[labnetwork] Calibration for thermocouplers for Jipelec JetFirst300
Owain Clark
odc1n08 at soton.ac.uk
Mon Dec 11 11:56:58 EST 2023
You don't need the supplier; all the information is in the tool manual to recalibrate, and it is a fairly simple process to follow. T/C is a known type to 1300C and reads a temperature directly when in good contact with wafer. Pyrometer mV readings are measured by setting the TC control temp or power % manually. Then you fill in the C:mV look up file, and auto tune all heat zones of the tool for PID gains using the TC and then with the pyrometer control. When controlling the pyrometer directly via PLC for PID tuning, you must shift the mV value with a multiplier constant to hit the required temp. That's all in the manual.
If you want to set lamp bank gains using multiple TCs for wafer heating uniformity per zone, this is done as a last step manually with as many TC's in contact with the wafer as you like. Any TC over #1 is for information only, not process control.
This is quite an important process to master, if you change substrate material emissivity or even just anneal conditions sometimes it is worth recalibrating around the loop for best control.
You will get an empty TC PID file if someone writes it without having values in all the cells, either white or yellow - I forget which. You must read values out of the PLC to PIMS before you write a file to disk. Else it will all be blank like you see. A simple thing to write a few lines of code to prevent, but whoever coded PIMs was not big on that sort of error catching.
BR, O.
From: labnetwork <labnetwork-bounces at mtl.mit.edu> On Behalf Of Hüseyin Yagci
Sent: 11 December 2023 13:38
To: Hosler, Richard S <hosler0 at purdue.edu>; labnetwork at mtl.mit.edu
Subject: Re: [labnetwork] Calibration for thermocouplers for Jipelec JetFirst300
You don't often get email from yagcih at cardiff.ac.uk<mailto:yagcih at cardiff.ac.uk>. Learn why this is important<https://aka.ms/LearnAboutSenderIdentification>
CAUTION: This e-mail originated outside the University of Southampton.
All of the comments make sense. Unfortunately, our tool (or previus usage on it) does not.
The TC calibration file is just empty. Full of quotation marks and no actual data. The process engineer that signed the tool off is not here anymore and we don't know if this list was supposed to be empty (as if, the values are hardcoded into the tool/board) or it was overwritten with an empty reading accidentally.
We are trying to get in touch with the supplier, but hasn't been fruitful so far.
Many thanks for your valuable suggestions.
Huseyin Bilge Yagci
Cardiff University
________________________________
From: Hosler, Richard S <hosler0 at purdue.edu<mailto:hosler0 at purdue.edu>>
Sent: 06 December 2023 15:19
To: Hüseyin Yagci <YagciH at cardiff.ac.uk<mailto:YagciH at cardiff.ac.uk>>; labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu> <labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>>
Subject: Re: Calibration for thermocouplers for Jipelec JetFirst300
External email to Cardiff University - Take care when replying/opening attachments or links.
Nid ebost mewnol o Brifysgol Caerdydd yw hwn - Cymerwch ofal wrth ateb/agor atodiadau neu ddolenni.
I've had many struggles with our Jipelec Jetfirst 200C over the years, but most of the issues have to do with the pyrometer's reference table which is calibrated against the TC. I would check your TC calibration file to make sure that any corrections that may have been added to the previous TC aren't skewing the replacement's results. A decently stable TC probably won't have any adjustments at all in the table file. I also purchased a dry-block thermal calibrator to do some sanity checks on various thermocouples as well, up to at least 430C anyway.
We used to use the thin wire type PT018 TCs and those aren't made anymore in Jipelec's form factor. The newer ones are rigid, and while they last longer I feel that the response isn't as fast and they are more prone to contaminating the chamber during the higher temperatures when doing pyrometer re-calibration. The older PIMS software on my tool seems to be doing some weird linear interpolation on a decidedly non-linear response curve, and it's hard to get an accurate adjustment for the whole 500-1200C range.
Other problem spots that may be affecting your readings at higher temperatures would be contamination on your upper quartz window and material falling down the pyrometer well onto the BaF2 window above the sensor itself. The large window can be acid cleaned and the BaF2 window should just have the material gently removed.
Hope that helps.
Richard Hosler
Senior Research Engineer
Birck Nanotechnology Center | Room 2287C
Office: 765.494.4712 | Email: richhosler at purdue.edu<mailto:richhosler at purdue.edu>
Wiki: https://wiki.itap.purdue.edu/display/BNCWiki
iLabs: https://purdue.ilabsolutions.com/homepage/
________________________________
From: labnetwork <labnetwork-bounces at mtl.mit.edu<mailto:labnetwork-bounces at mtl.mit.edu>> on behalf of Hüseyin Yagci <YagciH at cardiff.ac.uk<mailto:YagciH at cardiff.ac.uk>>
Sent: Wednesday, December 6, 2023 7:53 AM
To: labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu> <labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>>
Subject: [labnetwork] Calibration for thermocouplers for Jipelec JetFirst300
You don't often get email from yagcih at cardiff.ac.uk<mailto:yagcih at cardiff.ac.uk>. Learn why this is important<https://aka.ms/LearnAboutSenderIdentification>
---- External Email: Use caution with attachments, links, or sharing data ----
Hello,
We have an ECM Jipelec JetFirst 300 RTA, capable of going to 1200C in our clean room. The system has a central thermocoupler (K-type TC) for calibration and readout, three peripheral read-only TCs and a pyrometer for high temperatures. The pyrometer reading is referenced to the central TC reading. Couple of weeks ago we noticed our contact performances varied across wafers and decided to change the central TC. The newly put TC reads up to 30C lower than the previous one, as well as all the peripheral TCs. The company does not seem to be checking the uniformity between the TCs they send, and we basically do not if the new TC is reading correctly or not. We are looking for a way to calibrate them in-house, without getting NPL involved. Did anybody encounter a similar problem before?
Many thanks,
Huseyin Bilge Yagci