[labnetwork] ZnS sputtering
Morrison, Richard H., Jr.
rmorrison at draper.com
Fri May 10 08:00:12 EDT 2013
That will most likely contaminate your system. It is also tricky to sputter as the Zn and S may disassociate. I would think that the target would be a powder press target, if so you will need to be very careful with heat generation. If the target gets hot it will crack and even spray particles over the machine.
I would not allow ZnS in a general use machine, too much risk to other users.
Rick
Draper Laboratory
Group Leader Microfabrication Operations
555 Technology Square
Cambridge Ma, 02139-3563
www.draper.com
rmorrison at draper.com
W 617-258-3420
C 508-930-3461
From: labnetwork-bounces at mtl.mit.edu [mailto:labnetwork-bounces at mtl.mit.edu] On Behalf Of raghavan
Sent: Friday, May 10, 2013 1:53 AM
To: labnetwork at mtl.mit.edu
Subject: [labnetwork] ZnS sputtering
Dear all
We have a request from a tool user to sputter ZnS in our sputter tool. Our sputter tool is a share tool used by many users. I would like to know from experienced users if ZnS sputtering needs any extra precautions to be taken? Does it contaminate the chamber?
Thank you
Best wishes
Raghavan
*************************************************
Dr.Vijayaraghavan
Technology Manager
National Nanofabrication Centre
Centre for Nano Science and Engineering
Indian institute of Science ( IISc)
Bangalore - 560 012
India
Ph: 09663304316
http://sindhu.ece.iisc.ernet.in/nanofab/twikii/bin/view/Main/WebHome
***************************************************
--
This message has been scanned for viruses and
dangerous content by MailScanner<http://www.mailscanner.info/>, and is
believed to be clean.
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mtl.mit.edu/pipermail/labnetwork/attachments/20130510/e88ae348/attachment.html>
More information about the labnetwork
mailing list