[labnetwork] Regarding : Aluminium E-Beam Evaporation
Youry Borisenkov
yb2471 at columbia.edu
Mon Aug 29 14:55:20 EDT 2022
Hi Chandan,
Here at Columbia we use a Fabmate crucible for Al e-beam depositions. We
suggest using e-beam to our users only when they tried the thermal
evaporation and if didn't work for their needs.
The crucible does break often (I would assume every 3-5 depositions) and
the only thing that looks helpful so far is ramping up and down slowly (you
mentioned 5 min, this is what we do for both up and down ramps).
It would be helpful to get a summary of your findings :)
Thanks,
Youry
On Mon, Aug 29, 2022 at 8:12 AM Chandan H B <chandanachar95 at gmail.com>
wrote:
> Dear Lab network community,
>
> We are currently facing an issue in the deposition of Aluminium in one of
> our Electron beam evaporation tools, we tried intermetallic, fab mate,
> copper, glassy coated graphite, and graphite crucibles for the evaporation.
> It seems that none of them are working out.
> We see crucible gets broken at the initial deposition quite often.
>
> Are we missing out on any parameters unchecked?
> Kindly recommend us a few parameters or solutions for the same.
>
> Any suggestions/Inputs are highly appreciated.
> Thanks in advance!
>
> Regards,
> CHANDAN H B
> THIN FILM ENGINEER
>
> P.S: Here are a few parameters that are provided to the tool.
> Power: 10KW
> Voltage: 10KV constant
> Current: Variable
> Rise1 & Soak1: 5min & 1min /8min & 2min
> Rise1 Power: 4% (for intermetallic crucible)
> Rise2 & Soak2: 5min & 1min /8min & 2min
> Rise2 Power: 8% (for intermetallic crucible)
> Ramp down: 5min
> Rate of deposition: 0.1nm/sec
> Beam Pattern: Spot Beam at the center of the crucible
> Crucible Volume: 20cc
> Material fill %: As recommended 67-75%
>
>
> _______________________________________________
> labnetwork mailing list
> labnetwork at mtl.mit.edu
>
> https://urldefense.proofpoint.com/v2/url?u=https-3A__mtl.mit.edu_mailman_listinfo.cgi_labnetwork&d=DwICAg&c=009klHSCxuh5AI1vNQzSO0KGjl4nbi2Q0M1QLJX9BeE&r=-s6JocV5E0wqGBIx8T7r2r5AeZ7Q3e53LV38GalxD2Y&m=aQwcGmt0_urp7RDhNMOEuY0d9YGSPAEzS5xAEcrltEfcPc9mJbagWZ-atwt8m6SD&s=i22peDdbqIIv0AdKsS5Sv09mYNeEylBECw2f8MKvieM&e=
>
>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20220829/41d29a1f/attachment.html>
More information about the labnetwork
mailing list