[labnetwork] FW: Gas Monitor
Dan Woodie
daniel.woodie at cornell.edu
Mon Jan 31 16:15:07 EST 2011
Bob / Mac,
Zellweger was the original manufacturer of the Lifeline II electrochemical cells that we just had to replace. They were acquired by Honeywell shortly after our install (2004 or so). DOD Technologies is a different company representing product from a Japanese manufacturer. Our current experience with Honeywell matches Bob's description.
Dan
Dan Woodie
Lab Use Manager
From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Robert M. Hamilton
Sent: Monday, January 31, 2011 1:33 PM
To: labnetwork at mtl.mit.edu
Subject: Re: [labnetwork] FW: Gas Monitor
At UC Bekeley we have had experience with Zellweger tape-based systems. We chose their system because a tape-based system does not require the same level of annual service to verify and calibrate each individual gas detector as chemical sensors.
Having said this, we did not choose Zellweger for our new NanoLab. I found them slow in their response, expensive and not willing to support legacy systems. I believe Honeywell acquired or reps. Zellweger but have not had recent interactions with them.
Bob Hamilton
--
Robert M. Hamilton
University of CA at Berkeley
Rm 520 Sutardja Dai Hall
Berkeley, CA 94720-1754
bob at eecs.berkeley.edu<mailto:bob at eecs.berkeley.edu>
(e-mail preferred)
510-809-8600
510-325-7557 (My personal cell)
On 1/31/2011 7:49 AM, Mac Hathaway wrote:
Hey there, Floyd,
This is Mac Hathaway, from Harvard Center for Nanoscale Systems. At CNS, We use MST and Midas sensors (Honeywell), which both seem to work pretty well. The MST has a sensor cartridge right at the monitoring point, and the Midas can do remote sensing, i.e. the sensor can be some distance from the monitored point, drawing a continuous gas sample via poly or teflon tube, up to ~100 ft away depending on the gas.
There is another group that is breaking in, some guys from Zellweger (used to be MDA, old makers of Chemcassettes). They are called DOD Technologies.
Our sensor guys, EERC out of Rochester, NY are quite good, and would happy to help you spec out your new sensors, or even help you put a new system together. They helped us with our initial setup, and are currently doing the bump tests several times a year.
Mac Hathaway
Harvard CNS
_______________________________________________
labnetwork mailing list
labnetwork at mtl.mit.edu<mailto:labnetwork at mtl.mit.edu>
https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork
--
Robert M. Hamilton
University of CA at Berkeley
Rm 520 Sutardja Dai Hall
Berkeley, CA 94720-1754
bob at eecs.berkeley.edu<mailto:bob at eecs.berkeley.edu>
(e-mail preferred)
510-809-8600
510-325-7557 (My personal cell)
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20110131/c62e4ae6/attachment.html>
More information about the labnetwork
mailing list