[labnetwork] Toxic Gas Monitoring Systems
David A. Bunzow
dabunzow at lbl.gov
Wed Aug 8 16:51:53 EDT 2012
Hi Jerry,
We have two systems at The Molecular Foundry, an older Honeywell Analytics
(about 6-7 years old and a newer system (1-2 years old).
They operate similarly to what you describe, except ours can be read remotely in
locations immediately outside the lab enclosure where sensor are located, as
well as at kiosk-type video panels that allow the FD to see what is in alarm,
where its located, and the detection level (the FD likes this since they can
make a determination of how to proceed without visiting the detection location).
Hope this helps...
David A. Bunzow
User Facilities Program Manager
The Molecular Foundry
Materials Science Division
Lawrence Berkeley National Laboratory
1 Cyclotron Road MS 67-3207
Berkeley, CA 94720
Office: 510-486-4574
FAX: 510-486-7424
Cell: 510-542-1747
On 8/8/2012 10:13 AM, Bowser, Jerry wrote:
>
> Hello All,
>
> I wanted to survey the group for information regarding the type and brands of
> toxic gas monitoring control systems in use.
>
> Here at the CNST, we use various gas sensors (about 55) connected to a custom
> PLC based system that is just about at capacity. If a gas is detected, the
> system shuts down gas cabinets, sounds alarms, and alerts our fire department.
> The major shortfall of the system is that the detection level can be read at
> the sensor in the detection zone but is not displayed at the control interface.
>
> Before we decide on upgrading or replacing our control system, I thought I
> would see what others are currently using. Thanks for your input.
>
> Jerry
>
> *******************************************
>
> Jerry Bowser
>
> NanoFab Operations Group
>
> Center for Nanoscale Science and Technology
>
> National Institute of Standards and Technology
>
> Phone: (301) 975-8187
>
>
>
> _______________________________________________
> labnetwork mailing list
> labnetwork at mtl.mit.edu
> https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20120808/ee6e13d7/attachment.html>
More information about the labnetwork
mailing list