[labnetwork] CDA pressure booster?

Chang, Long lvchang at Central.UH.EDU
Tue Oct 4 10:58:24 EDT 2022


We use a nitrogen booster, see attached. The booster is powered by facility nitrogen at 70 psi and automatically keeps a reservoir cylinder at 1,500 psi. We tap that cylinder for 90psi and 500 psi to run our Nanoimprinter.

[cid:F443BF5E-FD0D-4ADE-9340-BE1D7CBB59B9]
On Oct 4, 2022, at 9:32 AM, Iulian Codreanu <codreanu at udel.edu<mailto:codreanu at udel.edu>> wrote:

Dear Lab Network,

The new ion mill I expect early next year requires 160 psig CDA, which is more than what my nominally 100 psig CDA system can provide.

I am writing to ask for advice on what the best approach for providing 160 psig may be. I would like to avoid installing a compressor outside of the cleanroom and having to penetrate the cleanroom envelope.

Thank you very much,

Iulian

--
iulian Codreanu, Ph.D.
Director, Nanofabrication Facility
University of Delaware
Harker ISE Lab, Room 163
221 Academy Street
Newark, DE 19716
302-831-2784
https://urldefense.com/v3/__http://udnf.udel.edu__;!!LkSTlj0I!CGo6WvreHfIvidUWGDgQ_Ud-XdcdkIHoRtorxMUGRVETt1WuDY9fdN9ncRORjXx9OP7qkOANXM_5MjyJTAWRl1o$

_______________________________________________
labnetwork mailing list
labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>
https://urldefense.com/v3/__https://mtl.mit.edu/mailman/listinfo.cgi/labnetwork__;!!LkSTlj0I!CGo6WvreHfIvidUWGDgQ_Ud-XdcdkIHoRtorxMUGRVETt1WuDY9fdN9ncRORjXx9OP7qkOANXM_5MjyJVzWCpAI$

-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20221004/b06b9953/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: Screen Shot 2022-10-04 at 9.53.56 AM.png
Type: image/png
Size: 9206970 bytes
Desc: Screen Shot 2022-10-04 at 9.53.56 AM.png
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20221004/b06b9953/attachment.png>


More information about the labnetwork mailing list