Hi Everyone,

My current supplier lists 24-36 months for the gases I use; they are 
unable to supply BCl3 at this time so there is no shelf life listed. I 
do not use B2H6.

I noticed that on the list Tom shared, H2 is listed at 12 months; that 
seems awfully short - my supplier lists 36 months but most of us 
probably regard that as being on the conservative side. Also, O2 and Ar 
seem to go bad in 3 years...

Happy holidays,

Iulian

iulian Codreanu, Ph.D.
Director, Nanofabrication Facility
University of Delaware
Harker ISE Lab, Room 163
221 Academy Street
Newark, DE 19716
302-831-2784
http://udnf.udel.edu

On 12/2/2022 10:46 AM, Tom Britton wrote:
>
> I will add in that this is from the “gas supplier”, and they may have 
> a vested interest in customer’s continuing to buy fresh gases. Also, 
> they could be leaning conservative as the quality of the gas that is 
> needed for 10nm HVM production and that which is needed for research 
> may be quite different.
>
> Happy holidays!
>
> Tom
>
> *From:* labnetwork <labnetwork-bounces at mtl.mit.edu> *On Behalf Of 
> *Albert William (Bill) Flounders
> *Sent:* Thursday, December 1, 2022 2:43 PM
> *To:* Collins, Deon <deonc69 at illinois.edu>
> *Cc:* labnetwork at mtl.mit.edu
> *Subject:* Re: [labnetwork] BCL3 question
>
> *[EXTERNAL EMAIL]*This message originated outside of your 
> organization. Remember to always use caution when opening attachments 
> or clicking on links.
>
> All,
>
> Tom Britton has provided a wide range of excellent info to many of us.
>
> I'm confident some supplier does recommend those storage limits and
>
> large fabs have no trouble meeting them.
>
> After review, we realized to meet these limits we would be disposing of
>
> almost all of our gases - so we are going to keep drinking the old milk
>
> regardless of what the carton says.
>
> The one exception I will note is diborane - which is the gas with the 
> shortest
>
> recommended storage time on this list. We routinely had clogged 
> injectors on
>
> an lpcvd system delivering B2H6 and bottle change often solved the issue.
>
> Polymerization of diborane is well documented and there's even a 
> thread from
>
> 2012 on lab net discussing the issue.
>
> Regards
>
> Bill Flounders
>
> UC Berkeley
>
> On Thu, Dec 1, 2022 at 1:02 PM Collins, Deon <deonc69 at illinois.edu> wrote:
>
>     I can tell you we personally had a BCL3 cylinder that was much,
>     much, older than what is listed on their standards. It performed
>     excellent till the day it was changed out three years ago. That
>     cylinder is still being used on one of our cleanroom process tools
>     with great results. Purity has an affect as well.
>
>     *From:* labnetwork <labnetwork-bounces at mtl.mit.edu> *On Behalf Of
>     *Tom Britton
>     *Sent:* Wednesday, November 30, 2022 10:56 AM
>     *To:* Morrison, Richard H., Jr <rmorrison at draper.com>;
>     labnetwork at mtl.mit.edu
>     *Subject:* Re: [labnetwork] BCL3 question
>
>     Hi Richard,
>
>     I got this a while back from I believe Air Liquide, but I’m not
>     100% positive of the source. Hopefully it helps.
>
>     Happy Holidays,
>
>     Tom
>
>     Tom Britton
>
>     Director of Sales
>
>     /Critical Systems, Inc./
>
>     Direct: 208.890.1417
>
>     Office: 877.572.5515
>
>     Skype: tombrittoncsi
>
>     /www.CriticalSystemsInc.com/
>     <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.criticalsystemsinc.com_&d=DwMFAg&c=euGZstcaTDllvimEN8b7jXrwqOf-v5A_CdpgnVfiiMM&r=D1j3aH4gf_B24MHGaru0vse-Gx_NllaIE8kLU3iLsM0&m=Zc1q2ymDPllIVSzjmSZUdSs5CpFTu1sKiLVG7DOoSNE&s=tJB4xuByIZ7Ff61fhBJGZ4iH3G9CczR1NFyf0xP_86A&e=>
>
>     //
>
>     logo for email signature png
>
>     **
>
>     */Note: 7 days per week Technical Support /*
>
>     /Phone: 1-888-218-6308/
>
>     /Email: //Fieldservices at criticalsystemsinc.com/
>     <mailto:Fieldservices at criticalsystemsinc.com>
>
>     *From:* labnetwork <labnetwork-bounces at mtl.mit.edu> *On Behalf Of
>     *Morrison, Richard H., Jr
>     *Sent:* Wednesday, November 30, 2022 7:53 AM
>     *To:* labnetwork at mtl.mit.edu
>     *Subject:* [labnetwork] BCL3 question
>
>     HI everyone,
>
>     Does anybody have an estimate on how long BCl3 last in a cylinder.
>     I have a cylinder of BCL3 install in 2013 and recently the etch
>     rate has dropped, does BCl3 go bad? The gas cabinet scale is not
>     in alarm.
>
>     Rick
>
>     Richard Morrison
>
>     PMTS
>
>     Draper Laboratory
>
>     555 Technology Square
>
>     Cambridge Ma  02139
>
>     Office: 617-258-3420
>
>     Cell: 508-930-3461
>
>     This email and any files transmitted with it are confidential and
>     intended solely for the use of the individual or entity to whom
>     they are addressed. If you have received this email in error
>     please notify the system manager. This message contains
>     confidential information and is intended only for the individual
>     named. If you are not the named addressee you should not
>     disseminate, distribute or copy this e-mail. Please notify the
>     sender immediately by e-mail if you have received this e-mail by
>     mistake and delete this e-mail from your system. If you are not
>     the intended recipient you are notified that disclosing, copying,
>     distributing or taking any action in reliance on the contents of
>     this information is strictly prohibited.'. If the disclaimer can't
>     be applied, attach the message to a new disclaimer message.
>
>     _______________________________________________
>     labnetwork mailing list
>     labnetwork at mtl.mit.edu
>     https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
>     <https://urldefense.proofpoint.com/v2/url?u=https-3A__mtl.mit.edu_mailman_listinfo.cgi_labnetwork&d=DwMFaQ&c=euGZstcaTDllvimEN8b7jXrwqOf-v5A_CdpgnVfiiMM&r=D1j3aH4gf_B24MHGaru0vse-Gx_NllaIE8kLU3iLsM0&m=7yb0Udu1VWDCbtkMy7TBVYuXjWRbUsrSCl6Zg5xuU0U&s=ZGh6sqjNZbZn5BU0TcLERG5Xk5pNKBlIdvWZeabK4og&e=>
>
> This email and any files transmitted with it are confidential and 
> intended solely for the use of the individual or entity to whom they 
> are addressed. If you have received this email in error please notify 
> the system manager. This message contains confidential information and 
> is intended only for the individual named. If you are not the named 
> addressee you should not disseminate, distribute or copy this e-mail. 
> Please notify the sender immediately by e-mail if you have received 
> this e-mail by mistake and delete this e-mail from your system. If you 
> are not the intended recipient you are notified that disclosing, 
> copying, distributing or taking any action in reliance on the contents 
> of this information is strictly prohibited.'. If the disclaimer can't 
> be applied, attach the message to a new disclaimer message.
>
> _______________________________________________
> labnetwork mailing list
> labnetwork at mtl.mit.edu
> https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20221202/d5d6dba1/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image001.png
Type: image/png
Size: 9501 bytes
Desc: not available
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20221202/d5d6dba1/attachment.png>


More information about the labnetwork mailing list