[labnetwork] gas room means of egress
Morrison, Richard H., Jr.
rmorrison at draper.com
Tue Nov 6 06:18:41 EST 2012
Hi,
AT Draper our gas room as 3 doors. One main locked to get in and 2 more emergency doors.
Rick
Draper Laboratory
Group Leader Microfabrication Operations
555 Technology Square
Cambridge Ma, 02139-3563
www.draper.com
rmorrison at draper.com
W 617-258-3420
C 508-930-3461
-----Original Message-----
From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of Iulian Codreanu
Sent: Monday, November 05, 2012 11:33 AM
To: Fab Network
Subject: [labnetwork] gas room means of egress
Dear Colleagues,
I am trying to influence a decision on a second means of egress from the gas room at the University of Delaware and your input is very much appreciated.
This is the room where the cabinets for toxic, flammable, and pyrophoric as well as some panels for inert gases will be located. The room is adjacent to and separated from the cleanroom by a fire rated wall (I was told it is at least two hours-rated). The code consultant states that a second means of egress is not needed for this H2 occupancy room but I feel it is a must.
How many doors does your gas room have?
Thank you very much.
Iulian
--
iulian Codreanu, Ph.D.
Director of Operations, UD Nanofab
University of Delaware
_______________________________________________
labnetwork mailing list
labnetwork at mtl.mit.edu
https://www-mtl.mit.edu/mailman/listinfo.cgi/labnetwork
More information about the labnetwork
mailing list