Hi Long,
We used a vacuum grease to attach the silicon wafer to the Al carrier for
DRIE etching in Utah nanofab . I was able to etch a 50x50um opneing through
a standard 4-inch silicon wafer. Releasing is time-consuming by soaking
overnight in xylene and final device came out pretty good. We can
discuss more.
Thanks
Hao-Chieh
On Tue, Apr 12, 2022 at 11:22 AM Kamal Yadav <kamal.yadav at gmail.com> wrote:
> Hi Long,
>
> Your PR will mostly char/ or burn, so cool grease is recommended but when
> you do a through silicon via, you may be exposing the cool grease to plasma
> and I am not sure what kind of damage it will bring to the devices/chamber.
> Of course exposing plasma to photoresist (if that works) is fine. Thanks
>
> On Tue, Apr 12, 2022 at 9:07 AM Hollingshead, Dave <
> hollingshead.19 at osu.edu> wrote:
>
>> Hi Long,
>>
>>
>>
>> We do this routinely with our DRIE system. As Jer mentioned, we typically
>> use small drops of santovac 5 oil to adhere the two wafers together to
>> provide thermal conduction. In our case we see etch rate changes or PR
>> “burning” without a good thermal conductor between the samples.
>>
>>
>>
>> -Dave
>>
>>
>>
>> *From:* labnetwork <labnetwork-bounces at mtl.mit.edu> * On Behalf Of *Jeremy
>> Upham
>> *Sent:* Monday, April 11, 2022 23:08
>> *Cc:* labnetwork at mtl.mit.edu
>> *Subject:* Re: [labnetwork] RIE through Silicon
>>
>>
>>
>> I think your plan could work, so I'm not sure that I see the problem In
>> my experience the carbon in the PR can contribute to the plasma chemistry
>> and affect the etch, so I tend to use vacuum grease as a thermally
>> conducting binder between
>>
>> I think your plan could work, so I'm not sure that I see the problem
>>
>>
>>
>> In my experience the carbon in the PR can contribute to the plasma
>> chemistry and affect the etch, so I tend to use vacuum grease as a
>> thermally conducting binder between the etched sample and the backing wafer
>> instead. Though your system may well work for your recipe.
>>
>>
>>
>> Hope that helps,
>>
>>
>>
>> Jer Upham PhD
>> Senior Lab Manager / Research Scientist
>> Quantum Photonics Group
>> University of Ottawa
>> Tel.: 1 (613) 562-5800 ext 7325
>>
>>
>>
>>
>>
>> On Mon, Apr 11, 2022 at 6:54 PM Chang, Long <lvchang at central.uh.edu>
>> wrote:
>>
>> *Attention : courriel externe | external email*
>>
>> Hi Guys,
>>
>>
>>
>> I have an Oxford RIE with backside Helium. I want to etch through the
>> silicon wafer (380um thick). The largest pattern is a 1mm diameter circle.
>> My plan is to ride the 4” sample wafer on a 4" carrier wafer with PR. Does
>> anyone have a good solution to this problem?
>>
>>
>>
>> Thanks,
>>
>> Long
>>
>>
>>
>> _______________________________________________
>> labnetwork mailing list
>> labnetwork at mtl.mit.edu
>> https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
>>
>
>
> --
> Thanks,
> Kamal
>
> _______________________________________________
> labnetwork mailing list
> labnetwork at mtl.mit.edu
> https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20220412/836f3f67/attachment.html>
More information about the labnetwork
mailing list