had success with
> times under sixteen hours. But we have not done a thorough study.
>
> Any input welcome.
>
> Thanks,
> Rob MacDonald
> GE Research
>
>
>
> Sent from my iPhone